[time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Anthony Roby
I've been reading a lot about ADEV and following the threads on the list, 
particularly Karen's in-flight thread. What I haven't come across is a simple 
explanation of the basic setup required to go about collecting the data.  John 
Miles referenced this page http://www.ke5fx.com/tpll.htm, and the simple setup 
at the bottom of the page looks like a reasonable place to start.  Seems that 
I'd need to acquire a phase detector and build or buy some filters and the amp. 
 I can probably figure that out, but how do I get the data into a PC?  Is there 
a basic hardware and software setup that someone could point me to or recommend?

Thanks

Anthony
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Azelio Boriani
Usually phase detectors (we prefer to call them Time Interval
Analysers) have data interfaces (RS232, GPIB, LAN). If you build your
own then the interface is up to you: usually an RS232 is the best
choice. The software to use: take a look at the Miles' TimeLab
www.ke5fx.com/timelab/readme.htm or
http://www.thegleam.com/ke5fx/. Of course, in case of the GPIB, you
also will need an adapter: GPIB/USB or GPIB/LAN but the less expensive
GPIB/RS232 maybe a good choice.

On Sat, Nov 1, 2014 at 2:29 PM, Anthony Roby ar...@antamy.com wrote:
 I've been reading a lot about ADEV and following the threads on the list, 
 particularly Karen's in-flight thread. What I haven't come across is a simple 
 explanation of the basic setup required to go about collecting the data.  
 John Miles referenced this page http://www.ke5fx.com/tpll.htm, and the simple 
 setup at the bottom of the page looks like a reasonable place to start.  
 Seems that I'd need to acquire a phase detector and build or buy some filters 
 and the amp.  I can probably figure that out, but how do I get the data into 
 a PC?  Is there a basic hardware and software setup that someone could point 
 me to or recommend?

 Thanks

 Anthony
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Anders Wallin
what you want to measure is a time-series of either frequency data or phase
data.
the simplest possible case for a beginner would be to have two clocks with
1-PPS (one pulse per second) outputs, and connect one clock to the
start-input and the other to the stop-input of a time-interval counter. If
you measure phase, keep the time-interval number to a 5...6 digit number
and you don't have to worry too much about the internal time-base of the
counter.


On Sat, Nov 1, 2014 at 3:29 PM, Anthony Roby ar...@antamy.com wrote:

 I've been reading a lot about ADEV and following the threads on the list,
 particularly Karen's in-flight thread. What I haven't come across is a
 simple explanation of the basic setup required to go about collecting the
 data.  John Miles referenced this page http://www.ke5fx.com/tpll.htm, and
 the simple setup at the bottom of the page looks like a reasonable place to
 start.  Seems that I'd need to acquire a phase detector and build or buy
 some filters and the amp.  I can probably figure that out, but how do I get
 the data into a PC?  Is there a basic hardware and software setup that
 someone could point me to or recommend?

 Thanks

 Anthony
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Magnus Danielson

Anthony,

On 11/01/2014 02:29 PM, Anthony Roby wrote:

I've been reading a lot about ADEV and following the threads on the list, 
particularly Karen's in-flight thread.
What I haven't come across is a simple explanation of the basic setup required 
to go about collecting the data.
John Miles referenced this page http://www.ke5fx.com/tpll.htm, and the simple 
setup at the bottom of the page looks
like a reasonable place to start.  Seems that I'd need to acquire a phase 
detector and build or buy some filters and
the amp.  I can probably figure that out, but how do I get the data into a PC?  
Is there a basic hardware and software
setup that someone could point me to or recommend?


The time-interval counter, such as HP5370 or SR620, get started (channel 
1) by a reference clock, such as 1 PPS and is then stopped (channel 2) 
by signal under test. The counter is typically read out through GPIB, 
even if some counters have serial interface and maybe even USB or 
Ethernet for really new (or retro-fitted), and the recommended path is 
to get a GPIB to USB interface for instance.

Then use John Miles TimeLab.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Bob Camp
Hi

The problem with setting up to measure any of this stuff is that it’s *very* 
dependent on the gear you have. There’s no big surprises below. It’s all “spend 
more money and have fewer things to figure out”. 

First you need a way to measure frequency out of your mixer (there are LOTS of 
ways …):

If you grab a HP 5334 or HP 5335 to measure the frequency or time, you need to 
get the data out via GPIB. That’s a project by it’s self.

If you grab a HP 53131 or CNT 90 to measure the frequency, you can use a serial 
port. One less project, way more money.

If you grab a HP 5371, you just push a button and it pops up an ADEV plot. Less 
money than a 53131 (usually). It’s a bit limited on the range it will do ADEV 
over.

If you grab a brand new Symmetricom Time Pod, you spend (gulp!!) a bit more. 
You can now do measurements easily over it’s frequency range. (If you can 
afford that option - can I come play at your house ?)

Yes each of these is it’s own little rabbit hole to wander down and each has 
it’s own issues. 

Now you need a reference:

GPSDO’s have their fans, they often have spurs and other “crud” on the output. 
They do have very good long term stability. 

OCXO’s are often pretty quiet and spur free. You need some way to calibrate 
them for really long term stuff.

Rb’s come in a wide range of sizes and costs. Some are better than others. 
Their ADEV at 1 second often not as good as an OCXO.

Cs standards and Hydrogen masers are (to me) in the same cost category as 
exotic test gear above. Fine for a company, not quite so easy for a basement 
guy.

Again each of these is a bit of a project by it’s self. Often it turns into a 
“a couple of these and a couple of those” sort of solution. You only know 
something is right if you have another one to compare it to. 

Now you have some gear. Let’s ignore building the mixer board and assume that’s 
taken care of already. 

Now you need software:

TimeLab is free and it will interface with many of the counters you might be 
using. It does all the math for you and puts up pretty plots. Highly 
recommended. There are other free software packages out there.

Stable32 is another commonly used program. You need to get the data out of the 
counter before you can use it. There are other paid programs out there, Stable 
32 is the best of the bunch.

An Excel spread sheet is indeed another option. The ADEV math is *not* very 
complicated. Excel can do it all very easily. 

Getting data out of the counter could be a terminal program sort of thing 
(serial port) or something more complex (GPIB). In both cases there are a 
number of programs to pick between. 

With software, things like what computer and operating system you have will 
influence your choices. I’m a Mac person, but I also run Linux, Windows, 
FreeBSD and a few other things as needed. Most people have a favorite ….

So, that’s the quick and dirty start to the “quick ADEV setup” process. There 
are:

4 counter like things X 4 reference ideas X (at least 4 mixer approaches) X 
(way more than) 4 software options =  256 paths you could follow. 

That’s quite a few, and many of them would be utter nonsense. Even pruning out 
the ones are unlikely to ever be followed, you still have lots.

Here’s one:

1) Drive your “known good” reference OCXO into a Minicircuits RPD-1 mixer. 

2) Drive your “device under test” into the other side of the RPD-1

3) Amplify and limit the output with a  couple of OP-37 op amps running off of 
+/-15V.
first stage is an amp with high pass and low pass sections
second stage is an inverter / limiter (back to back diodes in a 
normal op amp inverter).
third stage is same as the second
you might (or might not) want a fourth stage 

4) Feed the +/- 0.7 V  limited output into a 5335  (DC couple the input) in 
talk only mode

5) Hook up the 5335 to a National Instruments GPIB card on a PC

6) Run a Visual Basic routine to grab the outputs and write them to a text 
file. 

7) Process the text file with Excel. Use the ADEV formula from the original 
NIST papers. 

There have been a *lot* of OCXO’s tested that way in many different factories 
over a couple of decades. (Yes, it’s probably easier these days to substitute 
Time Lab for steps 6 and 7). 

Is it the ideal or perfect way to do it? Certainly not. It is one of many 
simple ways it can be done. Even this simple way has two forks in it (TimeLab 
or not). It also has a few features like a know good OCXO and the undocumented 
circuit on the output of the RPD-1. 

Quick, simple and pretty much useless to you. It’s my favorite way to do it, 
but that really should not make it your favorite.

What to do?

It’s really a multi part process. Break each section down and address it 
separately. You *do* have a goal in mind, but get each chunk running by it’s 
self. Do simple verifications each step of the way. Tie it all together at the 
end. Accept that you *will* spend 

Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Anthony Roby
Thanks for this.  I should have said that I have a Racal-Dana 1992 counter 
(with GPIB) and I have an Isotemp OCXO134-10, so its sounds like I just need 
the opamps, an RPD-1 and a GPIB-USB interface plus some software.  I'll do a 
bit more digging around and see if I can get the GPIB up and running in the 
next couple of weeks.

Anthony

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Bob Camp
Sent: Saturday, November 01, 2014 3:18 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Measuring ADEV for a beginner

Hi

The problem with setting up to measure any of this stuff is that it ??s *very* 
dependent on the gear you have. There ??s no big surprises below. It ??s all  
??spend more money and have fewer things to figure out ??. 

First you need a way to measure frequency out of your mixer (there are LOTS of 
ways  ? ):

If you grab a HP 5334 or HP 5335 to measure the frequency or time, you need to 
get the data out via GPIB. That ??s a project by it ??s self.

If you grab a HP 53131 or CNT 90 to measure the frequency, you can use a serial 
port. One less project, way more money.

If you grab a HP 5371, you just push a button and it pops up an ADEV plot. Less 
money than a 53131 (usually). It ??s a bit limited on the range it will do ADEV 
over.

If you grab a brand new Symmetricom Time Pod, you spend (gulp!!) a bit more. 
You can now do measurements easily over it ??s frequency range. (If you can 
afford that option - can I come play at your house ?)

Yes each of these is it ??s own little rabbit hole to wander down and each has 
it ??s own issues. 

Now you need a reference:

GPSDO ??s have their fans, they often have spurs and other  ??crud ?? on the 
output. They do have very good long term stability. 

OCXO ??s are often pretty quiet and spur free. You need some way to calibrate 
them for really long term stuff.

Rb ??s come in a wide range of sizes and costs. Some are better than others. 
Their ADEV at 1 second often not as good as an OCXO.

Cs standards and Hydrogen masers are (to me) in the same cost category as 
exotic test gear above. Fine for a company, not quite so easy for a basement 
guy.

Again each of these is a bit of a project by it ??s self. Often it turns into a 
 ??a couple of these and a couple of those ?? sort of solution. You only know 
something is right if you have another one to compare it to. 

Now you have some gear. Let ??s ignore building the mixer board and assume that 
??s taken care of already. 

Now you need software:

TimeLab is free and it will interface with many of the counters you might be 
using. It does all the math for you and puts up pretty plots. Highly 
recommended. There are other free software packages out there.

Stable32 is another commonly used program. You need to get the data out of the 
counter before you can use it. There are other paid programs out there, Stable 
32 is the best of the bunch.

An Excel spread sheet is indeed another option. The ADEV math is *not* very 
complicated. Excel can do it all very easily. 

Getting data out of the counter could be a terminal program sort of thing 
(serial port) or something more complex (GPIB). In both cases there are a 
number of programs to pick between. 

With software, things like what computer and operating system you have will 
influence your choices. I ??m a Mac person, but I also run Linux, Windows, 
FreeBSD and a few other things as needed. Most people have a favorite  ? .

So, that ??s the quick and dirty start to the  ??quick ADEV setup ?? process. 
There are:

4 counter like things X 4 reference ideas X (at least 4 mixer approaches) X 
(way more than) 4 software options =  256 paths you could follow. 

That ??s quite a few, and many of them would be utter nonsense. Even pruning 
out the ones are unlikely to ever be followed, you still have lots.

Here ??s one:

1) Drive your  ??known good ?? reference OCXO into a Minicircuits RPD-1 mixer. 

2) Drive your  ??device under test ?? into the other side of the RPD-1

3) Amplify and limit the output with a  couple of OP-37 op amps running off of 
+/-15V.
first stage is an amp with high pass and low pass sections
second stage is an inverter / limiter (back to back diodes in a 
normal op amp inverter).
third stage is same as the second
you might (or might not) want a fourth stage 

4) Feed the +/- 0.7 V  limited output into a 5335  (DC couple the input) in 
talk only mode

5) Hook up the 5335 to a National Instruments GPIB card on a PC

6) Run a Visual Basic routine to grab the outputs and write them to a text 
file. 

7) Process the text file with Excel. Use the ADEV formula from the original 
NIST papers. 

There have been a *lot* of OCXO ??s tested that way in many different factories 
over a couple of decades. (Yes, it ??s probably easier these days to substitute 
Time Lab for steps 6 and 7). 

Is it the ideal

Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Anthony Roby
Thanks - seems that I should be able to do this with my Racal-Dana 1992 counter.

Anthony

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Magnus 
Danielson
Sent: Saturday, November 01, 2014 1:29 PM
To: time-nuts@febo.com
Cc: mag...@rubidium.se
Subject: Re: [time-nuts] Measuring ADEV for a beginner

Anthony,

On 11/01/2014 02:29 PM, Anthony Roby wrote:
 I've been reading a lot about ADEV and following the threads on the list, 
 particularly Karen's in-flight thread.
 What I haven't come across is a simple explanation of the basic setup 
 required to go about collecting the data.
 John Miles referenced this page http://www.ke5fx.com/tpll.htm, and the 
 simple setup at the bottom of the page looks like a reasonable place 
 to start.  Seems that I'd need to acquire a phase detector and build 
 or buy some filters and the amp.  I can probably figure that out, but how do 
 I get the data into a PC?  Is there a basic hardware and software setup that 
 someone could point me to or recommend?

The time-interval counter, such as HP5370 or SR620, get started (channel
1) by a reference clock, such as 1 PPS and is then stopped (channel 2) by 
signal under test. The counter is typically read out through GPIB, even if some 
counters have serial interface and maybe even USB or Ethernet for really new 
(or retro-fitted), and the recommended path is to get a GPIB to USB interface 
for instance.
Then use John Miles TimeLab.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Hal Murray

ar...@antamy.com said:
 I'll do a bit more digging around and see if I can get the GPIB up and
 running in the next couple of weeks.

I've been happy with the ProLogic GPIB-USB gizmo.
  https://www.sparkfun.com/products/549

It took some fiddling to get going, but I like chasing that sort of thing.
  http://www.megapathdsl.net/~hmurray/hacks/probe.c


-- 
These are my opinions.  I hate spam.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Measuring ADEV for a beginner

2014-11-01 Thread Magnus Danielson

That will work as a starting-point.

When you look at your ADEV plot, you will notice a 1/tau curve for the 
lower taus, that is due to your counters limitations. If you need to go 
below that, you need a better counter, but for the moment you should 
start believe the plot as it flattens out, that has more to do with your 
signal, unless that is the same as your reference or a very high 
measurement floor of the counter.


A good test is to split your reference into the start and stop inputs, 
and then measure the amount of noise you have. Preferably with a 
slightly longer cable to the stop channel. This will give you the noise 
floor of the counter, for that signaltype and trigger-point.


There are ways to improve things if your noise is higher than the 
single-shot resolution, as you read out the ADEV at tau=1s.


Cheers,
Magnus

On 11/01/2014 09:43 PM, Anthony Roby wrote:

Thanks - seems that I should be able to do this with my Racal-Dana 1992 counter.

Anthony

-Original Message-
From: time-nuts [mailto:time-nuts-boun...@febo.com] On Behalf Of Magnus 
Danielson
Sent: Saturday, November 01, 2014 1:29 PM
To: time-nuts@febo.com
Cc: mag...@rubidium.se
Subject: Re: [time-nuts] Measuring ADEV for a beginner

Anthony,

On 11/01/2014 02:29 PM, Anthony Roby wrote:

I've been reading a lot about ADEV and following the threads on the list, 
particularly Karen's in-flight thread.
What I haven't come across is a simple explanation of the basic setup required 
to go about collecting the data.
John Miles referenced this page http://www.ke5fx.com/tpll.htm, and the
simple setup at the bottom of the page looks like a reasonable place
to start.  Seems that I'd need to acquire a phase detector and build
or buy some filters and the amp.  I can probably figure that out, but how do I 
get the data into a PC?  Is there a basic hardware and software setup that 
someone could point me to or recommend?


The time-interval counter, such as HP5370 or SR620, get started (channel
1) by a reference clock, such as 1 PPS and is then stopped (channel 2) by 
signal under test. The counter is typically read out through GPIB, even if some 
counters have serial interface and maybe even USB or Ethernet for really new 
(or retro-fitted), and the recommended path is to get a GPIB to USB interface 
for instance.
Then use John Miles TimeLab.

Cheers,
Magnus
___
time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.