Hello Randy,

your stability measurements of 732A versus 3458A depend on the integration time. It is not possible that way to distinguish, which of both instruments contribute more noise.

For NPLC 100, you may expect around 0.1ppm noise, i.e 1µV/10V, at NPLC 1000, maybe 0.01ppm. (mains frequency for NPLC determination normally is measured automatically in the 3458A)

For noise figures, see DCV datasheet of 3458A, or have a look at real measurements in the legendary LTZ1000 blog here, should be similar what you may find with the 732A:

http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/180/
http://www.eevblog.com/forum/projects/ultra-precision-reference-ltz1000/270/

The 732A may also be less noisy, if you measure its outputs running on batteries only, and disconnect the mains.


There is no reasonable explanation for the observed drift of the 1V and 1.018V outputs.

These voltages are derived by simple resistive dividers, and any leakage currents should influence the measurement instantaneously and statically. And both outputs should behave exactly the same, looking at the schematic.

If the 10V range of the 3458A is really stable, i.e. 1µV noise at 10V from the 732A, then simply measure the 732As 1V outputs on this same 10V range on the 3458A, which will still resolve 100nV. Maybe, these 1V will also be stable in this configuration, and then your 3458A has some trouble with its 1V range.

Then, you may check the guard function, if there's a difference with the switch set to "open", or to "Low", and connecting guard to the 732A or not .
Maybe you have some odd ground loops in your cabling.

Well, maybe thermoelectric voltages may be a problem in the 1V ranges.. hope you use good connections.. blank copper cables may be oxydized, for example, giving 500µV/°C vs. copper.

Maybe the 1.000V jack is not ok internally? Some degradation or bad solder joint somewhere?

Does this drift stabilize somewhere after some time?

Frank