Hi

The 10811 is very unusual in that the EFC is not bypassed. It’s still got a 
modulation bandwidth.  The important part is that the EFC has an inherent 
“lowpass" due to FM to PM conversion. 

The EFC line on the 10811 should be bypassed to ground at the socket with at 
least a 1000 pf cap.  Yes that’s not in the little manual. It’s the only way to 
knock out all of the RF tune effect on the EFC line. Of course if you are 
driving it with an op-amp that will terminate. 

Bob

On Aug 7, 2014, at 10:08 PM, Bob Stewart  wrote:

> Hi Bob,
> 
> I hadn't even considered a filter in the OCXO.  This isn't a 10811, but 
> that's the OCXO I have a schematic of, so I'll assume that's the benchmark.  
> Following the EFC in, it looks like it goes to a 100K resistor and then tees 
> to the 100pf varicap and a 15pf to the xtal.  Other caps are attached as 
> well, but it doesn't look like it's bypassed to ground anywhere along the EFC 
> line.  I see that there's another 100K to a 6.4V reference with a 6.8uf cap.  
> So, that means that the EFC line ranges from +6.4 to -6.4?  I haven't worked 
> out the time constant, but that wouldn't seem to apply for a 10MHz signal 
> riding on the EFC voltage.
> 
> 
> Like I said, I don't have a 10811 on my GPSDO.  It's my faithful Trimble 
> 34310-T.  Still, I would imagine that they at least looked at HP's design.
> 
> As to Hal's comment about probe pickup.  I was careful to specify the X10 
> position of the probe.  In the X1 position there was a signal that wasn't 
> visible in X10, but should have been.  So, I assumed that was some sort of 
> induced signal.  I'm using a generic cheap Chinese probe available on ebay.
> 
> 
> Bob
> 
> 
> 
> ________________
> From: Bob Camp 
> To: Discussion of precise time and frequency measurement  
> Sent: Thursday, August 7, 2014 6:11 PM
> Subject: Re: [time-nuts] Effects of noise on EFC line? - Resolved
> 
> 
> Hi
> 
> Your EFC line is probably bypassed internally to the OCXO. A 3db modulation 
> bandwidth beyond 1 KHz is unlikely. A modulation bandwidth below 100 Hz is 
> quite possible. 
> 
> Next thing to consider is that the EFC does FM on the OCXO. Phase noise is PM 
> modulation. FM is 1/Fmod relative to PM. If I go up a decade in frequency 
> with constant FM, my PM sideband will go down by 20 db. Yes that’s for small 
> modulation indexes. That’s very likely the case if we are dealing with noise. 
> 
> You can calculate exactly what PM sideband in dbc you will get from a 1 Hz 
> tone at 1 mV p-p on your EFC. From that you can pretty quickly work out what 
> this or that number of microvolts will do at this or that frequency.  The 
> answer is normally that the noise you have from a reasonable regulator or 
> op-amp isn’t a big deal. 
> 
> Bob
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Effects of noise on EFC line? - Resolved

2014-08-07 Thread Bob Stewart
Hi Bob,

I hadn't even considered a filter in the OCXO.  This isn't a 10811, but that's 
the OCXO I have a schematic of, so I'll assume that's the benchmark.  Following 
the EFC in, it looks like it goes to a 100K resistor and then tees to the 100pf 
varicap and a 15pf to the xtal.  Other caps are attached as well, but it 
doesn't look like it's bypassed to ground anywhere along the EFC line.  I see 
that there's another 100K to a 6.4V reference with a 6.8uf cap.  So, that means 
that the EFC line ranges from +6.4 to -6.4?  I haven't worked out the time 
constant, but that wouldn't seem to apply for a 10MHz signal riding on the EFC 
voltage.


Like I said, I don't have a 10811 on my GPSDO.  It's my faithful Trimble 
34310-T.  Still, I would imagine that they at least looked at HP's design.

As to Hal's comment about probe pickup.  I was careful to specify the X10 
position of the probe.  In the X1 position there was a signal that wasn't 
visible in X10, but should have been.  So, I assumed that was some sort of 
induced signal.  I'm using a generic cheap Chinese probe available on ebay.


Bob




 From: Bob Camp 
To: Discussion of precise time and frequency measurement  
Sent: Thursday, August 7, 2014 6:11 PM
Subject: Re: [time-nuts] Effects of noise on EFC line? - Resolved
 

Hi

Your EFC line is probably bypassed internally to the OCXO. A 3db modulation 
bandwidth beyond 1 KHz is unlikely. A modulation bandwidth below 100 Hz is 
quite possible. 

Next thing to consider is that the EFC does FM on the OCXO. Phase noise is PM 
modulation. FM is 1/Fmod relative to PM. If I go up a decade in frequency with 
constant FM, my PM sideband will go down by 20 db. Yes that’s for small 
modulation indexes. That’s very likely the case if we are dealing with noise. 

You can calculate exactly what PM sideband in dbc you will get from a 1 Hz tone 
at 1 mV p-p on your EFC. From that you can pretty quickly work out what this or 
that number of microvolts will do at this or that frequency.  The answer is 
normally that the noise you have from a reasonable regulator or op-amp isn’t a 
big deal. 

Bob
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] Effects of noise on EFC line? - Resolved

2014-08-07 Thread Alex Pummer



he is on the right truck, just look around in your testing environment, 
do you have shielded test set up, common ground for all the test gears ?


 On 8/7/2014 3:58 PM, Hal Murray wrote:

b...@evoria.net said:

So, I may throw another cap on it, but it seems to be clean down to what I
can measure at the OCXO on my old Tek 455 with an X10 probe.

Another thing to consider when chasing that sort of problem: How much are you
picking up with your scope probe and/or its ground wire?

The classic home brew better-probe is to use a chunk of 50 ohm coax feeding
into scope with a 50 ohm terminator.  (You get 2 of them if you cut a readily
available connectorized cable in half.)  Insert 950 ohms at the probe end to
get higher input impedance at the cost of a 20:1 divider.
   http://www.signalintegrity.com/Pubs/straight/probes.htm

It usually requires soldering at the DUT.  For me, it also requires a mental
shift.  You have to think of your scope probe as a consumable rather than
something that won't wear out if you take care of it.  A chunk of coax will
last a long time.  Just cut off another 1/2 inch when the end gets too
mangled.




___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Effects of noise on EFC line? - Resolved

2014-08-07 Thread Bob Camp
Hi

Your EFC line is probably bypassed internally to the OCXO. A 3db modulation 
bandwidth beyond 1 KHz is unlikely. A modulation bandwidth below 100 Hz is 
quite possible. 

Next thing to consider is that the EFC does FM on the OCXO. Phase noise is PM 
modulation. FM is 1/Fmod relative to PM. If I go up a decade in frequency with 
constant FM, my PM sideband will go down by 20 db. Yes that’s for small 
modulation indexes. That’s very likely the case if we are dealing with noise. 

You can calculate exactly what PM sideband in dbc you will get from a 1 Hz tone 
at 1 mV p-p on your EFC. From that you can pretty quickly work out what this or 
that number of microvolts will do at this or that frequency.  The answer is 
normally that the noise you have from a reasonable regulator or op-amp isn’t a 
big deal. 

Bob


On Aug 7, 2014, at 6:58 PM, Hal Murray  wrote:

> 
> b...@evoria.net said:
>> So, I may throw another cap on it, but it seems to be clean down to what I
>> can measure at the OCXO on my old Tek 455 with an X10 probe. 
> 
> Another thing to consider when chasing that sort of problem: How much are you 
> picking up with your scope probe and/or its ground wire?
> 
> The classic home brew better-probe is to use a chunk of 50 ohm coax feeding 
> into scope with a 50 ohm terminator.  (You get 2 of them if you cut a readily 
> available connectorized cable in half.)  Insert 950 ohms at the probe end to 
> get higher input impedance at the cost of a 20:1 divider.
>  http://www.signalintegrity.com/Pubs/straight/probes.htm
> 
> It usually requires soldering at the DUT.  For me, it also requires a mental 
> shift.  You have to think of your scope probe as a consumable rather than 
> something that won't wear out if you take care of it.  A chunk of coax will 
> last a long time.  Just cut off another 1/2 inch when the end gets too 
> mangled.
> 
> 
> -- 
> These are my opinions.  I hate spam.
> 
> 
> 
> ___
> time-nuts mailing list -- time-nuts@febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Effects of noise on EFC line? - Resolved

2014-08-07 Thread Hal Murray

b...@evoria.net said:
> So, I may throw another cap on it, but it seems to be clean down to what I
> can measure at the OCXO on my old Tek 455 with an X10 probe. 

Another thing to consider when chasing that sort of problem: How much are you 
picking up with your scope probe and/or its ground wire?

The classic home brew better-probe is to use a chunk of 50 ohm coax feeding 
into scope with a 50 ohm terminator.  (You get 2 of them if you cut a readily 
available connectorized cable in half.)  Insert 950 ohms at the probe end to 
get higher input impedance at the cost of a 20:1 divider.
  http://www.signalintegrity.com/Pubs/straight/probes.htm

It usually requires soldering at the DUT.  For me, it also requires a mental 
shift.  You have to think of your scope probe as a consumable rather than 
something that won't wear out if you take care of it.  A chunk of coax will 
last a long time.  Just cut off another 1/2 inch when the end gets too 
mangled.


-- 
These are my opinions.  I hate spam.



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] Effects of noise on EFC line? - Resolved

2014-08-06 Thread Bob Stewart
Thanks to everyone for their comments, and especially to Bill, WB6BNQ for cast 
a jaundiced eye on my schematic.

The problem turned out to be in the design of my 5V rail on the PCB.  There is 
a 2.5 inch run from the 5V regulator over to the hex driver (74HCT365) that 
outputs two signals of either 1MHz or 5MHz, as well as four 10MHz outputs.  
About .5" before the hex driver, I tapped off 5V to power both a thermistor and 
the output op-amp.  I hadn't counted on the impedance of about 2.5" of trace at 
10MHz, and the noise the hex driver would put on it.

What I did was cut the tee at both ends, and put a  jumper from  the thermistor 
& op-amp directly to the 5V regulator.  The 10MHz noise is now gone.  There is 
some noise from the DAC output that looks to be at the PIC's clock frequency, 
but that was filtered out by the voltage divider and .1uf cap at the EFC pin of 
the OCXO.

I did learn something interesting, though.  The Vref output of the OCXO 
(Trimble 34310-T) has about 4mv of 10MHz on it.  That was also cleaned up by 
the voltage divider/corrector.  So, I may throw another cap on it, but it seems 
to be clean down to what I can measure at the OCXO on my old Tek 455 with an 
X10 probe.


Bob




 From: Bob Stewart 
To: Time Nuts  
Sent: Tuesday, August 5, 2014 1:36 PM
Subject: [time-nuts] Effects of noise on EFC line?
 

I've run into a noise problem on the EFC line of my GPSDO engine at the 
frequency of the oscillator.  I've traced the source down to the 74HCT365 I'm 
using to output the 1(or 5)MHz and 10MHz signals.  When I pull it, the EFC 
quietens down a lot.  I'm seeing about 50mv of 10MHz noise at the output of the 
op-amp that feeds the EFC voltage divider at the OCXO.  The voltage divider is 
corrected by the VRef from the OCXO with a simple circuit using temp-co'ed 
resistors.  On the 0.1uf cap at the OCXO's EFC pin, I'm seeing about 5mv of 
10MHz signal.


I've considered switching the HCT out for a 74LS365, assuming my drive levels 
are compatible.  Unfortunately, I don't have one "in stock", and "I'm way out 
of my pay grade", as they say.  I've also thought about putting a 100uh 
inductor in series with the EFC line.  I wonder if I'll have to isolate the 
74xx365 chip's VCC through an inductor?  Any thoughts?


I'm also wondering what the impact of this level of on-frequency noise will be? 
 Is the impact somewhat mitigated, since it's at oscillator frequency?  I don't 
have anything better than an HP 8558B to look at the output of the board.

I'm not quite ready to generally share my schematic with the list, but I can 
make individual exceptions.


Bob - AE6RV
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.