Steve,

Have you found an obvious downside to a shorter HPF cutoff of, say, 200
seconds?  Would the HCP FIX training data still apply or would the
classifier need to be retrained?

-Keith


On Thu, Oct 6, 2016 at 12:43 PM, Xu, Junqian <junqian...@mssm.edu> wrote:

>
> > On Oct 4, 2016, at 9:03 PM, Ely, Benjamin <benjamin....@mssm.edu> wrote:
> >
> > Thanks Steve, that's good to keep in mind. Our acquisition is a single
> "HCP-like" 15 minute run at MB6, 2.1mm isotropic resolution, TR=1s, AP
> phase encoding, 32-channel head coil on a 3T Skyra; hopefully that gives us
> a similar temporal profile.
>
> It’s not much about the acquisition protocol (though in the multiband era,
> MB is now related to scanner transmitter stability), but rather the scanner
> stability itself.
>
> > Sounds like I should compare our temporal stability against the HCP's -
> is there a measure you recommend?
>
> HCP Connectom Skyra scanner has quite small temporal drift and a very
> linear trend (specific to the gradient and body coil hardware
> characteristics), which a typical Skyra can’t match. To determine what
> detrending cutoff you should use for your site-specific data, you could run
> a 5-10min fBIRN phantom scan with your fMRI protocol and look at the
> scanner stability.
>