I would like to share with you my thoughts on two important points that 
concern a little all attempts to modify and upgrade the equipment.

   In general I am very reluctant to modify both the electrically and 
mechanically, in this case, the HP5065A. Modifying, even completely a board, I 
agree if there is a significant improvement, trying to use the existing pinout 
and, in the need for additional connection points, use additional connectors on 
the board. As regards the space on the pcb, smd components can be used whenever 
possible or doughter boards.

   Remove boards, transformers or anything else I find reluctant, I would 
prefer to have the possibility of being able to go back to the original 
configuration if necessary.
   The area occupied by the optional batteries, which I think almost nobody 
uses, can be used for new electronics.

   Using a switching power supply with better performance I agree and since 
there is already a dedicated DC input on the back of the 5065A I would prefer 
to use the one to connect to an external box that contains the new power supply 
and the management of a backup DC input. This eliminates the need to dismantle 
anything inside the HP5065A.

   Another important point is that of the certainty of the results of a change. 
I mean that most hobbyists who have a 5065A, including me, do not have the 
opportunity to measure the proposed improvement effects, first of all because 
they do not have a reference such as an HMaser available, nor even such a 
refined measuring system to appreciate the improvements made. I want to 
remember that between zero and 8kseconds the GPS system (e.g. HP GPSDO) in our 
laboratories has an Adev higher than that of the HP5065A and therefore the 
measure we do in that range is that of GPS, not that of rubidium.

   For this reason I invite all those who dedicate themselves to these very 
interesting changes proposed to test the results obtained in depth and to share 
them with us with numerical and graphic elements. I take this opportunity to 
thank them in advance for their scientific help.

   I want to specify that this is my point of view, it is not a rule and not 
necessarily shared by other people.

   thankyou ,  Luciano

   Luciano P. S. Paramithiotti
   tim...@timeok.it
   www.timeok.it

   Da "time-nuts" time-nuts-boun...@lists.febo.com
   A "Discussion of precise time and frequency measurement" 
time-nuts@lists.febo.com
   Cc
   Data Tue, 4 Aug 2020 13:32:31 -0400
   Oggetto Re: [time-nuts] "The Penultimate HP5065 A15"
   Hi

   If a more extensive rebuild is in the works …..

   +/-20V is (as has been observed) not an ideal voltage for “modern” 
electronics.
   If you dig into each of the boards, there is a lot of “drop it down right 
now” regulation
   done on a board by board basis. More or less what might be there:

   A1 Synth
   A2 Battery Charger
   A3 60 MHz multiplier
   A4 100 KHz divider
   A5 Digital Divider (= clock)
   A6 1 MHz divider
   A7 AC amplifier
   A8 Phase detector
   A9 Integrator
   A10 OCXO
   A11 Rb temp control
   A12 Rb assembly
   A13 5 MHz buffer
   A14 Logic
   A15 Power supply
   A16 Power for clock
   A17 Terminal board
   A18 Jumpers ( = alt for A2)
   A19 Led Clock board

   For most uses, A4,A5,A6,A16,and A19 are not required. A2 is just a pair of 
diodes (A18) rather than
   a battery charger. A17 is more part of the wiring harness than anything 
else. Looking at what’s left:

   A1 synth, this seems to be a target for various replacement schemes. Right 
now, it has a bunch
   of positive voltage rails with some circuits running on 20V. Replacement 
likely would run on <= +12.

   A3 Multiplier. Again a target for replacement in some schemes. Same supply 
as A1 for replacement.
   Existing design runs +20 direct to a lot of circuits.

   A7 AC amplifier. Now runs +/-20V. Pretty much begs for a modern op-amp based 
replacement
   board. +/-12 probably is fine for that board. A *good* -15 would work for 
the existing board with minor
   mods.

   A8 Phase detector. Replacement probably is all digital. Now runs +20V A9 
Later version of the board runs +/-15. Probably would work fine a good +/-12 
with minor mods

   A10 If it’s a 10811, it’s going to need > +18 for the heater and +12 for the 
OCXO. There are other
   “at least as good” parts that work fine on +12.

   A11 Unless you want to redo the heater windings on A12, you are stuck with 
+20 to +30V. Rest of the board
   sort of begs for a modern op-amp approach.

   A12 Lamp assembly is the only load (other than heater windings and C field). 
It does run on +20V.

   A13 +20V taken to +9 for everything on the board. Simple mod to run on +12 
(or +15 or +10 …). Replacement
   likely runs on +12

   A14 If the upstream boards get changed, this likely does as well. Sort of 
begs for a $1 MCU and a handful of resistors
   as a replacement.

   A15 ( the topic of discussion)

   So, there are two “customers” for -20V. Both would be happy with a fairly 
good -15V instead. If the replacement
   OCXO for A10 tunes 0 to 5V, the need for a negative supply becomes a bit 
unclear.

   There are a couple of places you can’t easily get around 20 to 30V. The 
heater windings on the physics
   package are the biggie. The Rb lamp driver is probably not worth messing 
with.

   If you keep A1, A3, A7, or A8, they will need the existing +20V. +12 or +15 
makes more sense for their
   replacements. A13 could easily be modified and run on +12 or +15.

   ====

   So why all this long winded yack?

   The ultimate need for +20 is really pretty small. The lamp still needs it. 
The heaters need something
   in that vicinity. They don’t need the super regulation or low noise that the 
lamp needs.

   The -20 probably does at least as well as a -15 V supply, even as the device 
sits right now.
   Long term -15 makes more sense.

   A bulk +12 (and maybe +5) likely take up the heavy lifting for most of the 
boards once they are
   modified. Indeed they would be useful even on the existing boards.

   Does this change anything you do on the A15 right now? Maybe not. It is 
worth thinking a bit on though.

   Bob





   > On Aug 4, 2020, at 6:35 AM, Poul-Henning Kamp <p...@phk.freebsd.dk> wrote:
   >
   > --------
   > Magnus Danielson writes:
   >
   >> If that is the range you aim to improve, then I strongly recommend you
   >> do re-read the postings I did about phase noise in detector range, as I
   >> have showed that phase-noise get's mixed down to DC and increase the
   >> noise there. With very ugly and hand-wavey hack I was able to
   >> significantly reduce that noise essentially by cleaning up the
   >> phase-noise.
   >
   > Ohh, absolutely.
   >
   > My plan is to put a modulated DDS synth in to get a much more modern
   > detector system.
   >
   > But right now the zener on my a15 is better at measuring temperature
   > than holding voltage stable, so I'm addressing that first.
   >
   > See the 3rd graph here:
   >
   > http://phk.freebsd.dk/hacks/HP5065A/20150908_a15/
   >
   >>>> I would go for the LTZ1000 if you can. The LM399 has gone out of
   >>>> fashion for a reason: [...]
   >>> Yes, I know, being also a volt-nut :-)
   >>
   >> I have not yet jumped down that rabbit-hole as I am by far not done with
   >> the current one. :)
   >>
   >> I've been looking at the LTZ1000 and one day I may get some.
   >
   > You *really* want the version of it which is packaged in the attractive
   > and *incredibly* useful "HP3458A" cabinet :-)
   >
   > --
   > Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
   > p...@freebsd.org | TCP/IP since RFC 956
   > FreeBSD committer | BSD since 4.3-tahoe
   > Never attribute to malice what can adequately be explained by incompetence.
   >
   > _______________________________________________
   > time-nuts mailing list -- time-nuts@lists.febo.com
   > To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
   > and follow the instructions there.


   _______________________________________________
   time-nuts mailing list -- time-nuts@lists.febo.com
   To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
   and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/time-nuts_lists.febo.com
and follow the instructions there.

Reply via email to