Paula-
It's a hard problem to diagnose with out knowing what platform you are using, 
but here is what I have learned from our experience-

- We had the same trouble on our Ventana Benchmark and after much hair pulling 
and troubleshooting, found it to be the slides we were using. The only slides 
that Ventana recommends for use on their stainer are the Erie superfrost. This 
fixed the problem for us. We have since found that the Bond360 slides from 
TruScientific work equally as well.

- We also found that we had similar problems on Dako's Autostainer. This time 
it was due to the instrument not being completely level.  We also learned that 
when transferring the slides to buffer after the retrieval process, you need to 
be quick or they can dry out, causing artifact or patchy staining effect. This 
too was easily solved by using a tween buffer or by using different racks to 
transfer the slides.

I hope this is of help to you, I know how frustrating troubleshooting staining 
issues can be!

Jennifer Bull