Ruslan wrote:

Have you looked at the design of the HP/Agilent/Keysight E1750A (sine
timebase of various standard frequencies)...?

According to the specifications on the 5087A/E1750A/E1752A spec sheet, the 1750A is no better WRT phase noise than the 5087A, and the 5087A is distinctly ... mediocre ... in that regard (not helped at all by having an AGC loop). Phase noise was obviously not HP's first design priority.

Since the OP is seeking better PN performance than his first design, which already has 10dB less PN than the 5087A, studying these HP distribution amp circuits will probably not be very instructive.

That said, if you have service manuals for the 1750A or 1752A, I'm sure many owners would be grateful if you post them -- they seem to be hard to find. Didier's site is one good place:

<http://www.ko4bb.com/getsimple/index.php?id=manuals>

Best regards,

Charles


_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.