Hi Jonathan,

only patch 1 fixes a bug which is occurring on systems. In my test it happened
every one of about 10 - 15.000 measurements. But this depends on the interrupt
load of the system. With a high interrupt load there might be much more wrong
measurements.

Patch 2 is a performance optimization which prevents unneded reads. There is no
bug which is fixed by it.

I'll send out the patch set with more detailed commit messages.

Thanks,

Andreas


Jonathan Cameron <ji...@kernel.org> schrieb am So, 08. Sep 14:49:
> On Sat, 7 Sep 2019 12:18:00 +0200
> Andreas Klinger <a...@it-klinger.de> wrote:
> 
> > Fix bug in sampling function hx711_cycle() when interrupt occures while
> > PD_SCK is high. If PD_SCK is high for at least 60 us power down mode of
> > the sensor is entered which in turn leads to a wrong measurement.
> > 
> > Move query of DOUT at the latest point of time which is at the end of
> > PD_SCK low period.
> > 
> > Signed-off-by: Andreas Klinger <a...@it-klinger.de>
> 
> Hi Andreas,
> 
> One thing I'm not clear on from these is how much a 'fix' they
> are.  That just effects whether we mark them for stable / push them
> out as quickly as possible or not.  So has this been seen in 
> normal operation?
> 
> + please add fixes tags to the two fixes.
> 
> For patch 3, it's in the very low importance category so it may
> well get forgotten if these two go through the fixes tree
> (up to you to remind me!)
> 
> Thanks,
> 
> Jonathan
> 
> > ---
> >  drivers/iio/adc/hx711.c | 9 +++++----
> >  1 file changed, 5 insertions(+), 4 deletions(-)
> > 
> > diff --git a/drivers/iio/adc/hx711.c b/drivers/iio/adc/hx711.c
> > index 88c7fe15003b..0678964dbd21 100644
> > --- a/drivers/iio/adc/hx711.c
> > +++ b/drivers/iio/adc/hx711.c
> > @@ -101,13 +101,14 @@ struct hx711_data {
> >  static int hx711_cycle(struct hx711_data *hx711_data)
> >  {
> >     int val;
> > +   unsigned long flags;
> >  
> >     /*
> >      * if preempted for more then 60us while PD_SCK is high:
> >      * hx711 is going in reset
> >      * ==> measuring is false
> >      */
> > -   preempt_disable();
> > +   local_irq_save(flags);
> >     gpiod_set_value(hx711_data->gpiod_pd_sck, 1);
> >  
> >     /*
> > @@ -117,7 +118,6 @@ static int hx711_cycle(struct hx711_data *hx711_data)
> >      */
> >     ndelay(hx711_data->data_ready_delay_ns);
> >  
> > -   val = gpiod_get_value(hx711_data->gpiod_dout);
> >     /*
> >      * here we are not waiting for 0.2 us as suggested by the datasheet,
> >      * because the oscilloscope showed in a test scenario
> > @@ -125,7 +125,7 @@ static int hx711_cycle(struct hx711_data *hx711_data)
> >      * and 0.56 us for PD_SCK low on TI Sitara with 800 MHz
> >      */
> >     gpiod_set_value(hx711_data->gpiod_pd_sck, 0);
> > -   preempt_enable();
> > +   local_irq_restore(flags);
> >  
> >     /*
> >      * make it a square wave for addressing cases with capacitance on
> > @@ -133,7 +133,8 @@ static int hx711_cycle(struct hx711_data *hx711_data)
> >      */
> >     ndelay(hx711_data->data_ready_delay_ns);
> >  
> > -   return val;
> > +   /* sample as late as possible */
> > +   return gpiod_get_value(hx711_data->gpiod_dout);
> >  }
> >  
> >  static int hx711_read(struct hx711_data *hx711_data)
> 

-- 
Andreas Klinger
Grabenreith 27
84508 Burgkirchen
+49 8623 919966
a...@it-klinger.de
www.it-klinger.de