Has there yet been published in the court documents, a schematic of Rossi's
system showing the location of the pumps and flow gauge?

On Thu, Feb 2, 2017 at 7:38 AM, Brian Ahern <[email protected]> wrote:

> Yesterday I corrected the Rossi calculations. I failed to note the water
> was above 100C with no pressure to keep it in the liquid phase. The
> metering device cannot function with a compressible fluid. It will always
> measure higher values than measuring it as a single liquid phase at the
> input.
>
> Measuring the flow beyond the heating stage is OK if the output
> temperature is below  100C.  Allowing the temperature to exceed 100C is a
> surfire way to get inflated flow measurements.
>
> Rossi was warned about involving two phase fluid flow. He did it anyway
> because it is so easy the provide inflated values.
>
> I agree with Jed that this was the most ambiguous method possible.  Use
> the minimum power to get to 103 C and have your flow meters operate in a
> two phase mode that is guaranteed to over report flow rates due to the
> increased compressibility.
>
> Once again he selected the most ambiguous method .
>