Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-15 Thread Teck Kwang Choo
Hello again. Just want your opinion on a thought that I have. Due to the
absence of Mg-ferrite (MgFe2O4) K-edge XANES spectrum, I have decided to do
my Fe(II)/Fe(III) quantification with only maghemite and magnetite as
end-member standards.

Because Fe in MgFe2O4 exists entirely as Fe(III) and its structure is the
same as both magnetite and maghemite (inverse-spinel structure with Fe
(III) occupying both tetrahedral and octahedral sites), I am inferring that
its K-edge XANES spectrum would be similar to that of maghemite.

I am planning to take the quantified fraction of maghemite as a sum of the
contributions of maghemite and Mg-ferrite.

For instance, from Linear Combination Fitting:

Fraction of Magnetite=0.05, Fraction of Maghemite = 0.95;
Due to absence of Mg-ferrite spectrum, i will be taking (Fraction of
Maghemite) + (Fraction of Mg-ferrite)= 0.95

What are your thoughts on this?

Kind regards.

Teck Kwang


On 14 August 2013 18:46, Matthew Marcus mamar...@lbl.gov wrote:

 **
 Yup.  That's consistent.  That sort of thing is why every XANES paper
 should state clearly
 the basis of the energy calibration.  With Newville's database, you can
 often figure that
 out yourself because many of the spectra were done with a reference foil.
 mam
 - Original Message -

 *From:* Teck Kwang Choo teck.kwang.c...@monash.edu
 *To:* XAFS Analysis using Ifeffit ifeffit@millenia.cars.aps.anl.gov
 *Sent:* Wednesday, August 14, 2013 1:07 AM
 *Subject:* Re: [Ifeffit] Linear Combination Fitting using ATHENA

 Yes Matthew, I used 7112 eV for my Fe K-edge calibration. I am guessing
 that is the reason a ~+1 eV shift in your maghemite spectrum is required is
 required to fit mine. Both pre-edge features match in the sense that they
 are roughly bound by the same energy values.


 On 14 August 2013 17:30, Matthew Marcus mamar...@lbl.gov wrote:

 **
 That depends on the way you did your calibration.  If it really matches,
 including the
 pre-edge features, with a 1eV shift, then that's what you probably have.
 Did you take
 into account my calibration, which puts the Fe K-edge at 7110.75eV?  Some
 people use other
 values like 7112eV.
 mam

  - Original Message -
 *From:* Teck Kwang Choo teck.kwang.c...@monash.edu
 *To:* XAFS Analysis using Ifeffit ifeffit@millenia.cars.aps.anl.gov
  *Sent:* Tuesday, August 13, 2013 9:20 PM
 *Subject:* Re: [Ifeffit] Linear Combination Fitting using ATHENA

   Also Matthew, I have found that my sample spectrum is an almost exact
 replica of the maghemite spectrum, the only differences being the
 absorbance height of the edge-peak (highest peak), which is ok but more
 importantly
 I have found that there is a +1eV shift of my spectrum relative to the
 maghemite one, probably due to slight differences in beamline. Do you think
 it is alright to make a +1 eV shift to the maghemite spectrum (or -1eV
 shift to my sample spectrum)?

 Thanks.

 Kind regards.

 Teck Kwang


 On 14 August 2013 10:07, Teck Kwang Choo teck.kwang.c...@monash.eduwrote:

  Thanks Matthew for the spectra.

 Is it correct to use the spectra that was taken in transmission mode
 (the one you did) to fit those taken in fluorescence mode (my case)?

 Teck Kwang


 On 14 August 2013 01:37, Matthew Marcus mamar...@lbl.gov wrote:

 Here are reference spectra.  The calibration is defined with the Fe
 metal edge at 7110.75eV.  The maghemite spectrum has been corrected for 
 mild
 overabsorption by reference to magnetite, which is very similar.  The
 magnetite is synthetic Fe3O4 from Aldrich, and done in TEY, so no
 overabsorption.
 The maghemite is from oxidation of Fe oxide nanoparticles and has been
 verified by XRD and comparison with the ETH group's maghemite spectrum.
 mam


 On 8/13/2013 1:33 AM, Teck Kwang Choo wrote:

 Hi all,

 @Drew Latta: I was wondering why the end members should be magnetite
 and maghemite. Is it because they are both of inverse-spinel structure, 
 the
 difference only being that the latter has all Fe completely oxidized into
 Fe(III)? Would you be able to provide the standard spectra to me if that 
 is
 the case? Thanks very much!

 Previously I have only thought of using hematite (alpha-Fe2O3) and FeO
 as end-members, both representing Fe(III) and Fe(II) respectively without
 giving much thought to the structure Fe takes. I also had my doubts using
 magnetite (Fe3O4) as a standard as it is a mixture of both Fe (II) and Fe
 (III). But thanks to the insights Drew has given, I now think I should 
 take
 the Fe-coordination environment into account with the use of magnetite and
 maghemite as standards.

 @Alexandre and Matthew: I have obtained Mossbauer spectra of some of
 my samples but like Drew said, it is not straightforward at this point in
 time. Will see what I can do with the data I have!

 Thanks for all your responses! It is greatly appreciated! This forum
 has been more helpful than I thought!

 Teck Kwang


 On 13 August 2013 01:34, Matthew Marcus mamar...@lbl.gov

Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-14 Thread Teck Kwang Choo
Yes Matthew, I used 7112 eV for my Fe K-edge calibration. I am guessing
that is the reason a ~+1 eV shift in your maghemite spectrum is required is
required to fit mine. Both pre-edge features match in the sense that they
are roughly bound by the same energy values.


On 14 August 2013 17:30, Matthew Marcus mamar...@lbl.gov wrote:

 **
 That depends on the way you did your calibration.  If it really matches,
 including the
 pre-edge features, with a 1eV shift, then that's what you probably have.
 Did you take
 into account my calibration, which puts the Fe K-edge at 7110.75eV?  Some
 people use other
 values like 7112eV.
 mam

 - Original Message -
 *From:* Teck Kwang Choo teck.kwang.c...@monash.edu
 *To:* XAFS Analysis using Ifeffit ifeffit@millenia.cars.aps.anl.gov
 *Sent:* Tuesday, August 13, 2013 9:20 PM
 *Subject:* Re: [Ifeffit] Linear Combination Fitting using ATHENA

  Also Matthew, I have found that my sample spectrum is an almost exact
 replica of the maghemite spectrum, the only differences being the
 absorbance height of the edge-peak (highest peak), which is ok but more
 importantly
 I have found that there is a +1eV shift of my spectrum relative to the
 maghemite one, probably due to slight differences in beamline. Do you think
 it is alright to make a +1 eV shift to the maghemite spectrum (or -1eV
 shift to my sample spectrum)?

 Thanks.

 Kind regards.

 Teck Kwang


 On 14 August 2013 10:07, Teck Kwang Choo teck.kwang.c...@monash.eduwrote:

  Thanks Matthew for the spectra.

 Is it correct to use the spectra that was taken in transmission mode (the
 one you did) to fit those taken in fluorescence mode (my case)?

 Teck Kwang


 On 14 August 2013 01:37, Matthew Marcus mamar...@lbl.gov wrote:

 Here are reference spectra.  The calibration is defined with the Fe
 metal edge at 7110.75eV.  The maghemite spectrum has been corrected for mild
 overabsorption by reference to magnetite, which is very similar.  The
 magnetite is synthetic Fe3O4 from Aldrich, and done in TEY, so no
 overabsorption.
 The maghemite is from oxidation of Fe oxide nanoparticles and has been
 verified by XRD and comparison with the ETH group's maghemite spectrum.
 mam


 On 8/13/2013 1:33 AM, Teck Kwang Choo wrote:

 Hi all,

 @Drew Latta: I was wondering why the end members should be magnetite
 and maghemite. Is it because they are both of inverse-spinel structure, the
 difference only being that the latter has all Fe completely oxidized into
 Fe(III)? Would you be able to provide the standard spectra to me if that is
 the case? Thanks very much!

 Previously I have only thought of using hematite (alpha-Fe2O3) and FeO
 as end-members, both representing Fe(III) and Fe(II) respectively without
 giving much thought to the structure Fe takes. I also had my doubts using
 magnetite (Fe3O4) as a standard as it is a mixture of both Fe (II) and Fe
 (III). But thanks to the insights Drew has given, I now think I should take
 the Fe-coordination environment into account with the use of magnetite and
 maghemite as standards.

 @Alexandre and Matthew: I have obtained Mossbauer spectra of some of my
 samples but like Drew said, it is not straightforward at this point in
 time. Will see what I can do with the data I have!

 Thanks for all your responses! It is greatly appreciated! This forum
 has been more helpful than I thought!

 Teck Kwang


 On 13 August 2013 01:34, Matthew Marcus mamar...@lbl.gov mailto:
 mamar...@lbl.gov wrote:

 While it's true that Mossbauer is the gold standard for Fe valence
 determination, access to the technique isn't all that common, and I don't
 know
 of any facility that can do it on a micro scale.  Is there one?
  mam


 On 8/12/2013 7:01 AM, Alexandre dos Santos Anastacio wrote:

 Hello,

 maybe that fingerprinting technique would be Mossbauer
 spectroscopy. The Fe(II)/Fe(III) ratio comes easily and also there are lots
 of papers about Mg-ferrite, an example:

 10.1109/TMAG.2009.2018880 http://dx.doi.org/10.1109/__**
 TMAG.2009.2018880 http://dx.doi.org/10.1109/__TMAG.2009.2018880 
 http://dx.doi.org/10.1109/**TMAG.2009.2018880http://dx.doi.org/10.1109/TMAG.2009.2018880
 

 Alexandre.





 __**___
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**__gov mailto:
 ifef...@millenia.cars.**aps.anl.gov Ifeffit@millenia.cars.aps.anl.gov
 
 http://millenia.cars.aps.anl._**_gov/mailman/listinfo/ifeffit 
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
 

 __**___
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**__gov mailto:
 ifef...@millenia.cars.**aps.anl.gov Ifeffit@millenia.cars.aps.anl.gov
 
 http://millenia.cars.aps.anl._**_gov/mailman/listinfo/ifeffit 
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp

Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-13 Thread Teck Kwang Choo
Hi all,

@Drew Latta: I was wondering why the end members should be magnetite and
maghemite. Is it because they are both of inverse-spinel structure, the
difference only being that the latter has all Fe completely oxidized into
Fe(III)? Would you be able to provide the standard spectra to me if that is
the case? Thanks very much!

Previously I have only thought of using hematite (alpha-Fe2O3) and FeO as
end-members, both representing Fe(III) and Fe(II) respectively without
giving much thought to the structure Fe takes. I also had my doubts using
magnetite (Fe3O4) as a standard as it is a mixture of both Fe (II) and Fe
(III). But thanks to the insights Drew has given, I now think I should take
the Fe-coordination environment into account with the use of magnetite and
maghemite as standards.

@Alexandre and Matthew: I have obtained Mossbauer spectra of some of my
samples but like Drew said, it is not straightforward at this point in
time. Will see what I can do with the data I have!

Thanks for all your responses! It is greatly appreciated! This forum has
been more helpful than I thought!

Teck Kwang


On 13 August 2013 01:34, Matthew Marcus mamar...@lbl.gov wrote:

 While it's true that Mossbauer is the gold standard for Fe valence
 determination, access to the technique isn't all that common, and I don't
 know
 of any facility that can do it on a micro scale.  Is there one?
 mam


 On 8/12/2013 7:01 AM, Alexandre dos Santos Anastacio wrote:

 Hello,

 maybe that fingerprinting technique would be Mossbauer spectroscopy. The
 Fe(II)/Fe(III) ratio comes easily and also there are lots of papers about
 Mg-ferrite, an example:

 10.1109/TMAG.2009.2018880 
 http://dx.doi.org/10.1109/**TMAG.2009.2018880http://dx.doi.org/10.1109/TMAG.2009.2018880
 

 Alexandre.




 __**_
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

  __**_
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit




-- 
Teck Kwang Choo
PhD Student
Department of Chemical Engineering
Room 225, Building 36
Monash University
Mobile No.: 04-11489904
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-13 Thread Matthew Marcus

Here are reference spectra.  The calibration is defined with the Fe metal edge 
at 7110.75eV.  The maghemite spectrum has been corrected for mild
overabsorption by reference to magnetite, which is very similar.  The magnetite 
is synthetic Fe3O4 from Aldrich, and done in TEY, so no overabsorption.
The maghemite is from oxidation of Fe oxide nanoparticles and has been verified 
by XRD and comparison with the ETH group's maghemite spectrum.
mam

On 8/13/2013 1:33 AM, Teck Kwang Choo wrote:

Hi all,

@Drew Latta: I was wondering why the end members should be magnetite and 
maghemite. Is it because they are both of inverse-spinel structure, the 
difference only being that the latter has all Fe completely oxidized into 
Fe(III)? Would you be able to provide the standard spectra to me if that is the 
case? Thanks very much!

Previously I have only thought of using hematite (alpha-Fe2O3) and FeO as 
end-members, both representing Fe(III) and Fe(II) respectively without giving 
much thought to the structure Fe takes. I also had my doubts using magnetite 
(Fe3O4) as a standard as it is a mixture of both Fe (II) and Fe (III). But 
thanks to the insights Drew has given, I now think I should take the 
Fe-coordination environment into account with the use of magnetite and 
maghemite as standards.

@Alexandre and Matthew: I have obtained Mossbauer spectra of some of my samples 
but like Drew said, it is not straightforward at this point in time. Will see 
what I can do with the data I have!

Thanks for all your responses! It is greatly appreciated! This forum has been 
more helpful than I thought!

Teck Kwang


On 13 August 2013 01:34, Matthew Marcus mamar...@lbl.gov 
mailto:mamar...@lbl.gov wrote:

While it's true that Mossbauer is the gold standard for Fe valence 
determination, access to the technique isn't all that common, and I don't know
of any facility that can do it on a micro scale.  Is there one?
 mam


On 8/12/2013 7:01 AM, Alexandre dos Santos Anastacio wrote:

Hello,

maybe that fingerprinting technique would be Mossbauer spectroscopy. 
The Fe(II)/Fe(III) ratio comes easily and also there are lots of papers about 
Mg-ferrite, an example:

10.1109/TMAG.2009.2018880 http://dx.doi.org/10.1109/__TMAG.2009.2018880 
http://dx.doi.org/10.1109/TMAG.2009.2018880

Alexandre.




_
Ifeffit mailing list
ifef...@millenia.cars.aps.anl.__gov 
mailto:Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.__gov/mailman/listinfo/ifeffit 
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

_
Ifeffit mailing list
ifef...@millenia.cars.aps.anl.__gov 
mailto:Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.__gov/mailman/listinfo/ifeffit 
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit




--
Teck Kwang Choo
PhD Student
Department of Chemical Engineering
Room 225, Building 36
Monash University
Mobile No.: 04-11489904


___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

#%Name: Maghemite nano powder
#%Empirical formula: gamma-Fe3+2O3  /Webmineral
#%Mineral group:  oxides
#%Xtal system: cubic   
#%Edge: K
#%Sample prep: powder on Kapton tape
#%Beamline: ALS 10.3.2, HxV slits = 200x20 um
#%Det:  I0=N2, Fluorescence mode (Ge solid state 7elts)
#%Ref/notes:  Mineralogical Research Company, San Jose, CA-USA. XRD confirmed.
#%Temp: Room temp (28C)
#%Timedate: 6:20 AM 7/3/2007
#%Provider: Matthew A. Marcus, Sirine Fakra.
#%
7.0100066E+3 1.3008636E-3
7.0150081E+3 1.2676025E-3
7.0200096E+3 1.4012953E-3
7.0250111E+3 1.3687201E-3
7.0300126E+3 1.4827145E-3
7.0350141E+3 1.5567103E-3
7.0400156E+3 1.6321360E-3
7.0450172E+3 1.8282378E-3
7.0500187E+3 1.8787587E-3
7.0550202E+3 2.0697013E-3
7.0600217E+3 2.1766526E-3
7.0650232E+3 2.3976456E-3
7.0700247E+3 2.8091858E-3
7.0750262E+3 3.1121079E-3
7.0800277E+3 3.7051882E-3
7.0850292E+3 4.2254896E-3
7.0900307E+3 5.0080461E-3
7.0950323E+3 6.3616899E-3
7.1000338E+3 8.1449022E-3
7.1005339E+3 8.3578735E-3
7.1010341E+3 8.5095725E-3
7.1015342E+3 8.6762463E-3
7.1020344E+3 9.0579738E-3
7.1025345E+3 9.2831028E-3
7.1030347E+3 9.6193633E-3
7.1035348E+3 9.8628340E-3
7.1040350E+3 1.0180513E-2
7.1045351E+3 1.0386116E-2
7.1050353E+3 1.1052578E-2
7.1055355E+3 1.1529281E-2
7.1060356E+3 1.2050395E-2
7.1065358E+3 1.2378185E-2
7.1070359E+3 1.3171965E-2
7.1075361E+3 1.3657637E-2
7.1080362E+3 1.4476360E-2
7.1085364E+3 1.5504986E-2
7.1090365E+3 1.6594136E-2
7.1095367E+3 1.8050181E-2
7.1100368E+3 1.9940670E-2
7.1105370E+3 2.2958282E-2
7.1110371E+3 2.7349793E-2
7.1115373E+3 3.5642680E-2
7.1120374E+3 5.0382320E-2
7.1125376E+3 7.0824214E-2
7.1130377E+3 9.2282370E-2
7.1135379E+3 9.5804412E-2
7.1140380E+3 8.5603494E-2

Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-13 Thread María Elena Montero Cabrera
Hi all!
I am out of my place and I can not check it now, but I believe that in the
database of XAFS (or something like that) there are some spectra of
maghemites obtained  by Lytle long time ago.
Best regards
Maria Elena Montero Cabrera
El ago 13, 2013 3:38 a.m., Teck Kwang Choo teck.kwang.c...@monash.edu
escribió:

 Hi all,

 @Drew Latta: I was wondering why the end members should be magnetite and
 maghemite. Is it because they are both of inverse-spinel structure, the
 difference only being that the latter has all Fe completely oxidized into
 Fe(III)? Would you be able to provide the standard spectra to me if that is
 the case? Thanks very much!

 Previously I have only thought of using hematite (alpha-Fe2O3) and FeO as
 end-members, both representing Fe(III) and Fe(II) respectively without
 giving much thought to the structure Fe takes. I also had my doubts using
 magnetite (Fe3O4) as a standard as it is a mixture of both Fe (II) and Fe
 (III). But thanks to the insights Drew has given, I now think I should take
 the Fe-coordination environment into account with the use of magnetite and
 maghemite as standards.

 @Alexandre and Matthew: I have obtained Mossbauer spectra of some of my
 samples but like Drew said, it is not straightforward at this point in
 time. Will see what I can do with the data I have!

 Thanks for all your responses! It is greatly appreciated! This forum has
 been more helpful than I thought!

 Teck Kwang


 On 13 August 2013 01:34, Matthew Marcus mamar...@lbl.gov wrote:

 While it's true that Mossbauer is the gold standard for Fe valence
 determination, access to the technique isn't all that common, and I don't
 know
 of any facility that can do it on a micro scale.  Is there one?
 mam


 On 8/12/2013 7:01 AM, Alexandre dos Santos Anastacio wrote:

 Hello,

 maybe that fingerprinting technique would be Mossbauer spectroscopy. The
 Fe(II)/Fe(III) ratio comes easily and also there are lots of papers about
 Mg-ferrite, an example:

 10.1109/TMAG.2009.2018880 
 http://dx.doi.org/10.1109/**TMAG.2009.2018880http://dx.doi.org/10.1109/TMAG.2009.2018880
 

 Alexandre.




 __**_
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

  __**_
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit




 --
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904

 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-13 Thread Teck Kwang Choo
Also Matthew, I have found that my sample spectrum is an almost exact
replica of the maghemite spectrum, the only differences being the
absorbance height of the edge-peak (highest peak), which is ok but more
importantly
I have found that there is a +1eV shift of my spectrum relative to the
maghemite one, probably due to slight differences in beamline. Do you think
it is alright to make a +1 eV shift to the maghemite spectrum (or -1eV
shift to my sample spectrum)?

Thanks.

Kind regards.

Teck Kwang


On 14 August 2013 10:07, Teck Kwang Choo teck.kwang.c...@monash.edu wrote:

 Thanks Matthew for the spectra.

 Is it correct to use the spectra that was taken in transmission mode (the
 one you did) to fit those taken in fluorescence mode (my case)?

 Teck Kwang


 On 14 August 2013 01:37, Matthew Marcus mamar...@lbl.gov wrote:

 Here are reference spectra.  The calibration is defined with the Fe metal
 edge at 7110.75eV.  The maghemite spectrum has been corrected for mild
 overabsorption by reference to magnetite, which is very similar.  The
 magnetite is synthetic Fe3O4 from Aldrich, and done in TEY, so no
 overabsorption.
 The maghemite is from oxidation of Fe oxide nanoparticles and has been
 verified by XRD and comparison with the ETH group's maghemite spectrum.
 mam


 On 8/13/2013 1:33 AM, Teck Kwang Choo wrote:

 Hi all,

 @Drew Latta: I was wondering why the end members should be magnetite and
 maghemite. Is it because they are both of inverse-spinel structure, the
 difference only being that the latter has all Fe completely oxidized into
 Fe(III)? Would you be able to provide the standard spectra to me if that is
 the case? Thanks very much!

 Previously I have only thought of using hematite (alpha-Fe2O3) and FeO
 as end-members, both representing Fe(III) and Fe(II) respectively without
 giving much thought to the structure Fe takes. I also had my doubts using
 magnetite (Fe3O4) as a standard as it is a mixture of both Fe (II) and Fe
 (III). But thanks to the insights Drew has given, I now think I should take
 the Fe-coordination environment into account with the use of magnetite and
 maghemite as standards.

 @Alexandre and Matthew: I have obtained Mossbauer spectra of some of my
 samples but like Drew said, it is not straightforward at this point in
 time. Will see what I can do with the data I have!

 Thanks for all your responses! It is greatly appreciated! This forum has
 been more helpful than I thought!

 Teck Kwang


 On 13 August 2013 01:34, Matthew Marcus mamar...@lbl.gov mailto:
 mamar...@lbl.gov wrote:

 While it's true that Mossbauer is the gold standard for Fe valence
 determination, access to the technique isn't all that common, and I don't
 know
 of any facility that can do it on a micro scale.  Is there one?
  mam


 On 8/12/2013 7:01 AM, Alexandre dos Santos Anastacio wrote:

 Hello,

 maybe that fingerprinting technique would be Mossbauer
 spectroscopy. The Fe(II)/Fe(III) ratio comes easily and also there are lots
 of papers about Mg-ferrite, an example:

 10.1109/TMAG.2009.2018880 http://dx.doi.org/10.1109/__**
 TMAG.2009.2018880 http://dx.doi.org/10.1109/__TMAG.2009.2018880 
 http://dx.doi.org/10.1109/**TMAG.2009.2018880http://dx.doi.org/10.1109/TMAG.2009.2018880
 

 Alexandre.





 __**___
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**__gov mailto:
 ifef...@millenia.cars.**aps.anl.gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl._**_gov/mailman/listinfo/ifeffit 
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
 

 __**___
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**__gov mailto:ifef...@millenia.cars.
 **aps.anl.gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl._**_gov/mailman/listinfo/ifeffit 
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit
 





 --
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904


 __**_
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit




 --
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904




-- 
Teck Kwang Choo
PhD Student
Department of Chemical Engineering
Room 225, Building 36
Monash University
Mobile No.: 04-11489904

Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-12 Thread Alexandre dos Santos Anastacio
Hello,

maybe that fingerprinting technique would be Mossbauer spectroscopy. The
Fe(II)/Fe(III) ratio comes easily and also there are lots of papers about
Mg-ferrite, an example:

10.1109/TMAG.2009.2018880 http://dx.doi.org/10.1109/TMAG.2009.2018880

Alexandre.
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-12 Thread Latta, Drew E.
I would say that the Mossbauer isn't always quite straight forward, especially 
if the OP is working with nanoparticles. You have to find the right temperature 
where superparamagnetism is gone but the Verwey transition hasn't occurred to 
reliably fit the resulting Mossbauer spectra. (I'll assume from ignorance that 
that Mg-ferrites have Verwey transitions.)

I would think the two best end-members for the XAS LC fits are possibly 
magnetite (fully stoichiometric: Fe2+/Fe3+ = 0.5) and either maghemite 
(gamma-Fe2O3) or the fully oxidized Mg-ferrite. Which Fe2O3 are you using Teck 
Kwang Choo? If you are using the alpha-form (hematite) then I would suspect you 
might not get very good fits as the structural difference between maghemite and 
hematite significantly influence their XANES spectra.

I've had good success finding Fe2+/Fe3+ ratios in partially oxidized magnetites 
by using magnetite and maghemite as end-members in LC fits. So I think you 
should have reasonable success for the Mg-ferrites using the right end-members.

Contact me in person below, if you are interested in magnetite/maghemite 
standard spectra.

Best,

Drew Latta
---
Drew E. Latta, Ph.D.
Postdoctoral Associate

Molecular Environmental Science Group
Biosciences Division
Argonne National Laboratory
Building 203, Room E-105
9700 South Cass Avenue
Argonne, IL  60439

Phone: 1-630-252-3985
dla...@anl.govmailto:dla...@anl.gov
http://www.mesg.anl.gov
---



From: ifeffit-boun...@millenia.cars.aps.anl.gov 
[mailto:ifeffit-boun...@millenia.cars.aps.anl.gov] On Behalf Of Alexandre dos 
Santos Anastacio
Sent: Monday, August 12, 2013 9:01 AM
To: XAFS Analysis using Ifeffit
Subject: Re: [Ifeffit] Linear Combination Fitting using ATHENA

Hello,

maybe that fingerprinting technique would be Mossbauer spectroscopy. The 
Fe(II)/Fe(III) ratio comes easily and also there are lots of papers about 
Mg-ferrite, an example:

10.1109/TMAG.2009.2018880http://dx.doi.org/10.1109/TMAG.2009.2018880

Alexandre.

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-11 Thread Teck Kwang Choo
The database: http://cars.uchicago.edu/~newville/ModelLib/search.html does
not seem to have Ferrite.

The matrix of my sample is largely Mg-ferrite where Fe is in the form of Fe
(III). If you anyone has the spectra for Mg-ferrite, would it be possible
to provide it to me? Otherwise, I do not think my sample spectra would be
useable at all!

Thanks and kind regards.


On 10 August 2013 02:21, Matthew Marcus mamar...@lbl.gov wrote:

 I would guess that you need better standards.  In particular, FeO is not
 your typical Fe(II) mineral or oxide.  What is the system you're working on?
 I have an extensive collection of Fe XANES reference spectra, including a
 number of Fe(II) species.

 LCF gives you the fraction of Fe atoms included in the sample.  For
 example, suppose you had a sample which was genuinely a mix representable
 as FeO.Fe2O3.  Then, LCF would properly report 2/3 Fe2O3 + 1/3 FeO because
 2/3 of the Fe atoms are in the Fe2O3 fraction.
 mam


 On 8/8/2013 8:37 PM, Teck Kwang Choo wrote:

 Hello,

 I have been using ATHENA to perform Linear Combination Fitting (LCF) on
 my XANES spectrum to quantify the amounts of Fe (II) and Fe (III) in my
 samples using only FeO and Fe2O3 as standards. The results showed a
 majority of Fe (II), which is contrary to what was expected.

 This made me wonder if the LCF results from XANES is a molar percentage
 (mol %) or weight percentage (wt%).

 Could anyone enlighten me on this matter?

 Thanks!

 --
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904


 __**_
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

  __**_
 Ifeffit mailing list
 ifef...@millenia.cars.aps.anl.**gov Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.**gov/mailman/listinfo/ifeffithttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit




-- 
Teck Kwang Choo
PhD Student
Department of Chemical Engineering
Room 225, Building 36
Monash University
Mobile No.: 04-11489904
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-11 Thread Teck Kwang Choo
That's magnetite unfortunately!


On 12 August 2013 14:02, Scott Calvin scal...@sarahlawrence.edu wrote:

 Yes it does, Teck Kwang:
 http://millenia.cars.aps.anl.gov/cgi-bin/newville/Model-Search.cgi/ModelData/fe3o4_rt01_mar02.xmu

 --Scott Calvin
 Sarah Lawrence College

 On Aug 11, 2013, at 11:12 PM, Teck Kwang Choo teck.kwang.c...@monash.edu
 mailto:teck.kwang.c...@monash.edu wrote:

 The database: http://cars.uchicago.edu/~newville/ModelLib/search.html
 http://cars.uchicago.edu/%7Enewville/ModelLib/search.html does not seem
 to have Ferrite.

 The matrix of my sample is largely Mg-ferrite where Fe is in the form of
 Fe (III). If you anyone has the spectra for Mg-ferrite, would it be
 possible to provide it to me? Otherwise, I do not think my sample spectra
 would be useable at all!

 Thanks and kind regards.


 On 10 August 2013 02:21, Matthew Marcus mamar...@lbl.govmailto:
 mamar...@lbl.gov wrote:
 I would guess that you need better standards.  In particular, FeO is not
 your typical Fe(II) mineral or oxide.  What is the system you're working on?
 I have an extensive collection of Fe XANES reference spectra, including a
 number of Fe(II) species.

 LCF gives you the fraction of Fe atoms included in the sample.  For
 example, suppose you had a sample which was genuinely a mix representable
 as FeO.Fe2O3.  Then, LCF would properly report 2/3 Fe2O3 + 1/3 FeO because
 2/3 of the Fe atoms are in the Fe2O3 fraction.
 mam


 On 8/8/2013 8:37 PM, Teck Kwang Choo wrote:
 Hello,

 I have been using ATHENA to perform Linear Combination Fitting (LCF) on my
 XANES spectrum to quantify the amounts of Fe (II) and Fe (III) in my
 samples using only FeO and Fe2O3 as standards. The results showed a
 majority of Fe (II), which is contrary to what was expected.

 This made me wonder if the LCF results from XANES is a molar percentage
 (mol %) or weight percentage (wt%).

 Could anyone enlighten me on this matter?

 Thanks!

 --
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904


 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.govmailto:Ifeffit@millenia.cars.aps.anl.gov
 
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.govmailto:Ifeffit@millenia.cars.aps.anl.gov
 
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit



 --
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904
 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.govmailto:Ifeffit@millenia.cars.aps.anl.gov
 
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit




-- 
Teck Kwang Choo
PhD Student
Department of Chemical Engineering
Room 225, Building 36
Monash University
Mobile No.: 04-11489904
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-09 Thread Edmund Welter
Dear Teck Kwang Choo,

it is the ratio of the number of (in your case) iron atoms. This ratio
has no unit. You will therefore get the same number if you calculate
molar % ratio or weight % (Except you consider the weight of the missing
electron in Fe3+... :-). Now, if you weigh your standards and want to
calculate the number of Fe atoms you have to consider the weight of
the oxygen. That means, the ratio you get from LCF fitting of your XAS
spectra is NOT the ratio between mol of FeO and Fe2O3, but mol FeII+ to
mol FeIII+! However, if necessary or usefull you can always calculate
the latter from the FeII/FeIII ratio using the molar weights.

So, as a chemist I would say, stay consequently with the molar ratio and
you are on the safe side. You can always multiply the number of mol with
the molar weight, but you would only make things more complicated and
end up with the same ratio or, after multiplying with 100, percentage.

Concerning the deviation between your result and your expectation: I
would be more nervous about the choice of reference substances. If you
only use FeO and Fe2O3 your result is only correct if your sample
contains no other Fe component than these two. If it contains
significant amounts of any other Fe component (even other oxides) you
would better not draw any important conclusions from the result of your
analysis.

Best regards
Edmund Welter



On 09.08.2013 05:37, Teck Kwang Choo wrote:
 Hello,

 I have been using ATHENA to perform Linear Combination Fitting (LCF)
 on my XANES spectrum to quantify the amounts of Fe (II) and Fe (III)
 in my samples using only FeO and Fe2O3 as standards. The results
 showed a majority of Fe (II), which is contrary to what was expected.

 This made me wonder if the LCF results from XANES is a molar
 percentage (mol %) or weight percentage (wt%).

 Could anyone enlighten me on this matter?

 Thanks!

 -- 
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904


 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit

___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


Re: [Ifeffit] Linear Combination Fitting using ATHENA

2013-08-09 Thread Teck Kwang Choo
Thanks for your reply Edmund.

I agree with you that I might have not chosen the right standards for my
sample. In fact, XRD analysis have confirmed that Ferrite is the main form
of Fe in my sample but I did not have any ferrite standards during XANES
analysis. Was wondering if you happen to have any Ferrite XANES spectra by
any chance?

Thanks and kind regards.

Teck Kwang


On 9 August 2013 16:53, Edmund Welter edmund.wel...@desy.de wrote:

  Dear Teck Kwang Choo,

 it is the ratio of the number of (in your case) iron atoms. This ratio has
 no unit. You will therefore get the same number if you calculate molar %
 ratio or weight % (Except you consider the weight of the missing electron
 in Fe3+... :-). Now, if you weigh your standards and want to calculate the
 number of Fe atoms you have to consider the weight of the oxygen. That
 means, the ratio you get from LCF fitting of your XAS spectra is NOT the
 ratio between mol of FeO and Fe2O3, but mol FeII+ to mol FeIII+! However,
 if necessary or usefull you can always calculate the latter from the
 FeII/FeIII ratio using the molar weights.

 So, as a chemist I would say, stay consequently with the molar ratio and
 you are on the safe side. You can always multiply the number of mol with
 the molar weight, but you would only make things more complicated and end
 up with the same ratio or, after multiplying with 100, percentage.

 Concerning the deviation between your result and your expectation: I would
 be more nervous about the choice of reference substances. If you only use
 FeO and Fe2O3 your result is only correct if your sample contains no
 other Fe component than these two. If it contains significant amounts of
 any other Fe component (even other oxides) you would better not draw any
 important conclusions from the result of your analysis.

 Best regards
 Edmund Welter




 On 09.08.2013 05:37, Teck Kwang Choo wrote:

 Hello,

 I have been using ATHENA to perform Linear Combination Fitting (LCF) on my
 XANES spectrum to quantify the amounts of Fe (II) and Fe (III) in my
 samples using only FeO and Fe2O3 as standards. The results showed a
 majority of Fe (II), which is contrary to what was expected.

 This made me wonder if the LCF results from XANES is a molar percentage
 (mol %) or weight percentage (wt%).

  Could anyone enlighten me on this matter?

  Thanks!

 --
 Teck Kwang Choo
 PhD Student
 Department of Chemical Engineering
 Room 225, Building 36
 Monash University
 Mobile No.: 04-11489904


 ___
 Ifeffit mailing 
 listifef...@millenia.cars.aps.anl.govhttp://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit



 ___
 Ifeffit mailing list
 Ifeffit@millenia.cars.aps.anl.gov
 http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit




-- 
Teck Kwang Choo
PhD Student
Department of Chemical Engineering
Room 225, Building 36
Monash University
Mobile No.: 04-11489904
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit


[Ifeffit] Linear Combination Fitting using ATHENA

2013-08-08 Thread Teck Kwang Choo
Hello,

I have been using ATHENA to perform Linear Combination Fitting (LCF) on my
XANES spectrum to quantify the amounts of Fe (II) and Fe (III) in my
samples using only FeO and Fe2O3 as standards. The results showed a
majority of Fe (II), which is contrary to what was expected.

This made me wonder if the LCF results from XANES is a molar percentage
(mol %) or weight percentage (wt%).

Could anyone enlighten me on this matter?

Thanks!

-- 
Teck Kwang Choo
PhD Student
Department of Chemical Engineering
Room 225, Building 36
Monash University
Mobile No.: 04-11489904
___
Ifeffit mailing list
Ifeffit@millenia.cars.aps.anl.gov
http://millenia.cars.aps.anl.gov/mailman/listinfo/ifeffit