Thanks for that reply, Micahel. Exactly, what I plan to do for diffusion
protocol.

Just wondering though on your or anyone else's suggestion on running BOLD
with MB>2.

Thanks

Regards

--VM


On Wed, Dec 7, 2016 at 10:50 AM, Harms, Michael <mha...@wustl.edu> wrote:

>
> Hi,
> It appears that you are trying to replicate the basics of one of our
> Lifespan pilot protocols on a Skyra — i.e., 1.5 mm voxels with 20 total min
> of dMRI scanning.  But due to the longer TE and longer TR (resulting in
> fewer total directions), you are definitely going to have worse overall SNR
> than what we achieved.  We highly suggest that you collect some pilot data
> to assess the quality of your proposed dMRI protocol with your end-measures
> of interest, including possibly a comparison against a protocol with larger
> voxels (e.g., 1.7-1.8 mm voxels) that would recover some SNR.
>
> cheers,
> -MH
>
> --
> Michael Harms, Ph.D.
> -----------------------------------------------------------
> Conte Center for the Neuroscience of Mental Disorders
> Washington University School of Medicine
> Department of Psychiatry, Box 8134
> 660 South Euclid Ave. Tel: 314-747-6173 <(314)%20747-6173>
> St. Louis, MO  63110 Email: mha...@wustl.edu
>
> From: <hcp-users-boun...@humanconnectome.org> on behalf of "Glasser,
> Matthew" <glass...@wustl.edu>
> Date: Wednesday, December 7, 2016 at 11:49 AM
> To: "hcp-users@humanconnectome.org" <hcp-users@humanconnectome.org>
> Subject: [HCP-Users] FW: Skyra - BOLD and diffusion MRI protocols
>
>
>
> From: Matt Glasser <glass...@wustl.edu>
> Date: Wednesday, December 7, 2016 at 10:14 AM
> To: neuroimage analyst <neuroimage.anal...@gmail.com>
> Subject: Re: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>
> I wouldn’t acquire multiple repetitions of the same directions, but rather
> more unique directions.  If necessary you can acquire the DWIs with only a
> single phase encoding direction in an effort to get more unique diffusion
> directions, but it is important that you get phase reversed b0 images.
>
> As for the MB error, I don’t know how to fix it, but you should definitely
> be able to do better than MB=2 with a 32 channel head coil.  Hopefully
> someone else on the list knows how to fix this.
>
> Peace,
>
> Matt.
>
> From: neuroimage analyst <neuroimage.anal...@gmail.com>
> Date: Tuesday, December 6, 2016 at 8:54 PM
> To: Matt Glasser <glass...@wustl.edu>
> Subject: Re: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>
> Hi Matt,
>
> 32 channel head coil.
>
> I was able to run the diffusion sequence on a phantom when TR/TE was
> 7000/106 ms. Since the scan time with this tr is already 20 minutes, I was
> thinking to acquire just b0 with PA phase encoding and the other 64
> directions 2b data+b0  with AP. will that be okay? my concern is snr at te
> will not be very good and data may not be too great and may be atleast 2
> reps are desired?
>
> Thanks
>
> Regards
>
> -VM
>
> On Dec 6, 2016 6:46 PM, "Glasser, Matthew" <glass...@wustl.edu> wrote:
>
>> What coil are you using?
>>
>> Peace,
>>
>> Matt.
>>
>> From: <hcp-users-boun...@humanconnectome.org> on behalf of neuroimage
>> analyst <neuroimage.anal...@gmail.com>
>> Date: Tuesday, December 6, 2016 at 11:56 AM
>> To: "hcp-users@humanconnectome.org" <hcp-users@humanconnectome.org>
>> Subject: [HCP-Users] Skyra - BOLD and diffusion MRI protocols
>>
>> Hi, HCP Users and Developers.
>>
>> We have been trying to build a protocol for BOLD and DWI on our Skyra
>> (VE11C) using the latest release of HCP pulse sequences. But we
>> haven't been completely successful with what we wanted to achieve.
>>
>> A) BOLD: We were hoping to get a TR in the range of 750ms, res = 2mm3,
>> echo spacing (ES) approx 0.65. We were able to get to echo spacing of 0.69
>> and all the other parameters with MB = 8. However, the sequence doesn't run
>> and it gives us "Max amplitude overflow on gradient z axis" after MB factor
>> exceeds 2.  It appears to me that then we have to sacrifice TR and only run
>> with MB = 2. Is there anybody with a Skyra who has able to achieve what we
>> are hoping for and willing to share the protocol with us? OR if somebody
>> could guide us to resolve the error of amplitude overflow.
>>
>> B) DWI: The idea was to have res = 1.5mm3, 2b-values at 1000 and 2500, 64
>> directions with the best BW and ES achievable. The sequence runs for 7
>> minutes out of total 13 minutes and gives gradient power amplifier error.
>> On Michael Harms's suggestion, I adjusted flip angle to 78/160 at a TR of
>> 5500 ms with excite/refocus pulse duration of 3840 and 7680. there was a
>> pop up warning that RF is clipped and the maximum refocusing angle was 142
>> instead of 160. I placed a 64 directions diffusion vector and ran the
>> sequence under "free" mode. Again, I will appreciate if  there is anybody
>> with a Skyra who has able to achieve what we are hoping for and willing to
>> share the protocol with us, along with the diffusion vector sets? OR if
>> somebody could guide us to resolve the error.
>>
>> Thank you.
>>
>> Regards
>>
>> --VM
>>
>> _______________________________________________
>> HCP-Users mailing list
>> HCP-Users@humanconnectome.org
>> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>>
>>
>> ------------------------------
>>
>> The materials in this message are private and may contain Protected
>> Healthcare Information or other information of a sensitive nature. If you
>> are not the intended recipient, be advised that any unauthorized use,
>> disclosure, copying or the taking of any action in reliance on the contents
>> of this information is strictly prohibited. If you have received this email
>> in error, please immediately notify the sender via telephone or return mail.
>>
>
> ------------------------------
>
> The materials in this message are private and may contain Protected
> Healthcare Information or other information of a sensitive nature. If you
> are not the intended recipient, be advised that any unauthorized use,
> disclosure, copying or the taking of any action in reliance on the contents
> of this information is strictly prohibited. If you have received this email
> in error, please immediately notify the sender via telephone or return mail.
>
> _______________________________________________
> HCP-Users mailing list
> HCP-Users@humanconnectome.org
> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>
>
> ------------------------------
>
> The materials in this message are private and may contain Protected
> Healthcare Information or other information of a sensitive nature. If you
> are not the intended recipient, be advised that any unauthorized use,
> disclosure, copying or the taking of any action in reliance on the contents
> of this information is strictly prohibited. If you have received this email
> in error, please immediately notify the sender via telephone or return mail.
>
> _______________________________________________
> HCP-Users mailing list
> HCP-Users@humanconnectome.org
> http://lists.humanconnectome.org/mailman/listinfo/hcp-users
>

_______________________________________________
HCP-Users mailing list
HCP-Users@humanconnectome.org
http://lists.humanconnectome.org/mailman/listinfo/hcp-users

Reply via email to