Dear Ronaldo!

I think this should be very straight forward, Generally, after Ni-column
protein can be digested with TEV (1:50, or 1:100 ratio) either at room
temperature or at 4C depending on the stability of the protein. I had tried
this at room temperature, overnight cleavage at 1:100 (TEV: Substrate)
ratio. After TEV digest you can load it onto the His-trap column again by
reducing the concentration of Imidazole by half or 1/3rd of the original
concentration at which protein was eluted. TEV digested protein will remain
unbound to His-trap column and will be in flow through. TEV with his tag
will remain bound to Ni-column. Finally Imidazole can be removed during the
concentration process by buffer exchange.

Raj
On Wed, Sep 9, 2009 at 8:48 AM, <na...@icb.ufmg.br> wrote:

> Dear CCP4bb users (this is an off-topic question),
>
> We have a few proteins being expressed as
> HIS-TAG_(TEV_cleavage_site)_PROTEIN
> and we are about to initiate the purification steps.
>
> We have already used the HiTrap-Chelating columns from GE for the first
> purification step (affinity chromatography) and we would like to move
> forward
> with TEV digestion and a second purification step.
>
> We know these following steps are very protein-dependent, but we were
> wondering
> one could share his/her experience in the following steps: removal of
> imidazole,
> cleavage protocol and cleavage identification, second chromatography, etc.
>
> Any experience would be appreciated.
>
> Thanks in advance
>
> Ronaldo.
>



-- 
Rajendra Kumar Singh

Reply via email to