Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-09 Thread Dave M
Ulrich,
If you're interested in a very good 5MHz OXCO, then I recommend the MTI 
Milliren model 260. (see Ebay #
271915776504). The data sheet is available.from several places, including 
MTI's site.  The daily aging rate for it is spec'ed at 5e-11, thermal 
stability is 2e-10 over -30C to +70C.  That's roughly 10x better than the 
10811, and much cheaper.
The Z3812A GPSDO units both use the MTI OXCO, so you could buy 1 or 2 of the 
REF-0 units and get the good OXCO performance plus all the other stuff that 
surrounds it.

Cheers,
Dave M




KA2WEU--- via time-nuts wrote:
 How good or bad is the 10811 and which one on   ebay   is a better
 choice ?
 I am looking for a very goo 5 MHz crystal oscillator  with
 documentation .

 Thanks, Ulrich N1UL


 In a message dated 8/8/2015 7:16:37 P.M. Eastern Daylight Time,
 kb...@n1k.org writes:

 Hi

 Ok, so John’s observation was the correct one. The data  we are
 looking at
 is *not* the
 performance of the OCXO’s but the strange  behavior of the counter at
 short ADEV Tau’s.

 Sorry for my bashing your  poor 10811.

 Bob

 On Aug 8, 2015, at 12:25 PM,  tim...@timeok.it wrote:

 Hi all,

 I try to  ansver to all you:

 Luciano, how was the blue trace taken? Is  this from your DMTD
 project?
 If so, it's looking promising.
 The green  and magenta traces are definitely in the right ballpark
 for
 measurements on a  5370-class counter. At 3E-11 @ t=12s the magenta
 trace is optimistic but not  outrageously so, while the green trace
 looks exactly like I'd expect for a  typical 10811 measured on a
 5370. The observed noise is
 due entirely to the  counter until about t=200s. We see a glimpse of
 the 10811's typical ADEV at  about 250 seconds, just before either
 drift or ADEV uncertainty causes the  trace to turn upwards. A longer
 run would be needed to distinguish between  these two situations.

 -- john, KE5FX
 Miles Design  LLC

 r: no, all the measurements are taken using an  HP53132A  in
 frequency
 mode. The difference can be the gate time. Using 1  second the
 resolution is lower than using 2 Second that permit the max  counter
 resolution.

 Hi

 Well an  un-stated assumption of mine was that they all came from the
 same measurement  system and that it
 had the same floor under all  circumstances….

 Bob

 r: all the measurements  are under the same conditions except for the
 gate time of the counter.

 Hi Luciano,

 Can you give me the link to your ADEV  posting image about 10811 vs
 105
 oscillators?  I had it and now can't  seem to find it.  I wanted to
 look at your plots as I read John Miles'  comments.  Many thanks.

 I have two very high  performing HP10811-60109 OCXO units which I got
 from Corby.

  Many thanks.

 Jim Robbins
 N1JR

 r:  Jim, I will load soon some files on my site. The 105B I have is a
 fantastic  exception, unfortunately it have a defect, may be a bad
 solder inside  cause randomly a phase jump and return to the original
 phase trend, so all The  long term ADEV are distorted by this problem
 I have to fix.

  Tom VB, i will do all the cross measurements on 10811, 00105 and
 rubidium I  have and i will upload they but I need time to do this.
 Unfortunately  I have only the HP53132A as TIC and my best reference
 are
 four HP5065A. I  normally use the counter function because the TI
 function
 on 1 PPS have 100  time less resolution.
 Will be interesting doing also the Phase noise  tests. I will do it.
 please see:
 http://www.timeok.it/files/time_and_frequency_house_standard_201r.pdf

 Luciano
 www.timeok.it





 On Sat 08/08/15 02:23 , Bob Camp kb...@n1k.org  wrote:

 Hi

 Well an un-stated  assumption of mine was that they all came from
 the
 same
  measurement system and that it
 had the same floor under all  circumstances….

 Bob

 On  Aug 7, 2015, at 5:21 PM, John Miles  wrote:

 Hi

 If that data  is correct, then the 10811 you have is defective.
  Bob

 Well... some of the data is reasonable  for a scenario where a
 counter is being used to measure  OCXOs.

 Looking at the ADEV plot, I'd say the  blue trace (HP105B vs
 5065A) is
 the most questionable one if it  came from a standalone TIC or
 frequency counter, because 7E-12 @  t=1s isn't achievable with most
 counters under most circumstances.  A Wavecrest box can measure at
 that level if it's
 set
 up  _perfectly_ to take bursts of 100+ wrap-free averages within a
 small fraction of the t0 interval. It might also be doable with an
 HP 5370A/B under similar conditions, but I'd have less confidence
 that the
 averaging
 isn't distorting the measurement. So while It  looks like a valid
 measurement of an OCXO with some minor  crosstalk or other external
 interference, that may just be a  coincidence.

 Luciano, how was the blue trace  taken? Is this from your DMTD
 project?
 If so, it's looking  promising.

 The green and magenta traces are  definitely in the right ballpark
 for
 measurements on a 5370-class  counter. At 3E-11 @ t=12s the magenta
 trace is
 optimistic but not  

Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-09 Thread Bob Camp
Hi

All 10811 OCXO’s have 10 MHz 3rd overtone SC cut crystals in them. They are the 
first commercial (as opposed to military) OCXO to use the SC. The target was 
good 
phase noise rather than good short term stability. They did quite well for the 
era in 
terms of phase noise. 

Back then and now, better short term designs exist. Also, then and now, the odd 
part 
pops out of the batch that is a bit better than the rest. It is very rare to 
find
anybody with the ability to accurately test a good one who then decides to sell 
that OCXO.

Bob



 On Aug 8, 2015, at 7:21 PM, KA2WEU--- via time-nuts time-nuts@febo.com 
 wrote:
 
 How good or bad is the 10811 and which one on   ebay   is a better choice ? 
 I am looking for a very goo 5 MHz crystal oscillator  with documentation .
 
 Thanks, Ulrich N1UL 
 
 
 In a message dated 8/8/2015 7:16:37 P.M. Eastern Daylight Time,  
 kb...@n1k.org writes:
 
 Hi
 
 Ok, so John’s observation was the correct one. The data  we are looking at 
 is *not* the
 performance of the OCXO’s but the strange  behavior of the counter at short 
 ADEV Tau’s. 
 
 Sorry for my bashing your  poor 10811.
 
 Bob
 
 On Aug 8, 2015, at 12:25 PM,  tim...@timeok.it wrote:
 
 Hi all,
 
 I try to  ansver to all you:
 
 Luciano, how was the blue trace taken? Is  this from your DMTD project? 
 If so, it's looking promising.
 The green  and magenta traces are definitely in the right ballpark for 
 measurements on a  5370-class counter. At 3E-11 @ t=12s the magenta trace is 
 optimistic but not  outrageously so, while the green trace looks exactly like 
 I'd expect for a  typical 10811 measured on a 5370. The observed noise is 
 due entirely to the  counter until about t=200s. We see a glimpse of the 
 10811's typical ADEV at  about 250 seconds, just before either drift or ADEV 
 uncertainty causes the  trace to turn upwards. A longer run would be needed 
 to 
 distinguish between  these two situations.
 
 -- john, KE5FX
 Miles Design  LLC
 
 r: no, all the measurements are taken using an  HP53132A  in frequency 
 mode. The difference can be the gate time. Using 1  second the resolution is 
 lower than using 2 Second that permit the max  counter  resolution.
 
 Hi
 
 Well an  un-stated assumption of mine was that they all came from the 
 same measurement  system and that it
 had the same floor under all  circumstances….
 
 Bob
 
 r: all the measurements  are under the same conditions except for the 
 gate time of the counter.
 
 Hi Luciano,
 
 Can you give me the link to your ADEV  posting image about 10811 vs 105 
 oscillators?  I had it and now can't  seem to find it.  I wanted to look at 
 your plots as I read John Miles'  comments.  Many thanks.  
 
 I have two very high  performing HP10811-60109 OCXO units which I got 
 from Corby. 
 
 Many thanks.
 
 Jim Robbins
 N1JR
 
 r:  Jim, I will load soon some files on my site. The 105B I have is a 
 fantastic  exception, unfortunately it have a defect, may be a bad  solder 
 inside  cause randomly a phase jump and return to the original phase trend, 
 so 
 all The  long term ADEV are distorted by this problem I have to fix.
 
 Tom VB, i will do all the cross measurements on 10811, 00105 and 
 rubidium I  have and i will upload they but I need time to do this.
 Unfortunately  I have only the HP53132A as TIC and my best reference are 
 four HP5065A. I  normally use the counter function because the TI function 
 on 1 PPS have 100  time less resolution.
 Will be interesting doing also the Phase noise  tests. I will do it.
 please see:  
 http://www.timeok.it/files/time_and_frequency_house_standard_201r.pdf
 
 Luciano
 www.timeok.it
 
 
 
 
 
 On Sat 08/08/15 02:23 , Bob Camp kb...@n1k.org  wrote:
 
 Hi
 
 Well an un-stated  assumption of mine was that they all came from the 
 same
 measurement system and that it
 had the same floor under all  circumstances….
 
 Bob
 
 On  Aug 7, 2015, at 5:21 PM, John Miles  wrote:
 
 Hi
 
 If that data  is correct, then the 10811 you have is defective.
 Bob
 
 Well... some of the data is reasonable  for a scenario where a counter 
 is
 being used to measure  OCXOs.
 
 Looking at the ADEV plot, I'd say the  blue trace (HP105B vs 5065A) is
 the most questionable one if it  came from a standalone TIC or frequency
 counter, because 7E-12 @  t=1s isn't achievable with most counters under
 most circumstances.  A Wavecrest box can measure at that level if it's 
 set
 up  _perfectly_ to take bursts of 100+ wrap-free averages within a  small
 fraction of the t0 interval. It might also be doable with an  HP 5370A/B
 under similar conditions, but I'd have less confidence  that the 
 averaging
 isn't distorting the measurement. So while It  looks like a valid
 measurement of an OCXO with some minor  crosstalk or other external
 interference, that may just be a  coincidence.
 
 Luciano, how was the blue trace  taken? Is this from your DMTD project?
 If so, it's looking  promising.
 
 The green and magenta traces are  definitely in 

Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-08 Thread timeok
Hi all,

I try to ansver to all you:

Luciano, how was the blue trace taken? Is this from your DMTD project? If so, 
it's looking promising.
The green and magenta traces are definitely in the right ballpark for 
measurements on a 5370-class counter. At 3E-11 @ t=12s the magenta trace is 
optimistic but not outrageously so, while the green trace looks exactly like 
I'd expect for a typical 10811 measured on a 5370. The observed noise is due 
entirely to the counter until about t=200s. We see a glimpse of the 10811's 
typical ADEV at about 250 seconds, just before either drift or ADEV uncertainty 
causes the trace to turn upwards. A longer run would be needed to distinguish 
between these two situations.

-- john, KE5FX
Miles Design LLC

r: no, all the measurements are taken using an HP53132A  in frequency mode. The 
difference can be the gate time. Using 1 second the resolution is lower than 
using 2 Second that permit the max counter  resolution.

Hi

Well an un-stated assumption of mine was that they all came from the same 
measurement system and that it
had the same floor under all circumstances….

Bob

r: all the measurements are under the same conditions except for the gate time 
of the counter.

Hi Luciano,

Can you give me the link to your ADEV posting image about 10811 vs 105 
oscillators?  I had it and now can't seem to find it.  I wanted to look at your 
plots as I read John Miles' comments.  Many thanks.  

I have two very high performing HP10811-60109 OCXO units which I got from 
Corby. 

Many thanks.

Jim Robbins
N1JR

r: Jim, I will load soon some files on my site. The 105B I have is a fantastic 
exception, unfortunately it have a defect, may be a bad  solder inside cause 
randomly a phase jump and return to the original phase trend, so all The long 
term ADEV are distorted by this problem I have to fix.

Tom VB, i will do all the cross measurements on 10811, 00105 and rubidium I 
have and i will upload they but I need time to do this.
Unfortunately I have only the HP53132A as TIC and my best reference are four 
HP5065A. I normally use the counter function because the TI function on 1 PPS 
have 100 time less resolution.
Will be interesting doing also the Phase noise tests. I will do it.
please see: 
http://www.timeok.it/files/time_and_frequency_house_standard_201r.pdf

Luciano
www.timeok.it





On Sat 08/08/15 02:23 , Bob Camp kb...@n1k.org wrote:

 Hi
 
 Well an un-stated assumption of mine was that they all came from the same
 measurement system and that it
 had the same floor under all circumstances….
 
 Bob
 
  On Aug 7, 2015, at 5:21 PM, John Miles  wrote:
 
  Hi
 
  If that data is correct, then the 10811 you have is defective.
  Bob
 
  Well... some of the data is reasonable for a scenario where a counter is
 being used to measure OCXOs.
 
  Looking at the ADEV plot, I'd say the blue trace (HP105B vs 5065A) is
 the most questionable one if it came from a standalone TIC or frequency
 counter, because 7E-12 @ t=1s isn't achievable with most counters under
 most circumstances. A Wavecrest box can measure at that level if it's set
 up _perfectly_ to take bursts of 100+ wrap-free averages within a small
 fraction of the t0 interval. It might also be doable with an HP 5370A/B
 under similar conditions, but I'd have less confidence that the averaging
 isn't distorting the measurement. So while It looks like a valid
 measurement of an OCXO with some minor crosstalk or other external
 interference, that may just be a coincidence.
 
  Luciano, how was the blue trace taken? Is this from your DMTD project?
 If so, it's looking promising.
 
  The green and magenta traces are definitely in the right ballpark for
 measurements on a 5370-class counter. At 3E-11 @ t=12s the magenta trace is
 optimistic but not outrageously so, while the green trace looks exactly
 like I'd expect for a typical 10811 measured on a 5370. The observed noise
 is due entirely to the counter until about t=200s. We see a glimpse of the
 10811's typical ADEV at about 250 seconds, just before either drift or ADEV
 uncertainty causes the trace to turn upwards. A longer run would be needed
 to distinguish between these two situations.
 
  -- john, KE5FX
  Miles Design LLC
 
 
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [1]
  and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [2]
 and follow the instructions there.
 
 
 
 Links:
 --
 [1]
 http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
 ilman/listinfo/time-nuts[2]
 http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
 ilman/listinfo/time-nuts
 
Message sent via Atmail Open - http://atmail.org/

Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-08 Thread Bob Camp
Hi

Ok, so John’s observation was the correct one. The data we are looking at is 
*not* the
performance of the OCXO’s but the strange behavior of the counter at short ADEV 
Tau’s. 

Sorry for my bashing your poor 10811.

Bob

 On Aug 8, 2015, at 12:25 PM, tim...@timeok.it wrote:
 
 Hi all,
 
 I try to ansver to all you:
 
 Luciano, how was the blue trace taken? Is this from your DMTD project? If so, 
 it's looking promising.
 The green and magenta traces are definitely in the right ballpark for 
 measurements on a 5370-class counter. At 3E-11 @ t=12s the magenta trace is 
 optimistic but not outrageously so, while the green trace looks exactly like 
 I'd expect for a typical 10811 measured on a 5370. The observed noise is due 
 entirely to the counter until about t=200s. We see a glimpse of the 10811's 
 typical ADEV at about 250 seconds, just before either drift or ADEV 
 uncertainty causes the trace to turn upwards. A longer run would be needed to 
 distinguish between these two situations.
 
 -- john, KE5FX
 Miles Design LLC
 
 r: no, all the measurements are taken using an HP53132A  in frequency mode. 
 The difference can be the gate time. Using 1 second the resolution is lower 
 than using 2 Second that permit the max counter  resolution.
 
 Hi
 
 Well an un-stated assumption of mine was that they all came from the same 
 measurement system and that it
 had the same floor under all circumstances….
 
 Bob
 
 r: all the measurements are under the same conditions except for the gate 
 time of the counter.
 
 Hi Luciano,
 
 Can you give me the link to your ADEV posting image about 10811 vs 105 
 oscillators?  I had it and now can't seem to find it.  I wanted to look at 
 your plots as I read John Miles' comments.  Many thanks.  
 
 I have two very high performing HP10811-60109 OCXO units which I got from 
 Corby. 
 
 Many thanks.
 
 Jim Robbins
 N1JR
 
 r: Jim, I will load soon some files on my site. The 105B I have is a 
 fantastic exception, unfortunately it have a defect, may be a bad  solder 
 inside cause randomly a phase jump and return to the original phase trend, so 
 all The long term ADEV are distorted by this problem I have to fix.
 
 Tom VB, i will do all the cross measurements on 10811, 00105 and rubidium I 
 have and i will upload they but I need time to do this.
 Unfortunately I have only the HP53132A as TIC and my best reference are four 
 HP5065A. I normally use the counter function because the TI function on 1 PPS 
 have 100 time less resolution.
 Will be interesting doing also the Phase noise tests. I will do it.
 please see: 
 http://www.timeok.it/files/time_and_frequency_house_standard_201r.pdf
 
 Luciano
 www.timeok.it
 
 
 
 
 
 On Sat 08/08/15 02:23 , Bob Camp kb...@n1k.org wrote:
 
 Hi
 
 Well an un-stated assumption of mine was that they all came from the same
 measurement system and that it
 had the same floor under all circumstances….
 
 Bob
 
 On Aug 7, 2015, at 5:21 PM, John Miles  wrote:
 
 Hi
 
 If that data is correct, then the 10811 you have is defective.
 Bob
 
 Well... some of the data is reasonable for a scenario where a counter is
 being used to measure OCXOs.
 
 Looking at the ADEV plot, I'd say the blue trace (HP105B vs 5065A) is
 the most questionable one if it came from a standalone TIC or frequency
 counter, because 7E-12 @ t=1s isn't achievable with most counters under
 most circumstances. A Wavecrest box can measure at that level if it's set
 up _perfectly_ to take bursts of 100+ wrap-free averages within a small
 fraction of the t0 interval. It might also be doable with an HP 5370A/B
 under similar conditions, but I'd have less confidence that the averaging
 isn't distorting the measurement. So while It looks like a valid
 measurement of an OCXO with some minor crosstalk or other external
 interference, that may just be a coincidence.
 
 Luciano, how was the blue trace taken? Is this from your DMTD project?
 If so, it's looking promising.
 
 The green and magenta traces are definitely in the right ballpark for
 measurements on a 5370-class counter. At 3E-11 @ t=12s the magenta trace is
 optimistic but not outrageously so, while the green trace looks exactly
 like I'd expect for a typical 10811 measured on a 5370. The observed noise
 is due entirely to the counter until about t=200s. We see a glimpse of the
 10811's typical ADEV at about 250 seconds, just before either drift or ADEV
 uncertainty causes the trace to turn upwards. A longer run would be needed
 to distinguish between these two situations.
 
 -- john, KE5FX
 Miles Design LLC
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [1]
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [2]
 and follow the instructions 

Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-08 Thread KA2WEU--- via time-nuts
How good or bad is the 10811 and which one on   ebay   is a better choice ? 
I am looking for a very goo 5 MHz crystal oscillator  with documentation .
 
Thanks, Ulrich N1UL 
 
 
In a message dated 8/8/2015 7:16:37 P.M. Eastern Daylight Time,  
kb...@n1k.org writes:

Hi

Ok, so John’s observation was the correct one. The data  we are looking at 
is *not* the
performance of the OCXO’s but the strange  behavior of the counter at short 
ADEV Tau’s. 

Sorry for my bashing your  poor 10811.

Bob

 On Aug 8, 2015, at 12:25 PM,  tim...@timeok.it wrote:
 
 Hi all,
 
 I try to  ansver to all you:
 
 Luciano, how was the blue trace taken? Is  this from your DMTD project? 
If so, it's looking promising.
 The green  and magenta traces are definitely in the right ballpark for 
measurements on a  5370-class counter. At 3E-11 @ t=12s the magenta trace is 
optimistic but not  outrageously so, while the green trace looks exactly like 
I'd expect for a  typical 10811 measured on a 5370. The observed noise is 
due entirely to the  counter until about t=200s. We see a glimpse of the 
10811's typical ADEV at  about 250 seconds, just before either drift or ADEV 
uncertainty causes the  trace to turn upwards. A longer run would be needed to 
distinguish between  these two situations.
 
 -- john, KE5FX
 Miles Design  LLC
 
 r: no, all the measurements are taken using an  HP53132A  in frequency 
mode. The difference can be the gate time. Using 1  second the resolution is 
lower than using 2 Second that permit the max  counter  resolution.
 
 Hi
 
 Well an  un-stated assumption of mine was that they all came from the 
same measurement  system and that it
 had the same floor under all  circumstances….
 
 Bob
 
 r: all the measurements  are under the same conditions except for the 
gate time of the counter.
  
 Hi Luciano,
 
 Can you give me the link to your ADEV  posting image about 10811 vs 105 
oscillators?  I had it and now can't  seem to find it.  I wanted to look at 
your plots as I read John Miles'  comments.  Many thanks.  
 
 I have two very high  performing HP10811-60109 OCXO units which I got 
from Corby. 
 
  Many thanks.
 
 Jim Robbins
 N1JR
 
 r:  Jim, I will load soon some files on my site. The 105B I have is a 
fantastic  exception, unfortunately it have a defect, may be a bad  solder 
inside  cause randomly a phase jump and return to the original phase trend, so 
all The  long term ADEV are distorted by this problem I have to fix.
 
  Tom VB, i will do all the cross measurements on 10811, 00105 and 
rubidium I  have and i will upload they but I need time to do this.
 Unfortunately  I have only the HP53132A as TIC and my best reference are 
four HP5065A. I  normally use the counter function because the TI function 
on 1 PPS have 100  time less resolution.
 Will be interesting doing also the Phase noise  tests. I will do it.
 please see:  
http://www.timeok.it/files/time_and_frequency_house_standard_201r.pdf
  
 Luciano
 www.timeok.it
 
 
 
  
 
 On Sat 08/08/15 02:23 , Bob Camp kb...@n1k.org  wrote:
 
 Hi
 
 Well an un-stated  assumption of mine was that they all came from the 
same
  measurement system and that it
 had the same floor under all  circumstances….
 
 Bob
 
 On  Aug 7, 2015, at 5:21 PM, John Miles  wrote:
  
 Hi
 
 If that data  is correct, then the 10811 you have is defective.
  Bob
 
 Well... some of the data is reasonable  for a scenario where a counter 
is
 being used to measure  OCXOs.
 
 Looking at the ADEV plot, I'd say the  blue trace (HP105B vs 5065A) is
 the most questionable one if it  came from a standalone TIC or frequency
 counter, because 7E-12 @  t=1s isn't achievable with most counters under
 most circumstances.  A Wavecrest box can measure at that level if it's 
set
 up  _perfectly_ to take bursts of 100+ wrap-free averages within a  small
 fraction of the t0 interval. It might also be doable with an  HP 5370A/B
 under similar conditions, but I'd have less confidence  that the 
averaging
 isn't distorting the measurement. So while It  looks like a valid
 measurement of an OCXO with some minor  crosstalk or other external
 interference, that may just be a  coincidence.
 
 Luciano, how was the blue trace  taken? Is this from your DMTD project?
 If so, it's looking  promising.
 
 The green and magenta traces are  definitely in the right ballpark for
 measurements on a 5370-class  counter. At 3E-11 @ t=12s the magenta 
trace is
 optimistic but not  outrageously so, while the green trace looks exactly
 like I'd  expect for a typical 10811 measured on a 5370. The observed 
noise
  is due entirely to the counter until about t=200s. We see a glimpse of  
the
 10811's typical ADEV at about 250 seconds, just before either  drift or 
ADEV
 uncertainty causes the trace to turn upwards. A  longer run would be 
needed
 to distinguish between these two  situations.
 
 -- john, KE5FX
  Miles Design LLC
 
 
  ___
 time-nuts  mailing list -- time-nuts@febo.com
 

Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread John Miles
 Hi
 
 If that data is correct, then the 10811 you have is defective.
 Bob

Well... some of the data is reasonable for a scenario where a counter is being 
used to measure OCXOs.  

Looking at the ADEV plot, I'd say the blue trace (HP105B vs 5065A) is the most 
questionable one if it came from a standalone TIC or frequency counter, because 
7E-12 @ t=1s isn't achievable with most counters under most circumstances.  A 
Wavecrest box can measure at that level if it's set up _perfectly_ to take 
bursts of 100+ wrap-free averages within a small fraction of the t0 interval.  
It might also be doable with an HP 5370A/B under similar conditions, but I'd 
have less confidence that the averaging isn't distorting the measurement.  So 
while It looks like a valid measurement of an OCXO with some minor crosstalk or 
other external interference, that may just be a coincidence.  

Luciano, how was the blue trace taken?  Is this from your DMTD project?  If so, 
it's looking promising.

The green and magenta traces are definitely in the right ballpark for 
measurements on a 5370-class counter.  At 3E-11 @ t=12s the magenta trace is 
optimistic but not outrageously so, while the green trace looks exactly like 
I'd expect for a typical 10811 measured on a 5370.  The observed noise is due 
entirely to the counter until about t=200s.  We see a glimpse of the 10811's 
typical  ADEV at about 250 seconds, just before either drift or ADEV 
uncertainty causes the trace to turn upwards.  A longer run would be needed to 
distinguish between these two situations.

-- john, KE5FX
Miles Design LLC


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread Bob Camp
Hi

Well an un-stated assumption of mine was that they all came from the same 
measurement system and that it 
had the same floor under all circumstances….

Bob

 On Aug 7, 2015, at 5:21 PM, John Miles j...@miles.io wrote:
 
 Hi
 
 If that data is correct, then the 10811 you have is defective.
 Bob
 
 Well... some of the data is reasonable for a scenario where a counter is 
 being used to measure OCXOs.  
 
 Looking at the ADEV plot, I'd say the blue trace (HP105B vs 5065A) is the 
 most questionable one if it came from a standalone TIC or frequency counter, 
 because 7E-12 @ t=1s isn't achievable with most counters under most 
 circumstances.  A Wavecrest box can measure at that level if it's set up 
 _perfectly_ to take bursts of 100+ wrap-free averages within a small fraction 
 of the t0 interval.  It might also be doable with an HP 5370A/B under similar 
 conditions, but I'd have less confidence that the averaging isn't distorting 
 the measurement.  So while It looks like a valid measurement of an OCXO with 
 some minor crosstalk or other external interference, that may just be a 
 coincidence.  
 
 Luciano, how was the blue trace taken?  Is this from your DMTD project?  If 
 so, it's looking promising.
 
 The green and magenta traces are definitely in the right ballpark for 
 measurements on a 5370-class counter.  At 3E-11 @ t=12s the magenta trace is 
 optimistic but not outrageously so, while the green trace looks exactly like 
 I'd expect for a typical 10811 measured on a 5370.  The observed noise is due 
 entirely to the counter until about t=200s.  We see a glimpse of the 10811's 
 typical  ADEV at about 250 seconds, just before either drift or ADEV 
 uncertainty causes the trace to turn upwards.  A longer run would be needed 
 to distinguish between these two situations.
 
 -- john, KE5FX
 Miles Design LLC
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread Mark Spencer
Hi Luciano.  That is interesting.  Thanks for sharing this.

I also have an HP105B with the old style oscillator.  (An auction site purchase 
from years ago.)   Yours seems at bit better than the one I have at tau's of 
approx 100 seconds or so.  I'm curious if you have any data for longer tau's ?


All the best Mark Spencer


 On Aug 6, 2015, at 8:37 PM, tim...@timeok.it wrote:
 
 
 here the ADEV of the two oscillator. I have added the Super performance of an 
 HP105B (old oscillator) bought in an Ham fest.
 
 Luciano
 
 
 
 On Fri 07/08/15 03:37 , Bob Camp kb...@n1k.org wrote:
 
 Hi
 
 HP 10811’s vary over almost a 100:1 range in terms of ADEV performance
 at short tau. The standard model is un-sealed so it has some issues
 with humidity when it is in storage for a long time. In both the case of
 the 10811 and the 105, their stability will improve as they are
 on power. In the case of long term storage, they both may continue to
 improve for more than a month.
 
 All of this makes any sort of comparison between the two models a bit
 difficult. About all you can say is that on a given day, this example of
 one
 was (or was not) more stable than that example of the other model.
 
 Bob
 
 On Aug 6, 2015, at 5:03 AM, tim...@timeok.it wrote:
 
 
 
 Hi,
 
 I have done some measurement comparing two HP OCXO, the HP10811 and the
 105.
 
 These two oscillator are mounted in the HP5065A rubidium Standard, the
 105 in the old models, the 10811 in newer, and for this purpose they are
 selected units.
 
 I have two HP5065A use the different oscillators. For the stability test
 I have set the two 5065A in Open loop so the OCXO are free running .
 I have done the measurements using as reference a third HP5065A in
 closed loop operation.
 
 Here the files show the frequency and phase difference.
 
 The 00105-6034 appear to be more stable as frequency than the 10811.
 
 comments?
 
 
 Luciano
 www.timeok.it [1]
 Message sent via Atmail Open -
 http://atmail.org/___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [3]
 and follow the instructions there.
 
 
 
 Links:
 --
 [1] http://webmail.timeok.it/parse.php?redirect=http://www.timeok.it
 [2]
 http://webmail.timeok.it/parse.php?redirect=http://atmail.org/%26lt%3BHP
 [3]
 http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
 ilman/listinfo/time-nuts
 Message sent via Atmail Open - http://atmail.org/
 ocxo.gif
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread Tom Van Baak
Hi Luciano,

Thanks for that plot. In general an ADEV plots shows more succinctly the 
differences between oscillators than the raw phase and frequency plots that you 
posted earlier. The jitter and wander of your phase plots give a hint, but an 
ADEV plot neatly summarizes all this with statistics.

About the 00105 vs. 10811 -- it's nearly impossible to make solid claims about 
one vs. the other. Oscillators of the same make/model vary a lot. Oscillators 
that are 20 or 30 or 40 years old may not behave the same way they did when 
they left the manufacturing line. A ham fest or eBay buy adds its own special 
mystery, for better or worse.

The SC-cut 10811 warms up quickly. But that feature is irrelevant for a 
frequency standard that you power up once and then leave running the rest of 
your life. Low daily frequency drift is important, unless you use the 
oscillator as part of a 5065 or 5061 or GPSDO. Then even drift rate is 
irrelevant.

The simple answer is -- just measure it. Don't rely on the name or make or 
model. It may be an order of magnitude better than original spec. Or it might 
be an order of magnitude worse.

My dream would be to solicit a hundred ADEV plots of 00105 and 10811 from all 
time nuts and then make an informative plot or histogram. But still even that 
plot would not predict the performance of a random OCXO that you pick up from a 
ham fest.

/tvb


- Original Message - 
From: tim...@timeok.it
To: Discussion of precise time and frequency measurement time-nuts@febo.com
Sent: Thursday, August 06, 2015 8:37 PM
Subject: Re: [time-nuts] HP10811 vs 00105 OCXO



here the ADEV of the two oscillator. I have added the Super performance of an 
HP105B (old oscillator) bought in an Ham fest.

Luciano



On Fri 07/08/15 03:37 , Bob Camp kb...@n1k.org wrote:

 Hi
 
 HP 10811’s vary over almost a 100:1 range in terms of ADEV performance
 at short tau. The standard model is un-sealed so it has some issues
 with humidity when it is in storage for a long time. In both the case of
 the 10811 and the 105, their stability will improve as they are
 on power. In the case of long term storage, they both may continue to
 improve for more than a month.
 
 All of this makes any sort of comparison between the two models a bit
 difficult. About all you can say is that on a given day, this example of
 one
 was (or was not) more stable than that example of the other model.
 
 Bob
 
  On Aug 6, 2015, at 5:03 AM, tim...@timeok.it wrote:
 
 
 
  Hi,
 
  I have done some measurement comparing two HP OCXO, the HP10811 and the
 105.
 
  These two oscillator are mounted in the HP5065A rubidium Standard, the
 105 in the old models, the 10811 in newer, and for this purpose they are
 selected units.
 
  I have two HP5065A use the different oscillators. For the stability test
 I have set the two 5065A in Open loop so the OCXO are free running .
  I have done the measurements using as reference a third HP5065A in
 closed loop operation.
 
  Here the files show the frequency and phase difference.
 
  The 00105-6034 appear to be more stable as frequency than the 10811.
 
  comments?
 
 
  Luciano

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread Bob Camp
Hi

The outer oven in the 3801 was put there for a very specific purpose:

When the 3801 was designed, they had a target specification that included a 
warmup from -40C.
The 10811 was never designed to work at these temperatures and simply could not 
meet the 
required warmup time. 

The “solution” that they came up with was to put a heater around the 10811 
(without modifying it 
internally) to get the case up to a fairly high temperature quickly. This 
allowed them to meet the 
-40C specification on their target sheet. As it turns out, it is very unclear 
if the 3801 ever went into
the sort of environment that would expose it to that sort of low end 
temperature.

HP published a number of papers on the 10811 and it’s successor. They go into 
great detail about 
the thermal optimization of the part and why you can easily have “to much of a 
good thing” and actually
make the oscillator worse.

Bob


 On Aug 7, 2015, at 3:14 AM, tim...@timeok.it wrote:
 
 
 Hi  Corby,
 
 I agree with you that if selected, there may be many differences between 
 objects of the same type. Unfortunately it is very difficult to know what is 
 selected and what is not. For example I found a 10811 mounted in a  Z3801A  
 who have the temperature set-point  completely wrong. I remember that TVB has 
 done tests on a number of Z3801s and the results were very different from one 
  to each other. In fact I would expect HP since it is bothered to put a 
 second oven in Z3801A, had selected the oscillators but it is not.
 
 I have not yet measured the 105 Phase noise, I have a 5MHz reference is -115 
 and -135 dBc respectively at 1 and 10 Hz.
 
 Luciano
 www.timeok.it
 
 
 
 
 On Thu 06/08/15 20:38 , cdel...@juno.com wrote:
 
 Luciano,
 
 I have found that selected 10811 units will offer significantly better
 short term stability as a stand alone or open loop oscillator over the
 105 style oscillator. (Mid parts in 10-13th from 1 to 100 Seconds)
 
 However with the 5065A locked the performance will be driven by the loop
 and both styles will give similar performance.
 
 Cheers,
 
 Corby
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [1]
 and follow the instructions there.
 
 
 
 Links:
 --
 [1]
 http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
 ilman/listinfo/time-nuts
 
 Message sent via Atmail Open - http://atmail.org/
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread Poul-Henning Kamp

In message 55454.1438931...@timeok.it, tim...@timeok.it writes:

 For example I found a 10811 mounted in a  Z3801A  who have the
 temperature set-point completely wrong. 

The turn-over temperature can shift rather dramatically if the
X-tal is subjected to large mechanical shocks.  This is
very relevant for 2nd-hand equipment, and doubly so if it has
been through the E-waste circuit via China.


-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread Bob Camp
Hi

If that data is correct, then the 10811 you have is defective.

Bob

 On Aug 6, 2015, at 11:37 PM, tim...@timeok.it wrote:
 
 
 here the ADEV of the two oscillator. I have added the Super performance of an 
 HP105B (old oscillator) bought in an Ham fest.
 
 Luciano
 
 
 
 On Fri 07/08/15 03:37 , Bob Camp kb...@n1k.org wrote:
 
 Hi
 
 HP 10811’s vary over almost a 100:1 range in terms of ADEV performance
 at short tau. The standard model is un-sealed so it has some issues
 with humidity when it is in storage for a long time. In both the case of
 the 10811 and the 105, their stability will improve as they are
 on power. In the case of long term storage, they both may continue to
 improve for more than a month.
 
 All of this makes any sort of comparison between the two models a bit
 difficult. About all you can say is that on a given day, this example of
 one
 was (or was not) more stable than that example of the other model.
 
 Bob
 
 On Aug 6, 2015, at 5:03 AM, tim...@timeok.it wrote:
 
 
 
 Hi,
 
 I have done some measurement comparing two HP OCXO, the HP10811 and the
 105.
 
 These two oscillator are mounted in the HP5065A rubidium Standard, the
 105 in the old models, the 10811 in newer, and for this purpose they are
 selected units.
 
 I have two HP5065A use the different oscillators. For the stability test
 I have set the two 5065A in Open loop so the OCXO are free running .
 I have done the measurements using as reference a third HP5065A in
 closed loop operation.
 
 Here the files show the frequency and phase difference.
 
 The 00105-6034 appear to be more stable as frequency than the 10811.
 
 comments?
 
 
 Luciano
 www.timeok.it [1]
 Message sent via Atmail Open -
 http://atmail.org/___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [3]
 and follow the instructions there.
 
 
 
 Links:
 --
 [1] http://webmail.timeok.it/parse.php?redirect=http://www.timeok.it
 [2]
 http://webmail.timeok.it/parse.php?redirect=http://atmail.org/%26lt%3BHP
 [3]
 http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
 ilman/listinfo/time-nuts
 
 Message sent via Atmail Open - 
 http://atmail.org/ocxo.gif___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread timeok

Hi  Corby,

I agree with you that if selected, there may be many differences between 
objects of the same type. Unfortunately it is very difficult to know what is 
selected and what is not. For example I found a 10811 mounted in a  Z3801A  who 
have the temperature set-point  completely wrong. I remember that TVB has done 
tests on a number of Z3801s and the results were very different from one  to 
each other. In fact I would expect HP since it is bothered to put a second oven 
in Z3801A, had selected the oscillators but it is not.

I have not yet measured the 105 Phase noise, I have a 5MHz reference is -115 
and -135 dBc respectively at 1 and 10 Hz.

Luciano
www.timeok.it




On Thu 06/08/15 20:38 , cdel...@juno.com wrote:

 Luciano,
 
 I have found that selected 10811 units will offer significantly better
 short term stability as a stand alone or open loop oscillator over the
 105 style oscillator. (Mid parts in 10-13th from 1 to 100 Seconds)
 
 However with the 5065A locked the performance will be driven by the loop
 and both styles will give similar performance.
 
 Cheers,
 
 Corby
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [1]
 and follow the instructions there.
 
 
 
 Links:
 --
 [1]
 http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
 ilman/listinfo/time-nuts
 
Message sent via Atmail Open - http://atmail.org/
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-07 Thread timeok

here the ADEV of the two oscillator. I have added the Super performance of an 
HP105B (old oscillator) bought in an Ham fest.

Luciano



On Fri 07/08/15 03:37 , Bob Camp kb...@n1k.org wrote:

 Hi
 
 HP 10811’s vary over almost a 100:1 range in terms of ADEV performance
 at short tau. The standard model is un-sealed so it has some issues
 with humidity when it is in storage for a long time. In both the case of
 the 10811 and the 105, their stability will improve as they are
 on power. In the case of long term storage, they both may continue to
 improve for more than a month.
 
 All of this makes any sort of comparison between the two models a bit
 difficult. About all you can say is that on a given day, this example of
 one
 was (or was not) more stable than that example of the other model.
 
 Bob
 
  On Aug 6, 2015, at 5:03 AM, tim...@timeok.it wrote:
 
 
 
  Hi,
 
  I have done some measurement comparing two HP OCXO, the HP10811 and the
 105.
 
  These two oscillator are mounted in the HP5065A rubidium Standard, the
 105 in the old models, the 10811 in newer, and for this purpose they are
 selected units.
 
  I have two HP5065A use the different oscillators. For the stability test
 I have set the two 5065A in Open loop so the OCXO are free running .
  I have done the measurements using as reference a third HP5065A in
 closed loop operation.
 
  Here the files show the frequency and phase difference.
 
  The 00105-6034 appear to be more stable as frequency than the 10811.
 
  comments?
 
 
  Luciano
  www.timeok.it [1]
  Message sent via Atmail Open -
 http://atmail.org/___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts [3]
  and follow the instructions there.
 
 
 
 Links:
 --
 [1] http://webmail.timeok.it/parse.php?redirect=http://www.timeok.it
 [2]
 http://webmail.timeok.it/parse.php?redirect=http://atmail.org/%26lt%3BHP
 [3]
 http://webmail.timeok.it/parse.php?redirect=https://www.febo.com/cgi-bin/ma
 ilman/listinfo/time-nuts
 
Message sent via Atmail Open - http://atmail.org/___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

[time-nuts] HP10811 vs 00105 OCXO

2015-08-06 Thread cdelect
Luciano,

I have found that selected 10811 units will offer significantly better
short term stability as a stand alone or open loop oscillator over the
105 style oscillator. (Mid parts in 10-13th from 1 to 100 Seconds)

However with the 5065A locked the performance will be driven by the loop
and both styles will give similar performance.

Cheers,

Corby

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] HP10811 vs 00105 OCXO

2015-08-06 Thread Bob Camp
Hi

HP 10811’s vary over almost a 100:1 range in terms of ADEV performance at short 
tau. The standard model is un-sealed so it has some issues 
with humidity when it is in storage for a long time. In both the case of the 
10811 and the 105, their stability will improve as they are 
on power. In the case of long term storage, they both may continue to improve 
for more than a month. 

All of this makes any sort of comparison between the two models a bit 
difficult. About all you can say is that on a given day, this example of one
was (or was not) more stable than that example of the other model. 

Bob



 On Aug 6, 2015, at 5:03 AM, tim...@timeok.it wrote:
 
 
 
 Hi,
 
 I have done some measurement comparing  two HP OCXO, the HP10811 and the 105.
 
 These two oscillator are mounted in the HP5065A rubidium Standard, the 105 in 
 the old models, the 10811 in newer, and for this purpose they are selected 
 units.
 
 I have two HP5065A use the different oscillators. For the stability test I 
 have set the two 5065A in Open loop so the OCXO are free running .
 I have done the measurements using as reference a third HP5065A in closed 
 loop operation.
 
 Here the files show the frequency and phase difference.
 
 The 00105-6034 appear to be more stable as frequency than the 10811.
 
 comments?
 
 
 Luciano
 www.timeok.it
 Message sent via Atmail Open - http://atmail.org/HP 10811  00105 vs 
 HP5065A.gifHP 00105 6034 vs 
 HP5065A.gif___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.