Hello Time-Nuts,
I have an HP 5061A (relatively new chassis) with a very old Cs tube installed (working) and I’m trying to wrap my head around the set up and calibration of this unit, and just hoping for some help. As a little background, HP changed the Zeeman modulation frequency, C-field intensity, and A1 synthesizer frequency at serial number prefix 1336A. This is detailed in CHANGE 20 (pages 7-9 and 7-10) Model 5061A Manual Changes in the 5061A Manual, and covered in much more detail in HP Service Note 5061A-9 (document 05061-9061 6/80-02). OLD NEW Zeeman frequency 42.82kHz 53.53kHz C-field 61 milligauss 76 milligauss Synth. Frequency 12,631,771.6 Hz 12,631,772.5 Hz I can put the tube on frequency at the new synthesizer frequency by changing the C-field current (closer to the old C-field value, by changing A15R19). 1. The Zeeman modulation frequency is a complete mystery to me. Is it related to the value of the C-field that you are running? 2. If you are comparing the unit to GPS to calibrate it do you really care what the Zeeman frequency is? Adjust the C-field to put it on frequency and call it a day? I’m not sure if there was a physical change in the Cs tube as well. But HP mentions that the new values should have better stability. 3. Would I be better off changing the synthesizer frequency back to the old value and adjusting the C-field (I would think lower) to put the unit on frequency? My first thought is to leave well enough alone. Thanks for any insight. Skip Withrow _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.