Re: [time-nuts] Couple of questions for Racal 1992 owners

2009-08-27 Thread iov...@inwind.it
I have two 1992, and actually one has to terminate the output to see the real 
waveform.
They have both the 04E option, that I believe is a nice one. In long term 
measurements, I've found that mine have an aging rate in the range of 1x10E-10 
per day, or better (compared to Rb). The crystal is cut to 5 MHz, then doubled.
The time required to stabilize after power on is normal.
Antonio I8IOV


 My 1992 has a similarly ugly reference output - until I terminate it.  
 Then it's a fairly nice square wave between 0 and 1 volt.  Rise  fall 
 times are in the range of 5 ns.  The manual doesn't state what the wave 
 shape is, only that it's  600 mVp-p into 50 ohms. I have a brochure 
 that says TTL levels giving approx. 1Vp-p into 50 ohms.  My phase 
 measurements also have a similar amount of 'noise' to yours.
 
 Which timebase does yours have?  I've got option 4E (similar to 4B).  
 Although the timebase appears to be stable, I've noticed that the 
 counter drifts about 5e-9 during warmup, even when the oscillator has 
 been on standby for many days.  It takes a couple of hours to settle 
 down.  Does yours do anything similar?
 
 Ed
 
 John Green wrote:
  Just acquired a used 1992 off eBay. Excellent, almost like new with manual
  and even spare fuses. I noticed that the internal reference's waveform is
  more like a nasty sawtooth than a nice sine wave. Is that normal? Also, when
  I hook my GPSDO through a two way splitter and two different length cables
  and measure phase, I get a +\- 3 LSD ambiguity. Is that about right?
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.

 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Couple of questions for Racal 1992 owners

2009-08-27 Thread Lux, Jim (337C)
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
 Behalf Of Ed Palmer
 Sent: Thursday, August 27, 2009 12:06 PM
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Couple of questions for Racal 1992 owners
 
 Thanks for the info Antonio.  Racal didn't spec the warmup
 characteristics of the counter as a whole and it seemed a little odd
 that the warmup drift for the counter was as large as 10 days of
 oscillator drift.
 
 Ed

Intuitively, this wouldn't surprise me.  You set the OCXO up so that the 
temperature is somewhat higher than the highest expected ambient (so, say, it's 
at 60C) and pick a crystal that has a flat freq/temp characteristic near there. 
 So, when you turn on, your crystal is at, say, 25C, where the slope of the 
freq/temp curve is pretty steep.  I'll bet someone has some curves out there 
that could give a quick order of magnitude estimate...

I have a poor quality chart here that shows an AT cut with about 40 ppm change 
for a 50 degree C change, and an SC changes about 5 ppm over the same span.  
Since a decent oscillator has aging of small ppm/year, that's orders of 
magnitude bigger for the temperature effect.

Is this a SC or an AT cut?

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Couple of questions for Racal 1992 owners

2009-08-27 Thread Ed Palmer
Thanks for the info Antonio.  Racal didn't spec the warmup 
characteristics of the counter as a whole and it seemed a little odd 
that the warmup drift for the counter was as large as 10 days of 
oscillator drift.


Ed

iov...@inwind.it wrote:

I have two 1992, and actually one has to terminate the output to see the real 
waveform.
They have both the 04E option, that I believe is a nice one. In long term 
measurements, I've found that mine have an aging rate in the range of 1x10E-10 
per day, or better (compared to Rb). The crystal is cut to 5 MHz, then doubled.
The time required to stabilize after power on is normal.
Antonio I8IOV


  
My 1992 has a similarly ugly reference output - until I terminate it.  
Then it's a fairly nice square wave between 0 and 1 volt.  Rise  fall 
times are in the range of 5 ns.  The manual doesn't state what the wave 
shape is, only that it's  600 mVp-p into 50 ohms. I have a brochure 
that says TTL levels giving approx. 1Vp-p into 50 ohms.  My phase 
measurements also have a similar amount of 'noise' to yours.


Which timebase does yours have?  I've got option 4E (similar to 4B).  
Although the timebase appears to be stable, I've noticed that the 
counter drifts about 5e-9 during warmup, even when the oscillator has 
been on standby for many days.  It takes a couple of hours to settle 
down.  Does yours do anything similar?


Ed

John Green wrote:


Just acquired a used 1992 off eBay. Excellent, almost like new with manual
and even spare fuses. I noticed that the internal reference's waveform is
more like a nasty sawtooth than a nice sine wave. Is that normal? Also, when
I hook my GPSDO through a two way splitter and two different length cables
and measure phase, I get a +\- 3 LSD ambiguity. Is that about right?

  




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Couple of questions for Racal 1992 owners

2009-08-27 Thread WarrenS
I should of added 
Yes, It's  normal for phase measurement to be 3 counts OR anything else 
depending on the frequency you are measuring.
The unit measures Time delay to 1ns, and calculates Phase which can be any 
number depending on Freq.
Use Time interval not phase to see what the noise is. 


My typical  Dana 1992 performance with the good OXCO is: (sample size of ONE) 
 4 year average aging rate 0.2 e-10 /day  (Attached)
Day to Day uncertainty around 3 e-10 (same conditions each day, morning at same 
room temp, reading 1 min after turn-on from standby)
Turn on warm-up drift  First 15 sec =1e-10,  First 1 min = 2 e-10
Warm up 1+ hr about =1e-9
OSC stability throughout the day after 1 hr turn (due to changing line, room 
temp, etc about 5e-10)
Time interval noise measurement of less than 1 ns (no 1 count jitter about 1/2 
the time as the delay between signals change). 
(Using two Square wave inputs from separate low noise sources , the first is 
100 Hz the second is 100KHz, so max of 100us delay.)
Turn on data above is Always from standby, and When not using the unit,  it is 
always in standby, except for a couple of times per year when power fails.

For your ambiguity test to eliminate LOTS of possible external noise sources, 
try  feeding it a single good square wave in the range of 10 Hz to 10MHz, (Rise 
 fall times in the 5ns range)
Trigger on common, start slope to positive, stop slop set to neg, 
Use and the trigger level of each channel to cause fine delay adjustment of  + 
- 5ns between signals.
Mine has noise mostly under 1 ns count. i.e can adj the trigger level so jitter 
is less than or equal to 1ns.

I have not looked at OSC out for a long time, most data above is from measuring 
the 10 sec freq average of a 10MHz. GPS tracked Osc
 
ws
***
From: iov...@inwind.it


I have two 1992, and actually one has to terminate the output to see the real 
waveform.
They have both the 04E option, that I believe is a nice one. In long term 
measurements, I've found that mine have an aging rate in the range of 1x10E-10 
per day, or better (compared to Rb). The crystal is cut to 5 MHz, then doubled.
The time required to stabilize after power on is normal.
Antonio I8IOV

***
 My 1992 has a similarly ugly reference output - until I terminate it.  
 Then it's a fairly nice square wave between 0 and 1 volt.  Rise  fall 
 times are in the range of 5 ns.  The manual doesn't state what the wave 
 shape is, only that it's  600 mVp-p into 50 ohms. I have a brochure 
 that says TTL levels giving approx. 1Vp-p into 50 ohms.  My phase 
 measurements also have a similar amount of 'noise' to yours.
 
 Which timebase does yours have?  I've got option 4E (similar to 4B).  
 Although the timebase appears to be stable, I've noticed that the 
 counter drifts about 5e-9 during warmup, even when the oscillator has 
 been on standby for many days.  It takes a couple of hours to settle 
 down.  Does yours do anything similar?
 
 Ed
*
 
 John Green wrote:
  Just acquired a used 1992 off eBay. Excellent, almost like new with manual
  and even spare fuses. I noticed that the internal reference's waveform is
  more like a nasty sawtooth than a nice sine wave. Is that normal? Also, when
  I hook my GPSDO through a two way splitter and two different length cables
  and measure phase, I get a +\- 3 LSD ambiguity. Is that about right?
  ___





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Couple of questions for Racal 1992 owners

2009-08-27 Thread Ed Palmer



Lux, Jim (337C) wrote:

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Ed Palmer
Sent: Thursday, August 27, 2009 12:06 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Couple of questions for Racal 1992 owners

Thanks for the info Antonio.  Racal didn't spec the warmup
characteristics of the counter as a whole and it seemed a little odd
that the warmup drift for the counter was as large as 10 days of
oscillator drift.

Ed



Intuitively, this wouldn't surprise me.  You set the OCXO up so that the 
temperature is somewhat higher than the highest expected ambient (so, say, it's 
at 60C) and pick a crystal that has a flat freq/temp characteristic near there. 
 So, when you turn on, your crystal is at, say, 25C, where the slope of the 
freq/temp curve is pretty steep.  I'll bet someone has some curves out there 
that could give a quick order of magnitude estimate...

I have a poor quality chart here that shows an AT cut with about 40 ppm change 
for a 50 degree C change, and an SC changes about 5 ppm over the same span.  
Since a decent oscillator has aging of small ppm/year, that's orders of 
magnitude bigger for the temperature effect.

Is this a SC or an AT cut?

  
I agree with what you've said, but on the 1991/1992 counter, there's a 
standby function which keeps the oscillator hot, to avoid the issue 
you've stated above.  So the 10 days worth of drift when you turn the 
counter on is just due to the counter warming up, not the oscillator.


The oscillators that I have (opt. 4A and 4E) are both AT crystals.  This 
isn't stated anywhere that I can find, but the frequency at startup is 
in the range of 20ppm high which is consistent with an AT cut crystal.


Ed


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] Couple of questions for Racal 1992 owners

2009-08-26 Thread John Green
Just acquired a used 1992 off eBay. Excellent, almost like new with manual
and even spare fuses. I noticed that the internal reference's waveform is
more like a nasty sawtooth than a nice sine wave. Is that normal? Also, when
I hook my GPSDO through a two way splitter and two different length cables
and measure phase, I get a +\- 3 LSD ambiguity. Is that about right?
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Couple of questions for Racal 1992 owners

2009-08-26 Thread Ed Palmer
My 1992 has a similarly ugly reference output - until I terminate it.  
Then it's a fairly nice square wave between 0 and 1 volt.  Rise  fall 
times are in the range of 5 ns.  The manual doesn't state what the wave 
shape is, only that it's  600 mVp-p into 50 ohms. I have a brochure 
that says TTL levels giving approx. 1Vp-p into 50 ohms.  My phase 
measurements also have a similar amount of 'noise' to yours.


Which timebase does yours have?  I've got option 4E (similar to 4B).  
Although the timebase appears to be stable, I've noticed that the 
counter drifts about 5e-9 during warmup, even when the oscillator has 
been on standby for many days.  It takes a couple of hours to settle 
down.  Does yours do anything similar?


Ed

John Green wrote:

Just acquired a used 1992 off eBay. Excellent, almost like new with manual
and even spare fuses. I noticed that the internal reference's waveform is
more like a nasty sawtooth than a nice sine wave. Is that normal? Also, when
I hook my GPSDO through a two way splitter and two different length cables
and measure phase, I get a +\- 3 LSD ambiguity. Is that about right?
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
  


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.