My Racal-Dana 1992 is the same way; its time interval mode is limited
to the 1 nanosecond interpolated resolution of the counter.

Some counters support time interval averaging which will produce much
much higher resolution but often they have a minimum time interval.

If the transmission line to be measured is configured as a shorted
line, then the pulse width can be measured instead to determine the
line length and pulse width averaging is much more commonly supported.

On Sun, 7 May 2017 22:05:58 -0400, you wrote:

>Hi
>
>The 53131 and 53132 will get you more resolution. The TICC, the 5370, and the 
>SR620 will do even better. None
>of them will do as well as a really fast scope.
>
>Bob
>
>> On May 7, 2017, at 8:52 PM, Jerry Hancock <je...@hanler.com> wrote:
>> 
>> I was showing my son how we could measure the difference in cable lengths by 
>> using the velocity of light and cable velocity factor.  I used a scope to 
>> measure the offset and was then thinking the 5335 could do it more 
>> accurately, but I was wrong, as it only reports to the nanosecond.  I 
>> thought I had seen somewhere where people were getting higher resolution 
>> using software along with the 5335, no?
_______________________________________________
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.