>I need to run EMC tests on a piece of high-power medical equipment that 
>is used on an intermittent basis.  The run time is about 15 seconds to 
>2 minutes on-time, followed by about 10 minutes off-time.  This is 
>comparable to typical system usage.  This causes problems with EMC 
>testing, since you normally need a longer observation period.

What needs a longer observation period? In principle, you should test 
the product at as near its normal operating conditions as possible. The 
further you deviate from them, the less trustworthy your results are. In 
the worst case, a serious problem might be hidden.

Don't forget that the EMC assessment required by the new Directive gives 
you the opportunity to justify your decisions about test procedures and 
much else that is not directly in conflict with the standards you apply.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
Either we are causing global warming, in which case we may be able to stop it,
or natural variation is causing it, and we probably can't stop it. You choose!
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list.    Website:  http://www.ieee-pses.org/

To post a message to the list, send your e-mail to emc-p...@ieee.org

Instructions:  http://listserv.ieee.org/request/user-guide.html

List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:

     Scott Douglas           emcp...@ptcnh.net
     Mike Cantwell           mcantw...@ieee.org

For policy questions, send mail to:

     Jim Bacher:             j.bac...@ieee.org
     David Heald:           dhe...@gmail.com

All emc-pstc postings are archived and searchable on the web at:

    http://www.ieeecommunities.org/emc-pstc



Reply via email to