Brian,
 
Working voltage is determined in the same manner as 950, 1010, etc. A special
consideration is that you may have to consider the possibility of any patient
connection being earthed.
 
Table 16:
The first row titled “equivalent to Basic insulation between parts of
opposite polarity” is what 950 refers to as Operational insulation in the
mains circuit = insulation that may protect from a Risk of Fire but not does
not provide Risk of Shock protection. 
 
The second row is for Basic & Supplementary insulation. And the third row is
for Double & Reinforced insulation. Each row is split to provide the creepage
& the clearance requirement for each working voltage column. 
 
The weird letters in the second column (A-f, A-a1, etc.) relate to Appendix E
in the back of the standard. Appendix E has diagrams showing what constitutes
operational, basic, supplementary, reinforced, & double insulation. So if you
are unsure if a particular creepage/clearance distance is considered
o/b/s/r/d, you can use Appendix E. Just remember that the “601” standards
do not use the term “operational” insulation. 
 
Regards,
Bill Bisenius
E.D.& D.
bi...@productsafet.com
www.productsafeT.com <http://www.productsafet.com/> 
 

From: Brian O'Connell [mailto:boconn...@t-yuden.com]
Sent: Wednesday, August 13, 2003 6:00 PM
To: 'emc-p...@majordomo.ieee.org'
Subject: IE C60601-1 table 16
 
Good People 
Would someone please provide relevant clause that explains how to use this
table. Are the peak/dc and rms WV used seperately to determine clearance &
creepage, as in 60950?
or is the greater of the dc or ac WV used to determine both crp & clr ? 
Is the same measurement technique for determining WV as in 60950, 1010-1, etc
? 
thanks much 
Brian 