No one mentioned tempco, so I will.  Ideally you should do your
calibration at a temperature corresponding to the long term
average in your workshop.  If the crystal is in a piece of
equipment with a temperate rise, it should be accounted for,
and then going forward you have to leave the equipment powered
up 24/7.  The crystal is probably a tuning fork, meaning it
won't be AT cut.  It may have a substantial tempco around
room temp.  In which case that old time-nuts insult may apply:

"congratulations, nice thermometer."

Rick N6RK

BTW, I go back 48 years with crystals.

On 4/1/2022 8:30 AM, Bernd Neubig wrote:
Hi,

If you do not want to make it a time-nuts style research project, but just
look for a quick fix - here is a rule of thumb:
This kind of crystal usually has a trimming sensitivity of around -10
ppm/pF. This means, if you increase the value of both capacitors on either
side by 2 pF will increase the load capacitance by 1 pF and thus lower the
frequency by  about 10 ppm.
If you need to vary one cap slightly more than the other, (to get a finer
resolution), please use the one at the output side of the on-chip oscillator
stage.
Take care of the start-up margin (safety to get a reliable start-up after
power on). The larger the capacitors, the lower becomes the margin,
sometimes the margin drops rather quickly. Therefore, after having made the
changes, test the start-up behavior by switching on and off several times,
preferably with a slow voltage ramp.

Enjoy the crystal world (as I did for the last 45 years and still doing).

Bernd


-----Ursprüngliche Nachricht-----
Von: Dan Kemppainen [mailto:d...@irtelemetrics.com]
Gesendet: Freitag, 1. April 2022 14:47
An: time-nuts@lists.febo.com
Betreff: [time-nuts] 32.768Khz Crystal Trimming

Hi,

I've got a 32.768Khz (USA number format) crystal on a RTCC oscillator of a
small micro, and it's running fast. Around 10 seconds per day or so.
This is a bit more than an order of magnitude more than the datasheet
states.

The 9 seconds per day error should be a good measurement. The RTCC is
running a 1 second counter, and that's being compared to a 1 second counter
derived by clocking the micro from a 10Mhz EXT clock reference.
This is consistent between multiple copies of the board.

I'm assuming, the C1/C2 load capacitors to ground needs to be higher in
value to trim that oscillator closer to the correct frequency. Is this
correct? Any quick back of the napkin calculations how much additional load
capacitance would be needed?

For Ref, this is the crystal:
https://abracon.com/Resonators/ABS06.pdf
ABS06-32.768KHZ-1-T

Thanks,
Dan


_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an
email to time-nuts-le...@lists.febo.com To unsubscribe, go to and follow the
instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an 
email to time-nuts-le...@lists.febo.com
To unsubscribe, go to and follow the instructions there.

_______________________________________________
time-nuts mailing list -- time-nuts@lists.febo.com -- To unsubscribe send an 
email to time-nuts-le...@lists.febo.com
To unsubscribe, go to and follow the instructions there.

Reply via email to