[PSES] FW: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread ce-test, qualified testing bv - Gert Gremmen
 

Several EMC antennas have AF step performance of 1 dB per 5 MHz or so at
certain frequencies.

 

If the antenna would have a linear behavior, 2 points would be enough

as all other frequencies could be linearly interpolated.

 

But my particular Chase BILOG 6111A has a bump at 14.1 dB at 258 MHz

and at both  250 MHz and 266 MHz  the antennafactor is 13.0 dB.

This is what I read from the calibration graph.

 

The manufacturer however, also has calibration (numeric) data provided
as discrete points at

comfortable round numbers of 250 (12.9) and 300 MHz. (13.6)

These points are in general used for creating a interpolation curve.

 

Linear interpolation using the table provides a antennafactor of 13.1 dB
at 258 MHz

which is a full -1 dB off from what I read from the graph. The bump has

virtually disappeared. Nice result with a 0.2 dB receiver.

 

That is why the output level change and required accuracy 

should be the driving factor in creating a calibration table.

It can easily be automated during a sweep.

 

What complicates this further is that not all manufacturers provide

data about their interpolation system. My RS receiver uses quadratic
spline

(I suppose Lagrange).

 

My RS FSP only uses LIN interpolation (with log option) for the
transducers.

 

 

 

 

 

Regards,

Ing.  Gert Gremmen, BSc

 

 

 

g.grem...@cetest.nl

www.cetest.nl


Kiotoweg 363

3047 BG Rotterdam

T 31(0)104152426
F 31(0)104154953

 

Before printing, think about the environment. 

 

 

Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Ken Javor
Verzonden: Tuesday, December 06, 2011 11:12 PM
Aan: EMC-PSTC@LISTSERV.IEEE.ORG
Onderwerp: Re: [PSES] Calibration practice for EMI test transducers

 

This discussion has gone far a field from the original post.  I'm
assuming some sort of scan has been run and a continuous sweep is
available. The question is, how densely does that sweep need to be
digitized? Not as a practical matter, but as a matter of compliance with
standards or standard practice.

What manner of antenna would have performance as cited below where at
100 MHz the antenna factor is 10 dB and at 120 MHz it is 12 dB, but at
110 MHz it could be 20 dB? Let's generalize the question to any closely
spaced frequencies.

Let's look at the types of antennas available.

30-200 MHz: A half-wave tuned dipole is nowhere near that sharp.
Neither is a biconical.

200 - 1000 MHz:  A half-wave tuned dipole is nowhere near that sharp.
Neither is a logperiodic, log-spiral, nor a Yagi.

1 GHz+:  Logperiodic, log-spiral nor pyramidal horns act the way
surmised.

My conclusion, and the point here is to invite discussion, not close it
out, is that no test-type antenna is a high Q device.  Antennas can
have arbitrarily high gains, depending on construction, but the high
gain is a geometrical quality, not a high quality factor in the
frequency domain. 
 
Ken Javor

Phone: (256) 650-5261



From: ralph.mcdiar...@schneider-electric.com
Date: Tue, 6 Dec 2011 11:25:43 -0800
To: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
Cc: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Calibration practice for EMI test transducers


Yet how far would you take that argument?  If it's say 11.1db at 110MHz
and 11.5dB at 112MHz, would you still worry about 
factor at 111MHz?  At some point surly this all becomes rather silly.

_ 

Ralph McDiarmid  |  Schneider Electric  |  Renewable Energies Business
|   CANADA  |   Regulatory Compliance Engineering



From: ce-test, qualified testing bv - Gert Gremmen
g.grem...@cetest.nl 
To: EMC-PSTC@LISTSERV.IEEE.ORG 
Date: 12/05/2011 10:06 PM 
Subject: Re: [PSES] Calibration practice for EMI test transducers 






If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate. 
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver is +/- 0.5 dB you should register a
calibration
 each time the difference between previous and current sample is 0.5 dB.
Then the total error will be slightly more than 0.5 dB.
There are more sophisticated interpolation methods such as 
cubic spline and polynomial  interpolation, and the error item
is subject to higher mathematics.
Play with it at:

http://terpconnect.umd.edu/~petersd/interp.html
http://terpconnect.umd.edu/~petersd/interp.html
http://terpconnect.umd.edu/~petersd/interp.html  

or at http://www.ibiblio.org/e-notes/Splines/Lagrange.htm
http

Re: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread Ken Javor
The quotation below is a little out-of-context. In the previous statement in
that post, it says,  ...the point ... is that no test-type antenna is a
high ³Q² device.

It is true that a tuned loop can be a very high Q antenna, but these are
designed solely for communications. No one uses such antennas for EMI
testing, precisely because of the sharp tuning involved.

And if one were to do so, presumably with a knowledge of the antenna's
behavior, the antenna factor would be properly characterized at the
frequency (ies) of operation, and not missed because the tuned frequency was
interpolated between two untuned frequencies.
 
Ken Javor

Phone: (256) 650-5261


 From: John Woodgate j...@jmwa.demon.co.uk
 Date: Wed, 7 Dec 2011 04:52:25 +
 To: EMC-PSTC@LISTSERV.IEEE.ORG
 Subject: Re: [PSES] Calibration practice for EMI test transducers
 
 In message cb03ef45.c9421%ken.ja...@emccompliance.com, dated Tue, 6
 Dec 2011, Ken Javor ken.ja...@emccompliance.com writes:
 
  Antennas can have arbitrarily high gains, depending on construction,
 but the high gain is a geometrical quality, not a high quality factor
 in the frequency domain.
 
 True for electric antennas, I think, but not for magnetic. Tuned loops
 and ferrite-rod antennas can have high Q. CISPR 16 includes at least one
 magnetic antenna, although it may well not be high Q.
 -- 
 OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
 John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK
 Some people who are peeling the finch of the financial crisis are thinking of
 biting a rook.
 
 -
 
 This message is from the IEEE Product Safety Engineering Society emc-pstc
 discussion list. To post a message to the list, send your e-mail to
 emc-p...@ieee.org
 
 All emc-pstc postings are archived and searchable on the web at:
 http://www.ieee-pses.org/emc-pstc.html
 
 Attachments are not permitted but the IEEE PSES Online Communities site at
 http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used
 formats), large files, etc.
 
 Website:  http://www.ieee-pses.org/
 Instructions:  http://listserv.ieee.org/request/user-guide.html
 List rules: http://www.ieee-pses.org/listrules.html
 
 For help, send mail to the list administrators:
 Scott Douglas emcp...@radiusnorth.net
 Mike Cantwell mcantw...@ieee.org
 
 For policy questions, send mail to:
 Jim Bacher:  j.bac...@ieee.org
 David Heald: dhe...@gmail.com

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread Cortland Richmond
An untuned loop should exhibit no resonances in its usable range.  See 
http://www.ets-lindgren.com/page/charts.cfm?i=6511 .
Also, A Novel Standard Loop Antenna for Antenna Calibration in the MF 
and HF Bands, ieeexplore.ieee.org/iel5/10111/32404/01513478.pdf


Amateurs have used home-made loops to get pretty accurate measurements 
of BPL emissions. http://vk1od.net/antenna/SmallUntunedSquareLoop/index.htm


One of his loops, 600mm square, has an AF flat within one dB from about 
5.5 to 30 MHz ( 
http://vk1od.net/antenna/SmallUntunedSquareLoop/Loop600.pdf ).



Cortland
KA5S

On 12/6/2011 11:52 PM, John Woodgate wrote:
True for electric antennas, I think, but not for magnetic. Tuned loops 
and ferrite-rod antennas can have high Q. CISPR 16 includes at least 
one magnetic antenna, although it may well not be high Q.



-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread Ken Javor
Now we get to the heart of the matter that prompted the original post.

That spike in the 137 cm tip-to-tip biconical used with a T1 dipole balun
(28 ­ 200 MHz) behavior might have been objectionable, but you wouldn¹t miss
it an analog sweep with even 1% frequency steps.  You would only miss it if
you calibrated every ten MHz as was done with the biconical antenna design
in MIL-STD-461A (1968 ­ when swept performance was difficult to do). There
is no need to use frequency steps commensurate with the sorts of BWs used
during EMI testing.  A 1 - 5 MHz BW would more than suffice to sweep an
antenna used above 30 MHz to ensure not missing anything.

The following quote is the rationale I have seen a few places with which I
take issue:

³Though the sweep may be continuous, if you can read amplitude only by
marking pixels, you need to insure .5 BW or less per pixel. That's straight
sampling theory, right?²

That is straight sampling theory applied to an unknown signal
characteristic.  If a continuous sweep using 1/2 BW steps as per sampling
theory has been performed, then discrete points for an antenna factor file
may be chosen by inspection so that in any particular frequency range either
of the two following conditions are met:

1. If the data is noisy, with no obvious trend, then frequency points should
be listed so that the antenna factor between any two listed points doesn¹t
change by more than 1 dB (or whatever delta is deemed necessary for adequate
accuracy). 
2. If a linear or log-linear approximation can be made over some frequency
interval, then only the two endpoints of that interval need be recorded,
with the understanding that between those two points the appropriate
interpolation will yield the desired accuracy. The interpolation has to be
one which the automated software can be instructed to perform on its
recorded antenna factor files.

One of these two approaches should work when deriving a set of discrete
points from a continuous sweep of any test transducer factor¹s performance.
The question is whether this is ³cast in stone² anywhere or left up to the
person doing the calibration, or the test engineer abstracting a subset of
data points for inclusion in his files.

 
Ken Javor

Phone: (256) 650-5261



From: Cortland Richmond k...@earthlink.net
Reply-To: k...@earthlink.net
Date: Tue, 06 Dec 2011 20:43:38 -0500
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Calibration practice for EMI test transducers

   We might *like* measuring antennas to be low-Q, but remember the bicon
shorting bars that were added some years ago? The skeleton bicons lacking
them inserted a rather sharp notch that had to be accounted for in
transducer tables. Baluns might go off from impact damage, too.  And so on.
I gave an ebay 50 MHz comb good past 2 GHz to a former employer who later
gave all their own EMC lab gear to an NRTL.
 
 But you still have to check.
 
 For that matter, I've seen coax factors programmed with GAIN at a narrow
range at microwave frequencies; that was caused by reflection from a crimped
cable*. So we do need to be able to spot them.
 
 *This was due to the ignorance of operators collecting the data, IMO. I
spoke to them and likely, they won't do it again.  But they believed the
SA/TG reading because they didn't know what they were measuring. That's
another thread.
 
 Though the sweep may be continuous, if you can read amplitude only by
marking pixels, you need to insure .5 BW or less per pixel. That's straight
sampling theory, right?  If your analyzer has a 1024 pixel wide screen you
can't SEE less than about .1% of the scan width and can rely on less. Does
it matter? Often it doesn't.
 
 Fun, isn't it?
 
 Cortland
 KA5S
 
 On 12/6/2011 5:11 PM, Ken Javor wrote:
  Re: [PSES] Calibration practice for EMI test transducers This discussion has
 gone far a field from the original post.  I¹m assuming some sort of scan has
 been run and a continuous sweep is available. The question is, how densely
 does that sweep need to be digitized? Not as a practical matter, but as a
 matter of compliance with standards or standard practice.
  
  What manner of antenna would have performance as cited below where at 100 MHz
 the antenna factor is 10 dB and at 120 MHz it is 12 dB, but at 110 MHz it
 could be 20 dB? Let¹s generalize the question to any closely spaced
 frequencies.
  
  Let¹s look at the types of antennas available.
  
  30-200 MHz: A half-wave tuned dipole is nowhere near that sharp.  Neither is
 a biconical.
  
  200 ­ 1000 MHz:  A half-wave tuned dipole is nowhere near that sharp.
 Neither is a logperiodic, log-spiral, nor a Yagi.
  
  1 GHz+:  Logperiodic, log-spiral nor pyramidal horns act the way surmised.
  
  My conclusion, and the point here is to invite discussion, not close it out,
 is that no test-type antenna is a high ³Q² device.  Antennas can have
 arbitrarily high gains, depending on construction, but the high gain is a
 geometrical quality, not a high quality factor

Re: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread John Woodgate
In message 4edf523d.4090...@earthlink.net, dated Wed, 7 Dec 2011, 
Cortland Richmond k...@earthlink.net writes:


An untuned loop should exhibit no resonances in its usable range.  See 
http://www.ets-lindgren.com/page/charts.cfm?i=6511 . 
Also, A Novel Standard Loop Antenna for Antenna Calibration in the MF 
and HF Bands, ieeexplore.ieee.org/iel5/10111/32404/01513478.pdf


Amateurs have used home-made loops to get pretty accurate measurements 
of BPL emissions. 
http://vk1od.net/antenna/SmallUntunedSquareLoop/index.htm 


One of his loops, 600mm square, has an AF flat within one dB from about 
5.5 to 30 MHz ( 
http://vk1od.net/antenna/SmallUntunedSquareLoop/Loop600.pdf ). 


Thank you; these are very interesting.
--
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK
Some people who are peeling the finch of the financial crisis are thinking of
biting a rook.

-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread Cortland Richmond
Yes, and in the real world that is what is done. But it has to meet the 
accuracy requirement of the test. For example, see MIL-STD461E 4.3.1  
(slightly snipped):


4.3.1  Measurement tolerances.
Unless otherwise stated for a particular measurement, the tolerance 
shall be as follows:

c. Amplitude, measurement receiver: ±2 dB
d. Amplitude, measurement system (includes measurement receivers, 
transducers, cables, and so forth): ±3 dB


If the receiver is already  ±2 dB we don't have much to work with.  
Calibration with X4 TUR might require an accuracy of 0.25 dB.   Might 
settle for less, no?


Sooner or later, someone will say their project passes by  0.5 dB;  SHIP IT.

Same old story.

Cortland
KA5S

On 12/7/2011 6:50 AM, Ken Javor wrote:

Now we get to the heart of the matter that prompted the original post.

That spike in the 137 cm tip-to-tip biconical used with a T1 dipole 
balun (28 -- 200 MHz) behavior might have been objectionable, but you 
wouldn't miss it an analog sweep with even 1% frequency steps.  You 
would only miss it if you calibrated every ten MHz as was done with 
the biconical antenna design in MIL-STD-461A (1968 -- when swept 
performance was difficult to do). There is no need to use frequency 
steps commensurate with the sorts of BWs used during EMI testing.  A 1 
- 5 MHz BW would more than suffice to sweep an antenna used above 30 
MHz to ensure not missing anything.


The following quote is the rationale I have seen a few places with 
which I take issue:


Though the sweep may be continuous, if you can read amplitude only 
by marking pixels, you need to insure .5 BW or less per pixel. That's 
straight sampling theory, right?


That is straight sampling theory applied to an unknown signal 
characteristic.  If a continuous sweep using 1/2 BW steps as per 
sampling theory has been performed, then discrete points for an 
antenna factor file may be chosen by inspection so that in any 
particular frequency range either of the two following conditions are met:


 1. If the data is noisy, with no obvious trend, then frequency points
should be listed so that the antenna factor between any two listed
points doesn't change by more than 1 dB (or whatever delta is
deemed necessary for adequate accuracy).
 2. If a linear or log-linear approximation can be made over some
frequency interval, then only the two endpoints of that interval
need be recorded, with the understanding that between those two
points the appropriate interpolation will yield the desired
accuracy. The interpolation has to be one which the automated
software can be instructed to perform on its recorded antenna
factor files.


One of these two approaches should work when deriving a set of 
discrete points from a continuous sweep of any test transducer 
factor's performance.
The question is whether this is cast in stone anywhere or left up to 
the person doing the calibration, or the test engineer abstracting a 
subset of data points for inclusion in his files.



Ken Javor

Phone: (256) 650-5261



**




-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread John Woodgate
In message 4edf7f60.9020...@earthlink.net, dated Wed, 7 Dec 2011, 
Cortland Richmond k...@earthlink.net writes:


Sooner or later, someone will say their project passes by  0.5 dB;  
SHIP IT.


It may be reprehensible, but it's realistic. The vast majority of cases 
of actual interference occur when the source is 10 dB or more stronger 
than the immunity of the victim can cope with. If that 'pass by 0.5 dB' 
is really a 'fail by 2.5 dB', the chances are no-one will ever know.

--
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK
Some people who are peeling the finch of the financial crisis are thinking of
biting a rook.

-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] FW: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread Ralph . McDiarmid
But how are those measurements made?  At what distance above the ground 
plane?   Surly the calibration setup should
match how the antenna is used during emissions measurement, otherwise 1dB 
here or there becomes meaningless.
_
 


Ralph McDiarmid  |   Schneider Electric   |  Renewable Energies Business | 
  CANADA  |   Regulatory Compliance Engineering




From:
ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
To:
EMC-PSTC@LISTSERV.IEEE.ORG
Date:
12/07/2011 12:59 AM
Subject:
[PSES] FW: [PSES] Calibration practice for EMI test transducers



 
Several EMC antennas have AF step performance of 1 dB per 5 MHz or so at 
certain frequencies.
 
If the antenna would have a linear behavior, 2 points would be enough
as all other frequencies could be linearly interpolated.
 
But my particular Chase BILOG 6111A has a bump at 14.1 dB at 258 MHz
and at both  250 MHz and 266 MHz  the antennafactor is 13.0 dB.
This is what I read from the calibration graph.
 
The manufacturer however, also has calibration (numeric) data provided as 
discrete points at
comfortable round numbers of 250 (12.9) and 300 MHz. (13.6)
These points are in general used for creating a interpolation curve.
 
Linear interpolation using the table provides a antennafactor of 13.1 dB 
at 258 MHz
which is a full -1 dB off from what I read from the graph. The bump has
virtually disappeared. Nice result with a 0.2 dB receiver.
 
That is why the output level change and required accuracy 
should be the driving factor in creating a calibration table.
It can easily be automated during a sweep.
 
What complicates this further is that not all manufacturers provide
data about their interpolation system. My RS receiver uses quadratic 
spline
(I suppose Lagrange).
 
My RS FSP only uses LIN interpolation (with log option) for the 
transducers.
 
 
 
 
 
Regards,

Ing.  Gert Gremmen, BSc
 
 
 
g.grem...@cetest.nl
www.cetest.nl

Kiotoweg 363
3047 BG Rotterdam
T 31(0)104152426
F 31(0)104154953
 
Before printing, think about the environment. 
 
 
Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Ken Javor
Verzonden: Tuesday, December 06, 2011 11:12 PM
Aan: EMC-PSTC@LISTSERV.IEEE.ORG
Onderwerp: Re: [PSES] Calibration practice for EMI test transducers
 
This discussion has gone far a field from the original post.  I’m assuming 
some sort of scan has been run and a continuous sweep is available. The 
question is, how densely does that sweep need to be digitized? Not as a 
practical matter, but as a matter of compliance with standards or standard 
practice.

What manner of antenna would have performance as cited below where at 100 
MHz the antenna factor is 10 dB and at 120 MHz it is 12 dB, but at 110 MHz 
it could be 20 dB? Let’s generalize the question to any closely spaced 
frequencies.

Let’s look at the types of antennas available.

30-200 MHz: A half-wave tuned dipole is nowhere near that sharp.  Neither 
is a biconical.

200 – 1000 MHz:  A half-wave tuned dipole is nowhere near that sharp. 
Neither is a logperiodic, log-spiral, nor a Yagi.

1 GHz+:  Logperiodic, log-spiral nor pyramidal horns act the way surmised.

My conclusion, and the point here is to invite discussion, not close it 
out, is that no test-type antenna is a high “Q” device.  Antennas can have 
arbitrarily high gains, depending on construction, but the high gain is a 
geometrical quality, not a high quality factor in the frequency domain. 
 
Ken Javor

Phone: (256) 650-5261

From: ralph.mcdiar...@schneider-electric.com
Date: Tue, 6 Dec 2011 11:25:43 -0800
To: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
Cc: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Calibration practice for EMI test transducers


Yet how far would you take that argument?  If it's say 11.1db at 110MHz 
and 11.5dB at 112MHz, would you still worry about 
factor at 111MHz?  At some point surly this all becomes rather silly.
_
 


Ralph McDiarmid  |  Schneider Electric  |  Renewable Energies Business  |  
CANADA  |   Regulatory Compliance Engineering



From: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl 

To: EMC-PSTC@LISTSERV.IEEE.ORG 
Date: 12/05/2011 10:06 PM 
Subject: Re: [PSES] Calibration practice for EMI test transducers 




If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate. 
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver

Re: [PSES] Calibration practice for EMI test transducers

2011-12-07 Thread Ralph . McDiarmid
Many times there seems a disconnect between what is required by a standard 
and what is needed for compatibility.
_
 


Ralph McDiarmid  |   Schneider Electric   |  Renewable Energies Business | 
  CANADA  |   Regulatory Compliance Engineering 




From:
John Woodgate j...@jmwa.demon.co.uk
To:
EMC-PSTC@LISTSERV.IEEE.ORG
Date:
12/07/2011 07:11 AM
Subject:
Re: [PSES] Calibration practice for EMI test transducers



In message 4edf7f60.9020...@earthlink.net, dated Wed, 7 Dec 2011, 
Cortland Richmond k...@earthlink.net writes:

Sooner or later, someone will say their project passes by  0.5 dB;  
SHIP IT.

It may be reprehensible, but it's realistic. The vast majority of cases 
of actual interference occur when the source is 10 dB or more stronger 
than the immunity of the victim can cope with. If that 'pass by 0.5 dB' 
is really a 'fail by 2.5 dB', the chances are no-one will ever know.
-- 
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK
Some people who are peeling the finch of the financial crisis are thinking 
of
biting a rook.

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in 
well-used formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com

__
This email has been scanned by the Symantec Email Security.cloud service.
__



-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-06 Thread Cortland Richmond

Hello Gert,

It was Ken Javor who was asking, but I will note that ETS-Lindgren probe 
documents specify an accuracy of +/- 2 dB transfer impedance, which for 
*formal* calibration requires as you say 0.5 dB accuracy. It also takes 
precision 50 Ohm test fixtures, better loads that I used, and a better 
instrument, too.  Call my process a sanity check, adequate only to get 
an ad-hoc setup back to readings that more closely approximate what was 
seen at outside test houses.  That's what the customer wanted.


However, any deviation from a smooth response may be detected fairly 
easily, and if I saw a jog in a current probe function I would 
recommend replacement. I did recommend the customer get their analyzer 
and probe calibrated.



Regards,

Cortland Richmond
KA5S

On 12/6/2011 1:03 AM, ce-test, qualified testing bv - Gert Gremmen wrote:

If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate.
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver is ± 0.5 dB you should register a
calibration
  each time the difference between previous and current sample is 0.5 dB.
Then the total error will be slightly more than 0.5 dB.
There are more sophisticated interpolation methods such as
cubic spline and polynomial  interpolation, and the error item
is subject to higher mathematics.
Play with it at:

http://terpconnect.umd.edu/~petersd/interp.html

or athttp://www.ibiblio.org/e-notes/Splines/Lagrange.htm

you can actually move the points on the latter page.

Gert Gremmen


-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-06 Thread Ralph . McDiarmid
Yet how far would you take that argument?  If it's say 11.1db at 110MHz 
and 11.5dB at 112MHz, would you still worry about 
factor at 111MHz?  At some point surly this all becomes rather silly.
_
 


Ralph McDiarmid  |   Schneider Electric   |  Renewable Energies Business | 
  CANADA  |   Regulatory Compliance Engineering




From:
ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
To:
EMC-PSTC@LISTSERV.IEEE.ORG
Date:
12/05/2011 10:06 PM
Subject:
Re: [PSES] Calibration practice for EMI test transducers



If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate. 
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver is +/- 0.5 dB you should register a
calibration
 each time the difference between previous and current sample is 0.5 dB.
Then the total error will be slightly more than 0.5 dB.
There are more sophisticated interpolation methods such as 
cubic spline and polynomial  interpolation, and the error item
is subject to higher mathematics.
Play with it at:

http://terpconnect.umd.edu/~petersd/interp.html

or at http://www.ibiblio.org/e-notes/Splines/Lagrange.htm

you can actually move the points on the latter page.

Gert Gremmen


-Oorspronkelijk bericht-
Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Cortland
Richmond
Verzonden: dinsdag 6 december 2011 3:01
Aan: emc-p...@ieee.org
Onderwerp: RE: Calibration practice for EMI test transducers

I recently had occasion to verify on an 8591EM generic factors for an 
EMCO current clamp during pretest, and used its tracking generator to 
see if with a short wire the result was a straight line on the screen. 
That simply called for paying attention to 1/BW.


Cortland
KA5S

On 12/5/2011 5:33 PM, Ken Javor wrote:
If an antenna, current probe or other transducer is going to be
calibrated
over its frequency range of operation, what determines the step sizes
between measured frequencies, or if swept, what determines the density
of
test points reported?

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site
at http://product-compliance.oc.ieee.org/ can be used for graphics (in
well-used formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in 
well-used formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com

__
This email has been scanned by the Symantec Email Security.cloud service.
__



-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used

Re: [PSES] Calibration practice for EMI test transducers

2011-12-06 Thread Ken Javor
This discussion has gone far a field from the original post.  I¹m assuming
some sort of scan has been run and a continuous sweep is available. The
question is, how densely does that sweep need to be digitized? Not as a
practical matter, but as a matter of compliance with standards or standard
practice.

What manner of antenna would have performance as cited below where at 100
MHz the antenna factor is 10 dB and at 120 MHz it is 12 dB, but at 110 MHz
it could be 20 dB? Let¹s generalize the question to any closely spaced
frequencies.

Let¹s look at the types of antennas available.

30-200 MHz: A half-wave tuned dipole is nowhere near that sharp.  Neither is
a biconical.

200 ­ 1000 MHz:  A half-wave tuned dipole is nowhere near that sharp.
Neither is a logperiodic, log-spiral, nor a Yagi.

1 GHz+:  Logperiodic, log-spiral nor pyramidal horns act the way surmised.

My conclusion, and the point here is to invite discussion, not close it out,
is that no test-type antenna is a high ³Q² device.  Antennas can have
arbitrarily high gains, depending on construction, but the high gain is a
geometrical quality, not a high quality factor in the frequency domain.
 
Ken Javor

Phone: (256) 650-5261



From: ralph.mcdiar...@schneider-electric.com
Date: Tue, 6 Dec 2011 11:25:43 -0800
To: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
Cc: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Calibration practice for EMI test transducers


Yet how far would you take that argument?  If it's say 11.1db at 110MHz and
11.5dB at 112MHz, would you still worry about
factor at 111MHz?  At some point surly this all becomes rather silly.

_ 

Ralph McDiarmid  |  Schneider Electric   |  Renewable Energies Business  |
CANADA  |   Regulatory Compliance Engineering



From: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
To: EMC-PSTC@LISTSERV.IEEE.ORG
Date: 12/05/2011 10:06 PM
Subject: Re: [PSES] Calibration practice for EMI test transducers




If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate.
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver is +/- 0.5 dB you should register a
calibration
 each time the difference between previous and current sample is 0.5 dB.
Then the total error will be slightly more than 0.5 dB.
There are more sophisticated interpolation methods such as
cubic spline and polynomial  interpolation, and the error item
is subject to higher mathematics.
Play with it at:

http://terpconnect.umd.edu/~petersd/interp.html
http://terpconnect.umd.edu/~petersd/interp.html

or at http://www.ibiblio.org/e-notes/Splines/Lagrange.htm
http://www.ibiblio.org/e-notes/Splines/Lagrange.htm

you can actually move the points on the latter page.

Gert Gremmen


-Oorspronkelijk bericht-
Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org mailto:emc-p...@ieee.org
] Namens Cortland
Richmond
Verzonden: dinsdag 6 december 2011 3:01
Aan: emc-p...@ieee.org
Onderwerp: RE: Calibration practice for EMI test transducers

I recently had occasion to verify on an 8591EM generic factors for an
EMCO current clamp during pretest, and used its tracking generator to
see if with a short wire the result was a straight line on the screen.
That simply called for paying attention to 1/BW.


Cortland
KA5S

On 12/5/2011 5:33 PM, Ken Javor wrote:
If an antenna, current probe or other transducer is going to be
calibrated
over its frequency range of operation, what determines the step sizes
between measured frequencies, or if swept, what determines the density
of
test points reported?

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site
at http://product-compliance.oc.ieee.org/
http://product-compliance.oc.ieee.org/ can be used for graphics (in
well-used formats), large files, etc.

Website:  http://www.ieee-pses.org/ http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html
http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott

Re: [PSES] Calibration practice for EMI test transducers

2011-12-06 Thread Bill Owsley
Original post answer from here - COST.




 From: Ken Javor ken.ja...@emccompliance.com
To: EMC-PSTC@LISTSERV.IEEE.ORG 
Sent: Tuesday, December 6, 2011 5:11 PM
Subject: Re: [PSES] Calibration practice for EMI test transducers
 

Re: [PSES] Calibration practice for EMI test transducers 
This discussion has gone far a field from the original post.  I’m assuming some 
sort of scan has been run and a continuous sweep is available. The question is, 
how densely does that sweep need to be digitized? Not as a practical matter, 
but as a matter of compliance with standards or standard practice.

What manner of antenna would have performance as cited below where at 100 MHz 
the antenna factor is 10 dB and at 120 MHz it is 12 dB, but at 110 MHz it could 
be 20 dB? Let’s generalize the question to any closely spaced frequencies.

Let’s look at the types of antennas available.

30-200 MHz: A half-wave tuned dipole is nowhere near that sharp.  Neither is a 
biconical.

200 – 1000 MHz:  A half-wave tuned dipole is nowhere near that sharp.  Neither 
is a logperiodic, log-spiral, nor a Yagi.

1 GHz+:  Logperiodic, log-spiral nor pyramidal horns act the way surmised.

My conclusion, and the point here is to invite discussion, not close it out, is 
that no test-type antenna is a high “Q” device.  Antennas can have arbitrarily 
high gains, depending on construction, but the high gain is a geometrical 
quality, not a high quality factor in the frequency domain. 
 
Ken Javor

Phone: (256) 650-5261



From: ralph.mcdiar...@schneider-electric.com
Date: Tue, 6 Dec 2011 11:25:43 -0800
To: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
Cc: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Calibration practice for EMI test transducers


Yet how far would you take that argument?  If it's say 11.1db at 110MHz and 
11.5dB at 112MHz, would you still worry about 
factor at 111MHz?  At some point surly this all becomes rather silly.
_

Ralph McDiarmid |  Schneider Electric  |  Renewable Energies Business |   
CANADA  |   Regulatory Compliance Engineering



From:ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
To:EMC-PSTC@LISTSERV.IEEE.ORG
Date:12/05/2011 10:06 PM
Subject:Re: [PSES] Calibration practice for EMI test transducers




If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate. 
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver is +/- 0.5 dB you should register a
calibration
 each time the difference between previous and current sample is 0.5 dB.
Then the total error will be slightly more than 0.5 dB.
There are more sophisticated interpolation methods such as 
cubic spline and polynomial  interpolation, and the error item
is subject to higher mathematics.
Play with it at:

http://terpconnect.umd.edu/~petersd/interp.htmlhttp://terpconnect.umd.edu/~petersd/interp.html
 

or at 
http://www.ibiblio.org/e-notes/Splines/Lagrange.htmhttp://www.ibiblio.org/e-notes/Splines/Lagrange.htm
 

you can actually move the points on the latter page.

Gert Gremmen


-Oorspronkelijk bericht-
Van: emc-p...@ieee.org [mailto:emc-p...@ieee.orgmailto:emc-p...@ieee.org ] 
Namens Cortland
Richmond
Verzonden: dinsdag 6 december 2011 3:01
Aan: emc-p...@ieee.org
Onderwerp: RE: Calibration practice for EMI test transducers

I recently had occasion to verify on an 8591EM generic factors for an 
EMCO current clamp during pretest, and used its tracking generator to 
see if with a short wire the result was a straight line on the screen. 
That simply called for paying attention to 1/BW.


Cortland
KA5S

On 12/5/2011 5:33 PM, Ken Javor wrote:
If an antenna, current probe or other transducer is going to be
calibrated
over its frequency range of operation, what determines the step sizes
between measured frequencies, or if swept, what determines the density
of
test points reported?

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.htmlhttp://www.ieee-pses.org/emc-pstc.html 

Attachments are not permitted but the IEEE PSES Online Communities site
at 
http://product-compliance.oc.ieee.org/http://product

Re: [PSES] Calibration practice for EMI test transducers

2011-12-06 Thread Cortland Richmond
We might *like* measuring antennas to be low-Q, but remember the bicon 
shorting bars that were added some years ago? The skeleton bicons 
lacking them inserted a rather sharp notch that had to be accounted for 
in transducer tables. Baluns might go off from impact damage, too.  And 
so on.  I gave an ebay 50 MHz comb good past 2 GHz to a former employer 
who later gave all their own EMC lab gear to an NRTL.


But you still have to check.

For that matter, I've seen coax factors programmed with GAIN at a narrow 
range at microwave frequencies; that was caused by reflection from a 
crimped cable*. So we do need to be able to spot them.


*This was due to the ignorance of operators collecting the data, IMO. I 
spoke to them and likely, they won't do it again.  But they believed the 
SA/TG reading because they didn't know what they were measuring. That's 
another thread.


Though the sweep may be continuous, if you can read amplitude only by 
marking pixels, you need to insure .5 BW or less per pixel. That's 
straight sampling theory, right?  If your analyzer has a 1024 pixel wide 
screen you can't SEE less than about .1% of the scan width and can rely 
on less. Does it matter? Often it doesn't.


Fun, isn't it?

Cortland
KA5S

On 12/6/2011 5:11 PM, Ken Javor wrote:
This discussion has gone far a field from the original post.  I'm 
assuming some sort of scan has been run and a continuous sweep is 
available. The question is, how densely does that sweep need to be 
digitized? Not as a practical matter, but as a matter of compliance 
with standards or standard practice.


What manner of antenna would have performance as cited below where at 
100 MHz the antenna factor is 10 dB and at 120 MHz it is 12 dB, but at 
110 MHz it could be 20 dB? Let's generalize the question to any 
closely spaced frequencies.


Let's look at the types of antennas available.

30-200 MHz: A half-wave tuned dipole is nowhere near that sharp. 
 Neither is a biconical.


200 -- 1000 MHz:  A half-wave tuned dipole is nowhere near that sharp. 
 Neither is a logperiodic, log-spiral, nor a Yagi.


1 GHz+:  Logperiodic, log-spiral nor pyramidal horns act the way surmised.

My conclusion, and the point here is to invite discussion, not close 
it out, is that no test-type antenna is a high Q device.  Antennas 
can have arbitrarily high gains, depending on construction, but the 
high gain is a geometrical quality, not a high quality factor in the 
frequency domain.





-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-06 Thread John Woodgate
In message cb03ef45.c9421%ken.ja...@emccompliance.com, dated Tue, 6 
Dec 2011, Ken Javor ken.ja...@emccompliance.com writes:


 Antennas can have arbitrarily high gains, depending on construction, 
but the high gain is a geometrical quality, not a high quality factor 
in the frequency domain.


True for electric antennas, I think, but not for magnetic. Tuned loops 
and ferrite-rod antennas can have high Q. CISPR 16 includes at least one 
magnetic antenna, although it may well not be high Q.

--
OOO - Own Opinions Only. Try www.jmwa.demon.co.uk and www.isce.org.uk
John Woodgate, J M Woodgate and Associates, Rayleigh, Essex UK
Some people who are peeling the finch of the financial crisis are thinking of
biting a rook.

-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-06 Thread Ralph . McDiarmid
I sure hope that standard practice and practicality (pardon me)  go 
hand-in-hand.  If not, the committees members 
need to give their head a shake.
_
 


Ralph McDiarmid  |   Schneider Electric   |  Renewable Energies Business | 
  CANADA  |   Regulatory Compliance Engineering 




From:
Ken Javor ken.ja...@emccompliance.com
To:
EMC-PSTC@LISTSERV.IEEE.ORG
Date:
12/06/2011 02:14 PM
Subject:
Re: [PSES] Calibration practice for EMI test transducers



This discussion has gone far a field from the original post.  I’m assuming 
some sort of scan has been run and a continuous sweep is available. The 
question is, how densely does that sweep need to be digitized? Not as a 
practical matter, but as a matter of compliance with standards or standard 
practice.

What manner of antenna would have performance as cited below where at 100 
MHz the antenna factor is 10 dB and at 120 MHz it is 12 dB, but at 110 MHz 
it could be 20 dB? Let’s generalize the question to any closely spaced 
frequencies.

Let’s look at the types of antennas available.

30-200 MHz: A half-wave tuned dipole is nowhere near that sharp.  Neither 
is a biconical.

200 – 1000 MHz:  A half-wave tuned dipole is nowhere near that sharp. 
Neither is a logperiodic, log-spiral, nor a Yagi.

1 GHz+:  Logperiodic, log-spiral nor pyramidal horns act the way surmised.

My conclusion, and the point here is to invite discussion, not close it 
out, is that no test-type antenna is a high “Q” device.  Antennas can have 
arbitrarily high gains, depending on construction, but the high gain is a 
geometrical quality, not a high quality factor in the frequency domain. 
 
Ken Javor

Phone: (256) 650-5261


From: ralph.mcdiar...@schneider-electric.com
Date: Tue, 6 Dec 2011 11:25:43 -0800
To: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl
Cc: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Calibration practice for EMI test transducers


Yet how far would you take that argument?  If it's say 11.1db at 110MHz 
and 11.5dB at 112MHz, would you still worry about 
factor at 111MHz?  At some point surly this all becomes rather silly.
_
 


Ralph McDiarmid  |  Schneider Electric  |  Renewable Energies Business  |  
CANADA  |   Regulatory Compliance Engineering



From: ce-test, qualified testing bv - Gert Gremmen g.grem...@cetest.nl 

To: EMC-PSTC@LISTSERV.IEEE.ORG 
Date: 12/05/2011 10:06 PM 
Subject: Re: [PSES] Calibration practice for EMI test transducers 



If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate. 
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver is +/- 0.5 dB you should register a
calibration
 each time the difference between previous and current sample is 0.5 dB.
Then the total error will be slightly more than 0.5 dB.
There are more sophisticated interpolation methods such as 
cubic spline and polynomial  interpolation, and the error item
is subject to higher mathematics.
Play with it at:

http://terpconnect.umd.edu/~petersd/interp.html 
http://terpconnect.umd.edu/~petersd/interp.html 

or at http://www.ibiblio.org/e-notes/Splines/Lagrange.htm 
http://www.ibiblio.org/e-notes/Splines/Lagrange.htm 

you can actually move the points on the latter page.

Gert Gremmen


-Oorspronkelijk bericht-
Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org 
mailto:emc-p...@ieee.org ] Namens Cortland
Richmond
Verzonden: dinsdag 6 december 2011 3:01
Aan: emc-p...@ieee.org
Onderwerp: RE: Calibration practice for EMI test transducers

I recently had occasion to verify on an 8591EM generic factors for an 
EMCO current clamp during pretest, and used its tracking generator to 
see if with a short wire the result was a straight line on the screen. 
That simply called for paying attention to 1/BW.


Cortland
KA5S

On 12/5/2011 5:33 PM, Ken Javor wrote:
If an antenna, current probe or other transducer is going to be
calibrated
over its frequency range of operation, what determines the step sizes
between measured frequencies, or if swept, what determines the density
of
test points reported?

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html 
http

[PSES] Calibration practice for EMI test transducers

2011-12-05 Thread Ken Javor
If an antenna, current probe or other transducer is going to be calibrated
over its frequency range of operation, what determines the step sizes
between measured frequencies, or if swept, what determines the density of
test points reported?

Thank you,
 
Ken Javor

Phone: (256) 650-5261

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-05 Thread Cortland Richmond
I recently had occasion to verify on an 8591EM generic factors for an 
EMCO current clamp during pretest, and used its tracking generator to 
see if with a short wire the result was a straight line on the screen. 
That simply called for paying attention to 1/BW.



Cortland
KA5S

On 12/5/2011 5:33 PM, Ken Javor wrote:
If an antenna, current probe or other transducer is going to be calibrated
over its frequency range of operation, what determines the step sizes
between measured frequencies, or if swept, what determines the density of
test points reported?

-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com


Re: [PSES] Calibration practice for EMI test transducers

2011-12-05 Thread ce-test, qualified testing bv - Gert Gremmen
If at 100 MHz your sensor indicates during calibration 10 dB
and at 120 MHz 12 dB, how are you going to know the value at
110 MHz. It maight be 20 as well as 0 dB, you don't not know as you did
not measure/ calibrate. 
This requires some knowledge about the behavior of the sensor,
and the 1/BW is and indication of that.
There is another aspect in EMI measurements.

As the measuring receiver interpolates
between the calibrated samples, the max interpolation error
(lin interpolation) is 50% of the vertical differences between
calibration points.

If the measuring receiver is +/- 0.5 dB you should register a
calibration
 each time the difference between previous and current sample is 0.5 dB.
Then the total error will be slightly more than 0.5 dB.
There are more sophisticated interpolation methods such as 
cubic spline and polynomial  interpolation, and the error item
is subject to higher mathematics.
Play with it at:

http://terpconnect.umd.edu/~petersd/interp.html

or at http://www.ibiblio.org/e-notes/Splines/Lagrange.htm

you can actually move the points on the latter page.

Gert Gremmen


-Oorspronkelijk bericht-
Van: emc-p...@ieee.org [mailto:emc-p...@ieee.org] Namens Cortland
Richmond
Verzonden: dinsdag 6 december 2011 3:01
Aan: emc-p...@ieee.org
Onderwerp: RE: Calibration practice for EMI test transducers

I recently had occasion to verify on an 8591EM generic factors for an 
EMCO current clamp during pretest, and used its tracking generator to 
see if with a short wire the result was a straight line on the screen. 
That simply called for paying attention to 1/BW.


Cortland
KA5S

On 12/5/2011 5:33 PM, Ken Javor wrote:
If an antenna, current probe or other transducer is going to be
calibrated
over its frequency range of operation, what determines the step sizes
between measured frequencies, or if swept, what determines the density
of
test points reported?

-

This message is from the IEEE Product Safety Engineering Society
emc-pstc discussion list. To post a message to the list, send your
e-mail to emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site
at http://product-compliance.oc.ieee.org/ can be used for graphics (in
well-used formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 
emc-p...@ieee.org

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas emcp...@radiusnorth.net
Mike Cantwell mcantw...@ieee.org

For policy questions, send mail to:
Jim Bacher:  j.bac...@ieee.org
David Heald: dhe...@gmail.com