Bill,
A 5071A will never go solid-green "continuous operation" on its own.
That solid-green state is set manually from the blinking-green state, by
design.
This prevents you from seeing green, having a short power failure
overnight, and then you come back in the morning and see green and you
think all is well. The 5061A and 5065A have a similar fail safe.
The yellow LED means attention, as in something's wrong and it did not
warm up or lock. The internal diagnostics are pretty good so if it gives
up there's a good reason. The error code is available on the front panel
LCD or via SCPI over the RS232 port. Depending on the diagnostic log or
the text of fatal error it's pretty clear when there's no hope. In the
case of leaky tube or ion pump failure there's nothing wrong with trying
again and again and again and again until you give up hope.
On the other hand if all goes well during warm-up and lock is achieved,
yellow goes out on its own and you get blinking-green. It's up to you
then to manually set it to solid-green.
/tvb
On 7/5/2020 11:50 AM, Bill Notfaded wrote:
If the power is cycled on one of these will it always go into attention
mode yellow until cleared? Or will it go green continuous operation all in
it's own? If one of these has been on the shelf a while is it common for
one even with a good tube to report attention yellow for a while until tube
pumps down and eventually will the yellow clear on it's own?
Bill
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com