Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ...

2013-07-07 Thread Bob Smither
On 07/06/2013 07:55 AM, Adrian wrote:
> Mark,
>
> I wouldn't replace a 20V cap with an only 10V rated one, especially when the
> nominal working voltage is alrady 10V. That would be a very bad idea.
> You can always use caps with higher voltage ratings as replacements which is
> good for reliability, but don't do the other way.
>
> The exotic values HP was using (17.000 uF) are probably near impossible to
> find, except old stock (not so great with electrolytics) or you go to a
> manufacturer and get them custom made.

There are several on ebay.com.  Search for "17,000 ufd."  Sprague 17,000 ufd at
20 Volts, CD 17,000 ufd at 30 Volts, etc.

-- 
Bob Smither, PhD   Circuit Concepts, Inc.
=
   No man's life, liberty, or property are safe while the legislature
   is in session.
--Mark Twain (1866)
=
smit...@c-c-i.com  http://www.C-C-I.Com  281-331-2744(office)  -4616(fax)

<>___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
Yes that is true in Australia, but one has to prove that it is the utility 
companies fault.
The main problem is we are right across the street for the main transformer for 
the area.
So, the start of the 240V run for the whole neighbourhood basically. 
Our switchbox meter has a voltage function that I checked this morning, the 
voltage is ~255-258V on all 3 phases.
I will give a call to power Co stating this is not no, but I don't like my 
chances.

Boo..


-marki



-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Ed Palmer
Sent: Sunday, 7 July 2013 9:29 AM
To: time-nuts@febo.com
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears to be reading o/k

Check with your power company.  They may be able to switch taps on the 
transformer to reduce the voltage.  I don't know what the situation is for you, 
but in some places power companies can be forced to repair/replace equipment 
that they fry due to faults in their system.

Ed

On 7/6/2013 5:14 PM, Mark C. Stephens wrote:
> I just checked the HP manual, most of the linear equipment is specified: 240V 
> +5/-10%
> So, I am a little over at 258V here ):
> I am going to unplug everything until I can get this sorted out.
> Can't afford to lose anything more, in both time and spares obtainability.
>
>
> marki
>
>
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
> Behalf Of Jim Lux
> Sent: Sunday, 7 July 2013 8:01 AM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
> normally, PB LEDS scannning and appears to be reading o/k
>
> On 7/6/13 2:46 PM, Mark C. Stephens wrote:
>> Hopefully HP Voltage "Derated" the Cap as well  so it can handle our
>> 250V here.. We are across the road from the main transformer for the
>> area so the voltage is highest at our place, I checked the meter box
>> this morning - it is 255-258V on all 3 phases, no wonder I am having
>> problems.
>>
>
> Is that nominally 240 phase to neutral/ 415 phase to phase.  typically, the 
> utility is allowed 10% tolerance.. I see that wikipedia says AS60038 calls 
> out 230V +10% -6% so that's 253 to 216
>
> One thing to be wary of is that the utility supplies voltages at, say, 120V 
> or 240V (in the US), but utilization equipment (motors, for
> instance) is labeled as 115V or 230V.  This accounts for the 2% voltage drop 
> in the branch circuit, among other things.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Ed Palmer
Check with your power company.  They may be able to switch taps on the 
transformer to reduce the voltage.  I don't know what the situation is 
for you, but in some places power companies can be forced to 
repair/replace equipment that they fry due to faults in their system.


Ed

On 7/6/2013 5:14 PM, Mark C. Stephens wrote:

I just checked the HP manual, most of the linear equipment is specified: 240V 
+5/-10%
So, I am a little over at 258V here ):
I am going to unplug everything until I can get this sorted out.
Can't afford to lose anything more, in both time and spares obtainability.


marki



-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Jim Lux
Sent: Sunday, 7 July 2013 8:01 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears to be reading o/k

On 7/6/13 2:46 PM, Mark C. Stephens wrote:

Hopefully HP Voltage "Derated" the Cap as well  so it can handle our
250V here.. We are across the road from the main transformer for the
area so the voltage is highest at our place, I checked the meter box
this morning - it is 255-258V on all 3 phases, no wonder I am having
problems.



Is that nominally 240 phase to neutral/ 415 phase to phase.  typically, the 
utility is allowed 10% tolerance.. I see that wikipedia says AS60038 calls out 
230V +10% -6% so that's 253 to 216

One thing to be wary of is that the utility supplies voltages at, say, 120V or 
240V (in the US), but utilization equipment (motors, for
instance) is labeled as 115V or 230V.  This accounts for the 2% voltage drop in 
the branch circuit, among other things.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
I just checked the HP manual, most of the linear equipment is specified: 240V 
+5/-10%
So, I am a little over at 258V here ):
I am going to unplug everything until I can get this sorted out.
Can't afford to lose anything more, in both time and spares obtainability.


marki



-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Jim Lux
Sent: Sunday, 7 July 2013 8:01 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears to be reading o/k

On 7/6/13 2:46 PM, Mark C. Stephens wrote:
> Hopefully HP Voltage "Derated" the Cap as well  so it can handle our 
> 250V here.. We are across the road from the main transformer for the 
> area so the voltage is highest at our place, I checked the meter box 
> this morning - it is 255-258V on all 3 phases, no wonder I am having 
> problems.
>


Is that nominally 240 phase to neutral/ 415 phase to phase.  typically, the 
utility is allowed 10% tolerance.. I see that wikipedia says AS60038 calls out 
230V +10% -6% so that's 253 to 216

One thing to be wary of is that the utility supplies voltages at, say, 120V or 
240V (in the US), but utilization equipment (motors, for
instance) is labeled as 115V or 230V.  This accounts for the 2% voltage drop in 
the branch circuit, among other things.



___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Jim Lux

On 7/6/13 2:46 PM, Mark C. Stephens wrote:

Hopefully HP Voltage "Derated" the Cap as well  so it can handle our
250V here.. We are across the road from the main transformer for the
area so the voltage is highest at our place, I checked the meter box
this morning - it is 255-258V on all 3 phases, no wonder I am having
problems.




Is that nominally 240 phase to neutral/ 415 phase to phase.  typically,
the utility is allowed 10% tolerance.. I see that wikipedia says AS60038
calls out 230V +10% -6% so that's 253 to 216

One thing to be wary of is that the utility supplies voltages at, say, 
120V or 240V (in the US), but utilization equipment (motors, for 
instance) is labeled as 115V or 230V.  This accounts for the 2% voltage 
drop in the branch circuit, among other things.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
Hopefully HP Voltage "Derated" the Cap as well  so it can handle our 250V here..
We are across the road from the main transformer for the area so the voltage is 
highest at our place, I checked the meter box this morning - it is 255-258V on 
all 3 phases, no wonder I am having problems.


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Orin Eman
Sent: Sunday, 7 July 2013 2:12 AM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears to be reading o/k

What's the specified tolerance on the 17,000 uF?  You probably have quite a 
range to work within.  If you are lucky, they specified something like
+80/-20% and the 29,000 uF might be OK (measure it).


On Sat, Jul 6, 2013 at 4:18 AM, Mark C. Stephens wrote:

> C3 (17000uf @ 10v) is open circuit.
> Where am I going to get one of those from.
>
> I need some advice here, Should I just use modern low ESR solder lug 
> type soldered directly onto the board?
> Or Should I try and locate a screw terminal type Capacitor?
>
> Obviously, I want to do the best for this timer as I rely on it 
> tremendously.
>
>
> -marki
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
> On Behalf Of Mark C. Stephens
> Sent: Saturday, 6 July 2013 8:47 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons 
> selecting normally, PB LEDS scannning and appears to be reading o/k
>
> Checked the PSU voltages on the PSU card next to the 10811.
> The -5.2 is reading -4.2.
> Took a look with the scope and instead of a nice flat DC I see a 4V 
> Peak to Peak spike @ 100Hz.
> Trying to track down a service manual.
> Any ideas on this one guys, if the Pass tranny was shorted, it would 
> be a lot more than 4.2V unless the overvoltage Zener is able to sink 
> all that current...
>
>
> -marki
>
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
> On Behalf Of Mark C. Stephens
> Sent: Saturday, 6 July 2013 8:18 PM
> To: time-nuts@febo.com
> Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
> normally, PB LEDS scannning and appears to be reading o/k
>
> I went to use my 5370B todayt and noticed the LEDS on the display are 
> slowly sort of pulsing bright one after another.
> Also all the push button LEDS are going on and off one by one.
> The strangest thing is that it appears to be working o/k via HPIB.
>
> I don't have extender boards but I do have a signature analyser.
>
> Could it be something silly like one of the PSU regulator has gone kaput?
> Has anyone seen this issue before?
>
>
> -marki
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Orin Eman
What's the specified tolerance on the 17,000 uF?  You probably have quite a
range to work within.  If you are lucky, they specified something like
+80/-20% and the 29,000 uF might be OK (measure it).


On Sat, Jul 6, 2013 at 4:18 AM, Mark C. Stephens wrote:

> C3 (17000uf @ 10v) is open circuit.
> Where am I going to get one of those from.
>
> I need some advice here, Should I just use modern low ESR solder lug type
> soldered directly onto the board?
> Or Should I try and locate a screw terminal type Capacitor?
>
> Obviously, I want to do the best for this timer as I rely on it
> tremendously.
>
>
> -marki
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
> Behalf Of Mark C. Stephens
> Sent: Saturday, 6 July 2013 8:47 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting
> normally, PB LEDS scannning and appears to be reading o/k
>
> Checked the PSU voltages on the PSU card next to the 10811.
> The -5.2 is reading -4.2.
> Took a look with the scope and instead of a nice flat DC I see a 4V Peak
> to Peak spike @ 100Hz.
> Trying to track down a service manual.
> Any ideas on this one guys, if the Pass tranny was shorted, it would be a
> lot more than 4.2V unless the overvoltage Zener is able to sink all that
> current...
>
>
> -marki
>
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
> Behalf Of Mark C. Stephens
> Sent: Saturday, 6 July 2013 8:18 PM
> To: time-nuts@febo.com
> Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting
> normally, PB LEDS scannning and appears to be reading o/k
>
> I went to use my 5370B todayt and noticed the LEDS on the display are
> slowly sort of pulsing bright one after another.
> Also all the push button LEDS are going on and off one by one.
> The strangest thing is that it appears to be working o/k via HPIB.
>
> I don't have extender boards but I do have a signature analyser.
>
> Could it be something silly like one of the PSU regulator has gone kaput?
> Has anyone seen this issue before?
>
>
> -marki
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ...

2013-07-06 Thread GandalfG8
You wouldn't even need a UPS, check Google for mains voltage  conditioners.
 
APC's kit, for example, should be available worldwide and they  make a 
range that, here in the UK anyway, starts for under 50  GBP.
 
Regards
 
Nigel
GM8PZR 
 
 
In a message dated 06/07/2013 13:26:06 GMT Daylight Time,  
ma...@non-stop.com.au writes:

I  reckon, the line voltage here is 250v and the equipment is set for 240V, 
that  extra 10V on the mains is why I am having so much equipment failure.
Also  the Heat sink on the 5370B got so hot I mounted a 5" fan across it to 
keep it  at a respectable temperature.

How can I drop the Mains to 240V, I have  a boat load of gear that needs to 
be powered concurrently.
(8566A, 8568B,  3585A, 5335A, 5370B, 8901A, etc, powered on together) we 
are starting to talk  some serious current there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ...

2013-07-06 Thread GandalfG8
Running load might not be quite as bad as you'd expect, switching them  all 
on at exactly the same time would probably cause a nice fat  pulse but 
perhaps that's a bit extreme:-), allowing an adequate margin  though is always 
a 
good idea.
 
If you could tap into the supply and get an idea of running  current that 
would be a good start, after that it's likely to come  down to an appropriate 
autotransformer, fixed or variable, one of the older  style saturable 
reactance type constant voltage transformers, or perhaps an  "online" UPS, 
sinusoidal to be preferred, that will handle the input  voltage and regulate 
the 
output.
 
The latter would probably be the best, or easiest anyway, option these  
day, and assuming it meets your voltage and current needs then the off mains  
hold up time, unless you do need that too, probably wouldn't be too  much of 
an issue so that could help keep the cost down.
 
Regards
 
Nigel
GM8PZR
 
 
 
In a message dated 06/07/2013 13:26:06 GMT Daylight Time,  
ma...@non-stop.com.au writes:

I  reckon, the line voltage here is 250v and the equipment is set for 240V, 
that  extra 10V on the mains is why I am having so much equipment failure.
Also  the Heat sink on the 5370B got so hot I mounted a 5" fan across it to 
keep it  at a respectable temperature.

How can I drop the Mains to 240V, I have  a boat load of gear that needs to 
be powered concurrently.
(8566A, 8568B,  3585A, 5335A, 5370B, 8901A, etc, powered on together) we 
are starting to talk  some serious current there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
Most of the Caps I have of that size are around ~1/2 the capacitance but twice 
the voltage.
I'll go check the dumpster (...err garage) for some old gear that's due to be 
scrapped.
I am going hate myself if I have to part mule a probably working 8568A/B for a 
some sort of PSU cap.
Yeah, I see your point about the cap voltage, that hadn't occurred in my 
electrolytic candidates to me for some reason!
I'll take that one on-board definitely :)

-marki

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Chuck Harris
Sent: Saturday, 6 July 2013 11:05 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears to be reading o/k

Mark C. Stephens wrote:
> Correct, C3 (17000uf @ 10v) is measuring 39uf :) I have a 29000@10V, 
> same dimensions, do you think the rectifiers can hack that big fella?

Without a blush!

Except for the 10V part.  (Which I missed on the first post)

You have to use a capacitor that is at least 2x the operating voltage, and is 
also greater than any peak voltages that could occur in the boundary cases, 
such as the power line being at the scope's rated maximum power line voltage.

-Chuck Harris
> 1uf is quite a bit more.
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ...

2013-07-06 Thread Mark C. Stephens
Yeah, I was looking at the local suppliers and comparing items.
They all seem to stop at 10,000uf for the "normal places".
I wonder what CR4 is, It's the -5.2 bridge before C3.
I'd better check it too, must be a reason for the second time that caps failed.
I suppose I could replace CR4 with a higher rated bridge and replace C3 with 
the 29,000uf
Then I'll have prey to the transformer gods for the 8VAC that supplies CR4 
doesn't go open circuit under the strain :)
HP had some odd values for things on occasion, but look at what they achieved 
for 1970-1990...


-marki




-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Adrian
Sent: Saturday, 6 July 2013 10:55 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB ...

Mark,

I wouldn't replace a 20V cap with an only 10V rated one, especially when the 
nominal working voltage is alrady 10V. That would be a very bad idea.
You can always use caps with higher voltage ratings as replacements which is 
good for reliability, but don't do the other way.

The exotic values HP was using (17.000 uF) are probably near impossible to 
find, except old stock (not so great with electrolytics) or you go to a 
manufacturer and get them custom made.
The closest commonly available standard value is 22.000 uF, voltage rating at 
least 20V (25...35V would be great). 15.000 uF would work as well if available. 
For higher reliability you may go for a 105°C model.

Don't care too much about the mounting style. For repeated outdoor use and 
transport you can fix the cap with hot glue and / or binding straps. 
Otherwise I hope your lab isn't moving and shaking enough for the cap to 
loosen... ;)

Adrian

Mark C. Stephens schrieb:
> Hi Nigel,
>
>
> The only screw type electro can find is 29000uf@10V. it's the same dimensions.
> Should I risk the strain on the rectifiers (another 10Kuf is rather a lot)?
> Without this timer I am dead in the water so I need to do the right thing 
> here...
>
> That's why I posted on the Agilent group too, I need to be sure that I do the 
> right thing!
>
> By the way, the failed electro measures 39uf :)
>
> I reckon, the line voltage here is 250v and the equipment is set for 240V, 
> that extra 10V on the mains is why I am having so much equipment failure.
> Also the Heat sink on the 5370B got so hot I mounted a 5" fan across it to 
> keep it at a respectable temperature.
>
> How can I drop the Mains to 240V, I have a boat load of gear that needs to be 
> powered concurrently.
> (8566A, 8568B, 3585A, 5335A, 5370B, 8901A, etc, powered on together) we are 
> starting to talk some serious current there.
>
>
> -marki
>
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
> On Behalf Of gandal...@aol.com
> Sent: Saturday, 6 July 2013 9:37 PM
> To: time-nuts@febo.com
> Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
> normally, PB ...
>
> As per my reply to the same question on the HP group, you've  identified the 
> problem so just go for a reasonably close-ish equivalent and  replace it!
>   
> Regards
>   
> Nigel
> GM8PZR
>   
>   
> In a message dated 06/07/2013 12:18:44 GMT Daylight Time, 
> ma...@non-stop.com.au writes:
>
> C3  (17000uf @ 10v) is open circuit.
> Where am I going to get one of those  from.
>
> I need some advice here, Should I just use modern low ESR solder  lug type 
> soldered directly onto the board?
> Or Should I try and locate a  screw terminal type Capacitor?
>
> Obviously, I want to do the best for  this timer as I rely on it tremendously.
>
>
> -marki
>
> -----Original Message-----
> From:  time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
> On Behalf Of  Mark C. Stephens
> Sent: Saturday, 6 July 2013 8:47 PM
> To: Discussion of  precise time and frequency measurement
> Subject: Re: [time-nuts] HP 5370B  Leds pulsing slowly, buttons 
> selecting normally, PB LEDS scannning and appears  to be reading o/k
>
> Checked the PSU voltages on the PSU card next to the  10811.
> The -5.2 is reading -4.2.
> Took a look with the scope and instead  of a nice flat DC I see a 4V Peak to 
> Peak spike @ 100Hz.
> Trying to track  down a service manual.
> Any ideas on this one guys, if the Pass tranny was  shorted, it would be a 
> lot more than 4.2V unless the overvoltage Zener is able  to sink all that 
> current...
>
>
> -marki
>
>
> -Original  Message-
> From: time-nuts-boun...@febo.com  [mailto:time-nuts-boun...@febo.com] 
> On Behalf Of Mark C. Stephens
> Sent:  Saturday, 6 Ju

Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Chuck Harris

Mark C. Stephens wrote:

Correct, C3 (17000uf @ 10v) is measuring 39uf :)
I have a 29000@10V, same dimensions, do you think the rectifiers can hack that 
big fella?


Without a blush!

Except for the 10V part.  (Which I missed on the first post)

You have to use a capacitor that is at least 2x the operating voltage,
and is also greater than any peak voltages that could occur in the
boundary cases, such as the power line being at the scope's rated
maximum power line voltage.

-Chuck Harris

1uf is quite a bit more.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Chuck Harris

Mark C. Stephens wrote:

Correct, C3 (17000uf @ 10v) is measuring 39uf :)
I have a 29000@10V, same dimensions, do you think the rectifiers can hack that 
big fella?


Without a blush!

-Chuck Harris

1uf is quite a bit more.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ...

2013-07-06 Thread Adrian

Mark,

I wouldn't replace a 20V cap with an only 10V rated one, especially when 
the nominal working voltage is alrady 10V. That would be a very bad idea.
You can always use caps with higher voltage ratings as replacements 
which is good for reliability, but don't do the other way.


The exotic values HP was using (17.000 uF) are probably near impossible 
to find, except old stock (not so great with electrolytics) or you go to 
a manufacturer and get them custom made.
The closest commonly available standard value is 22.000 uF, voltage 
rating at least 20V (25...35V would be great). 15.000 uF would work as 
well if available. For higher reliability you may go for a 105°C model.


Don't care too much about the mounting style. For repeated outdoor use 
and transport you can fix the cap with hot glue and / or binding straps. 
Otherwise I hope your lab isn't moving and shaking enough for the cap to 
loosen... ;)


Adrian

Mark C. Stephens schrieb:

Hi Nigel,


The only screw type electro can find is 29000uf@10V. it's the same dimensions.
Should I risk the strain on the rectifiers (another 10Kuf is rather a lot)?
Without this timer I am dead in the water so I need to do the right thing 
here...

That's why I posted on the Agilent group too, I need to be sure that I do the 
right thing!

By the way, the failed electro measures 39uf :)

I reckon, the line voltage here is 250v and the equipment is set for 240V, that 
extra 10V on the mains is why I am having so much equipment failure.
Also the Heat sink on the 5370B got so hot I mounted a 5" fan across it to keep 
it at a respectable temperature.

How can I drop the Mains to 240V, I have a boat load of gear that needs to be 
powered concurrently.
(8566A, 8568B, 3585A, 5335A, 5370B, 8901A, etc, powered on together) we are 
starting to talk some serious current there.


-marki


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of gandal...@aol.com
Sent: Saturday, 6 July 2013 9:37 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB ...

As per my reply to the same question on the HP group, you've  identified the 
problem so just go for a reasonably close-ish equivalent and  replace it!
  
Regards
  
Nigel

GM8PZR
  
  
In a message dated 06/07/2013 12:18:44 GMT Daylight Time, ma...@non-stop.com.au writes:


C3  (17000uf @ 10v) is open circuit.
Where am I going to get one of those  from.

I need some advice here, Should I just use modern low ESR solder  lug type 
soldered directly onto the board?
Or Should I try and locate a  screw terminal type Capacitor?

Obviously, I want to do the best for  this timer as I rely on it tremendously.


-marki

-Original Message-
From:  time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of  Mark C. Stephens
Sent: Saturday, 6 July 2013 8:47 PM
To: Discussion of  precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B  Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears  to be reading o/k

Checked the PSU voltages on the PSU card next to the  10811.
The -5.2 is reading -4.2.
Took a look with the scope and instead  of a nice flat DC I see a 4V Peak to 
Peak spike @ 100Hz.
Trying to track  down a service manual.
Any ideas on this one guys, if the Pass tranny was  shorted, it would be a lot 
more than 4.2V unless the overvoltage Zener is able  to sink all that current...


-marki


-Original  Message-
From: time-nuts-boun...@febo.com  [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Mark C. Stephens
Sent:  Saturday, 6 July 2013 8:18 PM
To: time-nuts@febo.com
Subject:  [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, 
PB LEDS  scannning and appears to be reading o/k

I went to use my 5370B todayt  and noticed the LEDS on the display are slowly 
sort of pulsing bright one  after another.
Also all the push button LEDS are going on and off one by  one.
The strangest thing is that it appears to be working o/k via  HPIB.

I don't have extender boards but I do have a signature  analyser.

Could it be something silly like one of the PSU regulator has  gone kaput?
Has anyone seen this issue  before?


-marki

___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
ti

Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
Correct, C3 (17000uf @ 10v) is measuring 39uf :)
I have a 29000@10V, same dimensions, do you think the rectifiers can hack that 
big fella?
1uf is quite a bit more.
Out of interest, this cap was replaced before, it is not a blue Sprague or 
Mallory but a grey Phillips Cap.


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Adrian
Sent: Saturday, 6 July 2013 9:19 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears to be reading o/k

sounds quite typical for a dried out electrolytic cap.
Look for the unregulated -10V that goes into the regulator circuit.
Most likely C3 on the power supply motherboard is dead.

Adrian

Mark C. Stephens schrieb:
> Checked the PSU voltages on the PSU card next to the 10811.
> The -5.2 is reading -4.2.
> Took a look with the scope and instead of a nice flat DC I see a 4V Peak to 
> Peak spike @ 100Hz.
> Trying to track down a service manual.
> Any ideas on this one guys, if the Pass tranny was shorted, it would be a lot 
> more than 4.2V unless the overvoltage Zener is able to sink all that 
> current...
>
>
> -marki
>
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] 
> On Behalf Of Mark C. Stephens
> Sent: Saturday, 6 July 2013 8:18 PM
> To: time-nuts@febo.com
> Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
> normally, PB LEDS scannning and appears to be reading o/k
>
> I went to use my 5370B todayt and noticed the LEDS on the display are slowly 
> sort of pulsing bright one after another.
> Also all the push button LEDS are going on and off one by one.
> The strangest thing is that it appears to be working o/k via HPIB.
>
> I don't have extender boards but I do have a signature analyser.
>
> Could it be something silly like one of the PSU regulator has gone kaput?
> Has anyone seen this issue before?
>
>
> -marki
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ...

2013-07-06 Thread Mark C. Stephens
Hi Nigel, 


The only screw type electro can find is 29000uf@10V. it's the same dimensions.
Should I risk the strain on the rectifiers (another 10Kuf is rather a lot)?
Without this timer I am dead in the water so I need to do the right thing 
here...

That's why I posted on the Agilent group too, I need to be sure that I do the 
right thing!

By the way, the failed electro measures 39uf :)

I reckon, the line voltage here is 250v and the equipment is set for 240V, that 
extra 10V on the mains is why I am having so much equipment failure.
Also the Heat sink on the 5370B got so hot I mounted a 5" fan across it to keep 
it at a respectable temperature.

How can I drop the Mains to 240V, I have a boat load of gear that needs to be 
powered concurrently.
(8566A, 8568B, 3585A, 5335A, 5370B, 8901A, etc, powered on together) we are 
starting to talk some serious current there.


-marki


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of gandal...@aol.com
Sent: Saturday, 6 July 2013 9:37 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB ...

As per my reply to the same question on the HP group, you've  identified the 
problem so just go for a reasonably close-ish equivalent and  replace it!
 
Regards
 
Nigel
GM8PZR
 
 
In a message dated 06/07/2013 12:18:44 GMT Daylight Time, ma...@non-stop.com.au 
writes:

C3  (17000uf @ 10v) is open circuit.
Where am I going to get one of those  from.

I need some advice here, Should I just use modern low ESR solder  lug type 
soldered directly onto the board?
Or Should I try and locate a  screw terminal type Capacitor? 

Obviously, I want to do the best for  this timer as I rely on it tremendously.


-marki

-Original Message-
From:  time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of  Mark C. Stephens
Sent: Saturday, 6 July 2013 8:47 PM
To: Discussion of  precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B  Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears  to be reading o/k

Checked the PSU voltages on the PSU card next to the  10811.
The -5.2 is reading -4.2.
Took a look with the scope and instead  of a nice flat DC I see a 4V Peak to 
Peak spike @ 100Hz.
Trying to track  down a service manual.
Any ideas on this one guys, if the Pass tranny was  shorted, it would be a lot 
more than 4.2V unless the overvoltage Zener is able  to sink all that current...


-marki


-Original  Message-
From: time-nuts-boun...@febo.com  [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Mark C. Stephens
Sent:  Saturday, 6 July 2013 8:18 PM
To: time-nuts@febo.com
Subject:  [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, 
PB LEDS  scannning and appears to be reading o/k

I went to use my 5370B todayt  and noticed the LEDS on the display are slowly 
sort of pulsing bright one  after another.
Also all the push button LEDS are going on and off one by  one.
The strangest thing is that it appears to be working o/k via  HPIB.

I don't have extender boards but I do have a signature  analyser.

Could it be something silly like one of the PSU regulator has  gone kaput?
Has anyone seen this issue  before?


-marki

___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ...

2013-07-06 Thread GandalfG8
As per my reply to the same question on the HP group, you've  identified 
the problem so just go for a reasonably close-ish equivalent and  replace it!
 
Regards
 
Nigel
GM8PZR
 
 
In a message dated 06/07/2013 12:18:44 GMT Daylight Time,  
ma...@non-stop.com.au writes:

C3  (17000uf @ 10v) is open circuit.
Where am I going to get one of those  from.

I need some advice here, Should I just use modern low ESR solder  lug type 
soldered directly onto the board?
Or Should I try and locate a  screw terminal type Capacitor? 

Obviously, I want to do the best for  this timer as I rely on it  
tremendously.


-marki

-Original Message-
From:  time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
Behalf Of  Mark C. Stephens
Sent: Saturday, 6 July 2013 8:47 PM
To: Discussion of  precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B  Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears  to be reading o/k

Checked the PSU voltages on the PSU card next to the  10811.
The -5.2 is reading -4.2.
Took a look with the scope and instead  of a nice flat DC I see a 4V Peak 
to Peak spike @ 100Hz.
Trying to track  down a service manual.
Any ideas on this one guys, if the Pass tranny was  shorted, it would be a 
lot more than 4.2V unless the overvoltage Zener is able  to sink all that 
current...


-marki


-Original  Message-
From: time-nuts-boun...@febo.com  [mailto:time-nuts-boun...@febo.com] On 
Behalf Of Mark C. Stephens
Sent:  Saturday, 6 July 2013 8:18 PM
To: time-nuts@febo.com
Subject:  [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS  scannning and appears to be reading o/k

I went to use my 5370B todayt  and noticed the LEDS on the display are 
slowly sort of pulsing bright one  after another.
Also all the push button LEDS are going on and off one by  one.
The strangest thing is that it appears to be working o/k via  HPIB.

I don't have extender boards but I do have a signature  analyser.

Could it be something silly like one of the PSU regulator has  gone kaput?
Has anyone seen this issue  before?


-marki

___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions  there.


___
time-nuts  mailing list -- time-nuts@febo.com
To unsubscribe, go to  
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the  instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Adrian

sounds quite typical for a dried out electrolytic cap.
Look for the unregulated -10V that goes into the regulator circuit.
Most likely C3 on the power supply motherboard is dead.

Adrian

Mark C. Stephens schrieb:

Checked the PSU voltages on the PSU card next to the 10811.
The -5.2 is reading -4.2.
Took a look with the scope and instead of a nice flat DC I see a 4V Peak to 
Peak spike @ 100Hz.
Trying to track down a service manual.
Any ideas on this one guys, if the Pass tranny was shorted, it would be a lot 
more than 4.2V unless the overvoltage Zener is able to sink all that current...


-marki


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Mark C. Stephens
Sent: Saturday, 6 July 2013 8:18 PM
To: time-nuts@febo.com
Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, 
PB LEDS scannning and appears to be reading o/k

I went to use my 5370B todayt and noticed the LEDS on the display are slowly 
sort of pulsing bright one after another.
Also all the push button LEDS are going on and off one by one.
The strangest thing is that it appears to be working o/k via HPIB.

I don't have extender boards but I do have a signature analyser.

Could it be something silly like one of the PSU regulator has gone kaput?
Has anyone seen this issue before?


-marki

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Azelio Boriani
Electrolytic capacitors: check them on the PSU, usually they are the
first to look for.

On Sat, Jul 6, 2013 at 12:46 PM, Mark C. Stephens  wrote:
> Checked the PSU voltages on the PSU card next to the 10811.
> The -5.2 is reading -4.2.
> Took a look with the scope and instead of a nice flat DC I see a 4V Peak to 
> Peak spike @ 100Hz.
> Trying to track down a service manual.
> Any ideas on this one guys, if the Pass tranny was shorted, it would be a lot 
> more than 4.2V unless the overvoltage Zener is able to sink all that 
> current...
>
>
> -marki
>
>
> -Original Message-
> From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
> Behalf Of Mark C. Stephens
> Sent: Saturday, 6 July 2013 8:18 PM
> To: time-nuts@febo.com
> Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
> normally, PB LEDS scannning and appears to be reading o/k
>
> I went to use my 5370B todayt and noticed the LEDS on the display are slowly 
> sort of pulsing bright one after another.
> Also all the push button LEDS are going on and off one by one.
> The strangest thing is that it appears to be working o/k via HPIB.
>
> I don't have extender boards but I do have a signature analyser.
>
> Could it be something silly like one of the PSU regulator has gone kaput?
> Has anyone seen this issue before?
>
>
> -marki
>
> ___
> time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
C3 (17000uf @ 10v) is open circuit.
Where am I going to get one of those from.

I need some advice here, Should I just use modern low ESR solder lug type 
soldered directly onto the board?
Or Should I try and locate a screw terminal type Capacitor? 

Obviously, I want to do the best for this timer as I rely on it tremendously.


-marki

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Mark C. Stephens
Sent: Saturday, 6 July 2013 8:47 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting 
normally, PB LEDS scannning and appears to be reading o/k

Checked the PSU voltages on the PSU card next to the 10811.
The -5.2 is reading -4.2.
Took a look with the scope and instead of a nice flat DC I see a 4V Peak to 
Peak spike @ 100Hz.
Trying to track down a service manual.
Any ideas on this one guys, if the Pass tranny was shorted, it would be a lot 
more than 4.2V unless the overvoltage Zener is able to sink all that current...


-marki


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Mark C. Stephens
Sent: Saturday, 6 July 2013 8:18 PM
To: time-nuts@febo.com
Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, 
PB LEDS scannning and appears to be reading o/k

I went to use my 5370B todayt and noticed the LEDS on the display are slowly 
sort of pulsing bright one after another.
Also all the push button LEDS are going on and off one by one.
The strangest thing is that it appears to be working o/k via HPIB.

I don't have extender boards but I do have a signature analyser.

Could it be something silly like one of the PSU regulator has gone kaput?
Has anyone seen this issue before?


-marki

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
Checked the PSU voltages on the PSU card next to the 10811.
The -5.2 is reading -4.2.
Took a look with the scope and instead of a nice flat DC I see a 4V Peak to 
Peak spike @ 100Hz.
Trying to track down a service manual.
Any ideas on this one guys, if the Pass tranny was shorted, it would be a lot 
more than 4.2V unless the overvoltage Zener is able to sink all that current...


-marki


-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf 
Of Mark C. Stephens
Sent: Saturday, 6 July 2013 8:18 PM
To: time-nuts@febo.com
Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, 
PB LEDS scannning and appears to be reading o/k

I went to use my 5370B todayt and noticed the LEDS on the display are slowly 
sort of pulsing bright one after another.
Also all the push button LEDS are going on and off one by one.
The strangest thing is that it appears to be working o/k via HPIB.

I don't have extender boards but I do have a signature analyser.

Could it be something silly like one of the PSU regulator has gone kaput?
Has anyone seen this issue before?


-marki

___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


[time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k

2013-07-06 Thread Mark C. Stephens
I went to use my 5370B todayt and noticed the LEDS on the display are slowly 
sort of pulsing bright one after another.
Also all the push button LEDS are going on and off one by one.
The strangest thing is that it appears to be working o/k via HPIB.

I don't have extender boards but I do have a signature analyser.

Could it be something silly like one of the PSU regulator has gone kaput?
Has anyone seen this issue before?


-marki

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts