[Histonet] RE: Instrument Verification

2014-04-07 Thread Michael LaFriniere
I agree with Elizabeth, and do exactly what she outlines and I am a routine 
large AP lab.  What a great suggestion for classes at the State, Regional and 
NSH meetings! It's a great plug and well worth the mentioning Elizabeth! This 
is something everyone can benefit from the bench tech all the way up the ladder!

Appropriate Methodologies to Validate instrumentation is a much needed topic!

Michael
Michael R. LaFriniere, HT (ASCP) 
Executive Director
 

Capital Choice Pathology Laboratory
12041 Bournefield Way, Suite A . Silver Spring, MD 20904  
P: 240.471.3427 . F: 240.471.3401 . Cell 410-940-8844
michael.lafrini...@ccplab.com
 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Elizabeth 
Chlipala
Sent: Thursday, April 03, 2014 4:17 PM
To: Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

Hello All

Coming from a GLP environment this type of equipment validation is standard in 
our setting.  This is just my opinion but I think the CAP checklist is moving 
towards the type of equipment documentation that is already required in a GxP 
or ISO environment.  I always thought that instrument qualification (IQ) - 
operational qualification (OQ) and process qualification (PQ) or simply stated 
IQ/OQ/PQ were used only in GxP settings but you now see some of the larger 
clinical labs running these types of validations on their equipment and 
processes.  To me it does make sense that some type of equipment validation 
should be required whether it  is a two page document on the microtomes, 
waterbaths, etc. or complete IQ/OQ/PQ's  on major pieces of equipment such as 
tissue processors, immunostainers and IHC retrieval units.  I believe that all 
of these are important processes that should be completed in histology 
laboratories today.We are a GLP compliant lab and every single piece of 
equipment is calibrated and validated as designated in our Master Validation 
Plan.  IHC stainers and retrieval units should be validated, even our 
refrigerators and freezers are calibrated and validated.  Our pipettors are 
calibrated quarterly, and any piece of equipment that generates a weight or 
temperature is calibrated yearly.  

For example if you do not validate your IHC retrieval units how can you really 
tell if they reach the temperature that they are programmed to reach, does the 
temperature stay consistent through the retrieval process, did it retrieve for 
the time programmed?  The only way to determine this is to perform a 
validation.  How do you troubleshoot problems if you do not know if your 
instruments are performing to their specification without testing those 
specifications - that's what equipment validation is and that's why in my 
opinion its important.  

Histology laboratories are now responsible for running IHC that directly 
effects a patients treatment - meaning the numerous therapeutic and prognostic 
markers we routinely run now.  Validation is an important process especially if 
you are using image analysis for these markers.  I hate to say it but we better 
get used to it, because this is not going away.  

And now the shameless plug -  I will be giving a 90 minute lecture at the 
Florida State Meeting  
https://classic.regonline.com/custImages/24/241449/FSH2014OnlineProgram.pdf 
on this exact topic, so if you want to learn how to create a Master Validation 
Plan and learn how to perform a basic validation or a more detailed IQ/OQ and 
PQ and to what extent you need to validate a particular piece of equipment -  
sign up for the meeting plus there are lots of other great topics being 
presented too.

Liz 

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Premier Laboratory, LLC
PO Box 18592
Boulder, CO 80308
(303) 682-3949 office
(303) 682-9060 fax
(303) 881-0763 cell
l...@premierlab.com
www.premierlab.com

March 10, 2014 is Histotechnology Professionals Day

Ship to Address:

Premier Laboratory, LLC
1567 Skyway Drive, Unit E
Longmont, CO 80504


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Martha 
Ward-Pathology
Sent: Thursday, April 03, 2014 1:22 PM
To: Cynthia Robinson; Terri Braud; histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

I'm with you.   There really appears to be no value to this particular 
requirement.I would only be concerned with it if I had just purchased it, 
or moved it into our lab from another location.


 
Martha Ward, MT (ASCP) QIHC
Manager

Molecular Diagnostics Lab
Medical Center Boulevard  \  Winston-Salem, NC 27157 p 336.716.2109  \  f 
336.716.5890 mw...@wakehealth.edu  
 
 



-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Cynthia Robinson
Sent: Thursday, April 03

[Histonet] RE: Instrument Verification

2014-04-07 Thread Elizabeth Chlipala
The NSH also has a teleconference this week on the same topic.

NSH Webinar: The Basics of Verification and Validation
Date: 
10 Apr 2014
Contact Info: 
NSH Office: 443-535-4060 or hi...@nsh.org

The Basics of Verification and Validation (Quality Management Series)
Presented by: Kathy Dwyer, HT(ASCP), Quest Diagnostics  Debbie Siena, 
HT(ASCP)QIHC, Statlab Medical Products

CAP has redefined verification and validation in the 2013 CAP checklist . This 
webinar will discuss the new CAP definitions and how it affects the laboratory. 
Creating a laboratory verification and validation process will ensure accurate 
and reproducible results daily and during changes to methods and/or processes. 
A standardized verification and validation process will ensure excellent 
patient care and compliance with regulatory agencies.

This webinar will guide the attendees on how to develop a new process for 
verifying and validating equipment and instrumentation.

http://www.nsh.org/content/nsh-webinar-basics-verification-and-validation

Liz

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Premier Laboratory, LLC
PO Box 18592
Boulder, CO 80308
(303) 682-3949 office
(303) 682-9060 fax
(303) 881-0763 cell
l...@premierlab.com
www.premierlab.com

March 10, 2014 is Histotechnology Professionals Day

Ship to Address:

Premier Laboratory, LLC
1567 Skyway Drive, Unit E
Longmont, CO 80504


-Original Message-
From: Michael LaFriniere [mailto:michael.lafrini...@ccplab.com] 
Sent: Monday, April 07, 2014 11:42 AM
To: Elizabeth Chlipala; Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
histonet@lists.utsouthwestern.edu
Subject: RE: Instrument Verification

I agree with Elizabeth, and do exactly what she outlines and I am a routine 
large AP lab.  What a great suggestion for classes at the State, Regional and 
NSH meetings! It's a great plug and well worth the mentioning Elizabeth! This 
is something everyone can benefit from the bench tech all the way up the ladder!

Appropriate Methodologies to Validate instrumentation is a much needed topic!

Michael
Michael R. LaFriniere, HT (ASCP) 
Executive Director
 

Capital Choice Pathology Laboratory
12041 Bournefield Way, Suite A . Silver Spring, MD 20904  
P: 240.471.3427 . F: 240.471.3401 . Cell 410-940-8844
michael.lafrini...@ccplab.com
 

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Elizabeth 
Chlipala
Sent: Thursday, April 03, 2014 4:17 PM
To: Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

Hello All

Coming from a GLP environment this type of equipment validation is standard in 
our setting.  This is just my opinion but I think the CAP checklist is moving 
towards the type of equipment documentation that is already required in a GxP 
or ISO environment.  I always thought that instrument qualification (IQ) - 
operational qualification (OQ) and process qualification (PQ) or simply stated 
IQ/OQ/PQ were used only in GxP settings but you now see some of the larger 
clinical labs running these types of validations on their equipment and 
processes.  To me it does make sense that some type of equipment validation 
should be required whether it  is a two page document on the microtomes, 
waterbaths, etc. or complete IQ/OQ/PQ's  on major pieces of equipment such as 
tissue processors, immunostainers and IHC retrieval units.  I believe that all 
of these are important processes that should be completed in histology 
laboratories today.We are a GLP compliant lab and every single piece of 
equipment is calibrated and validated as designated in our Master Validation 
Plan.  IHC stainers and retrieval units should be validated, even our 
refrigerators and freezers are calibrated and validated.  Our pipettors are 
calibrated quarterly, and any piece of equipment that generates a weight or 
temperature is calibrated yearly.  

For example if you do not validate your IHC retrieval units how can you really 
tell if they reach the temperature that they are programmed to reach, does the 
temperature stay consistent through the retrieval process, did it retrieve for 
the time programmed?  The only way to determine this is to perform a 
validation.  How do you troubleshoot problems if you do not know if your 
instruments are performing to their specification without testing those 
specifications - that's what equipment validation is and that's why in my 
opinion its important.  

Histology laboratories are now responsible for running IHC that directly 
effects a patients treatment - meaning the numerous therapeutic and prognostic 
markers we routinely run now.  Validation is an important process especially if 
you are using image analysis for these markers.  I hate to say it but we better 
get used to it, because this is not going away.  

And now the shameless plug -  I will be giving a 90 minute lecture

RE: [Histonet] RE: Instrument Verification

2014-04-04 Thread Miller, Suzie
I must say, the first time I read ANP.23045 I had the same reaction you all 
did. After reading it over a couple of times, we started looking at it 
differently. The question is not about validation--- it is about verifying that 
the equipment will function and perform as intended. It doesn't have anything 
to do with protocols and procedures for staining, cutting, processing, etc. If 
you look at any service report for service performed on your instrument, it is 
likely that the service engineer has some sort of notation on the report that 
states the unit was run and monitored for performance and meets all specs.
To comply with the CAP question we merely wrote a procedure for 
Instrument-Equipment Performance Verification stating that 1)The verification 
of instrument/equipment function will be performed by a qualified service 
engineer upon installation, after scheduled preventative maintenance, major 
instrument repairs, or relocation. 2) Use of the instrument/equipment will 
resume after the verification of its performance has been successfully 
completed.3) The service engineer will provide the Histology Laboratory with a 
report indicating that performance function was tested and satisfactory.
Of course, we also have all service reports (which include the notation that 
the function checks were performed and acceptable) available for each 
instrument.

Suzie Miller, MLT ASCP
Senior Histotech

Mercy Health System of Maine Laboratory

-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
Sent: Thursday, April 03, 2014 7:52 PM
To: Elizabeth Chlipala; Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
histonet@lists.utsouthwestern.edu
Subject: RE: [Histonet] RE: Instrument Verification

I do understand and sympathize with the situation in many clinical labs with 
staff , sometimes barely enough to do the work and it is challenging to keep up 
with expanding documentation also.  I would like to meet the GLP, but do 
struggle to be as extensive in my documentation. I do try to get as close to 
the ISO standards as possible, just to cover myself. I agree with Elizabeth's 
post that this seems to be the direction CAP has been heading over the years.
I think that if you get new instruments, methodology, technology they will 
certainly want to see the more robust documentation. For example ( see the 
current CAP today on IHC validation), this will surely be the guideline of 
tomorrow
But for those older, in long use instruments and technology,  my opinion is 
that if you have documentation in line with what the checklist stipulated when 
it went into use, and also all PM, maintenance, and QC- and have documented any 
corrective actions, this will probably fly for now? What does everyone else 
think?




Joelle Weaver MAOM, HTL (ASCP) QIHC

 From: l...@premierlab.com
 To: mw...@wakehealth.edu; robin...@mercyhealth.com;
 tbr...@holyredeemer.com; histonet@lists.utsouthwestern.edu
 Date: Thu, 3 Apr 2014 14:16:58 -0600
 CC:
 Subject: [Histonet] RE: Instrument Verification

 Hello All

 Coming from a GLP environment this type of equipment validation is standard 
 in our setting.  This is just my opinion but I think the CAP checklist is 
 moving towards the type of equipment documentation that is already required 
 in a GxP or ISO environment.  I always thought that instrument qualification 
 (IQ) - operational qualification (OQ) and process qualification (PQ) or 
 simply stated IQ/OQ/PQ were used only in GxP settings but you now see some of 
 the larger clinical labs running these types of validations on their 
 equipment and processes.  To me it does make sense that some type of 
 equipment validation should be required whether it  is a two page document on 
 the microtomes, waterbaths, etc. or complete IQ/OQ/PQ's  on major pieces of 
 equipment such as tissue processors, immunostainers and IHC retrieval units.  
 I believe that all of these are important processes that should be completed 
 in histology laboratories today.We are a GLP compliant lab and every 
 single piece of equipment is calibrated and validated as designated in our 
 Master Validation Plan.  IHC stainers and retrieval units should be 
 validated, even our refrigerators and freezers are calibrated and validated.  
 Our pipettors are calibrated quarterly, and any piece of equipment that 
 generates a weight or temperature is calibrated yearly.

 For example if you do not validate your IHC retrieval units how can you 
 really tell if they reach the temperature that they are programmed to reach, 
 does the temperature stay consistent through the retrieval process, did it 
 retrieve for the time programmed?  The only way to determine this is to 
 perform a validation.  How do you troubleshoot problems if you do not know if 
 your instruments are performing to their specification without testing those 
 specifications - that's what

RE: [Histonet] RE: Instrument Verification

2014-04-04 Thread joelle weaver
Pretty much what is in place here for existing/established instruments. 




Joelle Weaver MAOM, HTL (ASCP) QIHC
 
 From: mill...@emhs.org
 To: joellewea...@hotmail.com; l...@premierlab.com; mw...@wakehealth.edu; 
 robin...@mercyhealth.com; tbr...@holyredeemer.com; 
 histonet@lists.utsouthwestern.edu
 Subject: RE: [Histonet] RE: Instrument Verification
 Date: Fri, 4 Apr 2014 10:08:54 +
 
 I must say, the first time I read ANP.23045 I had the same reaction you all 
 did. After reading it over a couple of times, we started looking at it 
 differently. The question is not about validation--- it is about verifying 
 that the equipment will function and perform as intended. It doesn't have 
 anything to do with protocols and procedures for staining, cutting, 
 processing, etc. If you look at any service report for service performed on 
 your instrument, it is likely that the service engineer has some sort of 
 notation on the report that states the unit was run and monitored for 
 performance and meets all specs.
 To comply with the CAP question we merely wrote a procedure for 
 Instrument-Equipment Performance Verification stating that 1)The verification 
 of instrument/equipment function will be performed by a qualified service 
 engineer upon installation, after scheduled preventative maintenance, major 
 instrument repairs, or relocation. 2) Use of the instrument/equipment will 
 resume after the verification of its performance has been successfully 
 completed.3) The service engineer will provide the Histology Laboratory with 
 a report indicating that performance function was tested and satisfactory.
 Of course, we also have all service reports (which include the notation that 
 the function checks were performed and acceptable) available for each 
 instrument.
 
 Suzie Miller, MLT ASCP
 Senior Histotech
 
 Mercy Health System of Maine Laboratory
 
 -Original Message-
 From: histonet-boun...@lists.utsouthwestern.edu 
 [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of joelle weaver
 Sent: Thursday, April 03, 2014 7:52 PM
 To: Elizabeth Chlipala; Martha Ward-Pathology; Cynthia Robinson; Terri Braud; 
 histonet@lists.utsouthwestern.edu
 Subject: RE: [Histonet] RE: Instrument Verification
 
 I do understand and sympathize with the situation in many clinical labs with 
 staff , sometimes barely enough to do the work and it is challenging to keep 
 up with expanding documentation also.  I would like to meet the GLP, but do 
 struggle to be as extensive in my documentation. I do try to get as close to 
 the ISO standards as possible, just to cover myself. I agree with Elizabeth's 
 post that this seems to be the direction CAP has been heading over the years.
 I think that if you get new instruments, methodology, technology they will 
 certainly want to see the more robust documentation. For example ( see the 
 current CAP today on IHC validation), this will surely be the guideline of 
 tomorrow
 But for those older, in long use instruments and technology,  my opinion is 
 that if you have documentation in line with what the checklist stipulated 
 when it went into use, and also all PM, maintenance, and QC- and have 
 documented any corrective actions, this will probably fly for now? What 
 does everyone else think?
 
 
 
 
 Joelle Weaver MAOM, HTL (ASCP) QIHC
 
  From: l...@premierlab.com
  To: mw...@wakehealth.edu; robin...@mercyhealth.com;
  tbr...@holyredeemer.com; histonet@lists.utsouthwestern.edu
  Date: Thu, 3 Apr 2014 14:16:58 -0600
  CC:
  Subject: [Histonet] RE: Instrument Verification
 
  Hello All
 
  Coming from a GLP environment this type of equipment validation is standard 
  in our setting.  This is just my opinion but I think the CAP checklist is 
  moving towards the type of equipment documentation that is already required 
  in a GxP or ISO environment.  I always thought that instrument 
  qualification (IQ) - operational qualification (OQ) and process 
  qualification (PQ) or simply stated IQ/OQ/PQ were used only in GxP settings 
  but you now see some of the larger clinical labs running these types of 
  validations on their equipment and processes.  To me it does make sense 
  that some type of equipment validation should be required whether it  is a 
  two page document on the microtomes, waterbaths, etc. or complete 
  IQ/OQ/PQ's  on major pieces of equipment such as tissue processors, 
  immunostainers and IHC retrieval units.  I believe that all of these are 
  important processes that should be completed in histology laboratories 
  today.We are a GLP compliant lab and every single piece of equipment is 
  calibrated and validated as designated in our Master Validation Plan.  IHC 
  stainers and retrieval units should be validated, even our refrigerators 
  and freezers are calibrated and validated.  Our pipettors are calibrated 
  quarterly, and any piece of equipment that generates a weight or 
  temperature is calibrated yearly.
 
  For example

[Histonet] RE: Instrument Verification

2014-04-03 Thread Terri Braud
I just received my midcycle CAP and for cryostat validation, we are
planning to cut and stain a piece of frozen tonsil and have the path
sign off on it. For the tissue processors, we will run a one minute test
program. I hope this will fly.  Is it just me, or is CAP insanely out of
control with new or modified regulations and policies for AP?

Terri L. Braud, HT(ASCP)
Anatomic Pathology Supervisor
Holy Redeemer Hospital Laboratory
1648 Huntingdon Pike
Meadowbrook, PA 19046
Ph: 215-938-3676
Fax: 215-938-3874

6. Validation of cryostat (Gloria Tharp)

Message: 6
Date: Thu, 3 Apr 2014 09:59:26 -0500
From: Gloria Tharp gth...@pcasoutheast.com
Could anyone tell me how you are handling the new CAP ANP.23045 question
on function and verification of equipment regarding a cryostat.
Gloria Tharp, BA, HTL(ASCP)
--

Message: 7
Date: Thu, 3 Apr 2014 15:26:17 +
From: Leann M. Murphy lmurp...@aultman.com
How is everyone validating the tissue processor for new CAP ANP.23045
question on function and verification of equipment?
LeAnn Murphy
Aultman Hospital
Canton, Ohio

-



CONFIDENTIALITY NOTICE:

This E-Mail is intended only for the use of the individual or entity to which
it was sent. It may contain information that is privileged and/or confidential,
and the use or disclosure of such information may also be restricted under 
applicable
federal and state law. If you received this communication in error, please do 
not
distribute any part of it or retain any copies, and delete the original E-Mail.
Please notify the sender of any error by E-Mail.

Thank you for your cooperation.


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Instrument Verification

2014-04-03 Thread Cynthia Robinson
I agree with you in that CAP is just looking for things to change and doesn't 
seem to be considering the change and decrease in staffing seen in clinical 
settings. Cryostat validation? Reallycut a slide after you have cleaned and 
pm'd the thing and go on. Good grief...I don't need any more paper and 
documentation on routine processes. As for tissue processors, I have 20 year 
old VIP's that have been running and producing specimens acceptably. I did 
validate them prior to being put in use but we didn't document like we do now. 
And I don't see the need to do it at this stage of use. We did do a very 
extensive validation on the Peloris we put into use last year and will going 
forward on new equipment.  To me the daily QC of stain should provide our 
'validation' of the process and include the processor. I am interested in 
others thoughts as well.

Thanks for allowing me to rant.

Cindi Robinson, HT(ASCP)
Mercy Medical Center-Sioux City
Dunes Medical Laboratories
350 W Anchor Drive
Dakota Dunes SD 57049




-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Terri Braud
Sent: Thursday, April 03, 2014 1:00 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

I just received my midcycle CAP and for cryostat validation, we are planning to 
cut and stain a piece of frozen tonsil and have the path sign off on it. For 
the tissue processors, we will run a one minute test program. I hope this will 
fly.  Is it just me, or is CAP insanely out of control with new or modified 
regulations and policies for AP?  

Terri L. Braud, HT(ASCP)
Anatomic Pathology Supervisor
Holy Redeemer Hospital Laboratory
1648 Huntingdon Pike
Meadowbrook, PA 19046
Ph: 215-938-3676
Fax: 215-938-3874

6. Validation of cryostat (Gloria Tharp)

Message: 6
Date: Thu, 3 Apr 2014 09:59:26 -0500
From: Gloria Tharp gth...@pcasoutheast.com Could anyone tell me how you are 
handling the new CAP ANP.23045 question on function and verification of 
equipment regarding a cryostat.
Gloria Tharp, BA, HTL(ASCP)
--

Message: 7
Date: Thu, 3 Apr 2014 15:26:17 +
From: Leann M. Murphy lmurp...@aultman.com How is everyone validating the 
tissue processor for new CAP ANP.23045 question on function and verification of 
equipment?
LeAnn Murphy
Aultman Hospital
Canton, Ohio

-



CONFIDENTIALITY NOTICE:

This E-Mail is intended only for the use of the individual or entity to which 
it was sent. It may contain information that is privileged and/or confidential, 
and the use or disclosure of such information may also be restricted under 
applicable federal and state law. If you received this communication in error, 
please do not distribute any part of it or retain any copies, and delete the 
original E-Mail.
Please notify the sender of any error by E-Mail.

Thank you for your cooperation.


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Instrument Verification

2014-04-03 Thread Martha Ward-Pathology
I'm with you.   There really appears to be no value to this particular 
requirement.I would only be concerned with it if I had just purchased it, 
or moved it into our lab from another location.


 
Martha Ward, MT (ASCP) QIHC
Manager

Molecular Diagnostics Lab
Medical Center Boulevard  \  Winston-Salem, NC 27157
p 336.716.2109  \  f 336.716.5890  
mw...@wakehealth.edu  
 
 



-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Cynthia Robinson
Sent: Thursday, April 03, 2014 3:03 PM
To: Terri Braud; histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

I agree with you in that CAP is just looking for things to change and doesn't 
seem to be considering the change and decrease in staffing seen in clinical 
settings. Cryostat validation? Reallycut a slide after you have cleaned and 
pm'd the thing and go on. Good grief...I don't need any more paper and 
documentation on routine processes. As for tissue processors, I have 20 year 
old VIP's that have been running and producing specimens acceptably. I did 
validate them prior to being put in use but we didn't document like we do now. 
And I don't see the need to do it at this stage of use. We did do a very 
extensive validation on the Peloris we put into use last year and will going 
forward on new equipment.  To me the daily QC of stain should provide our 
'validation' of the process and include the processor. I am interested in 
others thoughts as well.

Thanks for allowing me to rant.

Cindi Robinson, HT(ASCP)
Mercy Medical Center-Sioux City
Dunes Medical Laboratories
350 W Anchor Drive
Dakota Dunes SD 57049




-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Terri Braud
Sent: Thursday, April 03, 2014 1:00 PM
To: histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

I just received my midcycle CAP and for cryostat validation, we are planning to 
cut and stain a piece of frozen tonsil and have the path sign off on it. For 
the tissue processors, we will run a one minute test program. I hope this will 
fly.  Is it just me, or is CAP insanely out of control with new or modified 
regulations and policies for AP?  

Terri L. Braud, HT(ASCP)
Anatomic Pathology Supervisor
Holy Redeemer Hospital Laboratory
1648 Huntingdon Pike
Meadowbrook, PA 19046
Ph: 215-938-3676
Fax: 215-938-3874

6. Validation of cryostat (Gloria Tharp)

Message: 6
Date: Thu, 3 Apr 2014 09:59:26 -0500
From: Gloria Tharp gth...@pcasoutheast.com Could anyone tell me how you are 
handling the new CAP ANP.23045 question on function and verification of 
equipment regarding a cryostat.
Gloria Tharp, BA, HTL(ASCP)
--

Message: 7
Date: Thu, 3 Apr 2014 15:26:17 +
From: Leann M. Murphy lmurp...@aultman.com How is everyone validating the 
tissue processor for new CAP ANP.23045 question on function and verification of 
equipment?
LeAnn Murphy
Aultman Hospital
Canton, Ohio

-



CONFIDENTIALITY NOTICE:

This E-Mail is intended only for the use of the individual or entity to which 
it was sent. It may contain information that is privileged and/or confidential, 
and the use or disclosure of such information may also be restricted under 
applicable federal and state law. If you received this communication in error, 
please do not distribute any part of it or retain any copies, and delete the 
original E-Mail.
Please notify the sender of any error by E-Mail.

Thank you for your cooperation.


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet

___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


___
Histonet mailing list
Histonet@lists.utsouthwestern.edu
http://lists.utsouthwestern.edu/mailman/listinfo/histonet


[Histonet] RE: Instrument Verification

2014-04-03 Thread Elizabeth Chlipala
Hello All

Coming from a GLP environment this type of equipment validation is standard in 
our setting.  This is just my opinion but I think the CAP checklist is moving 
towards the type of equipment documentation that is already required in a GxP 
or ISO environment.  I always thought that instrument qualification (IQ) - 
operational qualification (OQ) and process qualification (PQ) or simply stated 
IQ/OQ/PQ were used only in GxP settings but you now see some of the larger 
clinical labs running these types of validations on their equipment and 
processes.  To me it does make sense that some type of equipment validation 
should be required whether it  is a two page document on the microtomes, 
waterbaths, etc. or complete IQ/OQ/PQ's  on major pieces of equipment such as 
tissue processors, immunostainers and IHC retrieval units.  I believe that all 
of these are important processes that should be completed in histology 
laboratories today.We are a GLP compliant lab and every single piece of 
equipment is calibrated and validated as designated in our Master Validation 
Plan.  IHC stainers and retrieval units should be validated, even our 
refrigerators and freezers are calibrated and validated.  Our pipettors are 
calibrated quarterly, and any piece of equipment that generates a weight or 
temperature is calibrated yearly.  

For example if you do not validate your IHC retrieval units how can you really 
tell if they reach the temperature that they are programmed to reach, does the 
temperature stay consistent through the retrieval process, did it retrieve for 
the time programmed?  The only way to determine this is to perform a 
validation.  How do you troubleshoot problems if you do not know if your 
instruments are performing to their specification without testing those 
specifications - that's what equipment validation is and that's why in my 
opinion its important.  

Histology laboratories are now responsible for running IHC that directly 
effects a patients treatment - meaning the numerous therapeutic and prognostic 
markers we routinely run now.  Validation is an important process especially if 
you are using image analysis for these markers.  I hate to say it but we better 
get used to it, because this is not going away.  

And now the shameless plug -  I will be giving a 90 minute lecture at the 
Florida State Meeting  
https://classic.regonline.com/custImages/24/241449/FSH2014OnlineProgram.pdf 
on this exact topic, so if you want to learn how to create a Master Validation 
Plan and learn how to perform a basic validation or a more detailed IQ/OQ and 
PQ and to what extent you need to validate a particular piece of equipment -  
sign up for the meeting plus there are lots of other great topics being 
presented too.

Liz 

Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
Premier Laboratory, LLC
PO Box 18592
Boulder, CO 80308
(303) 682-3949 office
(303) 682-9060 fax
(303) 881-0763 cell
l...@premierlab.com
www.premierlab.com

March 10, 2014 is Histotechnology Professionals Day

Ship to Address:

Premier Laboratory, LLC
1567 Skyway Drive, Unit E
Longmont, CO 80504


-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Martha 
Ward-Pathology
Sent: Thursday, April 03, 2014 1:22 PM
To: Cynthia Robinson; Terri Braud; histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

I'm with you.   There really appears to be no value to this particular 
requirement.I would only be concerned with it if I had just purchased it, 
or moved it into our lab from another location.


 
Martha Ward, MT (ASCP) QIHC
Manager

Molecular Diagnostics Lab
Medical Center Boulevard  \  Winston-Salem, NC 27157 p 336.716.2109  \  f 
336.716.5890 mw...@wakehealth.edu  
 
 



-Original Message-
From: histonet-boun...@lists.utsouthwestern.edu 
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Cynthia Robinson
Sent: Thursday, April 03, 2014 3:03 PM
To: Terri Braud; histonet@lists.utsouthwestern.edu
Subject: [Histonet] RE: Instrument Verification

I agree with you in that CAP is just looking for things to change and doesn't 
seem to be considering the change and decrease in staffing seen in clinical 
settings. Cryostat validation? Reallycut a slide after you have cleaned and 
pm'd the thing and go on. Good grief...I don't need any more paper and 
documentation on routine processes. As for tissue processors, I have 20 year 
old VIP's that have been running and producing specimens acceptably. I did 
validate them prior to being put in use but we didn't document like we do now. 
And I don't see the need to do it at this stage of use. We did do a very 
extensive validation on the Peloris we put into use last year and will going 
forward on new equipment.  To me the daily QC of stain should provide our 
'validation' of the process and include the processor. I am interested in 
others thoughts as well

RE: [Histonet] RE: Instrument Verification

2014-04-03 Thread joelle weaver
I do understand and sympathize with the situation in many clinical labs with 
staff , sometimes barely enough to do the work and it is challenging to keep up 
with expanding documentation also.  I would like to meet the GLP, but do 
struggle to be as extensive in my documentation. I do try to get as close to 
the ISO standards as possible, just to cover myself. I agree with Elizabeth's 
post that this seems to be the direction CAP has been heading over the years.  
I think that if you get new instruments, methodology, technology they will 
certainly want to see the more robust documentation. For example ( see the 
current CAP today on IHC validation), this will surely be the guideline of 
tomorrow
But for those older, in long use instruments and technology,  my opinion is 
that if you have documentation in line with what the checklist stipulated when 
it went into use, and also all PM, maintenance, and QC- and have documented any 
corrective actions, this will probably fly for now? What does everyone else 
think? 




Joelle Weaver MAOM, HTL (ASCP) QIHC
 
 From: l...@premierlab.com
 To: mw...@wakehealth.edu; robin...@mercyhealth.com; tbr...@holyredeemer.com; 
 histonet@lists.utsouthwestern.edu
 Date: Thu, 3 Apr 2014 14:16:58 -0600
 CC: 
 Subject: [Histonet] RE: Instrument Verification
 
 Hello All
 
 Coming from a GLP environment this type of equipment validation is standard 
 in our setting.  This is just my opinion but I think the CAP checklist is 
 moving towards the type of equipment documentation that is already required 
 in a GxP or ISO environment.  I always thought that instrument qualification 
 (IQ) - operational qualification (OQ) and process qualification (PQ) or 
 simply stated IQ/OQ/PQ were used only in GxP settings but you now see some of 
 the larger clinical labs running these types of validations on their 
 equipment and processes.  To me it does make sense that some type of 
 equipment validation should be required whether it  is a two page document on 
 the microtomes, waterbaths, etc. or complete IQ/OQ/PQ's  on major pieces of 
 equipment such as tissue processors, immunostainers and IHC retrieval units.  
 I believe that all of these are important processes that should be completed 
 in histology laboratories today.We are a GLP compliant lab and every 
 single piece of equipment is calibrated and validated as designated in our 
 Master Validation Plan.  IHC stainers and retrieval units should be 
 validated, even our refrigerators and freezers are calibrated and validated.  
 Our pipettors are calibrated quarterly, and any piece of equipment that 
 generates a weight or temperature is calibrated yearly.  
 
 For example if you do not validate your IHC retrieval units how can you 
 really tell if they reach the temperature that they are programmed to reach, 
 does the temperature stay consistent through the retrieval process, did it 
 retrieve for the time programmed?  The only way to determine this is to 
 perform a validation.  How do you troubleshoot problems if you do not know if 
 your instruments are performing to their specification without testing those 
 specifications - that's what equipment validation is and that's why in my 
 opinion its important.  
 
 Histology laboratories are now responsible for running IHC that directly 
 effects a patients treatment - meaning the numerous therapeutic and 
 prognostic markers we routinely run now.  Validation is an important process 
 especially if you are using image analysis for these markers.  I hate to say 
 it but we better get used to it, because this is not going away.  
 
 And now the shameless plug -  I will be giving a 90 minute lecture at the 
 Florida State Meeting  
 https://classic.regonline.com/custImages/24/241449/FSH2014OnlineProgram.pdf
  on this exact topic, so if you want to learn how to create a Master 
 Validation Plan and learn how to perform a basic validation or a more 
 detailed IQ/OQ and PQ and to what extent you need to validate a particular 
 piece of equipment -  sign up for the meeting plus there are lots of other 
 great topics being presented too.
 
 Liz 
 
 Elizabeth A. Chlipala, BS, HTL(ASCP)QIHC
 Premier Laboratory, LLC
 PO Box 18592
 Boulder, CO 80308
 (303) 682-3949 office
 (303) 682-9060 fax
 (303) 881-0763 cell
 l...@premierlab.com
 www.premierlab.com
 
 March 10, 2014 is Histotechnology Professionals Day
 
 Ship to Address:
 
 Premier Laboratory, LLC
 1567 Skyway Drive, Unit E
 Longmont, CO 80504
 
 
 -Original Message-
 From: histonet-boun...@lists.utsouthwestern.edu 
 [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Martha 
 Ward-Pathology
 Sent: Thursday, April 03, 2014 1:22 PM
 To: Cynthia Robinson; Terri Braud; histonet@lists.utsouthwestern.edu
 Subject: [Histonet] RE: Instrument Verification
 
 I'm with you.   There really appears to be no value to this particular 
 requirement.I would only be concerned with it if I had just purchased it, 
 or moved