Kurt,

It looks like incorrect step size (difference between actual step size and
one that is written in the file) or incorrect distance to the detector. The
later may incorrect step size if detector is PSD.

I have similar situation with much smaller difference in peak position but
the shift (difference between actual and written step size) varies from
measurement to measurement. Bruker is working on this but so far no
solution/cause is known.

Your case may be simpler if the difference is constant. Perhaps calibration
of detector or its distance to the sample is needed.

Best,

 

Peter Zavalij 



X-ray Crystallographic Center
University of Maryland

College Park, MD



Office: (301)405-1861
Lab: (301)405-3230

Fax: (301)314-9121



From: Kurt Leinenweber [mailto:ku...@asu.edu] 
Sent: Thursday, January 08, 2009 6:09 PM
To: Rietveld_l@ill.fr
Subject: error in peak positions that is linear in theta

 

Dear Rietvelders,

 

(I hope this list is appropriate for this question).  I am trying to set up
a new D8 diffractometer from Bruker, and have it set up with a Vario
monochromator focusing on a sample in a capillary.  Of course I hope to do
Rietveld refinement on the samples eventually.  But there is an error in two
theta of the peak positions on all samples, and the error is linear in two
theta.  At 40 degrees the error (measured minus expected) is about 0.020
degrees, at 80 degrees the error is about 0.040 degrees, etcetera.

 

Has anyone ever seen anything like this and do you know where it could come
from?

 

Thank you very much,

 

                                                            - Kurt

 

 

*********