RE: D of C for medical P.S.

2003-02-28 Thread boconn...@t-yuden.com
Makes sense to me, but we are not shipping to a customer on this side of the
pond... 

So the consensus is, that I write the power supply's D of C to conform to the
LVD, and cite EN 60601-1 (the standard referenced in the unit's agency
certifications and CB Report).

thanks much, 
Brian 


>Would my customer (manufacturer of the end-use medical device) have any 
>trouble in Europe when the D of C for the unit's power supply is lists the 
>LVD? 

>After the power supply arrives at your customer's facility, no European 
>authority will ever ask to see the power supply DOC.  Once the power supply 
>is installed in the end product, the only Declaration of Conformity of 
>interest to European authorities will be your customer's MDD declaration on 
>the overall product. 


>>Thanks for the ideas. But as the LVD specifically precludes medical
stuff, 
>>   how can I declare conformity? Has any NCB issued a "formal recomendation" 
>>for this particular combination? 

>A **power supply** is not a 'medical device', unless you intend it to be 
>used for ECT or electro-cautery or something similar. If it's a power 
>supply for any unspecified medical instrument, it's not, in itself, a 
>medical instrument. 




Re: D of C for medical P.S.

2003-02-26 Thread soundsu...@aol.com

>Would my customer (manufacturer of the end-use medical device) have any 
trouble in Europe when the D of C for the unit's power supply is lists the 
LVD?

After the power supply arrives at your customer's facility, no European 
authority will ever ask to see the power supply DOC.  Once the power supply 
is installed in the end product, the only Declaration of Conformity of 
interest to European authorities will be your customer's MDD declaration on 
the overall product. 




Greg Galluccio
www.productapprovals.com


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



Re: D of C for medical P.S.

2003-02-26 Thread John Woodgate

I read in !emc-pstc that boconn...@t-yuden.com wrote (in ) about 'D of C for medical
P.S.' on Wed, 26 Feb 2003:
>Thanks for the ideas. But as the LVD specifically precludes medical
stuff, 
>how can I declare conformity? Has any NCB issued a "formal recomendation" 
>for this particular combination?
>

A **power supply** is not a 'medical device', unless you intend it to be
used for ECT or electro-cautery or something similar. If it's a power
supply for any unspecified medical instrument, it's not, in itself, a
medical instrument.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk 
Interested in professional sound reinforcement and distribution? Then go to 
http://www.isce.org.uk
PLEASE do NOT copy news posts to me by E-MAIL!


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



RE: D of C for medical P.S.

2003-02-26 Thread boconn...@t-yuden.com
Gentlemen 

Thanks for the ideas. But as the LVD specifically precludes medical stuff, how
can I declare conformity? Has any NCB issued a "formal recomendation" for this
particular combination?

Would my customer (manufacturer of the end-use medical device) have any
trouble in Europe when the D of C for the unit's power supply is lists the LVD?

thanks much, 
Brian 

<3E5C9943.14190.33FECD@localhost>) about 'D of C for medical P.S.' on 
Wed, 26 Feb 2003: 
>You are right not issuing a D of C to the MDD, as your product is not a
medical 
>device or an 
>accessory ot a medical device. 
> 
>On the other hand I see no reason not to issue a D of C to the LVD.
Compliance 
>with a standard in 
>the list of Harmonised Standards is not the only way to demonstrate
compliance 
>with the LVD. It 
>does not take very much effort to draw up a comparison table showing that the 
>requirements for a 
>power supply in IEC 60601-1 are more stringent than say, IEC 60950 or IEC 
>61010-1. Using 
>engineering judgment on this basis, I believe that you can issue a D of C to
the 
>LVD, specifying 
>IEC 60601-1 as the applied standard. You can present your comparison table to 
>anyone who questions 
>this, demonstrating the engineering basis for your declaration. 

That's what I would recommend, but I don't know enough about 60601-1 to 
be definite. 
-- 
Regards, John Woodgate 




Re: D of C for medical P.S.

2003-02-26 Thread John Woodgate

I read in !emc-pstc that Jon Griver  wrote (in
<3E5C9943.14190.33FECD@localhost>) about 'D of C for medical P.S.' on
Wed, 26 Feb 2003:
>You are right not issuing a D of C to the MDD, as your product is not a
medical 
>device or an 
>accessory ot a medical device.
>
>On the other hand I see no reason not to issue a D of C to the LVD.
Compliance 
>with a standard in 
>the list of Harmonised Standards is not the only way to demonstrate
compliance 
>with the LVD. It 
>does not take very much effort to draw up a comparison table showing that the 
>requirements for a 
>power supply in IEC 60601-1 are more stringent than say, IEC 60950 or IEC 
>61010-1. Using 
>engineering judgment on this basis, I believe that you can issue a D of C to
the 
>LVD, specifying 
>IEC 60601-1 as the applied standard. You can present your comparison table to 
>anyone who questions 
>this, demonstrating the engineering basis for your declaration.

That's what I would recommend, but I don't know enough about 60601-1 to
be definite.
-- 
Regards, John Woodgate, OOO - Own Opinions Only. http://www.jmwa.demon.co.uk 
Interested in professional sound reinforcement and distribution? Then go to 
http://www.isce.org.uk
PLEASE do NOT copy news posts to me by E-MAIL!


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



Re: D of C for medical P.S.

2003-02-26 Thread soundsu...@aol.com

>From the end-user's point of view, the power supply must meet the 
requirements of EN 60601-1.  That is the standard that will be applied when 
the OEM product is evaluated.  Any declaration of conformity or CE marking 
that goes on the component will serve only one purpose, and that is to move 
the product through customs.   The LVD seems most appropriate.

I would be quite surprised if a customs agent saw either EN 60601 or EN 60950 
on the declaration and questioned it's applicability, and I would be further 
surprised (and perhaps alarmed) if a simple explanation didn't serve to clear 
up the matter.



Greg Galluccio
www.productapprovals.com


This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



AW: D of C for medical P.S.

2003-02-26 Thread Horst Haug

Brian,Jon,

we have done several investigations of medical PC in accordance to IEC60601.
The requirements of IEC60601-1  is different from IEC60950  mainly in
regards of spacing, dielectric testing, leakage and the insulation of the
interfaces. This has to be documented well in the IEC60601 report.  The
report should be in the CB -scheme format. (no CB certificate is required)

The EMC testing should be done to IEC60601-1-2.

There is no problem to declare a medical PC and issue a D . of . C
accordingly.
When you like to sell it as medical PC in Europe, then you will also require
D. of. C for a medical components.

With best regards
Horst Haug

INNOVA Product Service GmbH
Ampferweg 6
Germany 87677 Stottwang
Tel: (49) 8345-952727


Von: owner-emc-p...@majordomo.ieee.org
[mailto:owner-emc-p...@majordomo.ieee.org]Im Auftrag von Jon Griver
Gesendet: Mittwoch, 26. Februar 2003 09:39
An: emc-p...@majordomo.ieee.org
Betreff: Re: D of C for medical P.S.


Brian,

You are right not issuing a D of C to the MDD, as your product is not a
medical device or an
accessory ot a medical device.

On the other hand I see no reason not to issue a D of C to the LVD.
Compliance with a standard in
the list of Harmonised Standards is not the only way to demonstrate
compliance with the LVD. It
does not take very much effort to draw up a comparison table showing that
the requirements for a
power supply in IEC 60601-1 are more stringent than say, IEC 60950 or IEC
61010-1. Using
engineering judgment on this basis, I believe that you can issue a D of C to
the LVD, specifying
IEC 60601-1 as the applied standard. You can present your comparison table
to anyone who questions
this, demonstrating the engineering basis for your declaration.

Regards,

Jon Griver
http://www.601help.com


On 25 Feb 2003 at 13:40, boconn...@t-yuden.com wrote:

>
> Good people of PSTC
> One of our custom (component) SMPS is evaluated/certified to
IEC/EN60601-1. The LVD precludes
> medical stuff, and the MDD is ambiguous on "components" or "materials"
being scoped by the MDD.
> The customer has determined that indicated conformance should be the LVD.
If the unit's D of C
> claims conformance to the LVD, then there is no applicable standard,
within the scope of the LVD
> that can be stated.
> So what is the correct content of the unit's Declaration of Conformity?
Can anyone please cite
> normative OJ guidance?
> thanx much.
> R/S,
> Brian




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



Re: D of C for medical P.S.

2003-02-26 Thread Jon Griver

Brian,

You are right not issuing a D of C to the MDD, as your product is not a
medical device or an 
accessory ot a medical device.

On the other hand I see no reason not to issue a D of C to the LVD. Compliance
with a standard in 
the list of Harmonised Standards is not the only way to demonstrate compliance
with the LVD. It 
does not take very much effort to draw up a comparison table showing that the
requirements for a 
power supply in IEC 60601-1 are more stringent than say, IEC 60950 or IEC
61010-1. Using 
engineering judgment on this basis, I believe that you can issue a D of C to
the LVD, specifying 
IEC 60601-1 as the applied standard. You can present your comparison table to
anyone who questions 
this, demonstrating the engineering basis for your declaration.

Regards,

Jon Griver
http://www.601help.com


On 25 Feb 2003 at 13:40, boconn...@t-yuden.com wrote:

> 
> Good people of PSTC 
> One of our custom (component) SMPS is evaluated/certified to IEC/EN60601-1.
The LVD precludes 
> medical stuff, and the MDD is ambiguous on "components" or "materials" being
scoped by the MDD.
> The customer has determined that indicated conformance should be the LVD. If
the unit's D of C 
> claims conformance to the LVD, then there is no applicable standard, within
the scope of the LVD 
> that can be stated.
> So what is the correct content of the unit's Declaration of Conformity? Can
anyone please cite 
> normative OJ guidance?
> thanx much.
> R/S,
> Brian




This message is from the IEEE EMC Society Product Safety
Technical Committee emc-pstc discussion list.

Visit our web site at:  http://www.ewh.ieee.org/soc/emcs/pstc/

To cancel your subscription, send mail to:
 majord...@ieee.org
with the single line:
 unsubscribe emc-pstc

For help, send mail to the list administrators:
 Ron Pickard:  emc-p...@hypercom.com
 Dave Heald:   davehe...@attbi.com

For policy questions, send mail to:
 Richard Nute:   ri...@ieee.org
 Jim Bacher: j.bac...@ieee.org

Archive is being moved, we will announce when it is back on-line.
All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc



D of C for medical P.S.

2003-02-25 Thread boconn...@t-yuden.com
Good people of PSTC 

One of our custom (component) SMPS is evaluated/certified to IEC/EN60601-1.
The LVD precludes medical stuff, and the MDD is ambiguous on "components" or
"materials" being scoped by the MDD.

The customer has determined that indicated conformance should be the LVD. If
the unit's D of C claims conformance to the LVD, then there is no applicable
standard, within the scope of the LVD that can be stated.

So what is the correct content of the unit's Declaration of Conformity? Can
anyone please cite normative OJ guidance? 

thanx much. 

R/S, 