Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-19 Thread mike cook

Le 19 févr. 2014 à 01:05, Tom Knox a écrit :

 Thanks Tom and Bob, I have been thinking of contacting Agilent for some time. 
 I think they are a great company with some good products, but there are a few 
 real blind spots in some current products. I also have seen in the past a 
 genuine interest in listening. I would be willing to approach them if I could 
 enlist your help in addressing potential changes to improve the product. 
 Thanks;
 Thomas Knox
 
   
   If they are steering the VCXXO,OCXO from the Ext. Ref. , then they are in 
effect calibrating it. Why not remember the applied EFC when they get phase 
lock?  That can be applied when the internal timebase is selected. 
It couldn't be that they might lose the chance to sell a signal generator ;-), 
as calibration needs a square wave input, and the Ext. Ref In is ignored.

 
 From: li...@rtty.us
 Date: Tue, 18 Feb 2014 18:00:17 -0500
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,  
 but nobody in?
 
 Hi
 
 Well at least this got me digging a little. 
 
 If you grab a copy of the 53230A spec sheet and look under the external 
 reference input, it’s pretty well described. It will accept 1, 5,10 MHz as 
 an external reference. It will lock over a 1 ppm range with the XO option 
 and 0.1 ppm with the OCXO option. Based on that I’d guess they are still 
 using the same basic PLL approach as on the older counters (5335 era). 
 
 The “Microsoft Windows inside” sticker on the back of the counter was a bit 
 of a surprise ….

  No sticker on mine. 

 
 Bob
 
 On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com wrote:
 
 TomK,
 
 If anyone has technical contacts deep within Agilent, let's see if this 
 issue can be resolved. I would have bought a 53230A when it came out a few 
 years ago but my eval units showed this clock noise problem. That plus the 
 poor quality of the ref out made me think the designers were cutting 
 corners, or had little experience in metrology, or maybe they thought this 
 was ok for a bench instrument.
 
 Otherwise it's a really nice counter; the first one from Agilent than can 
 actually do ADEV properly (since it is a time stamping counter).
 
 I should dig out my old data and send it to you. Maybe as group we can help 
 them fix the problem.
 
 /tvb
 
 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal 
 oscillator out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal applied. 
 These questions were related to several 53132A's I have seen configured 
 with a small board back near the Ext Ref input (OPT H01 I think) that 
 appeared to Switch the internal reference out of circuit. Agilent would 
 not share information on the option. My question to Agilent is why sell an 
 option and be unwilling to say what it does or how your stock unit 
 functions?
 Thomas Knox
 
 
 
 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
  but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of 
 the 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is 
 dirtier than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL 
 actually makes it worse. This was very disappointing. The XO version of 
 the counter performed worse than the OCXO version even with a maser as 
 the ext reference. Did your reading of the schematic show a way to 
 directly use the ext ref, bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted 
 copy of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those 
 digital ones that can lock up to many inputs. You could feed 3. 
 MHz in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The 
 internal oscillator (or an internal oscillator) is phase locked to the 
 external input through a fairly narrow analog loop. The idea is to clean 
 up the crud on the standard line. 
 
 With no external reference, the PLL drops out and you go back to what 
 ever the local reference is. 
 
 Yes there’s a little more to it than that and no the circuit is not 
 exactly the same on every counter HP ever made. 
 
 Bob
 
 On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
 
 Hi Mike,
 
 The most likely answer is when you select external time base for an 
 input, it disables the connection for the internal oscillator.  The 
 external input signal is probably also routed straight 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-19 Thread Said Jackson
Mike,

They are already giving you another way to calibrate the unit, different from 
how you think they should have done it and you are pulling out the statist card 
and accusing them of being greedy capitalists?

Come on, thats backseat driving. Be happy they invested millions of their own 
money and put out a more or less affordable new counter in a market flooded 
with good low-cost used counters.

Bye,
Said

Sent From iPhone

On Feb 19, 2014, at 0:33, mike cook michael.c...@sfr.fr wrote:

 
 Le 19 févr. 2014 à 01:05, Tom Knox a écrit :
 
 Thanks Tom and Bob, I have been thinking of contacting Agilent for some 
 time. I think they are a great company with some good products, but there 
 are a few real blind spots in some current products. I also have seen in the 
 past a genuine interest in listening. I would be willing to approach them if 
 I could enlist your help in addressing potential changes to improve the 
 product. 
 Thanks;
 Thomas Knox
 
   If they are steering the VCXXO,OCXO from the Ext. Ref. , then they are in 
 effect calibrating it. Why not remember the applied EFC when they get phase 
 lock?  That can be applied when the internal timebase is selected. 
 It couldn't be that they might lose the chance to sell a signal generator 
 ;-), as calibration needs a square wave input, and the Ext. Ref In is ignored.
 
 
 From: li...@rtty.us
 Date: Tue, 18 Feb 2014 18:00:17 -0500
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,
 but nobody in?
 
 Hi
 
 Well at least this got me digging a little. 
 
 If you grab a copy of the 53230A spec sheet and look under the external 
 reference input, it’s pretty well described. It will accept 1, 5,10 MHz as 
 an external reference. It will lock over a 1 ppm range with the XO option 
 and 0.1 ppm with the OCXO option. Based on that I’d guess they are still 
 using the same basic PLL approach as on the older counters (5335 era). 
 
 The “Microsoft Windows inside” sticker on the back of the counter was a bit 
 of a surprise ….
 
  No sticker on mine. 
 
 
 Bob
 
 On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com 
 wrote:
 
 TomK,
 
 If anyone has technical contacts deep within Agilent, let's see if this 
 issue can be resolved. I would have bought a 53230A when it came out a few 
 years ago but my eval units showed this clock noise problem. That plus the 
 poor quality of the ref out made me think the designers were cutting 
 corners, or had little experience in metrology, or maybe they thought this 
 was ok for a bench instrument.
 
 Otherwise it's a really nice counter; the first one from Agilent than can 
 actually do ADEV properly (since it is a time stamping counter).
 
 I should dig out my old data and send it to you. Maybe as group we can 
 help them fix the problem.
 
 /tvb
 
 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal 
 oscillator out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal 
 applied. These questions were related to several 53132A's I have seen 
 configured with a small board back near the Ext Ref input (OPT H01 I 
 think) that appeared to Switch the internal reference out of circuit. 
 Agilent would not share information on the option. My question to Agilent 
 is why sell an option and be unwilling to say what it does or how your 
 stock unit functions?
 Thomas Knox
 
 
 
 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,  
   but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of 
 the 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is 
 dirtier than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL 
 actually makes it worse. This was very disappointing. The XO version of 
 the counter performed worse than the OCXO version even with a maser as 
 the ext reference. Did your reading of the schematic show a way to 
 directly use the ext ref, bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted 
 copy of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those 
 digital ones that can lock up to many inputs. You could feed 3. 
 MHz in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The 
 internal oscillator (or an internal oscillator) is phase locked to the 
 external input through a fairly narrow analog loop. The idea is to 
 clean up the crud on the standard line. 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-19 Thread mike cook
That was a wink, Said, not a howl... 


Le 19 févr. 2014 à 17:25, Said Jackson a écrit :

 Mike,
 
 They are already giving you another way to calibrate the unit, different from 
 how you think they should have done it and you are pulling out the statist 
 card and accusing them of being greedy capitalists?
 
 Come on, thats backseat driving. Be happy they invested millions of their own 
 money and put out a more or less affordable new counter in a market flooded 
 with good low-cost used counters.
 
 Bye,
 Said
 
 Sent From iPhone
 
 On Feb 19, 2014, at 0:33, mike cook michael.c...@sfr.fr wrote:
 
 
 Le 19 févr. 2014 à 01:05, Tom Knox a écrit :
 
 Thanks Tom and Bob, I have been thinking of contacting Agilent for some 
 time. I think they are a great company with some good products, but there 
 are a few real blind spots in some current products. I also have seen in 
 the past a genuine interest in listening. I would be willing to approach 
 them if I could enlist your help in addressing potential changes to improve 
 the product. 
 Thanks;
 Thomas Knox
 
  If they are steering the VCXXO,OCXO from the Ext. Ref. , then they are in 
 effect calibrating it. Why not remember the applied EFC when they get phase 
 lock?  That can be applied when the internal timebase is selected. 
 It couldn't be that they might lose the chance to sell a signal generator 
 ;-), as calibration needs a square wave input, and the Ext. Ref In is 
 ignored.
 
 
 From: li...@rtty.us
 Date: Tue, 18 Feb 2014 18:00:17 -0500
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,
 but nobody in?
 
 Hi
 
 Well at least this got me digging a little. 
 
 If you grab a copy of the 53230A spec sheet and look under the external 
 reference input, it’s pretty well described. It will accept 1, 5,10 MHz as 
 an external reference. It will lock over a 1 ppm range with the XO option 
 and 0.1 ppm with the OCXO option. Based on that I’d guess they are still 
 using the same basic PLL approach as on the older counters (5335 era). 
 
 The “Microsoft Windows inside” sticker on the back of the counter was a 
 bit of a surprise ….
 
 No sticker on mine. 
 
 
 Bob
 
 On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com 
 wrote:
 
 TomK,
 
 If anyone has technical contacts deep within Agilent, let's see if this 
 issue can be resolved. I would have bought a 53230A when it came out a 
 few years ago but my eval units showed this clock noise problem. That 
 plus the poor quality of the ref out made me think the designers were 
 cutting corners, or had little experience in metrology, or maybe they 
 thought this was ok for a bench instrument.
 
 Otherwise it's a really nice counter; the first one from Agilent than can 
 actually do ADEV properly (since it is a time stamping counter).
 
 I should dig out my old data and send it to you. Maybe as group we can 
 help them fix the problem.
 
 /tvb
 
 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal 
 oscillator out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal 
 applied. These questions were related to several 53132A's I have seen 
 configured with a small board back near the Ext Ref input (OPT H01 I 
 think) that appeared to Switch the internal reference out of circuit. 
 Agilent would not share information on the option. My question to 
 Agilent is why sell an option and be unwilling to say what it does or 
 how your stock unit functions?
 Thomas Knox
 
 
 
 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on, 
but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of 
 the 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is 
 dirtier than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL 
 actually makes it worse. This was very disappointing. The XO version of 
 the counter performed worse than the OCXO version even with a maser as 
 the ext reference. Did your reading of the schematic show a way to 
 directly use the ext ref, bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted 
 copy of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those 
 digital ones that can lock up to many inputs. You could feed 
 3. MHz in as a standard input as well as 0.5, 1, 2.5, 5, and 
 10 MHz. The internal oscillator (or an internal oscillator) is phase 
 locked to the 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-19 Thread Tom Knox

I hope I have not come off sounding like that Said, I simply would like to see 
a great product better,  I am hoping/committed to work with Agilent toward a 
better product if they are interested. And in the past I have found they are 
interested in our feedback. The 53132A was revolutionary in it's day, but with 
advances in time and freq there is now a market for a 14 or 15 digit counter. 
I am still attempting to individually characterize each item in my time and 
freq system and understand their strengths and weaknesses. And hope to learn 
more about the 53230A in the coming weeks. But TVB's comments in particular 
seemed consistent with my impressions so far.
I would welcome your thoughts on the 53230A.
Thanks;l
Thomas Knox



 CC: time-nuts@febo.com
 From: saidj...@aol.com
 Date: Wed, 19 Feb 2014 08:25:28 -0800
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
 but nobody in?
 
 Mike,
 
 They are already giving you another way to calibrate the unit, different from 
 how you think they should have done it and you are pulling out the statist 
 card and accusing them of being greedy capitalists?
 
 Come on, thats backseat driving. Be happy they invested millions of their own 
 money and put out a more or less affordable new counter in a market flooded 
 with good low-cost used counters.
 
 Bye,
 Said
 
 Sent From iPhone
 
 On Feb 19, 2014, at 0:33, mike cook michael.c...@sfr.fr wrote:
 
  
  Le 19 févr. 2014 à 01:05, Tom Knox a écrit :
  
  Thanks Tom and Bob, I have been thinking of contacting Agilent for some 
  time. I think they are a great company with some good products, but there 
  are a few real blind spots in some current products. I also have seen in 
  the past a genuine interest in listening. I would be willing to approach 
  them if I could enlist your help in addressing potential changes to 
  improve the product. 
  Thanks;
  Thomas Knox
  
If they are steering the VCXXO,OCXO from the Ext. Ref. , then they are in 
  effect calibrating it. Why not remember the applied EFC when they get phase 
  lock?  That can be applied when the internal timebase is selected. 
  It couldn't be that they might lose the chance to sell a signal generator 
  ;-), as calibration needs a square wave input, and the Ext. Ref In is 
  ignored.
  
  
  From: li...@rtty.us
  Date: Tue, 18 Feb 2014 18:00:17 -0500
  To: time-nuts@febo.com
  Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
   but nobody in?
  
  Hi
  
  Well at least this got me digging a little. 
  
  If you grab a copy of the 53230A spec sheet and look under the external 
  reference input, it’s pretty well described. It will accept 1, 5,10 MHz 
  as an external reference. It will lock over a 1 ppm range with the XO 
  option and 0.1 ppm with the OCXO option. Based on that I’d guess they are 
  still using the same basic PLL approach as on the older counters (5335 
  era). 
  
  The “Microsoft Windows inside” sticker on the back of the counter was a 
  bit of a surprise ….
  
   No sticker on mine. 
  
  
  Bob
  
  On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com 
  wrote:
  
  TomK,
  
  If anyone has technical contacts deep within Agilent, let's see if this 
  issue can be resolved. I would have bought a 53230A when it came out a 
  few years ago but my eval units showed this clock noise problem. That 
  plus the poor quality of the ref out made me think the designers were 
  cutting corners, or had little experience in metrology, or maybe they 
  thought this was ok for a bench instrument.
  
  Otherwise it's a really nice counter; the first one from Agilent than 
  can actually do ADEV properly (since it is a time stamping counter).
  
  I should dig out my old data and send it to you. Maybe as group we can 
  help them fix the problem.
  
  /tvb
  
  On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
  
  I have asked Agilent 
  if stock versions of the 53230A and 53132A switched the internal 
  oscillator out of circuit with an Ext Ref signal 
  applied. I thought 
  Agilent's engineer was intentionally vague but said the oscillators were
  indeed switched out of circuit on the counter with Ext Ref signal 
  applied. These questions were related to several 53132A's I have seen 
  configured with a small board back near the Ext Ref input (OPT H01 I 
  think) that appeared to Switch the internal reference out of circuit. 
  Agilent would not share information on the option. My question to 
  Agilent is why sell an option and be unwilling to say what it does or 
  how your stock unit functions?
  Thomas Knox
  
  
  
  From: t...@leapsecond.com
  Date: Mon, 17 Feb 2014 09:38:28 -1000
  To: time-nuts@febo.com
  Subject: Re: [time-nuts] strange behavior of 53230A or is the light 
  on,but nobody in?
  
  Bob,
  
  I'm wondering if you (or any else) has measured the PLL performance of 
  the 53230-series?
  
  I agree it will 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-19 Thread Said Jackson
Sorry early morning rant,

There are counters out there already that can do 14/15 digits: tsc5125A and the 
Miles box for example. Very difficult to get a reference into that counter that 
can match and provide that type of stability.

I am sure Agilent would love to hear our feedback probably as long as we don't 
accuse them of leaving out features purely as a profit motive.

Heck Apple sells $160 production cost iPads for $800 and doesn't even include a 
calculator app for free. People don't care and they end up with $100+ billion 
cash in the bank. I'd rather have Agilent charge a bit more and have them still 
around 10 years from now.

Bye,
Said 

Sent From iPhone

On Feb 19, 2014, at 10:08, Tom Knox act...@hotmail.com wrote:

 
 I hope I have not come off sounding like that Said, I simply would like to 
 see a great product better,  I am hoping/committed to work with Agilent 
 toward a better product if they are interested. And in the past I have found 
 they are interested in our feedback. The 53132A was revolutionary in it's 
 day, but with advances in time and freq there is now a market for a 14 or 15 
 digit counter. 
 I am still attempting to individually characterize each item in my time and 
 freq system and understand their strengths and weaknesses. And hope to learn 
 more about the 53230A in the coming weeks. But TVB's comments in particular 
 seemed consistent with my impressions so far.
 I would welcome your thoughts on the 53230A.
 Thanks;l
 Thomas Knox
 
 
 
 CC: time-nuts@febo.com
 From: saidj...@aol.com
 Date: Wed, 19 Feb 2014 08:25:28 -0800
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,
 but nobody in?
 
 Mike,
 
 They are already giving you another way to calibrate the unit, different 
 from how you think they should have done it and you are pulling out the 
 statist card and accusing them of being greedy capitalists?
 
 Come on, thats backseat driving. Be happy they invested millions of their 
 own money and put out a more or less affordable new counter in a market 
 flooded with good low-cost used counters.
 
 Bye,
 Said
 
 Sent From iPhone
 
 On Feb 19, 2014, at 0:33, mike cook michael.c...@sfr.fr wrote:
 
 
 Le 19 févr. 2014 à 01:05, Tom Knox a écrit :
 
 Thanks Tom and Bob, I have been thinking of contacting Agilent for some 
 time. I think they are a great company with some good products, but there 
 are a few real blind spots in some current products. I also have seen in 
 the past a genuine interest in listening. I would be willing to approach 
 them if I could enlist your help in addressing potential changes to 
 improve the product. 
 Thanks;
 Thomas Knox
 
  If they are steering the VCXXO,OCXO from the Ext. Ref. , then they are in 
 effect calibrating it. Why not remember the applied EFC when they get phase 
 lock?  That can be applied when the internal timebase is selected. 
 It couldn't be that they might lose the chance to sell a signal generator 
 ;-), as calibration needs a square wave input, and the Ext. Ref In is 
 ignored.
 
 
 From: li...@rtty.us
 Date: Tue, 18 Feb 2014 18:00:17 -0500
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
  but nobody in?
 
 Hi
 
 Well at least this got me digging a little. 
 
 If you grab a copy of the 53230A spec sheet and look under the external 
 reference input, it’s pretty well described. It will accept 1, 5,10 MHz 
 as an external reference. It will lock over a 1 ppm range with the XO 
 option and 0.1 ppm with the OCXO option. Based on that I’d guess they are 
 still using the same basic PLL approach as on the older counters (5335 
 era). 
 
 The “Microsoft Windows inside” sticker on the back of the counter was a 
 bit of a surprise ….
 
 No sticker on mine. 
 
 
 Bob
 
 On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com 
 wrote:
 
 TomK,
 
 If anyone has technical contacts deep within Agilent, let's see if this 
 issue can be resolved. I would have bought a 53230A when it came out a 
 few years ago but my eval units showed this clock noise problem. That 
 plus the poor quality of the ref out made me think the designers were 
 cutting corners, or had little experience in metrology, or maybe they 
 thought this was ok for a bench instrument.
 
 Otherwise it's a really nice counter; the first one from Agilent than 
 can actually do ADEV properly (since it is a time stamping counter).
 
 I should dig out my old data and send it to you. Maybe as group we can 
 help them fix the problem.
 
 /tvb
 
 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal 
 oscillator out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal 
 applied. These questions were related to several 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-19 Thread Magnus Danielson

On 2014-02-19 22:34, Said Jackson wrote:

Sorry early morning rant,

There are counters out there already that can do 14/15 digits: tsc5125A and the 
Miles box for example. Very difficult to get a reference into that counter that 
can match and provide that type of stability.

I am sure Agilent would love to hear our feedback probably as long as we don't 
accuse them of leaving out features purely as a profit motive.

Heck Apple sells $160 production cost iPads for $800 and doesn't even include a 
calculator app for free. People don't care and they end up with $100+ billion 
cash in the bank. I'd rather have Agilent charge a bit more and have them still 
around 10 years from now.


They won't be, not under that name at least. Unfortunatly they ended up 
with another name which will be hard to remember.


Cheers,
Magnus

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Tom Knox
I have asked Agilent 
if stock versions of the 53230A and 53132A switched the internal oscillator out 
of circuit with an Ext Ref signal 
applied. I thought 
Agilent's engineer was intentionally vague but said the oscillators were
indeed switched out of circuit on the counter with Ext Ref signal applied. 
These questions were related to several 53132A's I have seen configured with a 
small board back near the Ext Ref input (OPT H01 I think) that appeared to 
Switch the internal reference out of circuit. Agilent would not share 
information on the option. My question to Agilent is why sell an option and be 
unwilling to say what it does or how your stock unit functions?
Thomas Knox



 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
 but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of the 
 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is dirtier 
 than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL actually 
 makes it worse. This was very disappointing. The XO version of the counter 
 performed worse than the OCXO version even with a maser as the ext reference. 
 Did your reading of the schematic show a way to directly use the ext ref, 
 bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted copy 
 of the ref in.
 
 /tvb (i5s)
 
  On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
  
  Hi
  
  If you dig into the schematics (when they supplied them … ):
  
  The external reference goes into a phase detector. It’s one of those 
  digital ones that can lock up to many inputs. You could feed 3. MHz 
  in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The internal 
  oscillator (or an internal oscillator) is phase locked to the external 
  input through a fairly narrow analog loop. The idea is to clean up the crud 
  on the standard line. 
  
  With no external reference, the PLL drops out and you go back to what ever 
  the local reference is. 
  
  Yes there’s a little more to it than that and no the circuit is not exactly 
  the same on every counter HP ever made. 
  
  Bob
  
  On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
  
  Hi Mike,
  
  The most likely answer is when you select external time base for an input, 
  it disables the connection for the internal oscillator.  The external 
  input signal is probably also routed straight to the reference output jack.
  
  However, it would be good to read the manual, as they usually cover how 
  those connections work.  Otherwise, perhaps someone that owns one could 
  provide further insight.
  
  BillWB6BNQ
  
  mike cook wrote:
  
  Something that must be simple to explain, but that I can't get my head 
  round.
  
  I got a new 53230A.
  When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
  timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
  about it as the counter only has a TCXO internal oscillator. So I fired 
  up my PRS10 and after leaving that on for some time, connected it to  Ext 
  Ref. , changed to the ext time base and measured again. This time 
  10.000.000.00x. Then I switched the two references, measuring the PRS10 
  against the T-Bolt. Again I got 10MHz down to the 11th digit.
  All that looked good so I have been using it with either the PRS10 locked 
  to GPS, or the T-Bolt as the external time base.
  
  After leaving it on (but not inactive) for a month, I did an Autocal. No 
  problem.
  I was wondering if that would have changed the internal time base 
  frequency, but no, using that still gave similar figures to the above.
  
  So at that point I decided to measure the Internal TB against my 
  reference. So I connected the Int. Ref. Out to channel 1, connected my 
  PRS10 ref to Ext. Ref In, selected the EXT time base and found that the 
  count was 10MHz dead on?  I don't get that at all.
  
  in summary:
  DUT against internal TB counts  10MHz.To me that means that the 
  internal timebase is a bit fast. Is that assumption correct?
  DUT against Ext.Ref counts 10MHz
  Internal TB against Ext.Ref counts 10MHz.   If my assumption above is 
  correct, the count should be greater than 10MHz, no?
  
  Can anyone shed any light on that?
  
  
  
  
  
  
  
  
  
  
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
  
  ___
  time-nuts mailing list -- time-nuts@febo.com
  To unsubscribe, go to 
  https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
  and follow the instructions there.
  
  ___
  

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Tom Van Baak (lab)
TomK,

If anyone has technical contacts deep within Agilent, let's see if this issue 
can be resolved. I would have bought a 53230A when it came out a few years ago 
but my eval units showed this clock noise problem. That plus the poor quality 
of the ref out made me think the designers were cutting corners, or had little 
experience in metrology, or maybe they thought this was ok for a bench 
instrument.

Otherwise it's a really nice counter; the first one from Agilent than can 
actually do ADEV properly (since it is a time stamping counter).

I should dig out my old data and send it to you. Maybe as group we can help 
them fix the problem.

/tvb

 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal oscillator 
 out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal applied. 
 These questions were related to several 53132A's I have seen configured with 
 a small board back near the Ext Ref input (OPT H01 I think) that appeared to 
 Switch the internal reference out of circuit. Agilent would not share 
 information on the option. My question to Agilent is why sell an option and 
 be unwilling to say what it does or how your stock unit functions?
 Thomas Knox
 
 
 
 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,
 but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of the 
 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is dirtier 
 than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL 
 actually makes it worse. This was very disappointing. The XO version of the 
 counter performed worse than the OCXO version even with a maser as the ext 
 reference. Did your reading of the schematic show a way to directly use the 
 ext ref, bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted copy 
 of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those 
 digital ones that can lock up to many inputs. You could feed 3. MHz 
 in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The internal 
 oscillator (or an internal oscillator) is phase locked to the external 
 input through a fairly narrow analog loop. The idea is to clean up the crud 
 on the standard line. 
 
 With no external reference, the PLL drops out and you go back to what ever 
 the local reference is. 
 
 Yes there’s a little more to it than that and no the circuit is not exactly 
 the same on every counter HP ever made. 
 
 Bob
 
 On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
 
 Hi Mike,
 
 The most likely answer is when you select external time base for an input, 
 it disables the connection for the internal oscillator.  The external 
 input signal is probably also routed straight to the reference output jack.
 
 However, it would be good to read the manual, as they usually cover how 
 those connections work.  Otherwise, perhaps someone that owns one could 
 provide further insight.
 
 BillWB6BNQ
 
 mike cook wrote:
 
 Something that must be simple to explain, but that I can't get my head 
 round.
 
 I got a new 53230A.
 When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
 timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
 about it as the counter only has a TCXO internal oscillator. So I fired 
 up my PRS10 and after leaving that on for some time, connected it to  Ext 
 Ref. , changed to the ext time base and measured again. This time 
 10.000.000.00x. Then I switched the two references, measuring the PRS10 
 against the T-Bolt. Again I got 10MHz down to the 11th digit.
 All that looked good so I have been using it with either the PRS10 locked 
 to GPS, or the T-Bolt as the external time base.
 
 After leaving it on (but not inactive) for a month, I did an Autocal. No 
 problem.
 I was wondering if that would have changed the internal time base 
 frequency, but no, using that still gave similar figures to the above.
 
 So at that point I decided to measure the Internal TB against my 
 reference. So I connected the Int. Ref. Out to channel 1, connected my 
 PRS10 ref to Ext. Ref In, selected the EXT time base and found that the 
 count was 10MHz dead on?  I don't get that at all.
 
 in summary:
 DUT against internal TB counts  10MHz.To me that means that the 
 internal timebase is a bit fast. Is that assumption correct?
 DUT against Ext.Ref counts 10MHz
 Internal TB against 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Bob Camp
Hi

Well at least this got me digging a little. 

If you grab a copy of the 53230A spec sheet and look under the external 
reference input, it’s pretty well described. It will accept 1, 5,10 MHz as an 
external reference. It will lock over a 1 ppm range with the XO option and 0.1 
ppm with the OCXO option. Based on that I’d guess they are still using the same 
basic PLL approach as on the older counters (5335 era). 

The “Microsoft Windows inside” sticker on the back of the counter was a bit of 
a surprise ….

Bob

On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com wrote:

 TomK,
 
 If anyone has technical contacts deep within Agilent, let's see if this issue 
 can be resolved. I would have bought a 53230A when it came out a few years 
 ago but my eval units showed this clock noise problem. That plus the poor 
 quality of the ref out made me think the designers were cutting corners, or 
 had little experience in metrology, or maybe they thought this was ok for a 
 bench instrument.
 
 Otherwise it's a really nice counter; the first one from Agilent than can 
 actually do ADEV properly (since it is a time stamping counter).
 
 I should dig out my old data and send it to you. Maybe as group we can help 
 them fix the problem.
 
 /tvb
 
 On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
 
 I have asked Agilent 
 if stock versions of the 53230A and 53132A switched the internal oscillator 
 out of circuit with an Ext Ref signal 
 applied. I thought 
 Agilent's engineer was intentionally vague but said the oscillators were
 indeed switched out of circuit on the counter with Ext Ref signal applied. 
 These questions were related to several 53132A's I have seen configured with 
 a small board back near the Ext Ref input (OPT H01 I think) that appeared to 
 Switch the internal reference out of circuit. Agilent would not share 
 information on the option. My question to Agilent is why sell an option and 
 be unwilling to say what it does or how your stock unit functions?
 Thomas Knox
 
 
 
 From: t...@leapsecond.com
 Date: Mon, 17 Feb 2014 09:38:28 -1000
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,
 but nobody in?
 
 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of the 
 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is dirtier 
 than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL 
 actually makes it worse. This was very disappointing. The XO version of the 
 counter performed worse than the OCXO version even with a maser as the ext 
 reference. Did your reading of the schematic show a way to directly use the 
 ext ref, bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted 
 copy of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those 
 digital ones that can lock up to many inputs. You could feed 3. 
 MHz in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The 
 internal oscillator (or an internal oscillator) is phase locked to the 
 external input through a fairly narrow analog loop. The idea is to clean 
 up the crud on the standard line. 
 
 With no external reference, the PLL drops out and you go back to what ever 
 the local reference is. 
 
 Yes there’s a little more to it than that and no the circuit is not 
 exactly the same on every counter HP ever made. 
 
 Bob
 
 On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
 
 Hi Mike,
 
 The most likely answer is when you select external time base for an 
 input, it disables the connection for the internal oscillator.  The 
 external input signal is probably also routed straight to the reference 
 output jack.
 
 However, it would be good to read the manual, as they usually cover how 
 those connections work.  Otherwise, perhaps someone that owns one could 
 provide further insight.
 
 BillWB6BNQ
 
 mike cook wrote:
 
 Something that must be simple to explain, but that I can't get my head 
 round.
 
 I got a new 53230A.
 When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
 timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
 about it as the counter only has a TCXO internal oscillator. So I fired 
 up my PRS10 and after leaving that on for some time, connected it to  
 Ext Ref. , changed to the ext time base and measured again. This time 
 10.000.000.00x. Then I switched the two references, measuring the PRS10 
 against the T-Bolt. Again I got 10MHz down to the 11th digit.
 All that looked good so I have been using it with either the PRS10 
 locked to GPS, or the T-Bolt as the external time base.
 
 After leaving it on (but not inactive) for a month, I did an Autocal. No 
 problem.
 I 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-18 Thread Tom Knox
Thanks Tom and Bob, I have been thinking of contacting Agilent for some time. I 
think they are a great company with some good products, but there are a few 
real blind spots in some current products. I also have seen in the past a 
genuine interest in listening. I would be willing to approach them if I could 
enlist your help in addressing potential changes to improve the product. 
Thanks;
Thomas Knox



 From: li...@rtty.us
 Date: Tue, 18 Feb 2014 18:00:17 -0500
 To: time-nuts@febo.com
 Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
 but nobody in?
 
 Hi
 
 Well at least this got me digging a little. 
 
 If you grab a copy of the 53230A spec sheet and look under the external 
 reference input, it’s pretty well described. It will accept 1, 5,10 MHz as an 
 external reference. It will lock over a 1 ppm range with the XO option and 
 0.1 ppm with the OCXO option. Based on that I’d guess they are still using 
 the same basic PLL approach as on the older counters (5335 era). 
 
 The “Microsoft Windows inside” sticker on the back of the counter was a bit 
 of a surprise ….
 
 Bob
 
 On Feb 18, 2014, at 11:51 AM, Tom Van Baak (lab) t...@leapsecond.com wrote:
 
  TomK,
  
  If anyone has technical contacts deep within Agilent, let's see if this 
  issue can be resolved. I would have bought a 53230A when it came out a few 
  years ago but my eval units showed this clock noise problem. That plus the 
  poor quality of the ref out made me think the designers were cutting 
  corners, or had little experience in metrology, or maybe they thought this 
  was ok for a bench instrument.
  
  Otherwise it's a really nice counter; the first one from Agilent than can 
  actually do ADEV properly (since it is a time stamping counter).
  
  I should dig out my old data and send it to you. Maybe as group we can help 
  them fix the problem.
  
  /tvb
  
  On Feb 18, 2014, at 12:10 AM, Tom Knox act...@hotmail.com wrote:
  
  I have asked Agilent 
  if stock versions of the 53230A and 53132A switched the internal 
  oscillator out of circuit with an Ext Ref signal 
  applied. I thought 
  Agilent's engineer was intentionally vague but said the oscillators were
  indeed switched out of circuit on the counter with Ext Ref signal applied. 
  These questions were related to several 53132A's I have seen configured 
  with a small board back near the Ext Ref input (OPT H01 I think) that 
  appeared to Switch the internal reference out of circuit. Agilent would 
  not share information on the option. My question to Agilent is why sell an 
  option and be unwilling to say what it does or how your stock unit 
  functions?
  Thomas Knox
  
  
  
  From: t...@leapsecond.com
  Date: Mon, 17 Feb 2014 09:38:28 -1000
  To: time-nuts@febo.com
  Subject: Re: [time-nuts] strange behavior of 53230A or is the light on,   
   but nobody in?
  
  Bob,
  
  I'm wondering if you (or any else) has measured the PLL performance of 
  the 53230-series?
  
  I agree it will clean up the crud but this assumes the ext ref is 
  dirtier than the internal osc.
  
  What I found instead was that if you use a good external ref the PLL 
  actually makes it worse. This was very disappointing. The XO version of 
  the counter performed worse than the OCXO version even with a maser as 
  the ext reference. Did your reading of the schematic show a way to 
  directly use the ext ref, bypassing the noisy PLL?
  
  The other thing I found was that the ref out signal was a very polluted 
  copy of the ref in.
  
  /tvb (i5s)
  
  On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
  
  Hi
  
  If you dig into the schematics (when they supplied them … ):
  
  The external reference goes into a phase detector. It’s one of those 
  digital ones that can lock up to many inputs. You could feed 3. 
  MHz in as a standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The 
  internal oscillator (or an internal oscillator) is phase locked to the 
  external input through a fairly narrow analog loop. The idea is to clean 
  up the crud on the standard line. 
  
  With no external reference, the PLL drops out and you go back to what 
  ever the local reference is. 
  
  Yes there’s a little more to it than that and no the circuit is not 
  exactly the same on every counter HP ever made. 
  
  Bob
  
  On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
  
  Hi Mike,
  
  The most likely answer is when you select external time base for an 
  input, it disables the connection for the internal oscillator.  The 
  external input signal is probably also routed straight to the reference 
  output jack.
  
  However, it would be good to read the manual, as they usually cover how 
  those connections work.  Otherwise, perhaps someone that owns one could 
  provide further insight.
  
  BillWB6BNQ
  
  mike cook wrote:
  
  Something that must be simple to explain, but that I can't get my head 
  round.
  
  I got a new 53230A.
  When 

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread David C. Partridge
Did you disconnect the external reference from the 53230A before doing the
test?

If no, I'll bet that the external reference being connected overrides the
internal reference 

Regards,
David Partridge 

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread mike cook

Le 17 févr. 2014 à 13:49, gandal...@aol.com a écrit :

 Hi Michael
  
 Internal reference out is likely to be of the actual reference in use, ie 
 with an external reference connected that's what will be on the output 
 connector and would explain what you're seeing.
  
 Your initial test, using the T'bolt and internal reference should be giving 
 you the accurate measure of the internal reference.
  
  Thanks guys. That would explain it, but there is nothing I can see in the doc 
which confirms it. I will see if I have a 5MHz or 1MHz oscillator that I could 
use to check that. 

 Regards
  
 Nigel
 GM8PZR

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread Bob Camp
Hi

It locks up the internal reference to the external reference when the external 
reference is present. It’s the same behavior as the 
5334,5335,5345,5360,5370,5318x,and 5313x. 

Bob

On Feb 17, 2014, at 8:31 AM, mike cook michael.c...@sfr.fr wrote:

 
 Le 17 févr. 2014 à 13:49, gandal...@aol.com a écrit :
 
 Hi Michael
 
 Internal reference out is likely to be of the actual reference in use, ie 
 with an external reference connected that's what will be on the output 
 connector and would explain what you're seeing.
 
 Your initial test, using the T'bolt and internal reference should be giving 
 you the accurate measure of the internal reference.
 
  Thanks guys. That would explain it, but there is nothing I can see in the 
 doc which confirms it. I will see if I have a 5MHz or 1MHz oscillator that I 
 could use to check that. 
 
 Regards
 
 Nigel
 GM8PZR
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread Poul-Henning Kamp
In message ff58a3fc-3e6a-4b13-83e3-c1363a556...@sfr.fr, mike cook writes:

So at that point I decided to measure the Internal TB against my
reference. So I connected the Int. Ref. Out to channel 1, connected
my PRS10 ref to Ext. Ref In, selected the EXT time base and found
that the count was 10MHz dead on?  I don't get that at all.

On all the HP kit I have, ref out is the frequency used by the instrument,
so if you feed it an external reference, it is just a copy of that external
reference.


-- 
Poul-Henning Kamp   | UNIX since Zilog Zeus 3.20
p...@freebsd.org | TCP/IP since RFC 956
FreeBSD committer   | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread wb6bnq

Hi Mike,

The most likely answer is when you select external time base for an 
input, it disables the connection for the internal oscillator.  The 
external input signal is probably also routed straight to the reference 
output jack.


However, it would be good to read the manual, as they usually cover how 
those connections work.  Otherwise, perhaps someone that owns one could 
provide further insight.


BillWB6BNQ

mike cook wrote:


Something that must be simple to explain, but that I can't get my head round.

I got a new 53230A.
When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried about 
it as the counter only has a TCXO internal oscillator. So I fired up my PRS10 
and after leaving that on for some time, connected it to  Ext Ref. , changed to 
the ext time base and measured again. This time 10.000.000.00x. Then I switched 
the two references, measuring the PRS10 against the T-Bolt. Again I got 10MHz 
down to the 11th digit.
All that looked good so I have been using it with either the PRS10 locked to 
GPS, or the T-Bolt as the external time base.

After leaving it on (but not inactive) for a month, I did an Autocal. No 
problem.
I was wondering if that would have changed the internal time base frequency, 
but no, using that still gave similar figures to the above.

So at that point I decided to measure the Internal TB against my reference. So 
I connected the Int. Ref. Out to channel 1, connected my PRS10 ref to Ext. Ref 
In, selected the EXT time base and found that the count was 10MHz dead on?  
I don't get that at all.

in summary:
DUT against internal TB counts  10MHz.To me that means that the internal 
timebase is a bit fast. Is that assumption correct?
DUT against Ext.Ref counts 10MHz
Internal TB against Ext.Ref counts 10MHz.   If my assumption above is 
correct, the count should be greater than 10MHz, no?

Can anyone shed any light on that?










___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

 



___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread Bob Camp
Hi

If you dig into the schematics (when they supplied them … ):

The external reference goes into a phase detector. It’s one of those digital 
ones that can lock up to many inputs. You could feed 3. MHz in as a 
standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The internal oscillator 
(or an internal oscillator) is phase locked to the external input through a 
fairly narrow analog loop. The idea is to clean up the crud on the standard 
line. 

With no external reference, the PLL drops out and you go back to what ever the 
local reference is. 

Yes there’s a little more to it than that and no the circuit is not exactly the 
same on every counter HP ever made. 

Bob

On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:

 Hi Mike,
 
 The most likely answer is when you select external time base for an input, it 
 disables the connection for the internal oscillator.  The external input 
 signal is probably also routed straight to the reference output jack.
 
 However, it would be good to read the manual, as they usually cover how those 
 connections work.  Otherwise, perhaps someone that owns one could provide 
 further insight.
 
 BillWB6BNQ
 
 mike cook wrote:
 
 Something that must be simple to explain, but that I can't get my head round.
 
 I got a new 53230A.
 When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
 timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
 about it as the counter only has a TCXO internal oscillator. So I fired up 
 my PRS10 and after leaving that on for some time, connected it to  Ext Ref. 
 , changed to the ext time base and measured again. This time 10.000.000.00x. 
 Then I switched the two references, measuring the PRS10 against the T-Bolt. 
 Again I got 10MHz down to the 11th digit.
 All that looked good so I have been using it with either the PRS10 locked to 
 GPS, or the T-Bolt as the external time base.
 
 After leaving it on (but not inactive) for a month, I did an Autocal. No 
 problem.
 I was wondering if that would have changed the internal time base frequency, 
 but no, using that still gave similar figures to the above.
 
 So at that point I decided to measure the Internal TB against my reference. 
 So I connected the Int. Ref. Out to channel 1, connected my PRS10 ref to 
 Ext. Ref In, selected the EXT time base and found that the count was 10MHz 
 dead on?  I don't get that at all.
 
 in summary:
 DUT against internal TB counts  10MHz.To me that means that the 
 internal timebase is a bit fast. Is that assumption correct?
 DUT against Ext.Ref counts 10MHz
 Internal TB against Ext.Ref counts 10MHz.   If my assumption above is 
 correct, the count should be greater than 10MHz, no?
 
 Can anyone shed any light on that?
 
 
 
 
 
 
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.


Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread Tom Van Baak (lab)
Bob,

I'm wondering if you (or any else) has measured the PLL performance of the 
53230-series?

I agree it will clean up the crud but this assumes the ext ref is dirtier 
than the internal osc.

What I found instead was that if you use a good external ref the PLL actually 
makes it worse. This was very disappointing. The XO version of the counter 
performed worse than the OCXO version even with a maser as the ext reference. 
Did your reading of the schematic show a way to directly use the ext ref, 
bypassing the noisy PLL?

The other thing I found was that the ref out signal was a very polluted copy of 
the ref in.

/tvb (i5s)

 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those digital 
 ones that can lock up to many inputs. You could feed 3. MHz in as a 
 standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The internal oscillator 
 (or an internal oscillator) is phase locked to the external input through a 
 fairly narrow analog loop. The idea is to clean up the crud on the standard 
 line. 
 
 With no external reference, the PLL drops out and you go back to what ever 
 the local reference is. 
 
 Yes there’s a little more to it than that and no the circuit is not exactly 
 the same on every counter HP ever made. 
 
 Bob
 
 On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
 
 Hi Mike,
 
 The most likely answer is when you select external time base for an input, 
 it disables the connection for the internal oscillator.  The external input 
 signal is probably also routed straight to the reference output jack.
 
 However, it would be good to read the manual, as they usually cover how 
 those connections work.  Otherwise, perhaps someone that owns one could 
 provide further insight.
 
 BillWB6BNQ
 
 mike cook wrote:
 
 Something that must be simple to explain, but that I can't get my head 
 round.
 
 I got a new 53230A.
 When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
 timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
 about it as the counter only has a TCXO internal oscillator. So I fired up 
 my PRS10 and after leaving that on for some time, connected it to  Ext Ref. 
 , changed to the ext time base and measured again. This time 
 10.000.000.00x. Then I switched the two references, measuring the PRS10 
 against the T-Bolt. Again I got 10MHz down to the 11th digit.
 All that looked good so I have been using it with either the PRS10 locked 
 to GPS, or the T-Bolt as the external time base.
 
 After leaving it on (but not inactive) for a month, I did an Autocal. No 
 problem.
 I was wondering if that would have changed the internal time base 
 frequency, but no, using that still gave similar figures to the above.
 
 So at that point I decided to measure the Internal TB against my reference. 
 So I connected the Int. Ref. Out to channel 1, connected my PRS10 ref to 
 Ext. Ref In, selected the EXT time base and found that the count was 10MHz 
 dead on?  I don't get that at all.
 
 in summary:
 DUT against internal TB counts  10MHz.To me that means that the 
 internal timebase is a bit fast. Is that assumption correct?
 DUT against Ext.Ref counts 10MHz
 Internal TB against Ext.Ref counts 10MHz.   If my assumption above is 
 correct, the count should be greater than 10MHz, no?
 
 Can anyone shed any light on that?
 
 
 
 
 
 
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.

Re: [time-nuts] strange behavior of 53230A or is the light on, but nobody in?

2014-02-17 Thread Bob Camp
Hi

I have a 53230, but have not gotten around to looking at it’s PLL cleanup 
process.

Bob

On Feb 17, 2014, at 2:38 PM, Tom Van Baak (lab) t...@leapsecond.com wrote:

 Bob,
 
 I'm wondering if you (or any else) has measured the PLL performance of the 
 53230-series?
 
 I agree it will clean up the crud but this assumes the ext ref is dirtier 
 than the internal osc.
 
 What I found instead was that if you use a good external ref the PLL actually 
 makes it worse. This was very disappointing. The XO version of the counter 
 performed worse than the OCXO version even with a maser as the ext reference. 
 Did your reading of the schematic show a way to directly use the ext ref, 
 bypassing the noisy PLL?
 
 The other thing I found was that the ref out signal was a very polluted copy 
 of the ref in.
 
 /tvb (i5s)
 
 On Feb 17, 2014, at 7:04 AM, Bob Camp li...@rtty.us wrote:
 
 Hi
 
 If you dig into the schematics (when they supplied them … ):
 
 The external reference goes into a phase detector. It’s one of those digital 
 ones that can lock up to many inputs. You could feed 3. MHz in as a 
 standard input as well as 0.5, 1, 2.5, 5, and 10 MHz. The internal 
 oscillator (or an internal oscillator) is phase locked to the external input 
 through a fairly narrow analog loop. The idea is to clean up the crud on the 
 standard line. 
 
 With no external reference, the PLL drops out and you go back to what ever 
 the local reference is. 
 
 Yes there’s a little more to it than that and no the circuit is not exactly 
 the same on every counter HP ever made. 
 
 Bob
 
 On Feb 17, 2014, at 7:55 AM, wb6bnq wb6...@cox.net wrote:
 
 Hi Mike,
 
 The most likely answer is when you select external time base for an input, 
 it disables the connection for the internal oscillator.  The external input 
 signal is probably also routed straight to the reference output jack.
 
 However, it would be good to read the manual, as they usually cover how 
 those connections work.  Otherwise, perhaps someone that owns one could 
 provide further insight.
 
 BillWB6BNQ
 
 mike cook wrote:
 
 Something that must be simple to explain, but that I can't get my head 
 round.
 
 I got a new 53230A.
 When first using it, I measured my T-Bolt 10MHz using the internal 10MHz 
 timebase and it came up short of 10MHz, 9.999 998 5xx. I wasn't worried 
 about it as the counter only has a TCXO internal oscillator. So I fired up 
 my PRS10 and after leaving that on for some time, connected it to  Ext 
 Ref. , changed to the ext time base and measured again. This time 
 10.000.000.00x. Then I switched the two references, measuring the PRS10 
 against the T-Bolt. Again I got 10MHz down to the 11th digit.
 All that looked good so I have been using it with either the PRS10 locked 
 to GPS, or the T-Bolt as the external time base.
 
 After leaving it on (but not inactive) for a month, I did an Autocal. No 
 problem.
 I was wondering if that would have changed the internal time base 
 frequency, but no, using that still gave similar figures to the above.
 
 So at that point I decided to measure the Internal TB against my 
 reference. So I connected the Int. Ref. Out to channel 1, connected my 
 PRS10 ref to Ext. Ref In, selected the EXT time base and found that the 
 count was 10MHz dead on?  I don't get that at all.
 
 in summary:
 DUT against internal TB counts  10MHz.To me that means that the 
 internal timebase is a bit fast. Is that assumption correct?
 DUT against Ext.Ref counts 10MHz
 Internal TB against Ext.Ref counts 10MHz.   If my assumption above is 
 correct, the count should be greater than 10MHz, no?
 
 Can anyone shed any light on that?
 
 
 
 
 
 
 
 
 
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to 
 https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.
 ___
 time-nuts mailing list -- time-nuts@febo.com
 To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
 and follow the instructions there.

___
time-nuts mailing list -- time-nuts@febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.