A LONG time ago, I did my Master's thesis on alpha spectroscopy at nuclear 
power plants.  I did all my research at Crystal River.  I must admit I have not 
kept up with the technology changes since the late 1970's but I suspect the 
problems I had then might still exist now.  At that time we used Ortec units 
and the actual counting of the sample was relatively straightforward.  The 
problem was sample preparation because getting a very thin layer is critical.  
The standard at the time was electrodeposition but I was using direct 
filtration onto a Millipore filter.  Claude Sill (an early pioneer/father of 
alpha spec) was one of my mentors on this work and he was very helpful with 
sample preparation protocol.

All that said,  I suspect we may begin looking at alpha spec here at DCPP.  We 
have a number of challenges here that could be helped with alpha spec.  We have 
high radon backgrounds at times (tens of DAC) and low level 2 conditions 
(1000-3000) in Unit 2.

Happy Saturday All,

Linda


Linda M. Sewell, CHP
Principal Health Physicist
Pacific Gas & Electric Company
Diablo Canyon Power Plant
MS 119/1/117
PO Box 56
Avila Beach, CA 93424
                                                                                
                                                                
P: 805.545.4315 | F: 805.545.2618| [email protected]<mailto:[email protected]>



From: [email protected] [mailto:[email protected]] On Behalf Of 
Holmes, Stephen J
Sent: Friday, May 04, 2012 9:41 AM
To: '[email protected]'
Subject: Powernet: Benchmark: Alpha Spec

At Ginna, we are evaluating purchasing Alpha Spec equipment (such as the 
Canberra Alpha Analyst with Alpha Apex software). Part of our justification for 
this expense would be to reduce our reliance on independent laboratories for 
performing alpha characterization (i.e. per the EPRI alpha monitoring 
guidelines). We currently have Ludlum 3030Ps and MGPI ABPMs which use alpha 
spec in a limited manner to perform radon rejection, but no true alpha 
spectroscopy equipment. We would like to know what the rest of the industry is 
doing; any experience, comments, or feedback are appreciated.

Do you have Alpha Spec capabilities at your facility?


If so, what equipment do you use?


What do you use the equipment for (e.g. radon rejection or a quantitative alpha 
spectrum of individual nuclides)?



If using for alpha spectrum analysis, what energy resolution can you achieve?



What type of sample preparation is required?



Any other equipment strengths/limitations or lessons learned you would like to 
share?


Thanks,
-Steve

Stephen J Holmes, PE
Sr. Plant Health Physicist, CENG-RE Ginna NPP
[email protected]<mailto:[email protected]>
585-771-3577; 585-545-1661 (cell)


>>> This e-mail and any attachments are confidential, may contain legal,
professional or other privileged information, and are intended solely for the
addressee.  If you are not the intended recipient, do not use the information
in this e-mail in any way, delete this e-mail and notify the sender. -IP2

Reply via email to