Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-29 Thread Bob Camp
Hi

Would be nice to use something like the ADS1278. 

http://www.ti.com/product/ads1278

Lots of channels to cross correlate, and very little flicker noise.

Bob


On Nov 28, 2012, at 9:29 AM, Support HpW-Works.com supp...@hpw-works.com 
wrote:

 Hi Bruce,
 
 The equivalent phase noise (measured in dBc/Hz) is essentially independent 
 of
 the sample size.
 
 The dBc/Hz normalization is based on this (also the sample rate). While the 
 sample
 rate  sample size = FFT Bin size (resolution or filter band width) is used 
 to get
 the required correction factor of the spectrum to get the dBc / (1Hz) Y 
 scaling
 back.
 
 So increasing the sample size isnt particularly useful for reducing the 
 phase
 noise floor.
 
 By theory, yes... but we use a sound card with a lot of flicker noise on the 
 lower
 end, also we have the 10...20Hz low freq. cutoff due the usage of a servo / 
 single
 5V power. Also the raising noise below 100Hz (ADC serve  noise on the ADC 
 power /
 input circuit) limits the performance.
 
 In my simple test increase of the sample size reduced the noise floor in 
 better way
 than just using sample size with 1-2K and large averaging cycles.
 
 Keep in mind, Bin resolution is sample rate / sample size:
 
 Example: 
 
 - 32khz / 32678 = about 1Hz
 - 32khz / 1024  = about 31 Hz
 
 
 However with a sound card plus a mixer a somewhat lower number of samples
 should suffice since there is no carrier.
 
 
 Ideally the input circuit  ADC of the 3562A would be nice but with much 
 larger RAM
 buffer and ASIO interface O:)
 
 Hanspeter
 
 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf
 Of Bruce Griffiths
 Sent: Dienstag, 27. November 2012 12:31
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Best phase detector / mixer for 100MHz?
 
 Support HpW-Works.com wrote:
 Bruce,
 
 
 There's no evidence of a cross power spectrum function in this suite.
 One needs to be able to average at least 10,000 cross power spectra
 for some applications.
 
 In the PSD (power density)  PSP (power spectrum) there are cross
 power and cross power complex average implemented (selectable using the
 spectrum channel mixer)!
 Additional to this you may apply / add averaging of the resulting
 spectrum or use additional peak hold.
 
 Average of 10'000 cross points is a large count and often seen on 1-4k
 sample size.
 Better in my opinion is to use a higher sample size 32k-64k and then
 less averaging is required.
 
 The equivalent phase noise (measured in dBc/Hz) is essentially independent 
 of
 the sample size.
 So increasing the sample size isnt particularly useful for reducing the 
 phase
 noise floor.
 The increased frequency resolution achieved by increasing the sample size is
 only useful for measuring spurs.
 In the direct digital method of measuring phase noise a few terasamples (a 
 few
 gigasamples at baseband) need to be processed to achieve a sufficiently low
 instrument noise floor.
 However with a sound card plus a mixer a somewhat lower number of samples
 should suffice since there is no carrier.
 Just download the evaluation version with fully feature set.
 
 HpW
 
 
 
 Bruce
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-29 Thread Bill Fuqua

At 03:56 PM 11/29/2012 +, you wrote:
 By theory, yes... but we use a sound card with a lot of flicker noise 
on the lower
 end, also we have the 10...20Hz low freq. cutoff due the usage of a 
servo / single
 5V power. Also the raising noise below 100Hz (ADC serve  noise on the 
ADC power /

 input circuit) limits the performance.


One way to avoid the problem in the 1/f noise region is to avoid it.
Mix down to a few kHz and analyze you signal or if you must mix down
to DC then chop it to bring it up to a kHz or so and analyze it using
a program such as  spectrum lab or in software to internally mix back down
to DC using software digital mixer and then analyze.
 You can use a MOSFET analog switch to chop the signal.
  When you chop it you will have an carrier with upper and lower sidebands.
You can simply take the FFT data and either shift the bins making the 
carrier (chopping frequency)
the Zero or DC bin or even better yet add the upper and lower sidebands so 
that you get a 3db improvement
in S/N. I will be db since the sidebands are identical mirror of each 
other, ie voltage doubles when they
are added increasing the signal by 6db but noise introduced by electronics 
on upper and lower sidebands

are not coherent and only increases by 3db.

73
Bill wa4lav




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-29 Thread Support HpW-Works.com
Bill,

 One way to avoid the problem in the 1/f noise region is to avoid it.
 Mix down to a few kHz and analyze you signal

This is a good point! Using my SW (www.hpw-works.com) simple set the x-scaling 
to a
center frequency  bandwidth of your choice with enough space in the lower 
region.
This is implemented out of the box. It's like a measurement using the d-jitter 
but
with a costume center frequency. See also the pictures on my homepage with 
zooming
into the 0.005 Hz region using a large sample size. 

Required is that the source oscillators are stable...  

Hanspeter


 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf
 Of Bill Fuqua
 Sent: Donnerstag, 29. November 2012 18:02
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] Best phase detector / mixer for 100MHz?
 
 At 03:56 PM 11/29/2012 +, you wrote:
   By theory, yes... but we use a sound card with a lot of flicker
   noise
  on the lower
   end, also we have the 10...20Hz low freq. cutoff due the usage of a
  servo / single
   5V power. Also the raising noise below 100Hz (ADC serve  noise on
   the
  ADC power /
   input circuit) limits the performance.
 
 One way to avoid the problem in the 1/f noise region is to avoid it.
 Mix down to a few kHz and analyze you signal or if you must mix down to DC
 then chop it to bring it up to a kHz or so and analyze it using a program 
 such
 as  spectrum lab or in software to internally mix back down to DC using
 software digital mixer and then analyze.
   You can use a MOSFET analog switch to chop the signal.
When you chop it you will have an carrier with upper and lower sidebands.
 You can simply take the FFT data and either shift the bins making the carrier
 (chopping frequency) the Zero or DC bin or even better yet add the upper and
 lower sidebands so that you get a 3db improvement in S/N. I will be db since
 the sidebands are identical mirror of each other, ie voltage doubles when 
 they
 are added increasing the signal by 6db but noise introduced by electronics on
 upper and lower sidebands are not coherent and only increases by 3db.
 
 73
 Bill wa4lav
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-28 Thread Support HpW-Works.com
Hi Bruce,

 The equivalent phase noise (measured in dBc/Hz) is essentially independent of
 the sample size.

The dBc/Hz normalization is based on this (also the sample rate). While the 
sample
rate  sample size = FFT Bin size (resolution or filter band width) is used to 
get
the required correction factor of the spectrum to get the dBc / (1Hz) Y scaling
back.

 So increasing the sample size isnt particularly useful for reducing the phase
 noise floor.

By theory, yes... but we use a sound card with a lot of flicker noise on the 
lower
end, also we have the 10...20Hz low freq. cutoff due the usage of a servo / 
single
5V power. Also the raising noise below 100Hz (ADC serve  noise on the ADC 
power /
input circuit) limits the performance.

In my simple test increase of the sample size reduced the noise floor in better 
way
than just using sample size with 1-2K and large averaging cycles.

Keep in mind, Bin resolution is sample rate / sample size:

Example: 

- 32khz / 32678 = about 1Hz
- 32khz / 1024  = about 31 Hz


 However with a sound card plus a mixer a somewhat lower number of samples
 should suffice since there is no carrier.


Ideally the input circuit  ADC of the 3562A would be nice but with much larger 
RAM
buffer and ASIO interface O:)

Hanspeter

 -Original Message-
 From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On 
 Behalf
 Of Bruce Griffiths
 Sent: Dienstag, 27. November 2012 12:31
 To: Discussion of precise time and frequency measurement
 Subject: Re: [time-nuts] Best phase detector / mixer for 100MHz?
 
 Support HpW-Works.com wrote:
  Bruce,
 
 
  There's no evidence of a cross power spectrum function in this suite.
  One needs to be able to average at least 10,000 cross power spectra
  for some applications.
 
  In the PSD (power density)  PSP (power spectrum) there are cross
  power and cross power complex average implemented (selectable using the
 spectrum channel mixer)!
  Additional to this you may apply / add averaging of the resulting
  spectrum or use additional peak hold.
 
  Average of 10'000 cross points is a large count and often seen on 1-4k
 sample size.
  Better in my opinion is to use a higher sample size 32k-64k and then
  less averaging is required.
 
 The equivalent phase noise (measured in dBc/Hz) is essentially independent of
 the sample size.
 So increasing the sample size isnt particularly useful for reducing the phase
 noise floor.
 The increased frequency resolution achieved by increasing the sample size is
 only useful for measuring spurs.
 In the direct digital method of measuring phase noise a few terasamples (a 
 few
 gigasamples at baseband) need to be processed to achieve a sufficiently low
 instrument noise floor.
 However with a sound card plus a mixer a somewhat lower number of samples
 should suffice since there is no carrier.
  Just download the evaluation version with fully feature set.
 
  HpW
 
 
 
 Bruce
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com To unsubscribe, go to
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-28 Thread Bruce Griffiths

Support HpW-Works.com wrote:

Hi Bruce,

   

The equivalent phase noise (measured in dBc/Hz) is essentially independent of
the sample size.
   

The dBc/Hz normalization is based on this (also the sample rate). While the 
sample
rate  sample size = FFT Bin size (resolution or filter band width) is used to 
get
the required correction factor of the spectrum to get the dBc / (1Hz) Y scaling
back.

   

So increasing the sample size isnt particularly useful for reducing the phase
noise floor.
   

By theory, yes... but we use a sound card with a lot of flicker noise on the 
lower
end, also we have the 10...20Hz low freq. cutoff due the usage of a servo / 
single
5V power. Also the raising noise below 100Hz (ADC serve  noise on the ADC 
power /
input circuit) limits the performance.

In my simple test increase of the sample size reduced the noise floor in better 
way
than just using sample size with 1-2K and large averaging cycles.

Keep in mind, Bin resolution is sample rate / sample size:

Example:

- 32khz / 32678 = about 1Hz
- 32khz / 1024  = about 31 Hz
   
Your sample sizes are far too small to be particularly useful in 
measuring phase noise down to offsets of 1-10Hz or so,
Phase noise is highly coloured at such offsets necessitating the use of 
bin sizes substantially smaller than the lowest offset frequency of 
interest.
However decreasing the bin size beyond a small fraction of the lowest 
frequency of interest is counter productive in that one forgoes the 
effect of averaging to reduce the variance of the noise signal power 
within each bin.
There is a NIST paper on the effect of equivalent filter bandwidth on 
the accuracy of coloured noise measurements.


Indeed one can break the frequency range into a set of bands, the higher 
frequency bands having larger bin sizes and greater averaging and hence 
(lower bin noise signal power variance) than the lower frequency bands.


However for the purposes of spur identification using as small a bin 
size as possible can be useful.




   

However with a sound card plus a mixer a somewhat lower number of samples
should suffice since there is no carrier.
   


Ideally the input circuit  ADC of the 3562A would be nice but with much larger 
RAM
buffer and ASIO interface O:)

Hanspeter

   
Using a high end PC should allow real time signal processing with 
200ksps or greater, this is certainly the case for some external USB 
instruments.


Bruce


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Bruce Griffiths

Demian Martin wrote:

I asked Wenzel about mixers for phase noise measurement and they directed me
to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
tested any myself but it's a pretty solid recommendation I think.

I got this guy to add cross correlation to his FFT software suite:
http://www.hpw-works.com/  however I have not built hardware to try it. He
has a 30 day trial and its not expensive for what it is, $300.

There's no evidence of a cross power spectrum function in this suite.
One needs to be able to average at least 10,000 cross power spectra for 
some applications.


Bruce


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Adrian

Bruce Griffiths schrieb:

Richard (Rick) Karlquist wrote:



On 11/23/2012 9:38 AM, Bruce Griffiths wrote:


NIST have shown (at least at 10MHz) that the high level mixers they
tested are noisier than the ZRPD1.

Bruce


Do you have a citation to where they said that?
What you quoted doesn't make sense, at least, out of context.

We need to clarify phase detector sensitivity specs.
For conventional (IE 50 ohm) phase detectors, it is
apples vs apples to just go by the volts per radian number.
However, mixers like the ZRPD1 artificially triple the
voltage sensitivity by operating at 500 ohms, and using
transformers to connect to 50 ohm equipment.  Doing this
doesn't increase the possible signal to noise ratio.
Consider this thought experiment.  Build your best 500 ohm
phase detector and postamp.  Now replace with a 50 ohm
phase detector and connect 3 postamps in parallel.  It is
a wash.  Of course, you don't have to actually do this.
You can simply use an op amp like the LT1028 with very
low noise voltage.

To actually put a 500 ohm detector on a par with a 50 ohm
detector, the 500 ohm detector would need to use 3 diodes
in series compared to one in the 50 ohm case.  With only
one diode per arm, the maximum drive power utilization is
considerably lower.

Rick Karlquist N6RK


http://tf.boulder.nist.gov/general/pdf/2556.pdf

Where all the high level mixers measured have a higher phase noise 
than lower level mixers/phase detectors like the ZRPD1.


Bruce


I can't confirm their results, at least not when plugged into a HP 
11848A which loads the PD with 50 Ohms.


Too bad they are unclear about the applied power.
(1) Is it 'LO Power = maximum mixer power = 11 dBm', that is +11 dBm 
into both ports?
(2) Is it +17 dBm due to 'mixers with a nominal LO rating of +7 dBm and 
maximum rating of +17 dBm'?
Or, is it 50 mW as specified by Minicircuits, which would be the maximum 
mixer power as by (1)?


With +16...+17 dBm signals I'm getting a noise floor of -175 dBc/Hz 
which is quite poor but understandable as in order to drive it at 
nominal input power, I have to put 6 dB attenuators between the 
oscillators and the phase detector. With only 3 dB attenuation, I'm 
getting -178 dBc/Hz. I wasn't sure if the 50 mW spec refers to the total 
RF power into the ZRPD-1 or if that is the max. power into each input, 
so I didn't try +17 dBm. I would have expected to see -180...-181 dBc/Hz.


At that input level the 11848A internal PD gets me -179 dBc/Hz (no 
attenuators required).


A WJ M9GC was 1 dB worse than the internal PD (also w/o attenuators).

Surprisingly I've got the best results (-181) with a +23 dBm mixer (w/o 
attenuators of course).


I did not have +23 dBm signals at hands, but I don't see that the ZRPD-1 
can compete with the 11848A PD or an external +23 dBm mixer at input 
levels of +17 dBm and above.


Adrian




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Adrian

Bruce Griffiths schrieb:

Demian Martin wrote:
I asked Wenzel about mixers for phase noise measurement and they 
directed me

to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
tested any myself but it's a pretty solid recommendation I think.

I got this guy to add cross correlation to his FFT software suite:
http://www.hpw-works.com/  however I have not built hardware to try 
it. He

has a 30 day trial and its not expensive for what it is, $300.

There's no evidence of a cross power spectrum function in this suite.
One needs to be able to average at least 10,000 cross power spectra 
for some applications.


Bruce

Here is one that does cross spectrum and more: http://www.sigview.com/
A while ago I tried it with a Xonar Essence XT sound card and the 
results looked quite promising.


Then I got a 3562A to get to 100 kHz (don't follow me there! These can 
be challenging to fix if not impossible. PAL's and other unobtainable 
'field programmables' appear to die faster than anything else in that 
boxes. Finally, I needed three of them to get one working unit...)


Adrian




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Bruce Griffiths

Adrian wrote:

Bruce Griffiths schrieb:

Demian Martin wrote:
I asked Wenzel about mixers for phase noise measurement and they 
directed me

to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
tested any myself but it's a pretty solid recommendation I think.

I got this guy to add cross correlation to his FFT software suite:
http://www.hpw-works.com/  however I have not built hardware to try 
it. He

has a 30 day trial and its not expensive for what it is, $300.

There's no evidence of a cross power spectrum function in this suite.
One needs to be able to average at least 10,000 cross power spectra 
for some applications.


Bruce

Here is one that does cross spectrum and more: http://www.sigview.com/
A while ago I tried it with a Xonar Essence XT sound card and the 
results looked quite promising.


Then I got a 3562A to get to 100 kHz (don't follow me there! These can 
be challenging to fix if not impossible. PAL's and other unobtainable 
'field programmables' appear to die faster than anything else in that 
boxes. Finally, I needed three of them to get one working unit...)


Adrian

Tried that as well, how does one get a log y scale on the cross spectrum 
plot?


Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Adrian

Bruce Griffiths schrieb:

Adrian wrote:

Bruce Griffiths schrieb:

Demian Martin wrote:
I asked Wenzel about mixers for phase noise measurement and they 
directed me

to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
tested any myself but it's a pretty solid recommendation I think.

I got this guy to add cross correlation to his FFT software suite:
http://www.hpw-works.com/  however I have not built hardware to try 
it. He

has a 30 day trial and its not expensive for what it is, $300.

There's no evidence of a cross power spectrum function in this suite.
One needs to be able to average at least 10,000 cross power spectra 
for some applications.


Bruce

Here is one that does cross spectrum and more: http://www.sigview.com/
A while ago I tried it with a Xonar Essence XT sound card and the 
results looked quite promising.


Then I got a 3562A to get to 100 kHz (don't follow me there! These 
can be challenging to fix if not impossible. PAL's and other 
unobtainable 'field programmables' appear to die faster than anything 
else in that boxes. Finally, I needed three of them to get one 
working unit...)


Adrian

Tried that as well, how does one get a log y scale on the cross 
spectrum plot?


Bruce


With Sigview? I'm sorry my trial version has long expired and even after 
installing the latest version it's now requiring a license key...


Adrian



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Support HpW-Works.com
Bruce,

 There's no evidence of a cross power spectrum function in this suite.
 One needs to be able to average at least 10,000 cross power spectra for some
 applications.

In the PSD (power density)  PSP (power spectrum) there are cross power and 
cross
power complex average implemented (selectable using the spectrum channel mixer)!
Additional to this you may apply / add averaging of the resulting spectrum or 
use
additional peak hold. 

Average of 10'000 cross points is a large count and often seen on 1-4k sample 
size.
Better in my opinion is to use a higher sample size 32k-64k and then less 
averaging
is required. 

Just download the evaluation version with fully feature set.

HpW



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Bruce Griffiths

Adrian wrote:

Bruce Griffiths schrieb:

Adrian wrote:

Bruce Griffiths schrieb:

Demian Martin wrote:
I asked Wenzel about mixers for phase noise measurement and they 
directed me

to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx  I have not 
obtained or

tested any myself but it's a pretty solid recommendation I think.

I got this guy to add cross correlation to his FFT software suite:
http://www.hpw-works.com/  however I have not built hardware to 
try it. He

has a 30 day trial and its not expensive for what it is, $300.

There's no evidence of a cross power spectrum function in this suite.
One needs to be able to average at least 10,000 cross power spectra 
for some applications.


Bruce

Here is one that does cross spectrum and more: http://www.sigview.com/
A while ago I tried it with a Xonar Essence XT sound card and the 
results looked quite promising.


Then I got a 3562A to get to 100 kHz (don't follow me there! These 
can be challenging to fix if not impossible. PAL's and other 
unobtainable 'field programmables' appear to die faster than 
anything else in that boxes. Finally, I needed three of them to get 
one working unit...)


Adrian

Tried that as well, how does one get a log y scale on the cross 
spectrum plot?


Bruce


With Sigview? I'm sorry my trial version has long expired and even 
after installing the latest version it's now requiring a license key...


Adrian

The help file states that the cross power spectrum is the product of the 
individual spectra - this is only correct if the individual spectra are 
real.


Bruce


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Bruce Griffiths

Support HpW-Works.com wrote:

Bruce,

   

There's no evidence of a cross power spectrum function in this suite.
One needs to be able to average at least 10,000 cross power spectra for some
applications.
   

In the PSD (power density)  PSP (power spectrum) there are cross power and 
cross
power complex average implemented (selectable using the spectrum channel mixer)!
Additional to this you may apply / add averaging of the resulting spectrum or 
use
additional peak hold.

Average of 10'000 cross points is a large count and often seen on 1-4k sample 
size.
Better in my opinion is to use a higher sample size 32k-64k and then less 
averaging
is required.
   
The equivalent phase noise (measured in dBc/Hz) is essentially 
independent of the sample size.
So increasing the sample size isnt particularly useful for reducing the 
phase noise floor.
The increased frequency resolution achieved by increasing the sample 
size is only useful for measuring spurs.
In the direct digital method of measuring phase noise a few terasamples 
(a few gigasamples at baseband) need to be processed to achieve a 
sufficiently low instrument noise floor.
However with a sound card plus a mixer a somewhat lower number of 
samples should suffice since there is no carrier.

Just download the evaluation version with fully feature set.

HpW


   

Bruce


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-27 Thread Adrian

Bruce Griffiths schrieb:

Support HpW-Works.com wrote:

Bruce,


There's no evidence of a cross power spectrum function in this suite.
One needs to be able to average at least 10,000 cross power spectra 
for some

applications.
In the PSD (power density)  PSP (power spectrum) there are cross 
power and cross
power complex average implemented (selectable using the spectrum 
channel mixer)!
Additional to this you may apply / add averaging of the resulting 
spectrum or use

additional peak hold.

Average of 10'000 cross points is a large count and often seen on 
1-4k sample size.
Better in my opinion is to use a higher sample size 32k-64k and then 
less averaging

is required.
The equivalent phase noise (measured in dBc/Hz) is essentially 
independent of the sample size.
So increasing the sample size isnt particularly useful for reducing 
the phase noise floor.
The increased frequency resolution achieved by increasing the sample 
size is only useful for measuring spurs.
In the direct digital method of measuring phase noise a few 
terasamples (a few gigasamples at baseband) need to be processed to 
achieve a sufficiently low instrument noise floor.
However with a sound card plus a mixer a somewhat lower number of 
samples should suffice since there is no carrier.

Just download the evaluation version with fully feature set.

HpW



Bruce


A HP/Agilent VEE based solution would be much more flexible. There is a 
sound card driver available.
I just didn't get deep enough into digital signal processing before the 
3562A came...


Adrian


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread Anders Time
Thanks a lot guys for all the input!
The nist articles was a very interesting read. I have ordered the
minicircuits ZRPD-1 and will try to build the 2NA mixer to to see how
far that will take it. I probably also will build a low noise jfet preamp
to see if that will reduce the noise. But I probably have to go to build
myself a crosscorrelation system to be able to -180dBc with confidence. Is
there any cheap way into cross correlation measurement, using high quality
soundcard for example? Or do I have to buy one of these big old expensive
dual FFT analyzers?
/Anders

On Fri, Nov 23, 2012 at 4:42 PM, Anders Time anderst...@gmail.com wrote:

 I have been using an minicircuits mixer as a phase detector for measuring
 low frequency(5-10MHz) and it is usually good enough. But when I want to
 measure 100MHz the sensitivity of the mixer decreases a lot, so when I want
 to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
 out) oscillators the sensitivity is not good enough. I read in an old
 article by Walls, Stein et al(Design considerations in state-of-the-art
 signal processing and phase noise measurement system) that one can use two
 diodes in series in the double balanced mixer to increase the sensitivity.
 I tried this with some standard 1n5711 Schottky and the sensitivity is now
 1V/rad, but is there a easier way to do this? /Anders


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread Doug Parker
thanks Bob
 
Doug
 


 From: Anders Time anderst...@gmail.com
To: time-nuts@febo.com 
Sent: Monday, 26 November 2012, 9:45
Subject: Re: [time-nuts] Best phase detector / mixer for 100MHz?
  
Thanks a lot guys for all the input!
The nist articles was a very interesting read. I have ordered the
minicircuits ZRPD-1 and will try to build the 2NA mixer to to see how
far that will take it. I probably also will build a low noise jfet preamp
to see if that will reduce the noise. But I probably have to go to build
myself a crosscorrelation system to be able to -180dBc with confidence. Is
there any cheap way into cross correlation measurement, using high quality
soundcard for example? Or do I have to buy one of these big old expensive
dual FFT analyzers?
/Anders

On Fri, Nov 23, 2012 at 4:42 PM, Anders Time anderst...@gmail.com wrote:

 I have been using an minicircuits mixer as a phase detector for measuring
 low frequency(5-10MHz) and it is usually good enough. But when I want to
 measure 100MHz the sensitivity of the mixer decreases a lot, so when I want
 to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
 out) oscillators the sensitivity is not good enough. I read in an old
 article by Walls, Stein et al(Design considerations in state-of-the-art
 signal processing and phase noise measurement system) that one can use two
 diodes in series in the double balanced mixer to increase the sensitivity.
 I tried this with some standard 1n5711 Schottky and the sensitivity is now
 1V/rad, but is there a easier way to do this? /Anders


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread Demian Martin
I asked Wenzel about mixers for phase noise measurement and they directed me
to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
tested any myself but it's a pretty solid recommendation I think.

I got this guy to add cross correlation to his FFT software suite:
http://www.hpw-works.com/  however I have not built hardware to try it. He
has a 30 day trial and its not expensive for what it is, $300. I would use
it with an ESI Juli@ card http://www.esi-audio.com/products/julia/  ($150 or
less on eBay or elsewhere) since you can be pretty sure it won't do
something to mess with the measurement and it runs at 192 KHz sample rate
quite well. Around $450 for a cross correlation analyzer. Or you can go
shopping for an old FFT with the capability and potential service nightmare.


 
 Thanks a lot guys for all the input!
 The nist articles was a very interesting read. I have ordered the
 minicircuits ZRPD-1 and will try to build the 2NA mixer to to see how
 far that will take it. I probably also will build a low noise jfet preamp
 to see if that will reduce the noise. But I probably have to go to build
 myself a crosscorrelation system to be able to -180dBc with confidence.
Is
 there any cheap way into cross correlation measurement, using high quality
 soundcard for example? Or do I have to buy one of these big old expensive
 dual FFT analyzers?
 /Anders
**


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread Magnus Danielson

Anders,

On 11/26/2012 10:45 AM, Anders Time wrote:

Thanks a lot guys for all the input!
The nist articles was a very interesting read. I have ordered the
minicircuits ZRPD-1 and will try to build the 2NA mixer to to see how
far that will take it. I probably also will build a low noise jfet preamp
to see if that will reduce the noise. But I probably have to go to build
myself a crosscorrelation system to be able to-180dBc with confidence. Is
there any cheap way into cross correlation measurement, using high quality
soundcard for example? Or do I have to buy one of these big old expensive
dual FFT analyzers?


If you can handle the frequency limitation of the audio-card, you can do 
that. Cross-correlation using FFT isn't that hard to achieve, and good 
FFT libraries exist. FFTW for instance and a handfull lines of code do 
it all.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread Bob Camp
Hi

 or you could see if somebody with it already in their free time related
program would add a sound card or set of cards to the mix of devices they
support.

Bob

-Original Message-
From: time-nuts-boun...@febo.com [mailto:time-nuts-boun...@febo.com] On
Behalf Of Magnus Danielson
Sent: Monday, November 26, 2012 2:52 PM
To: time-nuts@febo.com
Subject: Re: [time-nuts] Best phase detector / mixer for 100MHz?

Anders,

On 11/26/2012 10:45 AM, Anders Time wrote:
 Thanks a lot guys for all the input!
 The nist articles was a very interesting read. I have ordered the
 minicircuits ZRPD-1 and will try to build the 2NA mixer to to see how
 far that will take it. I probably also will build a low noise jfet preamp
 to see if that will reduce the noise. But I probably have to go to build
 myself a crosscorrelation system to be able to-180dBc with confidence. Is
 there any cheap way into cross correlation measurement, using high quality
 soundcard for example? Or do I have to buy one of these big old expensive
 dual FFT analyzers?

If you can handle the frequency limitation of the audio-card, you can do 
that. Cross-correlation using FFT isn't that hard to achieve, and good 
FFT libraries exist. FFTW for instance and a handfull lines of code do 
it all.

Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread Jim Lux

On 11/26/12 10:11 AM, Demian Martin wrote:

I asked Wenzel about mixers for phase noise measurement and they directed me
to Marki Microwave as what they use:
http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
tested any myself but it's a pretty solid recommendation I think.



Isn't Marki making the old WJ mixers?  That seems to be what someone 
told me a few years ago: they basically have all the tooling and 
designs, etc.





___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread Rick Karlquist
Jim Lux wrote:
 On 11/26/12 10:11 AM, Demian Martin wrote:
 I asked Wenzel about mixers for phase noise measurement and they
 directed me
 to Marki Microwave as what they use:
 http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
 tested any myself but it's a pretty solid recommendation I think.


 Isn't Marki making the old WJ mixers?  That seems to be what someone
 told me a few years ago: they basically have all the tooling and
 designs, etc.

That's interesting, because MaCom Technology Solutions seems to
be making WJ models as well.  A few years back, there was another
company (I think the name started with an S) that seemed to have
inherited the WJ line.  The company currently called WJ has
nothing to do with these mixers at all.

Rick


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-26 Thread bg
 Jim Lux wrote:
 On 11/26/12 10:11 AM, Demian Martin wrote:
 I asked Wenzel about mixers for phase noise measurement and they
 directed me
 to Marki Microwave as what they use:
 http://www.markimicrowave.com/2770/Mixers.aspx  I have not obtained or
 tested any myself but it's a pretty solid recommendation I think.


 Isn't Marki making the old WJ mixers?  That seems to be what someone
 told me a few years ago: they basically have all the tooling and
 designs, etc.

 That's interesting, because MaCom Technology Solutions seems to
 be making WJ models as well.  A few years back, there was another
 company (I think the name started with an S) that seemed to have
 inherited the WJ line.  The company currently called WJ has
 nothing to do with these mixers at all.

 Rick

Did you see this information from Mini Circuits on WJ mixer replacements?

 
http://217.34.103.131/pdfs/Surface-Mount%20Mixers%20Are%20WJ%20Replacements.pdf

--

Björn


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-24 Thread Bruce Griffiths

Richard (Rick) Karlquist wrote:



On 11/23/2012 9:38 AM, Bruce Griffiths wrote:


NIST have shown (at least at 10MHz) that the high level mixers they
tested are noisier than the ZRPD1.

Bruce


Do you have a citation to where they said that?
What you quoted doesn't make sense, at least, out of context.

We need to clarify phase detector sensitivity specs.
For conventional (IE 50 ohm) phase detectors, it is
apples vs apples to just go by the volts per radian number.
However, mixers like the ZRPD1 artificially triple the
voltage sensitivity by operating at 500 ohms, and using
transformers to connect to 50 ohm equipment.  Doing this
doesn't increase the possible signal to noise ratio.
Consider this thought experiment.  Build your best 500 ohm
phase detector and postamp.  Now replace with a 50 ohm
phase detector and connect 3 postamps in parallel.  It is
a wash.  Of course, you don't have to actually do this.
You can simply use an op amp like the LT1028 with very
low noise voltage.

To actually put a 500 ohm detector on a par with a 50 ohm
detector, the 500 ohm detector would need to use 3 diodes
in series compared to one in the 50 ohm case.  With only
one diode per arm, the maximum drive power utilization is
considerably lower.

Rick Karlquist N6RK


http://tf.boulder.nist.gov/general/pdf/2556.pdf

Where all the high level mixers measured have a higher phase noise than 
lower level mixers/phase detectors like the ZRPD1.


Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-24 Thread Rick Karlquist
Bruce Griffiths wrote:
 Richard (Rick) Karlquist wrote:


 On 11/23/2012 9:38 AM, Bruce Griffiths wrote:

 NIST have shown (at least at 10MHz) that the high level mixers they
 tested are noisier than the ZRPD1.

 Bruce


 http://tf.boulder.nist.gov/general/pdf/2556.pdf

 Where all the high level mixers measured have a higher phase noise than
 lower level mixers/phase detectors like the ZRPD1.

 Bruce

This data is surprising.  I am wondering if it is being corrupted
by AM noise on the source?  They do not mention what the source is.
They could be measuring AM noise suppression differences in the
phase detectors.

Rick


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-24 Thread Bruce Griffiths

Rick Karlquist wrote:

Bruce Griffiths wrote:
   

Richard (Rick) Karlquist wrote:
 


On 11/23/2012 9:38 AM, Bruce Griffiths wrote:
   

NIST have shown (at least at 10MHz) that the high level mixers they
tested are noisier than the ZRPD1.

Bruce
 
   
   

http://tf.boulder.nist.gov/general/pdf/2556.pdf

Where all the high level mixers measured have a higher phase noise than
lower level mixers/phase detectors like the ZRPD1.

Bruce
 

This data is surprising.  I am wondering if it is being corrupted
by AM noise on the source?  They do not mention what the source is.
They could be measuring AM noise suppression differences in the
phase detectors.

Rick


   
Another potential issue is the relatively high input current noise of 
the IF amplifiers (see the paper's Reference 9).
In the cross correlation setup used the IF amps both see the noise 
produced by the sum of the IF amp input noise current flowing in the  
mixer output impedance.

This noise common to both IF amps isn't rejected by cross correlation.
It can be more effective to use IF amps with very low input noise 
current that may have a higher input voltage noise.
The effect of the amplifier input voltage noise is reduced by cross 
correlation.
For offset frequencies =10Hz a well designed low noise amplifier with a 
discrete JFET input stage can have comparable voltage noise to the IF 
amp used with much lower input current noise.


Bruce

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-23 Thread Richard (Rick) Karlquist

The old Watkins Johnson M9 series was the state of the art for
stacked diode mixers.  You can still get the M9E and M9H from
MaCom Technology Solutions.  The M9E is better, but only if you
have the 1/2 watt! of LO drive needed.  As you have done already,
it is probably possible to homebrew something like these.
There were a series of papers written by WJ people 20 years ago
or so in Microwave Journal or maybe it was Microwaves and RF
that explained all about these things.  These should be required
reading if you are going to homebrew.  Try to match the diodes
so you get low DC offset.  Some mixers also use resistors and
capacitors to assist the diodes; again read the WJ papers.

The horsepower race in phase detectors was somewhat rendered
unnecessary by the cross correlation techniques developed 10
or 15 years ago.  You can extend the effective noise floor by
dozens of dB this way.

Rick Karlquist N6RK

On 11/23/2012 7:42 AM, Anders Time wrote:

I have been using an minicircuits mixer as a phase detector for measuring
low frequency(5-10MHz) and it is usually good enough. But when I want to
measure 100MHz the sensitivity of the mixer decreases a lot, so when I want
to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
out) oscillators the sensitivity is not good enough. I read in an old
article by Walls, Stein et al(Design considerations in state-of-the-art
signal processing and phase noise measurement system) that one can use two
diodes in series in the double balanced mixer to increase the sensitivity.
I tried this with some standard 1n5711 Schottky and the sensitivity is now
1V/rad, but is there a easier way to do this? /Anders
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-23 Thread Alan Melia
Hi Anders, isn't this format exactly what is inside the high level mixers 
(spec'e +17dBm) from Minicircuits?

Alan
G3NYK

- Original Message - 
From: Anders Time anderst...@gmail.com

To: time-nuts@febo.com
Sent: Friday, November 23, 2012 3:42 PM
Subject: [time-nuts] Best phase detector / mixer for 100MHz?



I have been using an minicircuits mixer as a phase detector for measuring
low frequency(5-10MHz) and it is usually good enough. But when I want to
measure 100MHz the sensitivity of the mixer decreases a lot, so when I 
want

to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
out) oscillators the sensitivity is not good enough. I read in an old
article by Walls, Stein et al(Design considerations in state-of-the-art
signal processing and phase noise measurement system) that one can use two
diodes in series in the double balanced mixer to increase the sensitivity.
I tried this with some standard 1n5711 Schottky and the sensitivity is now
1V/rad, but is there a easier way to do this? /Anders
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there. 



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-23 Thread Bruce Griffiths


NIST have shown (at least at 10MHz) that the high level mixers they 
tested are noisier than the ZRPD1.


Bruce

Alan Melia wrote:
Hi Anders, isn't this format exactly what is inside the high level 
mixers (spec'e +17dBm) from Minicircuits?

Alan
G3NYK

- Original Message - From: Anders Time anderst...@gmail.com
To: time-nuts@febo.com
Sent: Friday, November 23, 2012 3:42 PM
Subject: [time-nuts] Best phase detector / mixer for 100MHz?


I have been using an minicircuits mixer as a phase detector for 
measuring

low frequency(5-10MHz) and it is usually good enough. But when I want to
measure 100MHz the sensitivity of the mixer decreases a lot, so when 
I want

to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
out) oscillators the sensitivity is not good enough. I read in an old
article by Walls, Stein et al(Design considerations in state-of-the-art
signal processing and phase noise measurement system) that one can 
use two
diodes in series in the double balanced mixer to increase the 
sensitivity.
I tried this with some standard 1n5711 Schottky and the sensitivity 
is now

1V/rad, but is there a easier way to do this? /Anders
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there. 



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-23 Thread Richard (Rick) Karlquist



On 11/23/2012 9:38 AM, Bruce Griffiths wrote:


NIST have shown (at least at 10MHz) that the high level mixers they
tested are noisier than the ZRPD1.

Bruce


Do you have a citation to where they said that?
What you quoted doesn't make sense, at least, out of context.

We need to clarify phase detector sensitivity specs.
For conventional (IE 50 ohm) phase detectors, it is
apples vs apples to just go by the volts per radian number.
However, mixers like the ZRPD1 artificially triple the
voltage sensitivity by operating at 500 ohms, and using
transformers to connect to 50 ohm equipment.  Doing this
doesn't increase the possible signal to noise ratio.
Consider this thought experiment.  Build your best 500 ohm
phase detector and postamp.  Now replace with a 50 ohm
phase detector and connect 3 postamps in parallel.  It is
a wash.  Of course, you don't have to actually do this.
You can simply use an op amp like the LT1028 with very
low noise voltage.

To actually put a 500 ohm detector on a par with a 50 ohm
detector, the 500 ohm detector would need to use 3 diodes
in series compared to one in the 50 ohm case.  With only
one diode per arm, the maximum drive power utilization is
considerably lower.

Rick Karlquist N6RK

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-23 Thread Alan Melia

Not unreasonable Bruce..no free lunches etc.  :-))
Alan
G3NYK


- Original Message - 
From: Bruce Griffiths bruce.griffi...@xtra.co.nz
To: Discussion of precise time and frequency measurement 
time-nuts@febo.com

Sent: Friday, November 23, 2012 5:38 PM
Subject: Re: [time-nuts] Best phase detector / mixer for 100MHz?




NIST have shown (at least at 10MHz) that the high level mixers they tested 
are noisier than the ZRPD1.


Bruce

Alan Melia wrote:
Hi Anders, isn't this format exactly what is inside the high level mixers 
(spec'e +17dBm) from Minicircuits?

Alan
G3NYK

- Original Message - From: Anders Time anderst...@gmail.com
To: time-nuts@febo.com
Sent: Friday, November 23, 2012 3:42 PM
Subject: [time-nuts] Best phase detector / mixer for 100MHz?


I have been using an minicircuits mixer as a phase detector for 
measuring

low frequency(5-10MHz) and it is usually good enough. But when I want to
measure 100MHz the sensitivity of the mixer decreases a lot, so when I 
want

to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
out) oscillators the sensitivity is not good enough. I read in an old
article by Walls, Stein et al(Design considerations in state-of-the-art
signal processing and phase noise measurement system) that one can use 
two
diodes in series in the double balanced mixer to increase the 
sensitivity.
I tried this with some standard 1n5711 Schottky and the sensitivity is 
now

1V/rad, but is there a easier way to do this? /Anders
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts

and follow the instructions there.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to 
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there. 



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-23 Thread Magnus Danielson

Hi Anders,

On 11/23/2012 04:42 PM, Anders Time wrote:

I have been using an minicircuits mixer as a phase detector for measuring
low frequency(5-10MHz) and it is usually good enough. But when I want to
measure 100MHz the sensitivity of the mixer decreases a lot, so when I want
to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
out) oscillators the sensitivity is not good enough. I read in an old
article by Walls, Stein et al(Design considerations in state-of-the-art
signal processing and phase noise measurement system) that one can use two
diodes in series in the double balanced mixer to increase the sensitivity.
I tried this with some standard 1n5711 Schottky and the sensitivity is now
1V/rad, but is there a easier way to do this? /Anders


Craig Nelson and friends over at NIST had a little different stab at 
mixer which might interest you:

http://tf.boulder.nist.gov/general/pdf/2554.pdf

This was intended for 5 MHz, but I am sure you would enjoy reading about 
it never the less.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Best phase detector / mixer for 100MHz?

2012-11-23 Thread Bruce Griffiths

Magnus Danielson wrote:

Hi Anders,

On 11/23/2012 04:42 PM, Anders Time wrote:
I have been using an minicircuits mixer as a phase detector for 
measuring

low frequency(5-10MHz) and it is usually good enough. But when I want to
measure 100MHz the sensitivity of the mixer decreases a lot, so when 
I want

to measure some really low noise 100MHz(Pascall -178dBc floor with 18dBm
out) oscillators the sensitivity is not good enough. I read in an old
article by Walls, Stein et al(Design considerations in state-of-the-art
signal processing and phase noise measurement system) that one can 
use two
diodes in series in the double balanced mixer to increase the 
sensitivity.
I tried this with some standard 1n5711 Schottky and the sensitivity 
is now

1V/rad, but is there a easier way to do this? /Anders


Craig Nelson and friends over at NIST had a little different stab at 
mixer which might interest you:

http://tf.boulder.nist.gov/general/pdf/2554.pdf

This was intended for 5 MHz, but I am sure you would enjoy reading 
about it never the less.


Cheers,
Magnus

They've subsequently used it at 10MHz, 20MHz and 40MHz in regenerative 
dividers.


Bruce


___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.