[time-nuts] HP59309A HP-IB Digital Clock Operating and Service Manual

2007-04-25 Thread Martin Fischer

   Hello Community,
   thank you all for your hints concerning the HP59309A Clock
   and your kind assistance with making available the Operating and
   Service Manual.
   My HP 59309A had a real problem: One of the input protection diodes
   of the EXT FREQ STD input was shot. Suspect the previous owner
   confused the EXT FREQ STD input with the line input...  ;-)
   Regards
   Martin
   mailto:[EMAIL PROTECTED]
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] SMB-F to SMA-F Cable

2007-04-25 Thread James R. Gorr

If no one stepped up yet, let me see what connectors I
have.  I know I got LMR-100 but not sure if I have the
connectors, but this will be the excuse I need to run
to the supply house and stock up...

Jamie

--- Jason Rabel [EMAIL PROTECTED] wrote:

 I figured I would make the request to you guys
 before I hit fleabay (they
 have some but I don't need a dozen - just one).
 
 Does anyone have a short SMB-F to SMA-F cable I
 could buy? This is to
 connect to a Trimble ACE-III receiver. I have some
 extra receivers too if
 someone would rather do a swap.
 
 Jason
 
 
 ___
 time-nuts mailing list
 time-nuts@febo.com

https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 


Fedora Core 4 Linux

__
Do You Yahoo!?
Tired of spam?  Yahoo! Mail has the best spam protection around 
http://mail.yahoo.com 

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


[time-nuts] FMT

2007-04-25 Thread Connie Marshall
For details of the next run see:
http://pages.suddenlink.net/k5cm/

Connie
K5CM
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


[time-nuts] GPS Disciplined Oscillators

2007-04-25 Thread Mike Feher
I have to show my ignorance here, because this has been bothering me for a
while, and, I wonder if there is a relatively simple answer. This question
has to do with frequency accuracy and stability only. Also, let's talk of
long term like 24 hours or more, so let's ignore phase noise and just
concern ourselves with long term accuracy/stability. If I have three
separate oscillators, let's say a Rubidium, an OCXO and a TCXO all with
EFC's capable of closing the loop to lock to GPS, what kind of absolute
frequency difference should I see amongst the three at any given time,
random times, or, over the entire test period. Let's also make it simple and
say all three are at 10 MHz nominal, unlocked to GPS. When locked, and
properly designed with a narrow loop filter, I would expect the long term
accuracy to be very close amongst all 3 oscillators. Certainly better than a
few parts in 10^-11. First, am I wrong in this assumption? In either case,
crystals, and even Rubidium cells do age, while at different rates, so, it
is possible, that if lock with GPS is lost for some reason, because the
oscillator may have drifted/aged out of loop range, it cannot be disciplined
again. I, for the time being, also assume that the EFC on all 3 oscillators
has a range wide enough to keep the oscillator locked even as it ages. Are
the narrow loop bandwidths and wide EFC ranges contradictory? So, to
reiterate the question, if I was clear enough, what kind of frequency
excursions should I anticipate to see amongst my three disciplined
oscillators in lets say 24 hours, or in a month. Assume GPS disciplining was
working all of that time (can I even assume that with aging?). BTW, how is
my assumption regarding the oscillators aging? If the oscillator basic
frequency determining element drifts out of lock range, during lock, will it
stay in lock? - Thanks in advance for any enlightenment - Mike  

 
 
Mike B. Feher, N4FS
89 Arnold Blvd.
Howell, NJ, 07731
732-886-5960
 
 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] GPS Disciplined Oscillators

2007-04-25 Thread Jason Rabel
Mike,

I just so happened to be reading through my TS2100 manual and came across a
bit of info that might be applicable to your question:

The info below is for the various oscillator options you can get, and while
the number can vary based on the exact oscillator used (and its age), it
should give you a good ballpark idea of the differences.

VCXO: 48 milliseconds/day long term flywheeling
OCXO: 5 milliseconds/day long term flywheeling
Rubidium: 6.5 microseconds/month long term flywheeling 


Jason


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] GPS Disciplined Oscillators

2007-04-25 Thread SAIDJACK
 
In a message dated 4/25/2007 09:50:38 Pacific Daylight Time,  
[EMAIL PROTECTED] writes:

So,  to
reiterate the question, if I was clear enough, what kind of  frequency
excursions should I anticipate to see amongst my three  disciplined
oscillators in lets say 24 hours, or in a month. Assume GPS  disciplining was
working all of that time (can I even assume that with  aging?). BTW, how is
my assumption regarding the oscillators  aging?


Hello Mike,
 
from my experience, good oscillators will age parts in 1E-08 per day or  
less. Bad ones can age much faster, some parts in 1E-06 in the first days even. 
 
That's a pretty significant variation in EFC voltage required to compensate  
that.
 
We have seen units that age significantly in the first 2 - 3 days, then  slow 
down, and after 6 months or so almost have no aging. We have seen units  
where the aging actually accelerates over time.
 
Aging is a very slow changing process after the crystal has run some days  
(second derivative is small), so GPS locking will usually easily compensate 
this 
 error.
 
There are oscillators that have popcorn noise (frequency jumps) that can be  
really annoying, and this effect is similar to rapid aging (from one second to 
 another): almost sudden the phase/frequency of the OCXO changes radically 
from  one second to another. This is usually caused by either dust particles 
leaving,  or landing on the crystal, the crystal cracking in it's holders, or 
 
radioactive particles hitting the crystal.
 
Also, thermal effects will usually require much larger shifts in EFC  voltage 
than aging, unless you are using a double-oven high quality OCXO.  Thermal 
effects on the required EFC voltage usually swamp aging effects and are  harder 
to deal with because they are more or less random, and not a nice  (almost) 
straight line on the EFC voltage.
 
bye,
Said 



** See what's free at http://www.aol.com.
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread SAIDJACK
 
In a message dated 4/24/2007 15:14:11 Pacific Daylight Time,  
[EMAIL PROTECTED] writes:

  It is unclear to me if the spurious is normal or not. The spurious of   
 my PRS10 is 60dB down the 10MHz so it is strange that you see  no  
 spurious.



Hi Henk,
 
try running your PRS10 from a Pb battery, with nothing else connected (no  
RS232 etc).
 
The plot I sent you was from a PRS running from a switching power supply,  
with RS-232 connected to some other stuff.
 
Also, try making the sine wave of the PRS10 into a nice, fast edge rate  
square-wave with a Fairchild NC7SZ04 driver inverter for example. The Wavecrest 
 
units don't work as well with sine waves as with square waves. On the SZ04, put 
 a 10nF cap in series to the input, and a 1 MEG resistor from it's input pin 
to  it's output pin. Feed the chip from a very low noise 5V power source. 
Insert a  40-50 Ohm resistor into the output path going to the coax. Put an  
attenuator/AC-coupler into the coax so the Wavecrest doesen't get  overloaded 
by 
the 2.5Vpp DC signal. Result: low noise comparator/buffer :)
 
bye,
Said



** See what's free at http://www.aol.com.
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


[time-nuts] Wavecrest noise measurement hardware

2007-04-25 Thread John Miles
As an aside, how do the Wavecrest machines work?  Do they just run the
signal into a low-jitter ADC with a high-quality clock and derive all the
timing information digitally, or is the box full of low-noise tunable
synthesizers, mixers, filters, and the usual stuff?

-- john, KE5FX


 Hi Henk,

 try running your PRS10 from a Pb battery, with nothing else
 connected (no
 RS232 etc).

 The plot I sent you was from a PRS running from a switching power
 supply,
 with RS-232 connected to some other stuff.

 Also, try making the sine wave of the PRS10 into a nice, fast edge rate
 square-wave with a Fairchild NC7SZ04 driver inverter for example.
 The Wavecrest
 units don't work as well with sine waves as with square waves. On
 the SZ04, put
  a 10nF cap in series to the input, and a 1 MEG resistor from
 it's input pin
 to  it's output pin. Feed the chip from a very low noise 5V power source.
 Insert a  40-50 Ohm resistor into the output path going to the
 coax. Put an
 attenuator/AC-coupler into the coax so the Wavecrest doesen't get
  overloaded by
 the 2.5Vpp DC signal. Result: low noise comparator/buffer :)



___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] GPS Disciplined Oscillators

2007-04-25 Thread Brooke Clarke
Hi Mike:

Nice to see you here.  This is my take on it, but there are others on this list 
that are much more knowledgeable.

The key to the performance of a GPSDO has to do with the Allan plots for the 
oscillator, GPS receiver and it's antenna  and . . . .  Also the Time Interval 
counter used in the GPSDO that compares the oscillator and the GPS 1 PPS is 
important.

The key idea is that when you overlay the GPS receiver's plot with the 
oscillator plot there will be only one point where they intersect.  That's what 
the loop time constant should be for the PLL.  That way for time intervals that 
are shorter you get the good performance of the xtal oscillator and for longer 
time intervals you get the long term stability of GPS.

Inside the GPSDO it somehow compares the oscillator output with the 1 PPS from 
GPS.  It's very important how that's done.  Suppose that it's done by dividing 
the oscillator's output down to 1 PPS and then measuring the time interval 
between the two pulses.  The one shot resolution of that TI counter is a point 
on the Allan plot at 1 second (because we're using 1 pulse/second).  You can 
draw a line through this point slopping down at 45 degrees to indicate what 
stability you will get if you average for longer times or if you make 
measurements at a faster pace.

If that line is below the GPS - Oscillator intersection point then all is well. 
  But it the TI line is above the intersection point then the TI counter is 
limiting the performance of the GPSDO.

Of all the GPSDOs kits out there It's my understanding that the Brooks Shera is 
the best.  But it was designed back when the 8 channel Motorola GPS receivers 
were the standard.  But now with the M12+T (or other newer 12 channel) 
receivers that have much better performance it's limited by the TI counter.

It's difficult to answer your questions in a general way.  One of the problems 
has to do with the sensitivity and range of the EFC input to the oscillator. 
On one hand you want to use a D/A converter that has very small steps in order 
to allow tuning the oscillator with steps of E-12 or smaller.  And if you know 
the direction of drift and set the manual adjustment so the EFC is very near 
one end of it's range you then have almost the full range available until the 
next manual adjustment is needed.  When you combine both of these it turns out 
that you can't buy a DAC with that many bits.  So you can see that it's hard to 
generalize about what the stability will be at 24 hours for different designs 
of GPSDO.

A well designed GPSDO would have long term performance that was about the same 
as the GPS receiver.  Tom has a number of plots on his web page, but this one 
compares a number of GPS receivers.
http://www.leapsecond.com/pages/3gps/gps-adev-mdev.gif
The Datum2000 is a GPSDO, not just a receiver.
It would be interesting to see what happens to these receivers if the test was 
extended.  I expect that they all would bottom out and go horizontal in after a 
few days.  But at different stability values.  Seeing that would answer your 
question in a better way.

I have some TF info at:
http://www.pacificsites.com/~brooke/timefreq.shtml

Have Fun,

Brooke Clarke, N6GCE
http://www.PRC68.com
http://www.pacificsites.com/~brooke/PRC68COM.shtml
http://www.precisionclock.com


Mike Feher wrote:
 I have to show my ignorance here, because this has been bothering me for a
 while, and, I wonder if there is a relatively simple answer. This question
 has to do with frequency accuracy and stability only. Also, let's talk of
 long term like 24 hours or more, so let's ignore phase noise and just
 concern ourselves with long term accuracy/stability. If I have three
 separate oscillators, let's say a Rubidium, an OCXO and a TCXO all with
 EFC's capable of closing the loop to lock to GPS, what kind of absolute
 frequency difference should I see amongst the three at any given time,
 random times, or, over the entire test period. Let's also make it simple and
 say all three are at 10 MHz nominal, unlocked to GPS. When locked, and
 properly designed with a narrow loop filter, I would expect the long term
 accuracy to be very close amongst all 3 oscillators. Certainly better than a
 few parts in 10^-11. First, am I wrong in this assumption? In either case,
 crystals, and even Rubidium cells do age, while at different rates, so, it
 is possible, that if lock with GPS is lost for some reason, because the
 oscillator may have drifted/aged out of loop range, it cannot be disciplined
 again. I, for the time being, also assume that the EFC on all 3 oscillators
 has a range wide enough to keep the oscillator locked even as it ages. Are
 the narrow loop bandwidths and wide EFC ranges contradictory? So, to
 reiterate the question, if I was clear enough, what kind of frequency
 excursions should I anticipate to see amongst my three disciplined
 oscillators in lets say 24 hours, or in a month. Assume GPS disciplining was
 working all of that 

Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread John Ackermann N8UR
[EMAIL PROTECTED] wrote:

 Also, try making the sine wave of the PRS10 into a nice, fast edge rate  
 square-wave with a Fairchild NC7SZ04 driver inverter for example. The 
 Wavecrest  
 units don't work as well with sine waves as with square waves. On the SZ04, 
 put 
  a 10nF cap in series to the input, and a 1 MEG resistor from it's input pin 
 to  it's output pin. Feed the chip from a very low noise 5V power source. 
 Insert a  40-50 Ohm resistor into the output path going to the coax. Put an  
 attenuator/AC-coupler into the coax so the Wavecrest doesen't get  overloaded 
 by 
 the 2.5Vpp DC signal. Result: low noise comparator/buffer :)

And on the opposite front, the TSC-5120 isn't very happy with a square 
wave input, particularly if harmonics fall within its input range.  The 
symptom is lots of spurs showing up in the phase noise plots.  Adding a 
low pass filter with an appropriate cutoff helps a lot.

John

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] Wavecrest noise measurement hardware

2007-04-25 Thread Magnus Danielson
From: John Miles [EMAIL PROTECTED]
Subject: [time-nuts] Wavecrest noise measurement hardware
Date: Wed, 25 Apr 2007 11:43:30 -0700
Message-ID: [EMAIL PROTECTED]

 As an aside, how do the Wavecrest machines work?  Do they just run the
 signal into a low-jitter ADC with a high-quality clock and derive all the
 timing information digitally, or is the box full of low-noise tunable
 synthesizers, mixers, filters, and the usual stuff?

It is a fairly traditional design in which they have a coarse clock (100 MHz
for the DTS-207X and 200 MHz for the SIA-3000), generate an error signal which
charge a capacitor and then A/D convert the accumulated voltage into digital.

They have a patent from the DTS-207X days which should give you more than
enought feeling of how it works. ;)

Cheers,
Magnus

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread SAIDJACK
 
In a message dated 4/25/2007 11:55:53 Pacific Daylight Time, [EMAIL PROTECTED]  
writes:

And  on the opposite front, the TSC-5120 isn't very happy with a square  
wave input, particularly if harmonics fall within its input  range.  The 
symptom is lots of spurs showing up in the phase  noise plots.  Adding a 
low pass filter with an appropriate cutoff  helps a lot.



Very true! Have seen these spurs on the 5210A as well. A sharp 13MHz or so  
cut-off filter will help that.
 
bye,
Said



** See what's free at http://www.aol.com.
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] Wavecrest noise measurement hardware

2007-04-25 Thread SAIDJACK
 
In a message dated 4/25/2007 11:45:41 Pacific Daylight Time, [EMAIL PROTECTED]  
writes:

As  an aside, how do the Wavecrest machines work?  Do they just run  the
signal into a low-jitter ADC with a high-quality clock and derive  all the
timing information digitally, or is the box full of low-noise  tunable
synthesizers, mixers, filters, and the usual  stuff?



Hi John,
 
Wavecrest's DTS-2070 and DTS-2075 manuals have a pretty decent discription  
including block diagrams of just how they do it, and why their approach is so  
revolutionary and so much better than anyone else's.

First off, they do true time-intervall (A to B) measurements with an  
interpolator with 800fs or less resolution.
 
So no phase noise measurement using the NIST mixer setup, or the TSC  
ADC-based cross-correlation approach (which are bandwidth limited by  design)
 
There are no mixers or filters, just fast ECL comparators with some delay  
lines (actual coax cables wound into loops etc), and the usual coarse capture  
(100MHz) and fine capture (ADC working on charge pump).
 
There are no bandwidth limits except the speed of the ECL gates/comparators  
and that's several GHz.
 
With this time intervall data, they do FFT's for phase noise, jitter  
analysis, histograms, and all sorts of other fancy stuff, including analog  
oscilloscope mode.
 
bye,
Said



** See what's free at http://www.aol.com.
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] GPS Disciplined Oscillators

2007-04-25 Thread Bill Hawkins
Mike,

If you have three different oscillators that are locked to GPS,
then any difference seen when you read your clock once a day
will be caused by the short term noise. When you take the noise
difference counts as a percentage of counts in a day, the result
is a smaller number than if the oscillators were not locked to
GPS.

Some people on this list are worried down to the phase noise
level because they are correlating events that have been timed
by clocks whose only link is GPS or a portable secondary time
standard. Primary standards tend not to be portable, but quantum
physics could change that.

Locking means that there is only a phase difference between GPS
and the oscillator, not a frequency difference. For 10 MHz, that's
100 nanoseconds or (much) better while locked. Said another way,
if GPS lock is maintained for some length of time, the maximum
difference at the end of that length of time is 100 nanoseconds,
plus the drift of GPS, if any.

Ageing is compensated by a control servo while the frequencies are
locked. The full name for lock is phase lock which means that any
phase error is restored to nominal by the control servo. The servo is
constrained by a long time constant filter so that it won't dither
around zero error, trying to follow every noise spike and always
being too late.

Losing lock means that the control servo has saturated, and can
no longer move in the direction required to maintain lock. At that
point, there is no controlled relationship between GPS and the
oscillator. The performance of the oscillator is then the performance
with no GPS available.

You can measure ageing by measuring the output of the controller.

If the loss of lock was due to some hiccup, then lock could be restored
but the clock counter will be wrong by the number of counts added or
dropped while lock was lost.

Hope that helps.

Bill Hawkins


-Original Message-
From: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On
Behalf Of Mike Feher
Sent: Wednesday, April 25, 2007 11:46 AM
To: 'Discussion of precise time and frequency measurement'
Subject: [time-nuts] GPS Disciplined Oscillators

I have to show my ignorance here, because this has been bothering me for
a while, and, I wonder if there is a relatively simple answer. This
question has to do with frequency accuracy and stability only. Also,
let's talk of long term like 24 hours or more, so let's ignore phase
noise and just concern ourselves with long term accuracy/stability. If I
have three separate oscillators, let's say a Rubidium, an OCXO and a
TCXO all with EFC's capable of closing the loop to lock to GPS, what
kind of absolute frequency difference should I see amongst the three at
any given time, random times, or, over the entire test period. Let's
also make it simple and say all three are at 10 MHz nominal, unlocked to
GPS. When locked, and properly designed with a narrow loop filter, I
would expect the long term accuracy to be very close amongst all 3
oscillators. Certainly better than a few parts in 10^-11. First, am I
wrong in this assumption? In either case, crystals, and even Rubidium
cells do age, while at different rates, so, it is possible, that if lock
with GPS is lost for some reason, because the oscillator may have
drifted/aged out of loop range, it cannot be disciplined again. I, for
the time being, also assume that the EFC on all 3 oscillators has a
range wide enough to keep the oscillator locked even as it ages. Are the
narrow loop bandwidths and wide EFC ranges contradictory? So, to
reiterate the question, if I was clear enough, what kind of frequency
excursions should I anticipate to see amongst my three disciplined
oscillators in lets say 24 hours, or in a month. Assume GPS disciplining
was working all of that time (can I even assume that with aging?). BTW,
how is my assumption regarding the oscillators aging? If the oscillator
basic frequency determining element drifts out of lock range, during
lock, will it
stay in lock? - Thanks in advance for any enlightenment - Mike  

 
 
Mike B. Feher, N4FS
89 Arnold Blvd.
Howell, NJ, 07731
732-886-5960
 
 


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


[time-nuts] Updated VLBI tutorial available on gpstime.com

2007-04-25 Thread Tom Clark, K3IO
Next week the VLBI community is having another TOW (Technical Operations
Workshop) at the Haystack Observatory NW of Boston. The audience for the
TOW is mainly the technicians at each site who drive the telescopes and
keep the hardware running. VLBI is crucially dependent on timing and
every station is equipped with at least one Hydrogen Maser. As usual, I
have been tapped as the timing teacher.

My PowerPoint class notes Timing for VLBI are now posted available on
http://gpstime.com/.

These notes include some fairly recent data on how well the new
real-time de-sawtooth hardware in Rick's CNS Clock II works as
compared with the software correction we have been using for many years
(the two agree to 300 psec RMS now that we learned how to tune the
programmable delay line).

73, Tom

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread Henk ten Pierick

Hi,

Thank you all for the frequency plots of the PRS10. It is very  
interesting to see these and find the differences.
In the mean time I have restarted my PRS10 and have seen that some  
spurious signals are large at start up and reduce after that. I have  
attached a screen shot of the hp8590 spectrum analyzer showing the  
spurious. The spurious seem to be harmonics of the 357MHz  
synthesizer, we see the 1st, 2nd, 3rd and 4th harmonic. (The marker  
is on the 3rd harmonic). The spurious between 500MHz and 700MHz are  
not yet explained. Around 357MHz are sidebands at 290kHz distance.


Does anyone have schematics?

Henk





___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread Poul-Henning Kamp

I'm going to ask a seemingly stupid question, but bear with me:

Are you sure those signals are really present on the output ?

The reason why I ask is that you are in a frequency territory where
EMI is both radiated and conducted so you have to be really careful
with your setup, grounding in particular, to make sure you do not
actually measure the PRS10 acting as an antenna.

The best way to ensure this, is to run your PRS10 off batteries,
Sealed Lead Acid for instance, and the only other connection to
the PRS10 should be the coax to the spectrum analyzer.

Poul-Henning

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
[EMAIL PROTECTED] | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] Updated VLBI tutorial available on gpstime.com

2007-04-25 Thread Brooke Clarke
Hi Tom:

I didn't see in the VLBI slides what was done to tune the delay line, but did 
see a hint that there's been a firmware upgrade for it.  Can you elaborate?

I like the Cursor Clock on your web page.

Have Fun,

Brooke Clarke
http://www.PRC68.com
http://www.pacificsites.com/~brooke/PRC68COM.shtml
http://www.precisionclock.com



Tom Clark, K3IO wrote:
 Next week the VLBI community is having another TOW (Technical Operations
 Workshop) at the Haystack Observatory NW of Boston. The audience for the
 TOW is mainly the technicians at each site who drive the telescopes and
 keep the hardware running. VLBI is crucially dependent on timing and
 every station is equipped with at least one Hydrogen Maser. As usual, I
 have been tapped as the timing teacher.
 
 My PowerPoint class notes Timing for VLBI are now posted available on
 http://gpstime.com/.
 
 These notes include some fairly recent data on how well the new
 real-time de-sawtooth hardware in Rick's CNS Clock II works as
 compared with the software correction we have been using for many years
 (the two agree to 300 psec RMS now that we learned how to tune the
 programmable delay line).
 
 73, Tom
 
 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread Henk ten Pierick
Hi Poul-Henning,

1. It is not a stupid question but a very valid one.

2. I was aware of the EMI possibility and tried ferrites on the  
supply and signal wires with no result.

3. I will try batteries but have to find enough of them.

4. On the spectrum analyzer, I checked the amplitude setting and 10dB  
change gave 10dB change of signal level. Also a mixer level change  
had no influence on the spurious.

Henk


On Apr 25, 2007, at 22:07, Poul-Henning Kamp wrote:


 I'm going to ask a seemingly stupid question, but bear with me:

 Are you sure those signals are really present on the output ?

 The reason why I ask is that you are in a frequency territory where
 EMI is both radiated and conducted so you have to be really careful
 with your setup, grounding in particular, to make sure you do not
 actually measure the PRS10 acting as an antenna.

 The best way to ensure this, is to run your PRS10 off batteries,
 Sealed Lead Acid for instance, and the only other connection to
 the PRS10 should be the coax to the spectrum analyzer.

 Poul-Henning

 -- 
 Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
 [EMAIL PROTECTED] | TCP/IP since RFC 956
 FreeBSD committer   | BSD since 4.3-tahoe
 Never attribute to malice what can adequately be explained by  
 incompetence.

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread SAIDJACK
 
In a message dated 4/25/2007 12:49:47 Pacific Daylight Time,  
[EMAIL PROTECTED] writes:

spurious. The spurious seem to be harmonics of the 357MHz   
synthesizer, we see the 1st, 2nd, 3rd and 4th harmonic. (The marker   
is on the 3rd harmonic). The spurious between 500MHz and 700MHz are   
not yet explained. Around 357MHz are sidebands at 290kHz  distance.

Does anyone have schematics?




Hi Henk,
 
if you can't fix the unit (seems it has much more spurs than my unit) then  
maybe a simple high-quality low ESR 100pF to ground after a 4.7 Ohm  series 
resistor on the 10MHz output may reduce the high-frequency  stuff a bit...
 
bye,
Said  



** See what's free at http://www.aol.com.
___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread Poul-Henning Kamp
In message [EMAIL PROTECTED], Henk ten Pi
erick writes:

2. I was aware of the EMI possibility and tried ferrites on the  
supply and signal wires with no result.

If your powersupply is not very high quality, you are almost certain
to create a ground-loop through the power-cords of the power-supply
and the spectrum analyzer.

That's why batteries are such a good thing: they don't offer a path
for ground loops.

If you have a isolating transformer, you can try that in the signal
cable between the oscilloscope and the PRS10, but it's not as good
as running off batteries.

-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
[EMAIL PROTECTED] | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] retry: PRS10 has spurious frequencies

2007-04-25 Thread Dave Brown
Henk
Do any of the spurious signals show on the SA with a search antenna 
(located in your lab environment)connected instead of the PRS10?
DaveB

- Original Message - 
From: Henk ten Pierick [EMAIL PROTECTED]
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com
Sent: Thursday, April 26, 2007 8:42 AM
Subject: Re: [time-nuts] retry: PRS10 has spurious frequencies


 Hi Poul-Henning,

 1. It is not a stupid question but a very valid one.

 2. I was aware of the EMI possibility and tried ferrites on the
 supply and signal wires with no result.

 3. I will try batteries but have to find enough of them.

 4. On the spectrum analyzer, I checked the amplitude setting and 
 10dB
 change gave 10dB change of signal level. Also a mixer level change
 had no influence on the spurious.

 Henk


 On Apr 25, 2007, at 22:07, Poul-Henning Kamp wrote:


 I'm going to ask a seemingly stupid question, but bear with me:

 Are you sure those signals are really present on the output ?

 The reason why I ask is that you are in a frequency territory where
 EMI is both radiated and conducted so you have to be really careful
 with your setup, grounding in particular, to make sure you do not
 actually measure the PRS10 acting as an antenna.

 The best way to ensure this, is to run your PRS10 off batteries,
 Sealed Lead Acid for instance, and the only other connection to
 the PRS10 should be the coax to the spectrum analyzer.

 Poul-Henning

 -- 
 Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
 [EMAIL PROTECTED] | TCP/IP since RFC 956
 FreeBSD committer   | BSD since 4.3-tahoe
 Never attribute to malice what can adequately be explained by
 incompetence.

 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


 ___
 time-nuts mailing list
 time-nuts@febo.com
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 

___
time-nuts mailing list
time-nuts@febo.com
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts


Re: [time-nuts] Updated VLBI tutorial available on gpstime.com

2007-04-25 Thread Tom Clark, K3IO
Brooke Clarke asked:
 I didn't see in the VLBI slides what was done to tune the delay
 line, but did see a hint that there's been a firmware upgrade for it. 
 Can you elaborate?
Rick's latest version uses a delay line with more per-bit precision (now
150 psec steps). We also found that Motorola (correction byte) and
Maxim/DSI (delay line bits) seemed to use different definitions of  one
nsec and Rick needed to sort out a scale difference (the reason my new
slides #29  30 are different from what's in our PTTI talk).

I think Rick got the cursor clock code from TvB.

73, Tom