RE: [Histonet] Re: Histonet Digest, Vol 102, Issue 26

2012-05-23 Thread Kuhnla, Melissa
Hello,
I also use Ventana Ultra instruments.  We have had our air share of
tissue adhesion issues as well. We currently have most of it under
control by doing the following:


1. We switched from standard plus slides to Fisherbrand Excel Adhesion
slides (cat #22-034-985). I believe they are meant for automated
staining, solutions of varying temp and pH, etc.  We did see an
improvement!!
2. We at all time handle our slides with gloves on.  We dip and dry
controls upside-down.  I was very clear to us that dipping and handling
the slide while cutting the control section altered the charge of the
rest of the slide.  As a rule of thumb...if we need to repeat things due
to tissue loss, we run it on a slide by itself and run a separate
control.  If a focus of tumor is very small or tissue is minimal to
begin with, a pathologist may request separate controls right from the
start.
3. We also altered some of our processing schedules.  At first we were
loosing a lot of small biopsies and FNAs.  We shortened our processing
and now they adhere fine!! 
4. We discontinued the use of recycled Xylene in the processor as well.

BTW, We routinely cut at four microns, dry vertical for 15 minutes and
bake at 60 for 90-120 minutes


We have this issue under such a microscope here that we know it will be
impossible to get rid of this all together.  Breast cases (and sometimes
skin) remain to be the only culprits.  These tissues just by nature may
always be a problem.  We now are contemplating running batched controls
for ER/PR cases so all those cases are on their own slides.
Hope this helps.
Melissa
-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of
Madeleine Huey
Sent: Wednesday, May 23, 2012 12:29 AM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Re: Histonet Digest, Vol 102, Issue 26

 I'm a Histotechnologist working in the Regional Hospital in Barrie,
 ON Canada. We are using the Ventana Ultra for our Immunohistochemistry
 (IHC). Since the end of February, we have been having issues with some
 tissues lifting off our positive (marked with +) charged slides. It
 seems to be mostly with the fatty and/or larger sections. We now dry
 our slides for one hour at room temperature (R.T.) and an additional
 hour at 60 degrees C. We cut our IHC sections at 4 um. Since we have
 tried 2 different types of + slides and will be trying another type of
 charged slide (from Newcomer this time) I was wondering if anyone has
 any other suggestions?
 I also have another question regarding a QC (quality control) issue.
 We use a multi-tissue control that is applied to the top of all our
 test slides for IHC. One of our paths commented that there is some
 positive staining in the smooth muscle nuclei of thenormal bowel when
 we are testing for Progesterone (PR). We are using a Heat Induce
 Epitope Retrieval (HEIR) of 36 minutes with CC1 (Ventana's proprietary
 buffer @ pH of 8.0-8.5) and a primary antibody incubation time of 16
 minutes with PR clone 1E2. (Ventana instrumentation provides
 pre-diluted antibodies and the user adjusts the concentration of the
 antibody by adjusting the time the primary antibody is incubated with
 the tissue).
 I am concerned about the implications of this staining and I have
 not been able to find a reference to this kind of unusual staining
 pattern. The bowel tissue that we are using as QC is from a 62 year
 old female patient. I was wondering if anyone has had any experience
 with this kind of staining and /or any references that I could use.

 Thanking you in advance,
 I look forward to your input,
 Nancy Cloughley-Gray MLT

Nancy,

Your problem is very common, due to fatty tissue is difficult to
process.

Here's my suggestion if you want to salvage your fatty tissue blocks
(presumably the fatty tissues are well fixed);

1) melt down the fatty tissue block  re-process

2) cut  put on charged slides (Fisher Scientific Ultra Stick
charged slides work the best)

3) bake slide for 30 min @ 60c

4) do antigen retrieval with Biocare's Diva for 3 min in Pascal or
Biocare pressure cooker (very important; do NOT use Triology  Cell
Marque pressure cooker).

5) do your ihc as usual

Three important elements; (1) well process tissue, (2) Biocare Diva AR
buffer,  (3) Pascal pressure cooker.

Keep us posted what's your outcome.

Madeleine Huey BS, HTL/QIHC (ASCP)
Supervisor - Pathology (histology  IPOX)
madelein...@elcaminohospital.org

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


The information in this e-mail, and any attachments therein, is confidential 
and for use by the intended addressee only. If this message is received by you 
in error please do not disseminate or read further. Please reply to the sender 
that you have received the message in error, then delete the message. Although 
Catholic Health 

[Histonet] Excellent opportunity: Histology Manager in Berkeley, CA

2012-05-23 Thread Kiranjit Grewal
Regional Laboratory Section Manager – Histology




Company Name: 
Kaiser Permanente
Approximate Salary: 
Not Specified

Location: 
Berkeley, California
Country: 
United States

Industry: 
Management
Position type: 
Full Time

Experience level: 
 5 years
Education level: 
Bachelor's Degree






ID
714


Short Description: 
Regional Laboratory Section Manager, Histology - Kaiser Permanente in Berkeley, 
California
Long Description: 
Success is measured one step at a time 

I hold myself to the highest standards. And believe in my ability to make a 
difference—one day at a time. As an employee of Kaiser Permanente, I have the 
freedom to put these beliefs into practice. Here, I have the resources and 
support I need to realize my potential. If you believe professional empowerment 
leads to personal satisfaction, this is the place to put your beliefs into 
practice. We have the following full-time position in Berkeley, California. 

Regional Laboratory Section Manager – Histology 

In this role, you will direct and control the day-to-day Regional Laboratory 
Histology department operations including all personnel activities, technical 
oversight, and work flow. You will participate in the design of the Regional 
Laboratory and Northern California Region integrated laboratory quality system; 
design and implement effective risk control processes; and will ensure 
compliance with regulatory and accreditation agencies’ rules and regulations. 
You will also lead in the research of new and/or improved test development 
methodologies by performing experimental testing procedures; validating 
effectiveness/feasibility for implementation; cost of procedures; preparing and 
submitting recommendations for change to laboratory management and other 
stakeholders (such as Chiefs of Pathologists). You will also participate in 
department, inter-department, inter-facility, and inter-regional level projects 
which help the regional laboratory achieve its
 goal of providing quality service and client support in a cost effective 
manner. In addition, you will serve as a primary liaison to RILIS/ITS for 
ongoing and new issues. 

Qualifications include: 
• At least five years of experience in a high-volume histology laboratory 
• Previous supervisory/managerial experience (usually three years) 
• A bachelor’s degree in biological sciences or a related field; a master’s 
degree in science is preferred 
• Additional courses in business administration and/or management preferred 
• Certification by the American Society for Clinical Pathologists 
• IHC laboratory operations experience preferred 
• IHC method optimization/validation experience and IHC instrumentation 
experience preferred 
• The ability to work in a Labor/Management Partnership environment 

We offer a highly competitive salary and an exceptional benefits package. For 
immediate consideration, please e-mail your resume to charlene.schaef...@kp.org 
or visit http://jobs.kp.org for complete qualifications and job submission 
details, referencing job number 133637. Principals only. EOE/AA Employer. 

If you would like to hear the Kaiser Permanente story as told by our employees, 
watch the videos at kp.org/jobs/video. Follow us on twitter.com/KPCareers or 
visit the KP Careers tab on facebook.com/KPThrive. 

This position supports Kaiser Permanente’s code of conduct and compliance by 
adhering to all laws and regulations, accreditation and licensure requirements, 
and internal policies and procedures. 

jobs.kp.org 

KAISER PERMANENTE 

Contact: charlene.schaef...@kp.org 

 
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] (no subject)

2012-05-23 Thread Podawiltz, Thomas
The last person I hired for a tech position, was just finishing her on-line 
program. I helped her complete that and then she was given two years to get her 
certification. This was all part of her hiring agreement, so she knew this 
coming in and completely understood that failing to achieve either one would 
affect her employment.



Tom Podawiltz HT (ASCP)
Histology Section Head/Laboratory Safety Officer. 
 
 

 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
Sent: Tuesday, May 22, 2012 1:43 PM
To: sherrian.mc...@va.gov; histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] (no subject)


That seems to be the unfortunate situation at this time...




Joelle Weaver MAOM, HTL (ASCP) QIHC
  Date: Tue, 22 May 2012 12:39:39 -0500
 From: sherrian.mc...@va.gov
 To: histonet@lists.utsouthwestern.edu
 Subject: [Histonet] (no subject)
 
 I agree and would like to add.  This is one scenario that I have seen
 many times, where hospitals  or wherever will hire histotechs without
 certifications .  I am thinking that saves them money and they still
 have a histotech.  I have seen good histotechs  that have no
 certification and likewise some bad ones with certification.  Lately I
 have seen these  schools turn out histotechs   ,  it seems with little
 encouragement to get certified.  If places will hire them without being
 certified,  there seems little incentive (unless you are self motivated
 for more money) to move on up to certification.
 
  
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet
THIS MESSAGE IS CONFIDENTIAL.  
This e-mail message and any attachments are proprietary and confidential 
information intended only for the use of the recipient(s) named above. If you 
are not the intended recipient, you may not print,distribute, or copy this 
message or any attachments.  If you have received this communication in error, 
please notify the sender by return e-mail and delete this message and any 
attachments from your computer. Any views or opinions expressed are solely 
those of the author and do not necessarily represent those of LRGHealthcare.


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] PVK stain

2012-05-23 Thread Perry, Margaret
We are in need of some help trouble shooting this stain.  Is it normal for the 
tissue nuclei to stain red?  If not what are we doing wrong?  The Chlamydia are 
staining OK.

Margaret Perry HT(ASCP)
Dept of Veterinary and  Biomedical services
Box 2175
South Dakota State University
Brookings SD 57007
605-688-5638

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


RE: [Histonet] (no subject)

2012-05-23 Thread joelle weaver

TomI compliment your approach, and I personally really like to see when 
employers do this. I think it provides opportunity, yet incentive. Two years 
seems especially reasonable. Most places that I have known who use this method 
for new-grad hires, have allowed 1 year. It is even better when they provide 
additional training and support to the person as they are preparing. Having 
been involved in an online program and clinical curricula in the past, I feel 
from my observations that continuity with work in the environment is essential 
for most in connecting the theory with execution. 



Joelle Weaver MAOM, HTL (ASCP) QIHC
  From: tpodawi...@lrgh.org
 To: joellewea...@hotmail.com; sherrian.mc...@va.gov; 
 histonet@lists.utsouthwestern.edu
 Date: Wed, 23 May 2012 10:45:52 -0400
 Subject: RE: [Histonet] (no subject)
 
 The last person I hired for a tech position, was just finishing her on-line 
 program. I helped her complete that and then she was given two years to get 
 her certification. This was all part of her hiring agreement, so she knew 
 this coming in and completely understood that failing to achieve either one 
 would affect her employment.
 
 
 
 Tom Podawiltz HT (ASCP)
 Histology Section Head/Laboratory Safety Officer. 
  
  
 
  
 
 -Original Message-
 From: histonet-boun...@lists.utsouthwestern.edu 
 [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
 Sent: Tuesday, May 22, 2012 1:43 PM
 To: sherrian.mc...@va.gov; histonet@lists.utsouthwestern.edu
 Subject: RE: [Histonet] (no subject)
 
 
 That seems to be the unfortunate situation at this time...
 
 
 
 
 Joelle Weaver MAOM, HTL (ASCP) QIHC
   Date: Tue, 22 May 2012 12:39:39 -0500
  From: sherrian.mc...@va.gov
  To: histonet@lists.utsouthwestern.edu
  Subject: [Histonet] (no subject)
  
  I agree and would like to add.  This is one scenario that I have seen
  many times, where hospitals  or wherever will hire histotechs without
  certifications .  I am thinking that saves them money and they still
  have a histotech.  I have seen good histotechs  that have no
  certification and likewise some bad ones with certification.  Lately I
  have seen these  schools turn out histotechs   ,  it seems with little
  encouragement to get certified.  If places will hire them without being
  certified,  there seems little incentive (unless you are self motivated
  for more money) to move on up to certification.
  
   
  
  ___
  Histonet mailing list
  Histonet@lists.utsouthwestern.edu
  http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
 THIS MESSAGE IS CONFIDENTIAL.  
 This e-mail message and any attachments are proprietary and confidential 
 information intended only for the use of the recipient(s) named above. If you 
 are not the intended recipient, you may not print,distribute, or copy this 
 message or any attachments.  If you have received this communication in 
 error, please notify the sender by return e-mail and delete this message and 
 any attachments from your computer. Any views or opinions expressed are 
 solely those of the author and do not necessarily represent those of 
 LRGHealthcare.
 
  
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Antibody titer

2012-05-23 Thread Courtney Pierce
When you titer a new antibody you go one below and one above the recommend 
dilution. If you find that the one above the dilution works do you go then out 
two more to make sure they don't work? This is a question in our lab right now.

Courtney Pierce
IHC Specialist
Quintiles
Translational RD - Oncology
Innovation
Navigating the new health

610 Oakmont Lane
Westmont, IL 60559

Office: + 630-203-6234
courtney.pie...@quintiles.com

clinical | commercial | consulting | capital


**  IMPORTANT--PLEASE READ  
This electronic message, including its attachments, is COMPANY CONFIDENTIAL
and may contain PROPRIETARY or LEGALLY PRIVILEGED information.  If you are 
not the intended recipient, you are hereby notified that any use, disclosure,
copying, or distribution of this message or any of the information included
in it is unauthorized and strictly prohibited.  If you have received this
message in error, please immediately notify the sender by reply e-mail and
permanently delete this message and its attachments, along with any copies
thereof. Thank you. 


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: HTL exam

2012-05-23 Thread Mayer,Toysha N
We used to have our students take the BOC just before graduation in the summer, 
but found that the stress of finals coupled with the BOC was a bit much.  
Especially for those who worked.  We do require that the students sign up for 
the exam (we do this in class as a group) and advise them to take it within 
about a month of graduation.  
Some of the students request extra tutoring for the exam and we try to 
accommodate that as well.
Employers that are changing their requirements want registered or eligible; or 
completion of an accredited program, usually.


Toysha N. Mayer, MBA, HT (ASCP)
Instructor, Education Coordinator
Program in Histotechnology
School of Health Professions
MD Anderson Cancer Center
(713) 563-3481
tnma...@mdanderson.org





--

Message: 2
Date: Tue, 22 May 2012 12:39:39 -0500
From: McAnn, Sherrian sherrian.mc...@va.gov
Subject: [Histonet] (no subject)
To: histonet@lists.utsouthwestern.edu
Message-ID:
61e2b58cecef384094a363989d47c090084e5...@vhav17msga2.v17.med.va.gov
Content-Type: text/plain; charset=us-ascii

I agree and would like to add.  This is one scenario that I have seen many 
times, where hospitals  or wherever will hire histotechs without certifications 
.  I am thinking that saves them money and they still have a histotech.  I 
have seen good histotechs  that have no certification and likewise some bad 
ones with certification.  Lately I
have seen these  schools turn out histotechs   ,  it seems with little
encouragement to get certified.  If places will hire them without being 
certified,  there seems little incentive (unless you are self motivated for 
more money) to move on up to certification.

 



--

Message: 3
Date: Tue, 22 May 2012 17:42:44 +
From: joelle weaver joellewea...@hotmail.com
Subject: RE: [Histonet] (no subject)
To: sherrian.mc...@va.gov, histonet@lists.utsouthwestern.edu
Message-ID: snt135-w565f62180596fc8cfb7f88d8...@phx.gbl
Content-Type: text/plain; charset=iso-8859-1


That seems to be the unfortunate situation at this time...




Joelle Weaver MAOM, HTL (ASCP) QIHC
  Date: Tue, 22 May 2012 12:39:39 -0500
 From: sherrian.mc...@va.gov
 To: histonet@lists.utsouthwestern.edu
 Subject: [Histonet] (no subject)
 
 I agree and would like to add.  This is one scenario that I have seen 
 many times, where hospitals  or wherever will hire histotechs without 
 certifications .  I am thinking that saves them money and they still 
 have a histotech.  I have seen good histotechs  that have no 
 certification and likewise some bad ones with certification.  Lately I
 have seen these  schools turn out histotechs   ,  it seems with little
 encouragement to get certified.  If places will hire them without 
 being certified,  there seems little incentive (unless you are self 
 motivated for more money) to move on up to certification.
 
  
 
 ___
 Histonet mailing list
 Histonet@lists.utsouthwestern.edu
 http://lists.utsouthwestern.edu/mailman/listinfo/histonet
  

*

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Re: Histonet Digest, Vol 102, Issue 28

2012-05-23 Thread Amos Brooks
Hi,
   I have done this. It works quite well. It would probably be best to use
an alkaline phosphatase detection since the brilliant red of the fast red
precipitate is more distinguishable from the brown DAB. (I am assuming you
used HRP-DAB detection of the Tunel.) You needn't worry about the left over
HRP since you'll be doing an antigen retrieval step which should wash them
all out.

Amos

On Wed, May 23, 2012 at 1:00 PM,
histonet-requ...@lists.utsouthwestern.eduwrote:

 Message: 16
 Date: Wed, 23 May 2012 08:03:17 -0700
 From: Alicia R. Lange arla...@medicine.nevada.edu
 Subject: [Histonet] RESTAINING WITH KI67
 To: histonet@lists.utsouthwestern.edu
 Message-ID:
9809822f80b1414c90eb4438713c896801345...@medx.medicine.nevada.edu
 Content-Type: text/plain;   charset=us-ascii

 Hello Histoland!



 Has anyone restained tunel slides with ki67? Since tunel stains
 apoptosis and ki67 proliferation, I was just going to restain with ki67.
 I wanted to get some input before I did this. Thanks!



 Alicia Lange

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


Re: [Histonet] Antibody titer

2012-05-23 Thread Richard Cartun
Yes, I always want to see the immunoreactivity drop-off (or even disappear) 
with a higher dilution(s).  Otherwise, you might be wasting antibody by using a 
dilution that is too concentrated.  Also, be careful with the recommended 
dilution.  That dilution is only applicable if you are using the exact same 
staining conditions.

Richard

Richard W. Cartun, MS, PhD
Director, Histology  Immunopathology
Director, Biospecimen Collection Programs
Assistant Director, Anatomic Pathology
Hartford Hospital
80 Seymour Street
Hartford, CT  06102
(860) 545-1596 Office
(860) 545-2204 Fax


 Courtney Pierce courtney.pie...@quintiles.com 5/23/2012 12:57 PM 
When you titer a new antibody you go one below and one above the recommend 
dilution. If you find that the one above the dilution works do you go then out 
two more to make sure they don't work? This is a question in our lab right now.

Courtney Pierce
IHC Specialist
Quintiles
Translational RD - Oncology
Innovation
Navigating the new health

610 Oakmont Lane
Westmont, IL 60559

Office: + 630-203-6234
courtney.pie...@quintiles.com 

clinical | commercial | consulting | capital


**  IMPORTANT--PLEASE READ  
This electronic message, including its attachments, is COMPANY CONFIDENTIAL
and may contain PROPRIETARY or LEGALLY PRIVILEGED information.  If you are 
not the intended recipient, you are hereby notified that any use, disclosure,
copying, or distribution of this message or any of the information included
in it is unauthorized and strictly prohibited.  If you have received this
message in error, please immediately notify the sender by reply e-mail and
permanently delete this message and its attachments, along with any copies
thereof. Thank you. 


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu 
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Microwave verification

2012-05-23 Thread Bustamante, Lin
Does anyone knows a protocol for Microwave verification? (CAP requirement).
Thank you.
Lin Bustamante
Central Texas Gastrointestinal Clinic
Histology laboratory.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RNA extraction and amplification of LCM rice grain tissue

2012-05-23 Thread Marlen Müller

Dear HistoNet Community,

Currently we are working on a procedure to dissect the grain of the rice 
plant via a PALM Laser Microdisection System. According to published 
work in the barley grain (Schiebold et al 2011) 2 kits for RNA 
extraction, purification and amplification were used (for mRNA 
extraction Dynabeads mRNA DIRECT MicroKit or for total RNA 
extractionRNeasy MicroKit and for RNA amplification CE Version 
ExpressArt mRNA amplification Nanokit). We will use the extracted and 
amplified RNA then for high-throughput RNA Sequencing.
Do you have any experience with RNA amplification for very small amount 
of tissues? Would you use the classical trizol method or should I rely 
on a commercial kit for extraction?

Thanks alot for sharing your knowledge!

Best regards,
Marlen

--
ETH Zürich
*Marlen Müller*
PhD Student
Group of Plantbiotechnology
Universitätstrasse 2
8092 Zürich
Switzerland
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Microwave verification

2012-05-23 Thread Weems, Joyce K.
Someone else asked about this as well.. hope the attachment of our very simple 
instructions is ok to send. Sor far this has worked for us...j

Here is CAP checklist:

ANP.27170 Microwave Usage Phase I
Microwave devices are used in accordance with manufacturer's instructions.
NOTE: Microwave devices should be used in accordance with manufacturer's 
instructions, unless
CAP requirements are more stringent.
Evidence of Compliance:
✓ Written procedure for microwave usage

ANP.28290 Microwave Monitoring Phase I
Microwave devices are at least annually monitored for reproducibility.
NOTE: “Reproducibility” is defined as consistency in diagnostic quality 
obtained from microwave
equipment and procedures. For some devices, reproducibility may be evaluated by 
monitoring the
temperatures of identical samples after microwave processing. For those 
microwave devices
(particularly those incorporated into histology processing equipment) that use
temperature-independent methods to evaluate reproducibility, the laboratory 
should have a written
procedure for monitoring reproducibility that follows instrument manufacturer's 
instructions.
Information on such procedures is given in the reference to this checklist 
requirement (see below).
The microwave device should be tested for radiation leakage if there is visible 
damage to the device.
Evidence of Compliance:
✓ Written procedure for monitoring the diagnostic quality of specimens 
processed using microwaves

ANP.28860 Microwave Container Venting Phase I
All containers used in microwave devices are vented.
NOTE: Venting of containers is necessary so that processing occurs at 
atmospheric pressure, to
prevent explosion. For procedures using pressure above that of the atmosphere, 
specialized
containers must be used, with strict adherence to manufacturer instructions.
Evidence of Compliance:
✓ Written procedure for the use of appropriately vented containers

ANP.29430 Microwave Venting Phase I
Microwave devices are properly vented.
NOTE: Microwave devices should be placed in an appropriate ventilation hood to 
contain airborne
chemical contaminants and potentially infectious agents. Before operation of 
the microwave device,
36 of 43
Emory/Saint Josephs, Inc
Main Laboratory
Anatomic Pathology Checklist 01.04.2012
flammable and corrosive reagents should be removed from the hood, to prevent 
fire or chemical
damage to the electronic components of the device. Microwave devices used 
outside a fume hood
should have an integral fume extractor that is certified by the manufacturer 
for use in a clinical
laboratory.
The effectiveness of ventilation should be monitored at least annually.
This checklist requirement does not apply if only non-hazardous reagents (and 
non-infectious
specimens) are used in the device (e.g. water, certain biological stains, 
paraffin sections). The
laboratory should consult the MSDS sheets received with reagents and stains to 
assist in determining
proper handling requirements and safe use.
This checklist item does not apply to microwave devices that are designed by 
the manufacturer to
operate without venting.
Evidence of Compliance:
✓ Records of annual evaluation of ventilation effectiveness



Joyce Weems
Pathology Manager
678-843-7376 Phone
678-843-7831 Fax
joyce.we...@emoryhealthcare.org



www.saintjosephsatlanta.org
5665 Peachtree Dunwoody Road
Atlanta, GA 30342

This e-mail, including any attachments is the property of Saint Joseph’s 
Hospital and is intended for the sole use of the intended recipient(s).  It may 
contain information that is privileged and confidential.  Any unauthorized 
review, use, disclosure, or distribution is prohibited. If you are not the 
intended recipient, please delete this message, and reply to the sender 
regarding the error in a separate email.


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Bustamante, Lin
Sent: Wednesday, May 23, 2012 2:42 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] Microwave verification

Does anyone knows a protocol for Microwave verification? (CAP requirement).
Thank you.
Lin Bustamante
Central Texas Gastrointestinal Clinic
Histology laboratory.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet



This e-mail message (including any attachments) is for the sole use of
the intended recipient(s) and may contain confidential and privileged
information. If the reader of this message is not the intended
recipient, you are hereby notified that any dissemination, distribution
or copying of this message (including any attachments) is strictly
prohibited.

If you have received this message in error, please contact
the sender by reply e-mail message and destroy all copies of the
original message (including attachments).

[Histonet] old equipment for parts

2012-05-23 Thread Shirley A. Powell
I have an old Tissue Tek III cryo console that only gets luke cool and a 
ThermoFisher slide dryer that died on me.  If anyone wants these for parts or 
boat anchors you can have them, you pay shipping.  I hate to trash them if the 
parts can be recycled.

sp

Shirley A. Powell, HT(ASCP)HTL, QIHC
Technical Director
Histology Curricular Support Laboratory
Mercer University School of Medicine
1550 College Street
Macon, GA  31207
478-301-2374 Lab
478-301-5489 Fax

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Microwave verification

2012-05-23 Thread Weems, Joyce K.
Ok - attachment didn't work. Whoever wants the procedure, just let me know it 
is ok to send separately! j

Joyce Weems
Pathology Manager
678-843-7376 Phone
678-843-7831 Fax
joyce.we...@emoryhealthcare.org



www.saintjosephsatlanta.org
5665 Peachtree Dunwoody Road
Atlanta, GA 30342

This e-mail, including any attachments is the property of Saint Joseph’s 
Hospital and is intended for the sole use of the intended recipient(s).  It may 
contain information that is privileged and confidential.  Any unauthorized 
review, use, disclosure, or distribution is prohibited. If you are not the 
intended recipient, please delete this message, and reply to the sender 
regarding the error in a separate email. 


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Weems, Joyce K.
Sent: Wednesday, May 23, 2012 4:27 PM
To: 'Bustamante, Lin'; histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Microwave verification

Someone else asked about this as well.. hope the attachment of our very simple 
instructions is ok to send. Sor far this has worked for us...j

Here is CAP checklist:

ANP.27170 Microwave Usage Phase I
Microwave devices are used in accordance with manufacturer's instructions.
NOTE: Microwave devices should be used in accordance with manufacturer's 
instructions, unless CAP requirements are more stringent.
Evidence of Compliance:
✓ Written procedure for microwave usage

ANP.28290 Microwave Monitoring Phase I
Microwave devices are at least annually monitored for reproducibility.
NOTE: “Reproducibility” is defined as consistency in diagnostic quality 
obtained from microwave equipment and procedures. For some devices, 
reproducibility may be evaluated by monitoring the temperatures of identical 
samples after microwave processing. For those microwave devices (particularly 
those incorporated into histology processing equipment) that use 
temperature-independent methods to evaluate reproducibility, the laboratory 
should have a written procedure for monitoring reproducibility that follows 
instrument manufacturer's instructions.
Information on such procedures is given in the reference to this checklist 
requirement (see below).
The microwave device should be tested for radiation leakage if there is visible 
damage to the device.
Evidence of Compliance:
✓ Written procedure for monitoring the diagnostic quality of specimens 
processed using microwaves

ANP.28860 Microwave Container Venting Phase I All containers used in microwave 
devices are vented.
NOTE: Venting of containers is necessary so that processing occurs at 
atmospheric pressure, to prevent explosion. For procedures using pressure above 
that of the atmosphere, specialized containers must be used, with strict 
adherence to manufacturer instructions.
Evidence of Compliance:
✓ Written procedure for the use of appropriately vented containers

ANP.29430 Microwave Venting Phase I
Microwave devices are properly vented.
NOTE: Microwave devices should be placed in an appropriate ventilation hood to 
contain airborne chemical contaminants and potentially infectious agents. 
Before operation of the microwave device,
36 of 43
Emory/Saint Josephs, Inc
Main Laboratory
Anatomic Pathology Checklist 01.04.2012
flammable and corrosive reagents should be removed from the hood, to prevent 
fire or chemical damage to the electronic components of the device. Microwave 
devices used outside a fume hood should have an integral fume extractor that is 
certified by the manufacturer for use in a clinical laboratory.
The effectiveness of ventilation should be monitored at least annually.
This checklist requirement does not apply if only non-hazardous reagents (and 
non-infectious
specimens) are used in the device (e.g. water, certain biological stains, 
paraffin sections). The laboratory should consult the MSDS sheets received with 
reagents and stains to assist in determining proper handling requirements and 
safe use.
This checklist item does not apply to microwave devices that are designed by 
the manufacturer to operate without venting.
Evidence of Compliance:
✓ Records of annual evaluation of ventilation effectiveness



Joyce Weems
Pathology Manager
678-843-7376 Phone
678-843-7831 Fax
joyce.we...@emoryhealthcare.org



www.saintjosephsatlanta.org
5665 Peachtree Dunwoody Road
Atlanta, GA 30342

This e-mail, including any attachments is the property of Saint Joseph’s 
Hospital and is intended for the sole use of the intended recipient(s).  It may 
contain information that is privileged and confidential.  Any unauthorized 
review, use, disclosure, or distribution is prohibited. If you are not the 
intended recipient, please delete this message, and reply to the sender 
regarding the error in a separate email.


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Bustamante, Lin
Sent: Wednesday, May 

[Histonet] RE: Microwave verification

2012-05-23 Thread Bustamante, Lin
Thank you so much.
You are all great.
Lin :-)


From: Weems, Joyce K. [joyce.we...@emoryhealthcare.org]
Sent: Wednesday, May 23, 2012 3:33 PM
To: Weems, Joyce K.; Bustamante, Lin; histonet@lists.utsouthwestern.edu
Subject: RE: Microwave verification

Ok - attachment didn't work. Whoever wants the procedure, just let me know it 
is ok to send separately! j

Joyce Weems
Pathology Manager
678-843-7376 Phone
678-843-7831 Fax
joyce.we...@emoryhealthcare.org



www.saintjosephsatlanta.org
5665 Peachtree Dunwoody Road
Atlanta, GA 30342

This e-mail, including any attachments is the property of Saint Joseph’s 
Hospital and is intended for the sole use of the intended recipient(s).  It may 
contain information that is privileged and confidential.  Any unauthorized 
review, use, disclosure, or distribution is prohibited. If you are not the 
intended recipient, please delete this message, and reply to the sender 
regarding the error in a separate email.


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Weems, Joyce K.
Sent: Wednesday, May 23, 2012 4:27 PM
To: 'Bustamante, Lin'; histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Microwave verification

Someone else asked about this as well.. hope the attachment of our very simple 
instructions is ok to send. Sor far this has worked for us...j

Here is CAP checklist:

ANP.27170 Microwave Usage Phase I
Microwave devices are used in accordance with manufacturer's instructions.
NOTE: Microwave devices should be used in accordance with manufacturer's 
instructions, unless CAP requirements are more stringent.
Evidence of Compliance:
✓ Written procedure for microwave usage

ANP.28290 Microwave Monitoring Phase I
Microwave devices are at least annually monitored for reproducibility.
NOTE: “Reproducibility” is defined as consistency in diagnostic quality 
obtained from microwave equipment and procedures. For some devices, 
reproducibility may be evaluated by monitoring the temperatures of identical 
samples after microwave processing. For those microwave devices (particularly 
those incorporated into histology processing equipment) that use 
temperature-independent methods to evaluate reproducibility, the laboratory 
should have a written procedure for monitoring reproducibility that follows 
instrument manufacturer's instructions.
Information on such procedures is given in the reference to this checklist 
requirement (see below).
The microwave device should be tested for radiation leakage if there is visible 
damage to the device.
Evidence of Compliance:
✓ Written procedure for monitoring the diagnostic quality of specimens 
processed using microwaves

ANP.28860 Microwave Container Venting Phase I All containers used in microwave 
devices are vented.
NOTE: Venting of containers is necessary so that processing occurs at 
atmospheric pressure, to prevent explosion. For procedures using pressure above 
that of the atmosphere, specialized containers must be used, with strict 
adherence to manufacturer instructions.
Evidence of Compliance:
✓ Written procedure for the use of appropriately vented containers

ANP.29430 Microwave Venting Phase I
Microwave devices are properly vented.
NOTE: Microwave devices should be placed in an appropriate ventilation hood to 
contain airborne chemical contaminants and potentially infectious agents. 
Before operation of the microwave device,
36 of 43
Emory/Saint Josephs, Inc
Main Laboratory
Anatomic Pathology Checklist 01.04.2012
flammable and corrosive reagents should be removed from the hood, to prevent 
fire or chemical damage to the electronic components of the device. Microwave 
devices used outside a fume hood should have an integral fume extractor that is 
certified by the manufacturer for use in a clinical laboratory.
The effectiveness of ventilation should be monitored at least annually.
This checklist requirement does not apply if only non-hazardous reagents (and 
non-infectious
specimens) are used in the device (e.g. water, certain biological stains, 
paraffin sections). The laboratory should consult the MSDS sheets received with 
reagents and stains to assist in determining proper handling requirements and 
safe use.
This checklist item does not apply to microwave devices that are designed by 
the manufacturer to operate without venting.
Evidence of Compliance:
✓ Records of annual evaluation of ventilation effectiveness



Joyce Weems
Pathology Manager
678-843-7376 Phone
678-843-7831 Fax
joyce.we...@emoryhealthcare.org



www.saintjosephsatlanta.org
5665 Peachtree Dunwoody Road
Atlanta, GA 30342

This e-mail, including any attachments is the property of Saint Joseph’s 
Hospital and is intended for the sole use of the intended recipient(s).  It may 
contain information that is privileged and confidential.  Any unauthorized 
review, use, disclosure, or distribution is prohibited. If you are not the