Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-21 Thread Elliott Mac-FME001
Hi Derek!

Thanks for the input / feedback - will look into the alternate way of 
demonstrating compliance to the traceability requirements. 

Here is the excerpt from TIA 603 D

1.4.5 Standard Atmospheric Conditions
The standard atmospheric condition is a temperature of 25 °C at an atmospheric
pressure of 1013 hPa (1013 mbar). Measurements, however, may be carried out at
any combination of temperature, pressure, and relative humidity within the
following limits:

Temperature: 20 °C to 35 °C
Relative Humidity: 45% to 75%
Atmospheric Pressure: 860 hPa to 1060 hPa (860 mbar to 1060 mbar)

We are accredited to this standard by A2LA. 

Think the main problem with the weather station seems to be the rain gage and a 
few of the other measurement capabilities of the device. 

Think can live with limited cal to accredited parameters and find a good lab 
out there. 

BTW - I get the "lazy" way but my experience has been that defining what it is 
that you need and developing procurement requirements that are followed by the 
accredited calibration labs can be somewhat of a challenge! Can't just say you 
want "accredited cal". I have learned a lot about metrology in the process!

Good to hear from you and I hope that you and yours have a great Holiday Season!

Best regards,
 
Mac Elliott
 
 [] General Public  
-Original Message-
From: Derek Walton [mailto:lfresea...@aol.com] 
Sent: Monday, December 19, 2011 6:39 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis 
Weather Station

Hi Mac,

NVLAP requires that you follow Annex B of NIST HB-150. An accredited 
calibration is not the only way, just the "lazy" way.

Are you sure you need to record these parameters? If your standard has 
no limits, then quite possibly you could argue out of recording them. 
I'm not a fan of doing things "just cos'"

Cheers,

Derek.

On 12/19/2011 12:39 PM, Pettit, Ghery wrote:
> Agreed that the instrument needs to be calibrated, but the degree of accuracy 
> doesn't need to be NWS grade (IMHO).  What you are demonstrating is that the 
> environmental conditions are within the range specified for the test.  And if 
> you have enough control you shouldn't be pushing the edges, so it become less 
> critical.  Just don't try ESD out in the open on a hot, humid day in the 
> south.  Nor try it in the foothills above Denver at any time.  The 
> atmospheric pressure will be too low.  Remember, the barometer must be 
> calibrated to read "absolute" pressure, not pressure corrected to sea level 
> like the weather guessers will give you.
>
> Ghery S. Pettit
>
>
> -Original Message-
> From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Brian Oconnell
> Sent: Monday, December 19, 2011 10:00 AM
> To: EMC-PSTC@LISTSERV.IEEE.ORG
> Subject: RE: Lab Needed for 17025 Accredited Calibration for Davis Weather 
> Station
>
> Transcat will cal stuff that does these parameters and does 17025. RH, temp,
> and atm pressure xdcrs are common stuff. What is wrong with these cal-lab
> people?
>
> Possible to remove transducers/sensors and send to cal lab to do tables?
> Most Davis systems are not NWS-grade, and the rated accuracies could allow a
> company verification process where altitude and temp correction is applied
> to local METAR data, temp sensor is verified with your recently calibrated
> instrument used in Type Tests, and the RH sensor can be verified using the
> chemical kits or calc using dry/wet bulb readings.
>
> Do not have a weather station at place of employment, but have several
> similar sensors connected to data logger to monitor lab conditions during
> all tests, and have never had any problem submitting this data to auditors.
>
> Brian
>
> -Original Message-
>> On 12/19/11, Elliott Mac-FME001  wrote:
>>> All
>>> Our cal guys are telling me that they are having a hard time finding a
>>> lab that can do 17025 accredited calibrations for our Davis Weather
>>> Monitor that we use on our OATS to measure temp / humidity / and
>>> barometric pressure during our testing.
>>>
>>> Any suggestions?
>>>
>>> Best regards,
>>> Mac Elliott
> -
> 
> This message is from the IEEE Product Safety Engineering Society emc-pstc 
> discussion list. To post a message to the list, send your e-mail 
> to
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities site at 
> http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
> formats), lar

Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread Derek Walton

Hi Mac,

NVLAP requires that you follow Annex B of NIST HB-150. An accredited 
calibration is not the only way, just the "lazy" way.


Are you sure you need to record these parameters? If your standard has 
no limits, then quite possibly you could argue out of recording them. 
I'm not a fan of doing things "just cos'"


Cheers,

Derek.

On 12/19/2011 12:39 PM, Pettit, Ghery wrote:

Agreed that the instrument needs to be calibrated, but the degree of accuracy doesn't 
need to be NWS grade (IMHO).  What you are demonstrating is that the environmental 
conditions are within the range specified for the test.  And if you have enough control 
you shouldn't be pushing the edges, so it become less critical.  Just don't try ESD out 
in the open on a hot, humid day in the south.  Nor try it in the foothills above Denver 
at any time.  The atmospheric pressure will be too low.  Remember, the barometer must be 
calibrated to read "absolute" pressure, not pressure corrected to sea level 
like the weather guessers will give you.

Ghery S. Pettit


-Original Message-
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Brian Oconnell
Sent: Monday, December 19, 2011 10:00 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Lab Needed for 17025 Accredited Calibration for Davis Weather 
Station

Transcat will cal stuff that does these parameters and does 17025. RH, temp,
and atm pressure xdcrs are common stuff. What is wrong with these cal-lab
people?

Possible to remove transducers/sensors and send to cal lab to do tables?
Most Davis systems are not NWS-grade, and the rated accuracies could allow a
company verification process where altitude and temp correction is applied
to local METAR data, temp sensor is verified with your recently calibrated
instrument used in Type Tests, and the RH sensor can be verified using the
chemical kits or calc using dry/wet bulb readings.

Do not have a weather station at place of employment, but have several
similar sensors connected to data logger to monitor lab conditions during
all tests, and have never had any problem submitting this data to auditors.

Brian

-Original Message-

On 12/19/11, Elliott Mac-FME001  wrote:

All
Our cal guys are telling me that they are having a hard time finding a
lab that can do 17025 accredited calibrations for our Davis Weather
Monitor that we use on our OATS to measure temp / humidity / and
barometric pressure during our testing.

Any suggestions?

Best regards,
Mac Elliott

-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas
Mike Cantwell

For policy questions, send mail to:
Jim Bacher:
David Heald:

-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas
Mike Cantwell

For policy questions, send mail to:
Jim Bacher:
David Heald:



-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to 

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread Pettit, Ghery
Agreed that the instrument needs to be calibrated, but the degree of accuracy 
doesn't need to be NWS grade (IMHO).  What you are demonstrating is that the 
environmental conditions are within the range specified for the test.  And if 
you have enough control you shouldn't be pushing the edges, so it become less 
critical.  Just don't try ESD out in the open on a hot, humid day in the south. 
 Nor try it in the foothills above Denver at any time.  The atmospheric 
pressure will be too low.  Remember, the barometer must be calibrated to read 
"absolute" pressure, not pressure corrected to sea level like the weather 
guessers will give you.

Ghery S. Pettit


-Original Message-
From: emc-p...@ieee.org [mailto:emc-p...@ieee.org] On Behalf Of Brian Oconnell
Sent: Monday, December 19, 2011 10:00 AM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: RE: Lab Needed for 17025 Accredited Calibration for Davis Weather 
Station

Transcat will cal stuff that does these parameters and does 17025. RH, temp,
and atm pressure xdcrs are common stuff. What is wrong with these cal-lab
people?

Possible to remove transducers/sensors and send to cal lab to do tables?
Most Davis systems are not NWS-grade, and the rated accuracies could allow a
company verification process where altitude and temp correction is applied
to local METAR data, temp sensor is verified with your recently calibrated
instrument used in Type Tests, and the RH sensor can be verified using the
chemical kits or calc using dry/wet bulb readings.

Do not have a weather station at place of employment, but have several
similar sensors connected to data logger to monitor lab conditions during
all tests, and have never had any problem submitting this data to auditors. 

Brian

-Original Message-
>
> On 12/19/11, Elliott Mac-FME001  wrote:
>> All
>> Our cal guys are telling me that they are having a hard time finding a
>> lab that can do 17025 accredited calibrations for our Davis Weather
>> Monitor that we use on our OATS to measure temp / humidity / and
>> barometric pressure during our testing.
>>
>> Any suggestions?
>>
>> Best regards,
>> Mac Elliott

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread Brian Oconnell
Transcat will cal stuff that does these parameters and does 17025. RH, temp,
and atm pressure xdcrs are common stuff. What is wrong with these cal-lab
people?

Possible to remove transducers/sensors and send to cal lab to do tables?
Most Davis systems are not NWS-grade, and the rated accuracies could allow a
company verification process where altitude and temp correction is applied
to local METAR data, temp sensor is verified with your recently calibrated
instrument used in Type Tests, and the RH sensor can be verified using the
chemical kits or calc using dry/wet bulb readings.

Do not have a weather station at place of employment, but have several
similar sensors connected to data logger to monitor lab conditions during
all tests, and have never had any problem submitting this data to auditors. 

Brian

-Original Message-
>
> On 12/19/11, Elliott Mac-FME001  wrote:
>> All
>> Our cal guys are telling me that they are having a hard time finding a
>> lab that can do 17025 accredited calibrations for our Davis Weather
>> Monitor that we use on our OATS to measure temp / humidity / and
>> barometric pressure during our testing.
>>
>> Any suggestions?
>>
>> Best regards,
>> Mac Elliott

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread Elliott Mac-FME001
Hi Ken

I didn't but do now - few people have directed us there

Thanks and thanks to all who have responded

This list is a great tool!

Best regards,
 
Mac Elliott
 
 [] General Public  

-Original Message-
From: IBM Ken [mailto:ibm...@gmail.com] 
Sent: Monday, December 19, 2011 11:16 AM
To: Elliott Mac-FME001
Cc: EMC-PSTC@listserv.ieee.org
Subject: Re: Lab Needed for 17025 Accredited Calibration for Davis
Weather Station

Ahh; if you were in the Northeast I had a couple of suggestions.  The
only place I've dealt with getting ISO 17025 accredited calibration in
Florida is Voltech (Fort Myers), but they probably only do their own
equipment.

I assume you know about Simco?
http://www.simco.com/locations/us_east/miami_fl/

-Ken A.
IBM

On 12/19/11, Elliott Mac-FME001  wrote:
> Hi Ken
>
> We are close to Fort Lauderdale, Florida
>
> Best regards,
>
> Mac Elliott
>
> [] General Public
>
> -Original Message-
> From: IBM Ken [mailto:ibm...@gmail.com]
> Sent: Monday, December 19, 2011 10:32 AM
> To: Elliott Mac-FME001
> Cc: EMC-PSTC@listserv.ieee.org
> Subject: Re: Lab Needed for 17025 Accredited Calibration for Davis
> Weather Station
>
> Where are you located?
>
>
> On 12/19/11, Elliott Mac-FME001  wrote:
>> All
>>
>>
>>
>> Our cal guys are telling me that they are having a hard time finding
a
>> lab that can do 17025 accredited calibrations for our Davis Weather
>> Monitor that we use on our OATS to measure temp / humidity / and
>> barometric pressure during our testing.
>>
>>
>>
>> Any suggestions?
>>
>>
>>
>> Best regards,
>>
>>
>>
>> Mac Elliott
>>
>>
>> [] General Public
>>
>>
>> -
>> 
>> This message is from the IEEE Product Safety Engineering Society
> emc-pstc
>> discussion list. To post a message to the list, send your e-mail to
>> 
>>
>> All emc-pstc postings are archived and searchable on the web at:
>> http://www.ieee-pses.org/emc-pstc.html
>>
>> Attachments are not permitted but the IEEE PSES Online Communities
> site at
>> http://product-compliance.oc.ieee.org/ can be used for graphics (in
>> well-used formats), large files, etc.
>>
>> Website:  http://www.ieee-pses.org/
>> Instructions:  http://listserv.ieee.org/request/user-guide.html
>> List rules: http://www.ieee-pses.org/listrules.html
>>
>> For help, send mail to the list administrators:
>> Scott Douglas 
>> Mike Cantwell 
>>
>> For policy questions, send mail to:
>> Jim Bacher:  
>> David Heald: 
>>
>
>
>

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread IBM Ken
Ahh; if you were in the Northeast I had a couple of suggestions.  The
only place I've dealt with getting ISO 17025 accredited calibration in
Florida is Voltech (Fort Myers), but they probably only do their own
equipment.

I assume you know about Simco?  http://www.simco.com/locations/us_east/miami_fl/

-Ken A.
IBM

On 12/19/11, Elliott Mac-FME001  wrote:
> Hi Ken
>
> We are close to Fort Lauderdale, Florida
>
> Best regards,
>
> Mac Elliott
>
> [] General Public
>
> -Original Message-
> From: IBM Ken [mailto:ibm...@gmail.com]
> Sent: Monday, December 19, 2011 10:32 AM
> To: Elliott Mac-FME001
> Cc: EMC-PSTC@listserv.ieee.org
> Subject: Re: Lab Needed for 17025 Accredited Calibration for Davis
> Weather Station
>
> Where are you located?
>
>
> On 12/19/11, Elliott Mac-FME001  wrote:
>> All
>>
>>
>>
>> Our cal guys are telling me that they are having a hard time finding a
>> lab that can do 17025 accredited calibrations for our Davis Weather
>> Monitor that we use on our OATS to measure temp / humidity / and
>> barometric pressure during our testing.
>>
>>
>>
>> Any suggestions?
>>
>>
>>
>> Best regards,
>>
>>
>>
>> Mac Elliott
>>
>>
>> [] General Public
>>
>>
>> -
>> 
>> This message is from the IEEE Product Safety Engineering Society
> emc-pstc
>> discussion list. To post a message to the list, send your e-mail to
>> 
>>
>> All emc-pstc postings are archived and searchable on the web at:
>> http://www.ieee-pses.org/emc-pstc.html
>>
>> Attachments are not permitted but the IEEE PSES Online Communities
> site at
>> http://product-compliance.oc.ieee.org/ can be used for graphics (in
>> well-used formats), large files, etc.
>>
>> Website:  http://www.ieee-pses.org/
>> Instructions:  http://listserv.ieee.org/request/user-guide.html
>> List rules: http://www.ieee-pses.org/listrules.html
>>
>> For help, send mail to the list administrators:
>> Scott Douglas 
>> Mike Cantwell 
>>
>> For policy questions, send mail to:
>> Jim Bacher:  
>> David Heald: 
>>
>
>
>

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread Elliott Mac-FME001
Hi Ken

We are close to Fort Lauderdale, Florida

Best regards,
 
Mac Elliott
 
[] General Public  

-Original Message-
From: IBM Ken [mailto:ibm...@gmail.com] 
Sent: Monday, December 19, 2011 10:32 AM
To: Elliott Mac-FME001
Cc: EMC-PSTC@listserv.ieee.org
Subject: Re: Lab Needed for 17025 Accredited Calibration for Davis
Weather Station

Where are you located?


On 12/19/11, Elliott Mac-FME001  wrote:
> All
>
>
>
> Our cal guys are telling me that they are having a hard time finding a
> lab that can do 17025 accredited calibrations for our Davis Weather
> Monitor that we use on our OATS to measure temp / humidity / and
> barometric pressure during our testing.
>
>
>
> Any suggestions?
>
>
>
> Best regards,
>
>
>
> Mac Elliott
>
>
> [] General Public
>
>
> -
> 
> This message is from the IEEE Product Safety Engineering Society
emc-pstc
> discussion list. To post a message to the list, send your e-mail to
> 
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities
site at
> http://product-compliance.oc.ieee.org/ can be used for graphics (in
> well-used formats), large files, etc.
>
> Website:  http://www.ieee-pses.org/
> Instructions:  http://listserv.ieee.org/request/user-guide.html
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
> Scott Douglas 
> Mike Cantwell 
>
> For policy questions, send mail to:
> Jim Bacher:  
> David Heald: 
>

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread IBM Ken
Where are you located?


On 12/19/11, Elliott Mac-FME001  wrote:
> All
>
>
>
> Our cal guys are telling me that they are having a hard time finding a
> lab that can do 17025 accredited calibrations for our Davis Weather
> Monitor that we use on our OATS to measure temp / humidity / and
> barometric pressure during our testing.
>
>
>
> Any suggestions?
>
>
>
> Best regards,
>
>
>
> Mac Elliott
>
>
> [] General Public
>
>
> -
> 
> This message is from the IEEE Product Safety Engineering Society emc-pstc
> discussion list. To post a message to the list, send your e-mail to
> 
>
> All emc-pstc postings are archived and searchable on the web at:
> http://www.ieee-pses.org/emc-pstc.html
>
> Attachments are not permitted but the IEEE PSES Online Communities site at
> http://product-compliance.oc.ieee.org/ can be used for graphics (in
> well-used formats), large files, etc.
>
> Website:  http://www.ieee-pses.org/
> Instructions:  http://listserv.ieee.org/request/user-guide.html
> List rules: http://www.ieee-pses.org/listrules.html
>
> For help, send mail to the list administrators:
> Scott Douglas 
> Mike Cantwell 
>
> For policy questions, send mail to:
> Jim Bacher:  
> David Heald: 
>

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


Re: [PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread Cortland Richmond
I know that Cole-Parmer offers cal services (at a substantial added 
cost) for its own instruments sold to the public. It stands to reason 
that Davis might also.


Cortland Richmond
KA5S

On 12/19/2011 9:09 AM, Elliott Mac-FME001 wrote:


All

Our cal guys are telling me that they are having a hard time finding a 
lab that can do 17025 accredited calibrations for our Davis Weather 
Monitor that we use on our OATS to measure temp / humidity / and 
barometric pressure during our testing.


Any suggestions?



-

This message is from the IEEE Product Safety Engineering Society emc-pstc discussion 
list. To post a message to the list, send your e-mail to 

All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 


[PSES] Lab Needed for 17025 Accredited Calibration for Davis Weather Station

2011-12-19 Thread Elliott Mac-FME001
All

 

Our cal guys are telling me that they are having a hard time finding a
lab that can do 17025 accredited calibrations for our Davis Weather
Monitor that we use on our OATS to measure temp / humidity / and
barometric pressure during our testing. 

 

Any suggestions? 

 

Best regards,

 

Mac Elliott


[] General Public  


-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieee-pses.org/emc-pstc.html

Attachments are not permitted but the IEEE PSES Online Communities site at 
http://product-compliance.oc.ieee.org/ can be used for graphics (in well-used 
formats), large files, etc.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: