Hi,
certainly the environmental effects are significant and it is precisely for this reason that acquisitions have been made simultaneously on two 5065, one modified and the other not. To verify the results the same measures were taken with another pair of 5065, one modified and the other not. This second test was carried out at another site 300Km from the first one. Both tests showed a better and more consistent medium/long term ADEV than the unmodified HP5065A. I want to clarify that these measures have the objective of verifying the medium / long term stability and not the short term stability of the Super version I agree it is improved.. Luciano Da "Bob kb8tq" kb...@n1k.org A "Luciano Paramithiotti" tim...@timeok.it,"Discussion of precise time and frequency measurement" time-nuts@febo.com Cc Data Tue, 5 Jun 2018 10:05:34 -0400 Oggetto Re: [time-nuts] Raw phase data of super-5065 Hi It is pretty well known that without “help” ( beyond the filter stuff on the super) the 5065 has fairly poor temperature stability and somewhat variable response to pressure and humidity. Taking two samples right off the production line likely would have given you different long term data as a result. If you are going to the trouble of watching these devices for a long time, monitor temperature, pressure and humidity. I suspect you will find correlations …. Bob > On Jun 5, 2018, at 5:44 AM, tim...@timeok.it wrote: > > > Hi Attila, > I am doing the long therm stability test on a couple of HP5065, one original A type, the second B or "super" compared with my best long therm stability, a GPSDO HPz3816A. > This test will be 30 days long. It is a second test, with a second couple of HP5065A/B, to verify the long therm stability of both the version. > > Up to now the data available for version B or "super" only concerned the improvement of phase noise. > > From the first results it seems that the HP5065A is more stable within 30days than the modified version with the filter. > It must be considered that the short-term stability of 5065 is better than the GPSDO / TICC up to 10-80k sec then, in this test configuration the valid data will be beyond these values. > The results will be ready within 28 days and I will be happy to share them in order to have everyone's opinion. > Regards, > Luciano > > > Da "time-nuts" time-nuts-boun...@febo.com > A "Discussion of precise time and frequency measurement" time-nuts@febo.com > Cc > Data Mon, 4 Jun 2018 10:11:50 +0200 > Oggetto [time-nuts] Raw phase data of super-5065 > Hoi zäme! > > Would one (preferably multiple) of the owners of a super-5065 be so kind > and share the raw phase data of an super-5065 against a stable reference > with me? The longer the data trace, the better. > > Thanks in advance! > > Attila Kinali > > -- > It is upon moral qualities that a society is ultimately founded. All > the prosperity and technological sophistication in the world is of no > use without that foundation. > -- Miss Matheson, The Diamond Age, Neil Stephenson > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. > _______________________________________________ > time-nuts mailing list -- time-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts > and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.