'Talk Only' mode responds to serial poll mode which is querying all instruments 
on the bus.   And it's primary use is in automated test systems 

You want to disable talk-only mode for normal interactive control of an 
instrument using GPIB

Content by Scott
Typos by Siri

> On Feb 23, 2017, at 1:37 AM, James Peroulas <ja...@peroulas.com> wrote:
> 
> Does anyone have experience using older HP equipment with the NI
> GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old
> piece of equipment? I'm very familiar with using SCPI commands over telnet
> but not so familiar with HPIB.
> 
> So far I've installed the NI 488.2 software which has successfully found
> both the GPIB-ENET/1000 box and the HP5328A connected to it. I've opened
> the NI-488.2 communicator which should allow me to send commands to the
> counter and read back results. However, anything I type in the "send
> string" returns a timeout.
> 
> The manual for the 5328A shows example commands such as:
> CMD"?U*","T"
> But it's not clear to me what portion (all?) is the string that I should be
> sending to the 5328a.
> 
> The 5328a is set to "Talk Only" mode via dip switch. I would expect that I
> should be able to just 'read' the results, but 'read' also times out...
> 
> Thanks for any pointers,
> James
> _______________________________________________
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Reply via email to