RE: [Histonet] RE: Instrument Verification

2014-04-04 Thread Miller, Suzie
I must say, the first time I read ANP.23045 I had the same reaction you all 
did. After reading it over a couple of times, we started looking at it 
differently. The question is not about validation--- it is about verifying that 
the equipment will function and perform as intended. It doesn't have anything 
to do with protocols and procedures for staining, cutting, processing, etc. If 
you look at any service report for service performed on your instrument, it is 
likely that the service engineer has some sort of notation on the report that 
states the unit was run and monitored for performance and meets all specs.
To comply with the CAP question we merely wrote a procedure for 
Instrument-Equipment Performance Verification stating that 1)The verification 
of instrument/equipment function will be performed by a qualified service 
engineer upon installation, after scheduled preventative maintenance, major 
instrument repairs, or relocation. 2) Use of the instrument/equipment will 
resume after the verification of its performance has been successfully 
completed.3) The service engineer will provide the Histology Laboratory with a 
report indicating that performance function was tested and satisfactory.
Of course, we also have all service reports (which include the notation that 
the function checks were performed and acceptable) available for each 
instrument.

Suzie Miller, MLT ASCP
Senior Histotech

Mercy Health System of Maine Laboratory

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
Sent: Thursday, April 03, 2014 7:52 PM
To: Elizabeth Chlipala; Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] RE: Instrument Verification

I do understand and sympathize with the situation in many clinical labs with 
staff , sometimes barely enough to do the work and it is challenging to keep up 
with expanding documentation also.  I would like to meet the GLP, but do 
struggle to be as extensive in my documentation. I do try to get as close to 
the ISO standards as possible, just to cover myself. I agree with Elizabeth's 
post that this seems to be the direction CAP has been heading over the years.
I think that if you get new instruments, methodology, technology they will 
certainly want to see the more robust documentation. For example ( see the 
current CAP today on IHC validation), this will surely be the guideline of 
tomorrow
But for those older, in long use instruments and technology,  my opinion is 
that if you have documentation in line with what the checklist stipulated when 
it went into use, and also all PM, maintenance, and QC- and have documented any 
corrective actions, this will probably fly for now? What does everyone else 
think?




Joelle Weaver MAOM, HTL (ASCP) QIHC

 From: l...@premierlab.com
 To: mw...@wakehealth.edu; robin...@mercyhealth.com;
 tbr...@holyredeemer.com; histonet@lists.utsouthwestern.edu
 Date: Thu, 3 Apr 2014 14:16:58 -0600
 CC:
 Subject: [Histonet] RE: Instrument Verification

 Hello All

 Coming from a GLP environment this type of equipment validation is standard 
 in our setting.  This is just my opinion but I think the CAP checklist is 
 moving towards the type of equipment documentation that is already required 
 in a GxP or ISO environment.  I always thought that instrument qualification 
 (IQ) - operational qualification (OQ) and process qualification (PQ) or 
 simply stated IQ/OQ/PQ were used only in GxP settings but you now see some of 
 the larger clinical labs running these types of validations on their 
 equipment and processes.  To me it does make sense that some type of 
 equipment validation should be required whether it  is a two page document on 
 the microtomes, waterbaths, etc. or complete IQ/OQ/PQ's  on major pieces of 
 equipment such as tissue processors, immunostainers and IHC retrieval units.  
 I believe that all of these are important processes that should be completed 
 in histology laboratories today.We are a GLP compliant lab and every 
 single piece of equipment is calibrated and validated as designated in our 
 Master Validation Plan.  IHC stainers and retrieval units should be 
 validated, even our refrigerators and freezers are calibrated and validated.  
 Our pipettors are calibrated quarterly, and any piece of equipment that 
 generates a weight or temperature is calibrated yearly.

 For example if you do not validate your IHC retrieval units how can you 
 really tell if they reach the temperature that they are programmed to reach, 
 does the temperature stay consistent through the retrieval process, did it 
 retrieve for the time programmed?  The only way to determine this is to 
 perform a validation.  How do you troubleshoot problems if you do not know if 
 your instruments are performing to their specification without testing those 
 specifications - that's what 

RE: [Histonet] RE: Instrument Verification

2014-04-04 Thread joelle weaver
Pretty much what is in place here for existing/established instruments. 




Joelle Weaver MAOM, HTL (ASCP) QIHC
 
 From: mill...@emhs.org
 To: joellewea...@hotmail.com; l...@premierlab.com; mw...@wakehealth.edu; 
 robin...@mercyhealth.com; tbr...@holyredeemer.com; 
 histonet@lists.utsouthwestern.edu
 Subject: RE: [Histonet] RE: Instrument Verification
 Date: Fri, 4 Apr 2014 10:08:54 +
 
 I must say, the first time I read ANP.23045 I had the same reaction you all 
 did. After reading it over a couple of times, we started looking at it 
 differently. The question is not about validation--- it is about verifying 
 that the equipment will function and perform as intended. It doesn't have 
 anything to do with protocols and procedures for staining, cutting, 
 processing, etc. If you look at any service report for service performed on 
 your instrument, it is likely that the service engineer has some sort of 
 notation on the report that states the unit was run and monitored for 
 performance and meets all specs.
 To comply with the CAP question we merely wrote a procedure for 
 Instrument-Equipment Performance Verification stating that 1)The verification 
 of instrument/equipment function will be performed by a qualified service 
 engineer upon installation, after scheduled preventative maintenance, major 
 instrument repairs, or relocation. 2) Use of the instrument/equipment will 
 resume after the verification of its performance has been successfully 
 completed.3) The service engineer will provide the Histology Laboratory with 
 a report indicating that performance function was tested and satisfactory.
 Of course, we also have all service reports (which include the notation that 
 the function checks were performed and acceptable) available for each 
 instrument.
 
 Suzie Miller, MLT ASCP
 Senior Histotech
 
 Mercy Health System of Maine Laboratory
 
 -Original Message-
 From: histonet-boun...@lists.utsouthwestern.edu 
 [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
 Sent: Thursday, April 03, 2014 7:52 PM
 To: Elizabeth Chlipala; Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
 histonet@lists.utsouthwestern.edu
 Subject: RE: [Histonet] RE: Instrument Verification
 
 I do understand and sympathize with the situation in many clinical labs with 
 staff , sometimes barely enough to do the work and it is challenging to keep 
 up with expanding documentation also.  I would like to meet the GLP, but do 
 struggle to be as extensive in my documentation. I do try to get as close to 
 the ISO standards as possible, just to cover myself. I agree with Elizabeth's 
 post that this seems to be the direction CAP has been heading over the years.
 I think that if you get new instruments, methodology, technology they will 
 certainly want to see the more robust documentation. For example ( see the 
 current CAP today on IHC validation), this will surely be the guideline of 
 tomorrow
 But for those older, in long use instruments and technology,  my opinion is 
 that if you have documentation in line with what the checklist stipulated 
 when it went into use, and also all PM, maintenance, and QC- and have 
 documented any corrective actions, this will probably fly for now? What 
 does everyone else think?
 
 
 
 
 Joelle Weaver MAOM, HTL (ASCP) QIHC
 
  From: l...@premierlab.com
  To: mw...@wakehealth.edu; robin...@mercyhealth.com;
  tbr...@holyredeemer.com; histonet@lists.utsouthwestern.edu
  Date: Thu, 3 Apr 2014 14:16:58 -0600
  CC:
  Subject: [Histonet] RE: Instrument Verification
 
  Hello All
 
  Coming from a GLP environment this type of equipment validation is standard 
  in our setting.  This is just my opinion but I think the CAP checklist is 
  moving towards the type of equipment documentation that is already required 
  in a GxP or ISO environment.  I always thought that instrument 
  qualification (IQ) - operational qualification (OQ) and process 
  qualification (PQ) or simply stated IQ/OQ/PQ were used only in GxP settings 
  but you now see some of the larger clinical labs running these types of 
  validations on their equipment and processes.  To me it does make sense 
  that some type of equipment validation should be required whether it  is a 
  two page document on the microtomes, waterbaths, etc. or complete 
  IQ/OQ/PQ's  on major pieces of equipment such as tissue processors, 
  immunostainers and IHC retrieval units.  I believe that all of these are 
  important processes that should be completed in histology laboratories 
  today.We are a GLP compliant lab and every single piece of equipment is 
  calibrated and validated as designated in our Master Validation Plan.  IHC 
  stainers and retrieval units should be validated, even our refrigerators 
  and freezers are calibrated and validated.  Our pipettors are calibrated 
  quarterly, and any piece of equipment that generates a weight or 
  temperature is calibrated yearly.
 
  For example if 

[Histonet] GI Biopsies

2014-04-04 Thread Campbell, Tasha M.
Does anyone have problems with their GI biopsies kind of shredding when
cutting?  I microwave process them in the Shurwave microwave processor.
I have been doing this for over a year and only the past couple months I
have noticed that almost every biopsy I cut has knife lines in it.  I
have to change my blade constantly.  I was told by a TBS technician that
if I ever noticed the tissue doing this, it meant that I did not have
all the alcohol out of the tissue, but does that sound like it would
cause the tissue to shred like that?  I don't use my reagents more than
2 times and I have even tried increasing the time and temp a little but
no help.  I have been doing them for a year and did not have a problem
and I haven't changed anything.  Any solutions or suggestions???
Thanks!

 

 

 

 

Tasha Campbell, B.S.,HTL(ASCP)

Frederick Gastroenterology Associates

310 W. 9th St.

Frederick, MD 21701

301-695-6800 ext. 144 (w)

304-685-9307 (c)

 

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] PFK's on muscle biopsies

2014-04-04 Thread Celebre Julia
Happy Friday !!

For years now our PFK staining has been inconsistent and I'm hoping someone 
wouldn't mind sharing their no fail PFK method so I can make my neuropath's 
happy again.??

Julia Celebre   Sr MLT
Anatomic Pathology
Hamilton Civic Hospitals
905-527-4322  ext 46179


This information is directed in confidence solely to the person named above and 
may not otherwise be distributed, copied or disclosed. Therefore, this 
information should be considered strictly confidential. If you have received 
this email in error, please notify the sender immediately via a return email 
for further direction. Thank you for your assistance.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] anti-FLAG antibody

2014-04-04 Thread Teri Johnson
Happy Friday!

I am looking for a source of anti-FLAG antibody that works in mouse FFPE 
tissues.
Bonus points if it work with Ventana/Roche instrumentation.

Teri Johnson
Manager, Histology
Genomics Institute for
Novartis Research
Foundation
San Diego, CA
858-332-4752

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] 45th Annual Illinois Society for Histotechnologists Meeting May 15 16

2014-04-04 Thread Lori Bellafiore
Hello,

It is time to register for the 45th Annual ISH SPring Symposium in
Bloomington, IL May 15  16, 2014.

Please see the following website for complete meeting details and to
register online.

http://www.illinoishistologysociety.org/meetings.html

THank you!
-- 
Lori Ann Bellafiore
*Pathology Education Specialist*
*Genentech*
217.508.9975
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Ventana and CoPath Inface (HL7 Stain Order Interface)

2014-04-04 Thread Vickroy, Jim
We currently use the ABT system of CoPath Plus and have three Ventana Ultra 
Stainers and two Benchmark Special Stainers.   We have been told that the HL7 
Stain Order Interface will save us a lot of clerical time.  We are trying to 
contact someone that already has this interface to see if it is worth the 
money.  Please let us know if you have the interface and your experience with 
it.  If possible maybe we could contact you and get some feedback.

Jim V

James Vickroy BS, HT(ASCP)

Surgical  and Autopsy Pathology Technical Supervisor
Memorial Medical Center
217-788-4046



This message (including any attachments) contains confidential information 
intended for a specific individual and purpose, and is protected by law. If you 
are not the intended recipient, you should delete this message. Any disclosure, 
copying, or distribution of this message, or the taking of any action based on 
it, is strictly prohibited.
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] Biocare intellipath

2014-04-04 Thread Deloris Carter
I'm looking for some honest feedback from anyone using the Biocare
Intellipath in a clinical setting. Pros, cons, etc.
Thanks,
Deloris Carter, HT(ASCP)
dels...@gmail.com
___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet