UL allows (2601-1, clause 700.1) allows DC for di-electric withstand
production tests. But, note that the CB TRF should indicate that the applied
test voltage for Type Tests was AC.

luck, 
Brian 


-----Original Message----- 
From: Jon Griver [ mailto:[email protected]] 
Sent: Saturday, March 13, 2004 11:16 AM 
To: Forum Safety-emc (Forum Safety-emc) 
Subject: Re: Dielectric test for 60601-1 

Pierre, 

According to the text of IEC 60601-1, the test should be performed with AC 
voltage, and does not give the option of performing the test with DC 
voltage, as is the case in other standards. 

The standard allows some components to be disconnected during the test, but 
primary-earth capacitors are not mentioned. 

Nevertheless, I believe that it may be possible to  use engineering judgment 
and test using DC voltage, giving the justification in the test report. 

I give as an example the data sheet at: 
http://www.ipdpower.com/pdfs/SRP-25.pdf 

This power supply range has UL approval to UL2601-1 and TUV approval to EN 
60601-1. In the "Application Information" in the data sheet, the following 
appears: 

"This power supply has been safety approved and final tested using a DC 
dielectric strength test. Please consult factory before performing AC 
dielectric strength test." 

I do not know what justification UL and TUV used, but they seem to have 
found a way to get around the problem. 

Regards, 

Jon Griver 