I have not looked at the output spectrum of the Rb, I kind of just assumed
it would be okay since it's the most stable oscillator that we have in our
lab. I can take a look at the spectrum and report back.

If I do see output spurs, do you have any suggestions / tips on cleaning up
the signal to the point where it would work as an external reference?

-Chris

On Tue, Sep 1, 2020 at 10:06 AM Marcus D. Leech <[email protected]>
wrote:

> On 09/01/2020 11:30 AM, Christopher Flood wrote:
>
> Shape is a sinusoid. Vrms going into a 1 megaohm scope is 1.4V, so I would
> expect that to be ~.7V going into a 50 ohm load (the SDR), which I believe
> is under the max power requirement for the external reference of the N310.
>
> -Chris
>
> The notes suggest that a square-wave will produce lower phase-noise, but a
> sinusoid is also acceptable, and your VP-P should also be
>   acceptable for the LMK04828 master clock PLL chip that is used in the
> N310.
>
> Have you looked at the output spectrum of your Rb clock?  The two that I
> have had to be "cleaned up" before being used, because
>   there are typically fairly loud output spurs on "raw" Rb clocks, and
> there's no input filtering on the N310.
>
>
>
> On Tue, Sep 1, 2020 at 9:26 AM Marcus D Leech <[email protected]>
> wrote:
>
>> What are the output parameters of your Rb reference? Signal amplitude and
>> shape, etc.
>>
>> Sent from my iPhone
>>
>> > On Sep 1, 2020, at 11:18 AM, Christopher Flood via USRP-users <
>> [email protected]> wrote:
>> >
>> > 
>> >   Hi all,
>> >
>> > Sorry if this is coming through twice - I sent it once before I
>> subscribed to the list, but didn't know if it went through or not! (I
>> didn't see it in the archives...)
>> >
>> > -------
>> >
>> > I'm having some trouble using an external reference with the Ettus
>> N310. The reference I am using is a FS725 Rubidium Frequency Standard.
>> >
>> > To test the difference between the internal clock and the external
>> reference, I transmitted a 10MHz signal from the Ettus N310 with and
>> without the external reference. Using just the internal clock of the SDR,
>> the 10MHz signal slowly wanders with respect to the 10MHz signal from the
>> Rubidium Frequency Standard as seen on an oscilloscope. This is to be
>> expected since the stability of the internal clock in the SDR is worse than
>> the Rubidium Frequency Standard.
>> >
>> > However, when I connect the 10MHz signal from the Rubidium to the
>> external reference of the N310 and tell the N310 to use the external
>> reference, the 10MHz signal that the N310 generates wanders much faster
>> with respect to the 10MHz signal from the Rubidium. This is opposite of
>> what I believe should happen, assuming the N310 is truly using the external
>> reference signal.
>> >
>> > Have any of you experienced this or have any tips as to what might be
>> going on?
>> >
>> > Thanks!
>> >
>> > -Chris
>> > _______________________________________________
>> > USRP-users mailing list
>> > [email protected]
>> >
>> http://secure-web.cisco.com/1nWDwD0LbO77QthlIKVs0gyhEI4cVA_IfTj4rvvFSL4qWjw_BKsjarpNVGUg1WSCn9mLxWoszHiy1EL4wQBcyjVSZlchr4oAdnfbUKsQurDlSMkQ3xes5nNVfRZbMo1K_tC57_jfBijc7k1IJEeJI4eGE5lp6Hvxm5TrWP7xYZJN3Rno2PLqMojzmmGUON3j_FwZI5DrM1IRJJOp4zugqAiDWJkDCb9_EmunfDHehAlAUGJMMDQ9fvhujOQ6RMQSye2luXmluRt_L10jyNP6kk-a7ojOTF1lgm_5ZcB9LpYVMPx8JEbY8_VIcwrlrC-n2XC4dn41AUQyHPf47I3DWWUSBZZGaTEgad3YzhpIvd0Mk3XAPpGfdyAPTH6kEW-xpRHOSnTArwXMSBHYc-Wy_utBs6L16LMlPST1Q65z9FcHKQbE-3AjZkQbUl--HGEeSu2HcZH89_ufKsucuflnBKQ/http%3A%2F%2Flists.ettus.com%2Fmailman%2Flistinfo%2Fusrp-users_lists.ettus.com
>>
>>
>
_______________________________________________
USRP-users mailing list
[email protected]
http://lists.ettus.com/mailman/listinfo/usrp-users_lists.ettus.com