Diablo Canyon - We have small check sources available at strategic locations to test the instrument response prior to use. The source is mounted in a jig (in most cases) and allows for a single count (or dose) rate response.
________________________________ From: [email protected] [mailto:[email protected]] On Behalf Of Liesl K Germann Sent: Sunday, February 26, 2012 12:34 PM To: [email protected] Subject: Powernet: EP Instrument Source Checks Do you apply the below ANSI N323A-1997 Section 4.8 statement regarding source response checks to your Emergency Planning instruments? 4.8 Source response check To ensure proper operation of the instrument between calibrations, each instrument (with the exception of neutron instruments and high-dose equivalent rate photon instruments) shall be checked with a source during operation at least daily or prior to each intermittent use, whichever is less frequent. If at any time the instrument response to the source differs from the reference reading by more than 20% (for any photon instrument the reading should be at least ten times background), the instrument shall be returned to the calibration facility for calibration or for maintenance, repair, and recalibration, as required. [ ...] Reference readings should be obtained for one point on each scale or decade used. Thank you! Liesl Germann Health Physicist Millstone Power Station 860-447-1791 Ext. 6511 [email protected]<mailto:[email protected]> CONFIDENTIALITY NOTICE: This electronic message contains information which may be legally confidential and/or privileged and does not in any case represent a firm ENERGY COMMODITY bid or offer relating thereto which binds the sender without an additional express written confirmation to that effect. The information is intended solely for the individual or entity named above and access by anyone else is unauthorized. If you are not the intended recipient, any disclosure, copying, distribution, or use of the contents of this information is prohibited and may be unlawful. If you have received this electronic transmission in error, please reply immediately to the sender that you have received the message in error, and delete it. Thank you.