Dear Mr. Peter
I also faced the same problem as Kurt faced, our diffractometer is the 
D8-advance model from Bruker... I took the scan using a PSD detector(Lynxeye) 
and got an increase of the error in peak position with increasing 2theta.. 
Again, when I extracted the dat file from the raw one.... i observed the 
rounded up values of 2theta(up to 2 decimal pt.) in the file. Now will the 
problem be optimized if I generate 2Theta values using a step which is rounded 
up upto a higher number of decimal point and replace them in place of my 2theta 
column? thanking you..

prasun



----- Original Message -----
From: "Peter Y. Zavalij" <pzava...@umd.edu>
Date: Friday, January 9, 2009 7:17 pm
Subject: RE: error in peak positions that is linear in theta

> Kurt,
> 
> It looks like incorrect step size (difference between actual step 
> size and
> one that is written in the file) or incorrect distance to the 
> detector. The
> later may incorrect step size if detector is PSD.
> 
> I have similar situation with much smaller difference in peak 
> position but
> the shift (difference between actual and written step size) varies 
> frommeasurement to measurement. Bruker is working on this but so 
> far no
> solution/cause is known.
> 
> Your case may be simpler if the difference is constant. Perhaps 
> calibrationof detector or its distance to the sample is needed.
> 
> Best,
> 
> 
> 
> Peter Zavalij 
> 
> 
> 
> X-ray Crystallographic Center
> University of Maryland
> 
> College Park, MD
> 
> 
> 
> Office: (301)405-1861
> Lab: (301)405-3230
> 
> Fax: (301)314-9121
> 
> 
> 
> From: Kurt Leinenweber [ku...@asu.edu] 
> Sent: Thursday, January 08, 2009 6:09 PM
> To: Rietveld_l@ill.fr
> Subject: error in peak positions that is linear in theta
> 
> 
> 
> Dear Rietvelders,
> 
> 
> 
> (I hope this list is appropriate for this question).  I am trying 
> to set up
> a new D8 diffractometer from Bruker, and have it set up with a Vario
> monochromator focusing on a sample in a capillary.  Of course I 
> hope to do
> Rietveld refinement on the samples eventually.  But there is an 
> error in two
> theta of the peak positions on all samples, and the error is 
> linear in two
> theta.  At 40 degrees the error (measured minus expected) is about 
> 0.020degrees, at 80 degrees the error is about 0.040 degrees, 
> etcetera.
> 
> 
> Has anyone ever seen anything like this and do you know where it 
> could come
> from?
> 
> 
> 
> Thank you very much,
> 
> 
> 
>                                                            - Kurt
> 
> 
> 
> 
> 
> *********
> 
> Kurt Leinenweber
> 
> Dept. of Chemistry and Biochemistry
> 
> Arizona State University
> 
> Tempe, AZ 85287-1604
> 
> 
> 
> phone (480)-965-8853
> 
> fax (480)-965-2747
> 
> 
> 
> ***********
> 
> 
> 
> 

Reply via email to