What I did recently for two new processors, conventional type-
I  did parallel trial slides of multiple tissue types ( same types as patients) 
 for fixation, morphology , processing artifacts for 9 programs.
I grossed them in and recorded fixation times, type, thickness,  overall 
dimensions. I ran on the test programs. Then I embedded and sectioned and 
evaluated the  results by microscopic review by techs & then the medical 
director of  the H & E stained sections for each program and tissue type. 
Looking at any autolysis, nuclear detail, poor dehydration, other processing 
problems in each set. 
 
Then I just made a simple evaluation sheet for any tissue processing related 
issues, with a number rating/scale for the results. Retained records of the 
validation runs and the stained sections used for validation. Defined 
acceptable tissue types  and dimensions for the processing programs in the SOP, 
 and then I just created back up/recovery procedure and reprocessing procedure 
and ran through those for comparison. When completed, I just compiled into 
validation summary report. 




Joelle Weaver MAOM, HTL (ASCP) QIHC
 
> From: lmurp...@aultman.com
> To: Histonet@lists.utsouthwestern.edu
> Date: Thu, 3 Apr 2014 15:26:17 +0000
> CC: 
> Subject: [Histonet] (no subject)
> 
> How is everyone validating the tissue processor for new CAP ANP.23045 
> question on function and verification of equipment?
> 
> 
> 
> LeAnn Murphy
> 
> Aultman Hospital
> 
> Canton, Ohio
> 
> 
> 
> 
> 
> 
> _______________________________________________
> Histonet mailing list
> Histonet@lists.utsouthwestern.edu
> http://lists.utsouthwestern.edu/mailman/listinfo/histonet
                                          