RE: Switching Transformer Safety Standards

2009-01-29 Thread Ted Eckert
My apologies, but I meant 61010-1, the end product standard.  (My fingers can
sometimes type faster than my brain can work.)

 

Ted Eckert

Compliance Engineer

Microsoft Corporation

ted.eck...@microsoft.com

 

The opinions expressed are my own and do not necessarily reflect those of my
employer.

 

 

 

From: peter merguerian [mailto:pmerguerian2...@yahoo.com] 
Sent: Wednesday, January 28, 2009 8:47 PM
To: Kunde, Brian; EMC-PSTC@LISTSERV.IEEE.ORG; Ted Eckert
Subject: RE: Switching Transformer Safety Standards

 

Brian,

 

I agree with Ted, but replace the standard with IEC/EN/UL 61010-1 (your end
product standard). There is no reason to abide by other standards unless the
61010-1 standard refers to another standard.

 

Best Regards,

 

Peter

 



--- On Wed, 1/28/09, Ted Eckert  wrote:

From: Ted Eckert 
Subject: RE: Switching Transformer Safety Standards
To: "Kunde, Brian" , 
"EMC-PSTC@LISTSERV.IEEE.ORG"

Date: Wednesday, January 28, 2009, 3:07 PM

Hello Brian,

 

These are some general guidelines and are not necessarily complete.  

 

You should be able to have the transformer evaluated under 60601 as part of
your product.  You will need to obey all of the creepage, clearance and
insulation thickness requirements of 60601.  You will also need to consider
any required electric strength test from primary to secondary.  Additionally, 
consider the maximum temperature expected when selecting your insulation
materials.

 

How you construct the transformer will depend on the type of bobbin you intend
to use.  I presume the bobbin selection will be based on a trade-off of cost,
size and efficiency.  If the transformer can be large an inefficient, the
safety will likely be easier to manage. 

 

I realize that a lot of what I stated may be obvious, but it covers a lot of
the basics of switch-mode transformer safety.  The devil is in the details. 
Verifying all of the required parameters on a small transformer can be
difficult as there can be many creepage paths.  With a 2 kW output power, I
presume that this wonʼt be a small transformer.

 

Ted Eckert

Compliance Engineer

Microsoft Corporation

ted.eck...@microsoft.com

 

The opinions expressed are my own and do not necessarily reflect those of my
employer.

 

From: Kunde, Brian [mailto:brian_ku...@lecotc.com] 
Sent: Wednesday, January 28, 2009 2:01 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Switching Transformer Safety Standards

 

Greetings Transformer Experts,

 

We need to design a custom switching transformer that runs at 25kHz, modulated
230VAC mains input and 4kVrms (.5A) output as part of a 4KV HV power supply.  

 

We want to design and test this transformer properly for safety so can anyone
give us a list of construction and/or safety standards to cover UL, CSA,
Europe and any international IEC requirements?

 

We are not experts in the world of specialty transformer safety and usually
rely on the manufacture for construction information but in this case we would
like to get more involved with the details and knowing what standards to
design to would be very helpful. 

 

The end product will be used in a piece of laboratory equipment.  (61010-1).   

 

Thank you and best regards,

 

The Other Brian

 

 

_ 

LECO Corporation Notice: This communication may contain confidential
information intended for the named recipient(s) only. If you received this by
mistake, please destroy it and notify us of the error. Thank you. 

-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
mailto:emc-p...@ieee.org> >

All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
<http://www.ieeecommunities.org/emc-pstc> 
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/ <http://www.ieee-pses.org/> 
Instructions: http://listserv.ieee.org/request/user-guide.html
<http://listserv.ieee.org/request/user-guide.html> 
List rules: http://www.ieee-pses.org/listrules.html
<http://www.ieee-pses.org/listrules.html>  

For help, send mail to the list administrators:
Scott Douglas mailto:emcp...@ptcnh.net> >
Mike Cantwell mailto:mcantw...@ieee.org> > 

For policy questions, send mail to:
Jim Bacher mailto:j.bac...@ieee.org> >
David Heald mailto:dhe...@gmail.com> > 

-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to
mailto:emc-p...@ieee.org> >

All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
<http://www.ieeecommunities.org/emc-pstc> 
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/ <h

RE: Switching Transformer Safety Standards

2009-01-29 Thread peter merguerian
Brian,
 
I agree with Ted, but replace the standard with IEC/EN/UL 61010-1 (your end
product standard). There is no reason to abide by other standards unless the
61010-1 standard refers to another standard.
 
Best Regards,
 
Peter
 


--- On Wed, 1/28/09, Ted Eckert  wrote:


From: Ted Eckert 
Subject: RE: Switching Transformer Safety Standards
To: "Kunde, Brian" , 
"EMC-PSTC@LISTSERV.IEEE.ORG"

Date: Wednesday, January 28, 2009, 3:07 PM



Hello Brian,

 

These are some general guidelines and are not necessarily complete.  

 

You should be able to have the transformer evaluated under 60601 as 
part of
your product.  You will need to obey all of the creepage, clearance and
insulation thickness requirements of 60601.  You will also need to consider
any required electric strength test from primary to secondary.  Additionally, 
consider the maximum temperature expected when selecting your insulation
materials.

 

How you construct the transformer will depend on the type of bobbin you
intend to use.  I presume the bobbin selection will be based on a trade-off of
cost, size and efficiency.  If the transformer can be large an inefficient,
the safety will likely be easier to manage. 

 

I realize that a lot of what I stated may be obvious, but it covers a 
lot of
the basics of switch-mode transformer safety.  The devil is in the details. 
Verifying all of the required parameters on a small transformer can be
difficult as there can be many creepage paths.  With a 2 kW output power, I
presume that this won¢t be a small transformer.

 

Ted Eckert

Compliance Engineer

Microsoft Corporation

ted.eck...@microsoft.com

 

The opinions expressed are my own and do not necessarily reflect those 
of my
employer.

 

From: Kunde, Brian [mailto:brian_ku...@lecotc.com] 
Sent: Wednesday, January 28, 2009 2:01 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
    Subject: Switching Transformer Safety Standards

 

Greetings Transformer Experts,

 

We need to design a custom switching transformer that runs at 25kHz,
modulated 230VAC mains input and 4kVrms (.5A) output as part of a 4KV HV power
supply.  

 

We want to design and test this transformer properly for safety so can 
anyone
give us a list of construction and/or safety standards to cover UL, CSA,
Europe and any international IEC requirements?

 

We are not experts in the world of specialty transformer safety and 
usually
rely on the manufacture for construction information but in this case we would
like to get more involved with the details and knowing what standards to
design to would be very helpful. 

 

The end product will be used in a piece of laboratory equipment.  
(61010-1). 
 

 

Thank you and best regards,

 

The Other Brian

 

 

_ 

LECO Corporation Notice: This communication may contain confidential
information intended for the named recipient(s) only. If you received this by
mistake, please destroy it and notify us of the error. Thank you. 

-

This message is from the IEEE Product Safety Engineering Society 
emc-pstc
discussion list. To post a message to the list, send your e-mail to


All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to 
that URL. 
Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 
For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell  
For policy questions, send mail to:
Jim Bacher 
David Heald  

-

This message is from the IEEE Product Safety Engineering Society 
emc-pstc
discussion list. To post a message to the list, send your e-mail to


All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to 
that URL. 
Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 
For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell  
For policy questions, send mail to:
Jim Bacher 

RE: Switching Transformer Safety Standards

2009-01-28 Thread Ted Eckert
Hello Brian,

 

These are some general guidelines and are not necessarily complete.  

 

You should be able to have the transformer evaluated under 60601 as part of
your product.  You will need to obey all of the creepage, clearance and
insulation thickness requirements of 60601.  You will also need to consider
any required electric strength test from primary to secondary.  Additionally, 
consider the maximum temperature expected when selecting your insulation
materials.

 

How you construct the transformer will depend on the type of bobbin you intend
to use.  I presume the bobbin selection will be based on a trade-off of cost,
size and efficiency.  If the transformer can be large an inefficient, the
safety will likely be easier to manage. 

 

I realize that a lot of what I stated may be obvious, but it covers a lot of
the basics of switch-mode transformer safety.  The devil is in the details. 
Verifying all of the required parameters on a small transformer can be
difficult as there can be many creepage paths.  With a 2 kW output power, I
presume that this won’t be a small transformer.

 

Ted Eckert

Compliance Engineer

Microsoft Corporation

ted.eck...@microsoft.com

 

The opinions expressed are my own and do not necessarily reflect those of my
employer.

 

From: Kunde, Brian [mailto:brian_ku...@lecotc.com] 
Sent: Wednesday, January 28, 2009 2:01 PM
To: EMC-PSTC@LISTSERV.IEEE.ORG
Subject: Switching Transformer Safety Standards

 

Greetings Transformer Experts,

 

We need to design a custom switching transformer that runs at 25kHz, modulated
230VAC mains input and 4kVrms (.5A) output as part of a 4KV HV power supply.  

 

We want to design and test this transformer properly for safety so can anyone
give us a list of construction and/or safety standards to cover UL, CSA,
Europe and any international IEC requirements?

 

We are not experts in the world of specialty transformer safety and usually
rely on the manufacture for construction information but in this case we would
like to get more involved with the details and knowing what standards to
design to would be very helpful. 

 

The end product will be used in a piece of laboratory equipment.  (61010-1).   

 

Thank you and best regards,

 

The Other Brian

 

 

_ 

LECO Corporation Notice: This communication may contain confidential
information intended for the named recipient(s) only. If you received this by
mistake, please destroy it and notify us of the error. Thank you. 

-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to


All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell  

For policy questions, send mail to:
Jim Bacher 
David Heald  

-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to


All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell  

For policy questions, send mail to:
Jim Bacher 
David Heald  

-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to


All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell  

For policy questions, send mail to:
Jim Bacher 
David Heald  




RE: Switching Transformer Safety Standards

2009-01-28 Thread Brian O'Connell
Insufficent information on end-use and input rating.

none, or one, or some, or all of following:

UL5085-x (binational std w/CSA66.x)
IEC/EN61558-x
UL1310/CSA223
 

From: emc-p...@ieee.org [mailto:emc-p...@ieee.org]On Behalf Of
Kunde, Brian
Sent: Wednesday, January 28, 2009 2:01 PM
To: EMC-PSTC (emc-p...@ieee.org)
Subject: Switching Transformer Safety Standards

Greetings Transformer Experts,
 
We need to design a custom switching transformer that runs at
25kHz, modulated 230VAC mains input and 4kVrms (.5A) output as
part of a 4KV HV power supply.  
 
We want to design and test this transformer properly for safety
so can anyone give us a list of construction and/or safety
standards to cover UL, CSA, Europe and any international IEC
requirements?
 
We are not experts in the world of specialty transformer safety
and usually rely on the manufacture for construction information
but in this case we would like to get more involved with the
details and knowing what standards to design to would be very
helpful. 
 
The end product will be used in a piece of laboratory equipment.
(61010-1).   
 
Thank you and best regards,
 
The Other Brian

-

This message is from the IEEE Product Safety Engineering Society emc-pstc 
discussion list. To post a message to the list, send your e-mail to 


All emc-pstc postings are archived and searchable on the web at:
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL.

Website:  http://www.ieee-pses.org/
Instructions:  http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell 

For policy questions, send mail to:
Jim Bacher:  
David Heald: 




Switching Transformer Safety Standards

2009-01-28 Thread Kunde, Brian
Greetings Transformer Experts,

 

We need to design a custom switching transformer that runs at 25kHz, modulated
230VAC mains input and 4kVrms (.5A) output as part of a 4KV HV power supply.  

 

We want to design and test this transformer properly for safety so can anyone
give us a list of construction and/or safety standards to cover UL, CSA,
Europe and any international IEC requirements?

 

We are not experts in the world of specialty transformer safety and usually
rely on the manufacture for construction information but in this case we would
like to get more involved with the details and knowing what standards to
design to would be very helpful. 

 

The end product will be used in a piece of laboratory equipment.  (61010-1).   

 

Thank you and best regards,

 

The Other Brian

 

 

_ 

LECO Corporation Notice: This communication may contain confidential
information intended for the named recipient(s) only. If you received this by
mistake, please destroy it and notify us of the error. Thank you. 
-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to


All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell  

For policy questions, send mail to:
Jim Bacher 
David Heald  

-

This message is from the IEEE Product Safety Engineering Society emc-pstc
discussion list. To post a message to the list, send your e-mail to


All emc-pstc postings are archived and searchable on the web at
http://www.ieeecommunities.org/emc-pstc
Graphics (in well-used formats), large files, etc. can be posted to that URL. 

Website: http://www.ieee-pses.org/
Instructions: http://listserv.ieee.org/request/user-guide.html
List rules: http://www.ieee-pses.org/listrules.html 

For help, send mail to the list administrators:
Scott Douglas 
Mike Cantwell  

For policy questions, send mail to:
Jim Bacher 
David Heald