I am wondering if you are getting air bubbles/pockets in the cassettes 
themselves.  Are you using mesh cassettes?  Are you loading the trays with 
cassettes into an empty retort and then starting the delayed program?  If so,  
air pockets could be forming in the cassette and around the tissue pieces.  
Since there is no p/v cycle running yet, those air pockets may be just sitting 
there and the exposed tissue pieces may be getting dried out.

I suggest that once the retort is full and in ambient mode, open retort and 
move the cassette trays around a bit and see if air bubbles are rising from the 
cassettes.

When the trays of cassettes sit on the bench in formalin waiting to go on the 
VIP,  there may be enough movement when moving the container, etc,  that any 
air pockets are disrupted and the tissues are  not exposed to air pockets.

good luck,
amysue ruppert
Marshfield Labs Histology
________________________________
From: histonet-requ...@lists.utsouthwestern.edu 
<histonet-requ...@lists.utsouthwestern.edu>
Sent: Monday, February 5, 2024 12:00 PM
To: histonet@lists.utsouthwestern.edu <histonet@lists.utsouthwestern.edu>
Subject: [EXTERNAL] Histonet Digest, Vol 243, Issue 4


CAUTION: This email originated from outside of the Marshfield Clinic Health 
System. Do not click links or open attachments unless you recognize the sender 
and know the content is safe.


Send Histonet mailing list submissions to
        histonet@lists.utsouthwestern.edu

To subscribe or unsubscribe via the World Wide Web, visit
        
https://urldefense.com/v3/__http://lists.utsouthwestern.edu/mailman/listinfo/histonet__;!!J7HzeEKFbK9hUUY!PGATEy43_xJX9a3K3Na-c7VKlFfxZ5567gwny4UONDtiVRyrbBI5aHBtxx6V0DTFWbIgFABz11BpPHqIM6HdmZ9BO2ad6vok37eGCo6_gU_9gOLs_lTLCQ$
or, via email, send a message with subject or body 'help' to
        histonet-requ...@lists.utsouthwestern.edu

You can reach the person managing the list at
        histonet-ow...@lists.utsouthwestern.edu

When replying, please edit your Subject line so it is more specific
than "Re: Contents of Histonet digest..."


Today's Topics:

   1. Re: Processing artifact - delayed start (Bacon, Charles)


----------------------------------------------------------------------

Message: 1
Date: Mon, 5 Feb 2024 13:07:41 +0000
From: "Bacon, Charles" <charles.ba...@baystatehealth.org>
To: Verizon wireless <someperso...@yahoo.com>,
        "histonet@lists.utsouthwestern.edu"
        <histonet@lists.utsouthwestern.edu>
Subject: Re: [Histonet] Processing artifact - delayed start
Message-ID:
        
<dm6pr02mb4265a07fd5e22f4d78a7c64a92...@dm6pr02mb4265.namprd02.prod.outlook.com>

Content-Type: text/plain; charset="utf-8"

We have had 2 reasons why we saw processing issues like this:

1. I recently found out on our VIP 5 they did not turn the level sensors on 
during install. These sensors are known to error so often, Sakura tells 
technicians to set the default to off. All the processor can sense is pressure 
and time. So it may be that the formalin is not filling all the way. You are 
noticing this on delayed runs, but on those runs are you often stacking trays? 
If so, isolate the cases in the upper trays and review.

2. We found a clogged line. This happens with the formalin lines if you don?t 
do the hot water flush often enough. This can be an even bigger issue if you 
recycle and re-buffer your formalin (we do not).

Good luck!

Chuck Bacon, HTL(ASCP)CM
Supervisor Histology
Baystate Medical Center


-----Original Message-----
From: Verizon wireless <someperso...@yahoo.com>
Sent: Friday, February 2, 2024 10:37 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Processing artifact - delayed start

Dear Histonet Members,

We have terrible processing artifact if tissue sits in the formalin-filled 
retort (at ambient temperature) for too long (more than 10-12 hours) before a 
delayed process starts. The longer the wait, the worse it looks. We have Tissue 
Tek VIP 5 processors, and we process luminal gastrointestinal biopsies 
exclusively. I've attached some photomicrographs of problem cases on the 
Histonet Images website (with the same topic title).
This artifact typically affects a few specimens per day (~2% or less), even 
though everything is done on the same processor; it may affect all tissue 
portions in a cassette or only some of the tissue in that cassette. Some tissue 
portions may only have the artifact on one half with the other half looking 
perfectly fine. The techs sometimes note the tissue feeling "crunchy" at the 
time of embedding and / or at the time of microtomy. These tissues tend to 
suffer greater chatter artifact and have trouble sticking to the slides. The 
sections look just as bad on recuts as the originals. Re-processing does not 
seem to help at all.
If the cassettes sit in formalin in a container outside of the processor for 
days before the processor is loaded (with subsequent immediate start), things 
look perfectly fine. When we have staff around to start the processor 
immediately upon loading cassettes and empty immediately upon completion, the 
tissue looks perfectly fine.

Our current processing program is as follows:

1. 10% Formalin, 30 minutes, ambient temp, p/v on 2. 10% Formalin, 30 minutes, 
ambient temp, p/v on 3. 65% Alcohol, 10 minutes, ambient temp, p/v on4. 80% 
Alcohol, 10 minutes, ambient temp, p/v on5. 95% Alcohol, 7 minutes, ambient 
temp, p/v on6. 95% Alcohol, 7 minutes, ambient temp, p/v on 7. 100% Alcohol, 10 
minutes, ambient temp, p/v on 8. 100% Alcohol, 10 minutes, ambient temp, p/v on 
9. Xylene, 15 minutes, ambient temp, p/v on 10. Xylene, 15 minutes, ambient 
temp, p/v on 11. Paraffin, 10 minutes, 60 degrees C, p/v on12. Paraffin, 10 
minutes, 60 degrees C, p/v on13. Paraffin, 15 minutes, 60 degrees C, p/v on14. 
Paraffin, 15 minutes, 60 degrees C, p/v on

Obviously there are times we absolutely need to use delayed start. I would 
greatly appreciate guidance, and I'll be happy to provide any other details 
that might be useful.
Sincerely,
Brian Quigley MDLaboratory Director of a GI Pathology Laboratory



------------------------------

Subject: Digest Footer

_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
https://urldefense.com/v3/__http://lists.utsouthwestern.edu/mailman/listinfo/histonet__;!!J7HzeEKFbK9hUUY!PGATEy43_xJX9a3K3Na-c7VKlFfxZ5567gwny4UONDtiVRyrbBI5aHBtxx6V0DTFWbIgFABz11BpPHqIM6HdmZ9BO2ad6vok37eGCo6_gU_9gOLs_lTLCQ$

------------------------------

End of Histonet Digest, Vol 243, Issue 4
****************************************

______________________________________________________________________
The contents of this message may contain private, protected and/or privileged 
information.  If you received this message in error, you should destroy the 
e-mail message and any attachments or copies, and you are prohibited from 
retaining, distributing, disclosing or using any information contained within.  
Please contact the sender and advise of the erroneous delivery by return e-mail 
or telephone.  Thank you for your cooperation.
_______________________________________________
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Reply via email to