Re: Strange peaks from grainy Si (data re-attached)

2007-01-26 Thread martijn . fransen
Dear Congwu, 

Maybe your Mo K-Alpha contribution gets a lot of stronger if you would 
range the kV settings of your X-ray generator. As a rule of thumb, you 
need an acceleration voltage that is about 4 x higher than the 
characteristic energy in keV. In other words, I would advise to run your 
source at 60 keV instead of 35 keV. Your Mo K-Alpha lines are probably 
very  low compared to the bremsstrahlung. 

With best regards, 
Martijn



Dr. Martijn J. Fransen
Product Manager XRD
PANalytical
Lelyweg 1
7602 EA  Almelo, The Netherlands
Phone ++31 546 534 345
Fax ++ 31 546 534 598



 PANalytical
The Analytical X-ray Company  

The information contained in this message is confidential and may be 
legally privileged. The message is intended solely for the addressee(s). 
If you are not the intended recipient, you are hereby notified that any 
use, dissemination, or reproduction is strictly prohibited and may be 
unlawful. If you are not the intended recipient, please contact the sender 
by return e-mail and destroy all copies of the original message.







gregor 
01/25/2007 05:06 PM
Please respond to
rietveld_l@ill.fr

To
rietveld_l@ill.fr
cc

Subject
Re: Strange peaks from grainy Si (data re-attached)






Dear Congwu, 
as you put it, it definitely seems that you don't get out the MoK 
radiation from your collimator. Control the alignment and make sure 
the beam hits your sample.

best

miguel


On 25 Jan 2007 at 9:55, [EMAIL PROTECTED] wrote:

 Hi, Miguel, the unit is the so-called moment transfer that is
 calculated as sin(theta)/lambda. When I did the simulation, I use the
 Zr K-emission as the first wavelength and the Mo-K-alpha as the
 second, because the spectrum I measured with an AMTREK spectrometer
 shows that the Zr line is higher than Mo line.
 
 Yes, with powdercell, you are able to see the peaks from other
 wavelength when the width of the calculated profile (# of FWHM) is
 large enough. the attached figure is another simulation with the
 parameters of: wavelenth 0.79Å, 0.44Å, I1/I2=0.5, Width of calc.
 profile: 4000FWHM. The 0.44angstrom can only come from the
 bremsstrahlung in my experiment if it is true. The problems bothering
 me are the measured peak is much wider than the simulation and the
 bremsstrahlung at 0.44angstron is ~1000times weaker than the main
 line.
 
 congwu
 
 
 Quoting [EMAIL PROTECTED]:
 
  Congwu,
  
  I'm getting confused with the units of your diffraction patterns,
  could
  
  you just indicate 2theta which is probably the primary observation
  without any assumption about lambda?
  
  Also, I've trouble with your simulated Si pattern: there should be
  no traces of Zr radiation generated peaks, and the absorption edge
  in the bremsstrahlung would probably be difficult to be detected at
  the current level of resolution. So I would prefer to see the
  simulation for
  
  plain MoKa radiation and the full peak height of the 111, 220 and
  311 reflections. Observed intensities are very low, are you sure you
  get the Mo radiation out of your collimator?
  
  Another problem is how you can manage to calibrate the detector
  distance using such broad peaks, but I don't know about the purpose
  of your experiment.
  
  BTW, ground Si wafers are fine, just control the grain size under a
  microscope to make sure there are no grains exceeding 5um or so
  (relatively easy to achieve with smalll sammple size). The rings
  actually look a bit grainy.
  
  best
  
  miguel
  
  
  On 24 Jan 2007 at 15:17, [EMAIL PROTECTED] wrote:
  
   I just did an estimation: assuming the unknown peak is the [111]
  peak,
   if it is another wavelength, it corresponds to a wavelength of
   0.44 angstrom (28keV), consistent to the spectrum of Mo with 35kV
  voltage.
   I do not think it could be the contamination of W, because the
  voltage
   is still far from the W characteristic line to activate it.
   
   probably it is worthy of doing a broad spectrum simulation. does
   anybody know where I can find the formulas that powdercell uses to
   calculate the pattern?
   
   btw I re-attached the data.
   
   congwu
   
   
   
   
   Quoting Andrew Payzant [EMAIL PROTECTED]:
   
Peter,

That is my guess too ­ either additional (weak) characteristic
  lines
such as Mo-K beta, or contaminant lines such as W L, or, as you
point out, some wavelength in the Bremmstrahlung that meets the
Bragg condition for some silicon crystallites.

However, I would not have expected a few large crystallites to
  yield
a continuous ring, as appears in his CCD image, so perhaps there
  is
another explanation.

Andrew
-- 
E. Andrew Payzant
Senior RD Staff Member
High Temperature Materials Laboratory

Oak Ridge National Laboratory
1 Bethel Valley Road
PO Box 2008, MS 6064
Building 4515, Room 113
Oak Ridge, TN, 37831-6064

ph: (865) 574-6538   FAX: (865) 574-3940
web: http://html.ornl.gov/dtpgrp/staff/payzant.html

Re: Strange peaks from grainy Si (data re-attached)

2007-01-25 Thread gregor
Congwu,

I'm getting confused with the units of your diffraction patterns, could 
you just indicate 2theta which is probably the primary observation 
without any assumption about lambda?

Also, I've trouble with your simulated Si pattern: there should be no 
traces of Zr radiation generated peaks, and the absorption edge in 
the bremsstrahlung would probably be difficult to be detected at the 
current level of resolution. So I would prefer to see the simulation for 
plain MoKa radiation and the full peak height of the 111, 220 and 311 
reflections. Observed intensities are very low, are you sure you get 
the Mo radiation out of your collimator?

Another problem is how you can manage to calibrate the detector 
distance using such broad peaks, but I don't know about the purpose 
of your experiment.

BTW, ground Si wafers are fine, just control the grain size under a 
microscope to make sure there are no grains exceeding 5um or so 
(relatively easy to achieve with smalll sammple size). The rings 
actually look a bit grainy.

best

miguel


On 24 Jan 2007 at 15:17, [EMAIL PROTECTED] wrote:

 I just did an estimation: assuming the unknown peak is the [111] peak,
 if it is another wavelength, it corresponds to a wavelength of 0.44
 angstrom (28keV), consistent to the spectrum of Mo with 35kV voltage.
 I do not think it could be the contamination of W, because the voltage
 is still far from the W characteristic line to activate it.
 
 probably it is worthy of doing a broad spectrum simulation. does
 anybody know where I can find the formulas that powdercell uses to
 calculate the pattern?
 
 btw I re-attached the data.
 
 congwu
 
 
 
 
 Quoting Andrew Payzant [EMAIL PROTECTED]:
 
  Peter,
  
  That is my guess too ­ either additional (weak) characteristic lines
  such as Mo-K beta, or contaminant lines such as W L, or, as you
  point out, some wavelength in the Bremmstrahlung that meets the
  Bragg condition for some silicon crystallites.
  
  However, I would not have expected a few large crystallites to yield
  a continuous ring, as appears in his CCD image, so perhaps there is
  another explanation.
  
  Andrew
  -- 
  E. Andrew Payzant
  Senior RD Staff Member
  High Temperature Materials Laboratory
  
  Oak Ridge National Laboratory
  1 Bethel Valley Road
  PO Box 2008, MS 6064
  Building 4515, Room 113
  Oak Ridge, TN, 37831-6064
  
  ph: (865) 574-6538   FAX: (865) 574-3940
  web: http://html.ornl.gov/dtpgrp/staff/payzant.html
  
  
  
  
  
  From: [EMAIL PROTECTED]
  Reply-To: rietveld_l@ill.fr
  Date: Wed, 24 Jan 2007 12:51:57 -0500
  To: rietveld_l@ill.fr
  Subject: Re: Strange peaks from grainy Si
  
  
  I'll mail you some fine Si powder if you send me your address.
  
  My best guess is that you have a relatively big lump of Si in your
  sample that happens to be lined up to make a bright spot from the
  bremstrahlung part of the spectrum.  It happens to meet some Si
  Bragg reflection condition for some wavelength in the brems.
  spectrum.
  
  ^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~
  Peter W. Stephens
  Professor, Department of Physics and Astronomy
  Stony Brook University
  Stony Brook, NY 11794-3800
  fax 631-632-8176
  
  
  
 
 

-- 
Miguel Gregorkiewitz
Dip Scienze della Terra, Università
via Laterino 8, I-53100 Siena, Europe
fon +39'0577'233810 fax 233938
email [EMAIL PROTECTED]






Re: Strange peaks from grainy Si (data re-attached)

2007-01-25 Thread Jonathan Wright

... The problems bothering me are
the measured peak is much wider than the simulation


Peak widths should be 400 microns (your beamsize) divided by 3.5 cm 
(your distance) in radians, which is about 180*0.4/3.5/PI = 6.5 degrees.


HTH,

Jon


Re: Strange peaks from grainy Si (data re-attached)

2007-01-25 Thread gregor
Dear Congwu, 
as you put it, it definitely seems that you don't get out the MoK 
radiation from your collimator. Control the alignment and make sure 
the beam hits your sample.

best

miguel


On 25 Jan 2007 at 9:55, [EMAIL PROTECTED] wrote:

 Hi, Miguel, the unit is the so-called moment transfer that is
 calculated as sin(theta)/lambda. When I did the simulation, I use the
 Zr K-emission as the first wavelength and the Mo-K-alpha as the
 second, because the spectrum I measured with an AMTREK spectrometer
 shows that the Zr line is higher than Mo line.
 
 Yes, with powdercell, you are able to see the peaks from other
 wavelength when the width of the calculated profile (# of FWHM) is
 large enough. the attached figure is another simulation with the
 parameters of: wavelenth 0.79Å, 0.44Å, I1/I2=0.5, Width of calc.
 profile: 4000FWHM. The 0.44angstrom can only come from the
 bremsstrahlung in my experiment if it is true. The problems bothering
 me are the measured peak is much wider than the simulation and the
 bremsstrahlung at 0.44angstron is ~1000times weaker than the main
 line.
 
 congwu
 
 
 Quoting [EMAIL PROTECTED]:
 
  Congwu,
  
  I'm getting confused with the units of your diffraction patterns,
  could
  
  you just indicate 2theta which is probably the primary observation
  without any assumption about lambda?
  
  Also, I've trouble with your simulated Si pattern: there should be
  no traces of Zr radiation generated peaks, and the absorption edge
  in the bremsstrahlung would probably be difficult to be detected at
  the current level of resolution. So I would prefer to see the
  simulation for
  
  plain MoKa radiation and the full peak height of the 111, 220 and
  311 reflections. Observed intensities are very low, are you sure you
  get the Mo radiation out of your collimator?
  
  Another problem is how you can manage to calibrate the detector
  distance using such broad peaks, but I don't know about the purpose
  of your experiment.
  
  BTW, ground Si wafers are fine, just control the grain size under a
  microscope to make sure there are no grains exceeding 5um or so
  (relatively easy to achieve with smalll sammple size). The rings
  actually look a bit grainy.
  
  best
  
  miguel
  
  
  On 24 Jan 2007 at 15:17, [EMAIL PROTECTED] wrote:
  
   I just did an estimation: assuming the unknown peak is the [111]
  peak,
   if it is another wavelength, it corresponds to a wavelength of
   0.44 angstrom (28keV), consistent to the spectrum of Mo with 35kV
  voltage.
   I do not think it could be the contamination of W, because the
  voltage
   is still far from the W characteristic line to activate it.
   
   probably it is worthy of doing a broad spectrum simulation. does
   anybody know where I can find the formulas that powdercell uses to
   calculate the pattern?
   
   btw I re-attached the data.
   
   congwu
   
   
   
   
   Quoting Andrew Payzant [EMAIL PROTECTED]:
   
Peter,

That is my guess too ­ either additional (weak) characteristic
  lines
such as Mo-K beta, or contaminant lines such as W L, or, as you
point out, some wavelength in the Bremmstrahlung that meets the
Bragg condition for some silicon crystallites.

However, I would not have expected a few large crystallites to
  yield
a continuous ring, as appears in his CCD image, so perhaps there
  is
another explanation.

Andrew
-- 
E. Andrew Payzant
Senior RD Staff Member
High Temperature Materials Laboratory

Oak Ridge National Laboratory
1 Bethel Valley Road
PO Box 2008, MS 6064
Building 4515, Room 113
Oak Ridge, TN, 37831-6064

ph: (865) 574-6538   FAX: (865) 574-3940
web: http://html.ornl.gov/dtpgrp/staff/payzant.html





From: [EMAIL PROTECTED]
Reply-To: rietveld_l@ill.fr
Date: Wed, 24 Jan 2007 12:51:57 -0500
To: rietveld_l@ill.fr
Subject: Re: Strange peaks from grainy Si


I'll mail you some fine Si powder if you send me your address.

My best guess is that you have a relatively big lump of Si in
your sample that happens to be lined up to make a bright spot
from the bremstrahlung part of the spectrum.  It happens to meet
some Si Bragg reflection condition for some wavelength in the
brems. spectrum.

^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~^~
Peter W. Stephens
Professor, Department of Physics and Astronomy
Stony Brook University
Stony Brook, NY 11794-3800
fax 631-632-8176



   
   
  
  -- 
  Miguel Gregorkiewitz
  Dip Scienze della Terra, Università
  via Laterino 8, I-53100 Siena, Europe
  fon +39'0577'233810 fax 233938
  email [EMAIL PROTECTED]
  
  
  
  
  
 
 

-- 
Miguel Gregorkiewitz
Dip Scienze della Terra, Università
via Laterino 8, I-53100 Siena, Europe
fon +39'0577'233810 fax 233938
email [EMAIL PROTECTED]






Re: Strange peaks from grainy Si (data re-attached)

2007-01-24 Thread wright
Congwu,

If you can move the detector the radius of the silicon rings will change
according to the distance and Bragg angles. Moving the sample might be easier
for the same check. Does your spurious ring move so that you can trace it back
to a point inside the sample? Could it come from the collimator or beamstop? 

What does the background look like when you remove the silicon sample and just
measure the sample holder? If you know the ring only appears when you put the
silicon sample then try checking some other materials. Aluminium foil is
normally easy to find, either in the kitchen or wrapped around a UHV experiment.
Table salt (NaCl) can also be worth a try. If they all give spurious rings
coming from the sample and the background is clean then it sounds a bit strange.

Good luck,

Jon








-
This mail sent through IMP: http://horde.org/imp/