This answer is consistent to my experience also. I've attached a log file of 
two 3458A that represents typical operation. I ALWAYS log internal temperature 
and usually display corrected difference from CAL-10 Temp. The SN:19542 
exhibits worse T.C. characteristics and also has an unusual -T.C. for the 10 
VDC range. The graph clearly shows why logging temp along with critical 
measurements is important.

Regards...
Don

-----Original Message-----
From: volt-nuts <[email protected]> On Behalf Of Andreas Jahn
Sent: Friday, November 01, 2019 2:38 AM
To: [email protected]
Subject: Re: [volt-nuts] Lab temperature stability required for X digit 
multimeters.

Hello,

a typical 3458A has a T.C. around +/- 0.2 ppm/K.
For highest stability I would keep the temperature 23 deg C +/- less than 1 K.
And keep the fan filter clean so that the internal temperature does not rise.

But in this case you should also control humidity.
With epoxy packaged voltage references I have seen 0.5 ppm / %rH output voltage 
change.
Any epoxy PCB used within a instrument should have a similar effect on 
components soldered to the PCB.

with best regards
Andreas
Am 01.11.2019 um 01:48 schrieb Dr. David Kirkby:
> I know before I ask this question that “it depends” will be part of 
> any answer. However, I will ask it anyway.
>
> How stable does a lab temperature need to be maintained, in order to 
> warrant using an 8.5 digit multimeter such as a 3458A? The same 
> question for 6.5 and 7.5 digit meters.
>
> Dave.

_______________________________________________
volt-nuts mailing list -- [email protected] To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Attachment: HP3458A 09095 & 19542 5-Day Stability Run.pdf
Description: Adobe PDF document

_______________________________________________
volt-nuts mailing list -- [email protected]
To unsubscribe, go to 
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Reply via email to