Hi,

I've been doing some hardware tests on the QC2 AD9914 DDS system. Many
things are working (tested using the runtime test mode):
* basic read/write of the registers, reset values match the datasheet
* 100000 cycles of ddstest
* selection of different slots on the backplane
* reset - a modified register returns to its default value
* FUD (aka IO_UPDATE) - the SYNC_CLK output is correctly toggling at
125MHz (3GHz/24), and I can turn it on and off using the "SYNC_CLK
enable" bit of CFR2 followed by FUD.

I'm clocking the DDS at 3GHz, using a SynthNV locked to the 10MHz
output generated by the ARTIQ SoC on USER_GPIO_P, itself locked to the
125MHz internal RTIO clock.

There are some problems though, which may be due to a hardware issue:

The DDS output is stuck at zero (including at the input of the ERA-4
amplifier), and SYNC_OUT is not toggling.

The current drawn on 1.8V (analog + digital) is abnormal: 138mA
immediately upon power up (before interaction with the FPGA) which goes
to 170mA after ddsinit. The current draw is not affected by FTW
programming, but setting all power-down bits in CFR1 diminishes it
significantly.
According to the datasheet, the 1.8V current should be >400mA. What is
the current drawn by other boards, immediately upon power up and while
the DDS is properly generating a signal?

Both 1.8V points on JP2 are powered. Current on 3.3V is ~500mA which is
consistent with the datasheet.

Could you:
* let me know what the normal 1.8V current values are.
* confirm that register read/writes work on your hardware (ddstest).
* test the rest of AD9914 support in ARTIQ as the problems I'm
experiencing might be due to a faulty DDS board.

Do you have ideas about what the problem could be?

ddsinit does this:
https://github.com/m-labs/artiq/blob/fba05531f456ba4d0f2b47d241bb039ab66ba930/soc/runtime/test_mode.c#L240