Hi Nigel,
The only screw type electro can find is 29000uf@10V. it's the same dimensions. Should I risk the strain on the rectifiers (another 10Kuf is rather a lot)? Without this timer I am dead in the water so I need to do the right thing here... That's why I posted on the Agilent group too, I need to be sure that I do the right thing! By the way, the failed electro measures 39uf :) I reckon, the line voltage here is 250v and the equipment is set for 240V, that extra 10V on the mains is why I am having so much equipment failure. Also the Heat sink on the 5370B got so hot I mounted a 5" fan across it to keep it at a respectable temperature. How can I drop the Mains to 240V, I have a boat load of gear that needs to be powered concurrently. (8566A, 8568B, 3585A, 5335A, 5370B, 8901A, etc, powered on together) we are starting to talk some serious current there. -marki -----Original Message----- From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf Of gandal...@aol.com Sent: Saturday, 6 July 2013 9:37 PM To: time-nuts@febo.com Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB ... As per my reply to the same question on the HP group, you've identified the problem so just go for a reasonably close-ish equivalent and replace it! Regards Nigel GM8PZR In a message dated 06/07/2013 12:18:44 GMT Daylight Time, ma...@non-stop.com.au writes: C3 (17000uf @ 10v) is open circuit. Where am I going to get one of those from. I need some advice here, Should I just use modern low ESR solder lug type soldered directly onto the board? Or Should I try and locate a screw terminal type Capacitor? Obviously, I want to do the best for this timer as I rely on it tremendously. -marki -----Original Message----- From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf Of Mark C. Stephens Sent: Saturday, 6 July 2013 8:47 PM To: Discussion of precise time and frequency measurement Subject: Re: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k Checked the PSU voltages on the PSU card next to the 10811. The -5.2 is reading -4.2. Took a look with the scope and instead of a nice flat DC I see a 4V Peak to Peak spike @ 100Hz. Trying to track down a service manual. Any ideas on this one guys, if the Pass tranny was shorted, it would be a lot more than 4.2V unless the overvoltage Zener is able to sink all that current... -marki -----Original Message----- From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On Behalf Of Mark C. Stephens Sent: Saturday, 6 July 2013 8:18 PM To: time-nuts@febo.com Subject: [time-nuts] HP 5370B Leds pulsing slowly, buttons selecting normally, PB LEDS scannning and appears to be reading o/k I went to use my 5370B todayt and noticed the LEDS on the display are slowly sort of pulsing bright one after another. Also all the push button LEDS are going on and off one by one. The strangest thing is that it appears to be working o/k via HPIB. I don't have extender boards but I do have a signature analyser. Could it be something silly like one of the PSU regulator has gone kaput? Has anyone seen this issue before? -marki _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there. _______________________________________________ time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts and follow the instructions there.