[volt-nuts] LTZ1000 project build
This whole discussion does not really apply to the LTZ1000 circuit, that is the bare 7.15..V output, w/o step up amplifier, as all drift parameters were attenuated by orders of magnitude, depending on the resistors, R1-R5. In the eevblog thread, several volt-nuts measured the real dependency, in contrast to the datsheet, which simply claims 1:100 attenuation for each of these resistors: R1 R2 R3 R4 R5 author 120 7 7 12-15k 1k -1/770 -1/250 -1/1400 +1/100 janaf -1/600 -1/238 -1/1000 +1/87 lars -1/656 TiN 1/714 1/250 1/ 1/105 bbs38hot -1/670 -1/238 -1/1184 +1/74 -1/75 Andreas Therefore, even humidity effects at most affect R4/R5, with 1/75 impact. I also think, that it's possible for volt-nuts to measure such residual effects, by assembling several references, keeping the parameter of question (e.g. humidity or temperature) constant for the master reference, varying this paramater for a DUT, and evaluating the drift by monitoring the difference between the references. Same goes for longterm drift, by using a group of many references (A man with one, two, three clocks..) In the end, maintaining ultra precise references at first requires to care for controlled/constant ambient parameters, than to put too much effort to design references, which are immune to all possible drift factors. Frank ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
On 4 Jun 2016 20:44, "Poul-Henning Kamp"wrote: > > > In message <57527f8a.6070...@yandex.com>, Charles Steinmetz writes: > > >As I have mentioned before, for many years I have put precision circuits > >that may be sensitive to humidity into gasketed metal boxes with fresh > >silica gel packets. > Only thing to remember is that it also reduces air circulation in the > box and therefore increases thermal gradients. I don't know how practical they are, but I have see humidity calibration standards (e.g 50%). These use a chemical reaction to keep a volume at a constant humidity. That might be a way of keeping the humidity constant, without introducing increasing thermal gradients. They might introduce a whole host of other problems, so such a method would need researching. If a new PCB was designed, it would be nice if there were pads to take the expensive Vishay foil resistors, in addition to a cheaper choice. Dave ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
In message <57527f8a.6070...@yandex.com>, Charles Steinmetz writes: >As I have mentioned before, for many years I have put precision circuits >that may be sensitive to humidity into gasketed metal boxes with fresh >silica gel packets. I have opened some of these after 2 and 3 decades, >and even if there are some small leaks, or a vent hole to equalize >pressure, the silica gel has still been fully dry. Only thing to remember is that it also reduces air circulation in the box and therefore increases thermal gradients. -- Poul-Henning Kamp | UNIX since Zilog Zeus 3.20 p...@freebsd.org | TCP/IP since RFC 956 FreeBSD committer | BSD since 4.3-tahoe Never attribute to malice what can adequately be explained by incompetence. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
Lars wrote: if you have a completely sealed box with varying temperatures the > relative humidity in the box changes if the humidity is not close to zero. As I have mentioned before, for many years I have put precision circuits that may be sensitive to humidity into gasketed metal boxes with fresh silica gel packets. I have opened some of these after 2 and 3 decades, and even if there are some small leaks, or a vent hole to equalize pressure, the silica gel has still been fully dry. This is an easy solution, and works extremely well. I second the suggestion that the main drift problem in a DIY voltage reference (apart from the reference itself) is the resistors that set the scaling amplifier gain (or the DAC or PWM drift, if you go that way). If you can live without scaling, by all means do so. Best regards, Charles ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
the issue of resistor drifts is mainly related to the amplifier (7 to 10V) gain resistors, the other resistor drifts do have a very limited impact on the output (see data sheet). > Gesendet: Freitag, 03. Juni 2016 um 23:47 Uhr > Von: "Lars Walenius" <lars.walen...@hotmail.com> > An: "volt-nuts@febo.com" <volt-nuts@febo.com> > Betreff: Re: [volt-nuts] LTZ1000 project build > > > My experience is that boxes makes the time constants for humidity longer but > does not take it away. Heating in boxes takes the change due to humidity down > a little but it doesn´t go away. I have had boxes with LM399 and amps for > 7-10v with wirewounds for more than 20 years and they have definitely had > seasonal variations. Also if you have a completely sealed box with varying > temperatures the relative humidity in the box changes if the humidity is not > close to zero. > > A 10V 2ppm ref on Ebay was said to have no humidity effect but my unit > definitely varied about 20ppm due to humidity during a long humidity test. > One guess I have is that some units have much longer time constants due to > better sealing but I am quite sure it doesn´t go away complete. Another > problem with humidity I have seen is that you may be cheated by humidity > changing over the season and cancelling the aging. This happened during my > first tests of AD587KRZ many years ago. During the first four months two > samples both had less than 2ppm drift and the humidity in the room changed > from 45 to 25%RH so they looked very promising. But during the next four > months when the humidity went up they drifted 15ppm. > > Probably the 5 and 2ppm/C versions is manufactured in the same way. The > humidity graph under severe humidity and temperature testing shows a 500pm > difference for your chosen resistors. So doesn´t say so much. Hopefully they > are below 1ppm/%RH with low loads. > > Lars > > Från: Ian Johnston<mailto:i...@ianjohnston.com> > Skickat: den 3 juni 2016 20:42 > Till: volt-nuts@febo.com<mailto:volt-nuts@febo.com> > Ämne: Re: [volt-nuts] LTZ1000 project build > > Hi, > > The product I designed to use the 2ppm/degC resistors is sealed in an > enclosure where the LM399AH etc live, so I don't expect much problem > with humidity. > > I use 2K2, 3.3K, 10K & 22K. > > It's the RG2012L series from Susumu:- > > http://datasheet.octopart.com/RG2012L-103-L-T05-Susumu-datasheet-13139546.pdf > http://www.susumu.co.jp/common/pdf/n_catalog_partition01_en.pdf > > The 2nd PDF above is mainly for the 5ppm versions but does contains > various graphs including humidity. > > Ian. > > > > On 03/06/2016 17:01, Lars Walenius wrote: > > Hi, > > > > Do you have any idea of the aging and humidity sensitivity of these 2ppm/C > > resistors? > > > > I have tested many resistors during the years and my opinion is that tempco > > is not the biggest problem for normal nuts use but of course may be if you > > sell a product. For a nut it is quite easy to check the tempco but not > > aging and humidity sensitivity. > > > > Of the resistors I have tested it is only hermetical sealed resistors that > > I don´t see a humidity sensitivity on. Even if they are bulk metal foil, > > wire wounds or metal films with low tempco´s they seem to have very varying > > humidity sensitivity if not sealed. Lower values of resistance (100-1k) > > normally seems to be better than 10-100kohm values that in all families may > > have up to 1-2ppm/%RH. > > Last year I bought wire wounds, to be used in a LTZ1000 based design, that > > were supposed to be insensitive to humidity but they were not. They were > > even worse than other brands of WW and had several months of time constant. > > After a while the manufacturer admitted that it was a problem in the > > manufacturing and they were humidity sensitive. So far I haven´t received a > > replacement. > > > > Of course if you have a resistor with 2ppm/%RH and a seasonal variation of > > 50%RH it is only maximum 1ppm seasonal variation on the output on the LTZ. > > That is not easy to measure for most of us nuts. > > > > Lars > > > > Från: i...@ianjohnston.com<mailto:i...@ianjohnston.com> > > Skickat: den 26 maj 2016 13:35 > > Till: il...@xdevs.com<mailto:il...@xdevs.com>; > > volt-nuts@febo.com<mailto:volt-nuts@febo.com> > > Ämne: Re: [volt-nuts] LTZ1000 project build > > > > Hi all, > > > > TiN, out of interest I spent ages fishing for low ppm/degC resistors for my > > own reference..ended up with 2ppm units from
Re: [volt-nuts] LTZ1000 project build
Could this be caused by ionic contamination of the printed circuit board? The quick test I have used for this is to blow on the board with a straw. In extreme cases solving it usually involved dropping the board into an ultrasonic cleaner with water and a bit of dish soap or TSP, then a rinse in pure (deionized, reverse osmosis, distilled, whatever) water, and then another cycle in the ultrasonic cleaner with pure water. Dry with an air compressor and then bake out in an oven. In production, we had to run boards through a dishwasher using detergent and pure water before loading them with parts. On Fri, 3 Jun 2016 23:47:05 +0200, you wrote: > >My experience is that boxes makes the time constants for humidity longer but >does not take it away. Heating in boxes takes the change due to humidity down >a little but it doesn´t go away. I have had boxes with LM399 and amps for >7-10v with wirewounds for more than 20 years and they have definitely had >seasonal variations. Also if you have a completely sealed box with varying >temperatures the relative humidity in the box changes if the humidity is not >close to zero. > >A 10V 2ppm ref on Ebay was said to have no humidity effect but my unit >definitely varied about 20ppm due to humidity during a long humidity test. One >guess I have is that some units have much longer time constants due to better >sealing but I am quite sure it doesn´t go away complete. Another problem with >humidity I have seen is that you may be cheated by humidity changing over the >season and cancelling the aging. This happened during my first tests of >AD587KRZ many years ago. During the first four months two samples both had >less than 2ppm drift and the humidity in the room changed from 45 to 25%RH so >they looked very promising. But during the next four months when the humidity >went up they drifted 15ppm. > >Probably the 5 and 2ppm/C versions is manufactured in the same way. The >humidity graph under severe humidity and temperature testing shows a 500pm >difference for your chosen resistors. So doesn´t say so much. Hopefully they >are below 1ppm/%RH with low loads. > >Lars ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
My experience is that boxes makes the time constants for humidity longer but does not take it away. Heating in boxes takes the change due to humidity down a little but it doesn´t go away. I have had boxes with LM399 and amps for 7-10v with wirewounds for more than 20 years and they have definitely had seasonal variations. Also if you have a completely sealed box with varying temperatures the relative humidity in the box changes if the humidity is not close to zero. A 10V 2ppm ref on Ebay was said to have no humidity effect but my unit definitely varied about 20ppm due to humidity during a long humidity test. One guess I have is that some units have much longer time constants due to better sealing but I am quite sure it doesn´t go away complete. Another problem with humidity I have seen is that you may be cheated by humidity changing over the season and cancelling the aging. This happened during my first tests of AD587KRZ many years ago. During the first four months two samples both had less than 2ppm drift and the humidity in the room changed from 45 to 25%RH so they looked very promising. But during the next four months when the humidity went up they drifted 15ppm. Probably the 5 and 2ppm/C versions is manufactured in the same way. The humidity graph under severe humidity and temperature testing shows a 500pm difference for your chosen resistors. So doesn´t say so much. Hopefully they are below 1ppm/%RH with low loads. Lars Från: Ian Johnston<mailto:i...@ianjohnston.com> Skickat: den 3 juni 2016 20:42 Till: volt-nuts@febo.com<mailto:volt-nuts@febo.com> Ämne: Re: [volt-nuts] LTZ1000 project build Hi, The product I designed to use the 2ppm/degC resistors is sealed in an enclosure where the LM399AH etc live, so I don't expect much problem with humidity. I use 2K2, 3.3K, 10K & 22K. It's the RG2012L series from Susumu:- http://datasheet.octopart.com/RG2012L-103-L-T05-Susumu-datasheet-13139546.pdf http://www.susumu.co.jp/common/pdf/n_catalog_partition01_en.pdf The 2nd PDF above is mainly for the 5ppm versions but does contains various graphs including humidity. Ian. On 03/06/2016 17:01, Lars Walenius wrote: > Hi, > > Do you have any idea of the aging and humidity sensitivity of these 2ppm/C > resistors? > > I have tested many resistors during the years and my opinion is that tempco > is not the biggest problem for normal nuts use but of course may be if you > sell a product. For a nut it is quite easy to check the tempco but not aging > and humidity sensitivity. > > Of the resistors I have tested it is only hermetical sealed resistors that I > don´t see a humidity sensitivity on. Even if they are bulk metal foil, wire > wounds or metal films with low tempco´s they seem to have very varying > humidity sensitivity if not sealed. Lower values of resistance (100-1k) > normally seems to be better than 10-100kohm values that in all families may > have up to 1-2ppm/%RH. > Last year I bought wire wounds, to be used in a LTZ1000 based design, that > were supposed to be insensitive to humidity but they were not. They were even > worse than other brands of WW and had several months of time constant. After > a while the manufacturer admitted that it was a problem in the manufacturing > and they were humidity sensitive. So far I haven´t received a replacement. > > Of course if you have a resistor with 2ppm/%RH and a seasonal variation of > 50%RH it is only maximum 1ppm seasonal variation on the output on the LTZ. > That is not easy to measure for most of us nuts. > > Lars > > Från: i...@ianjohnston.com<mailto:i...@ianjohnston.com> > Skickat: den 26 maj 2016 13:35 > Till: il...@xdevs.com<mailto:il...@xdevs.com>; > volt-nuts@febo.com<mailto:volt-nuts@febo.com> > Ämne: Re: [volt-nuts] LTZ1000 project build > > Hi all, > > TiN, out of interest I spent ages fishing for low ppm/degC resistors for my > own reference..ended up with 2ppm units from DigiKey. Yes, they are > expensive and send the BOM into the clouds!..so I know how you > feel! > > Ian. > > > ----- Original Message ----- > From: Illya Tsemenko [mailto:il...@xdevs.com] > To: volt-nuts@febo.com > Sent: Thu, 26 May 2016 18:42:54 +0800 > Subject: [volt-nuts] LTZ1000 project build > > I'm glad my little project got so much attention. Worth to mention few > points regarding my (or any other LTZ1000) unit: > > A. One should consider cost of whole BOM, not only LTZ chip. In my case > resistors from VPG exceed cost of LTZ1000 by far. > B. PCB I designed was targeted best to my own application need, where > compact size and direct output (whatever is coming out of LTZ) were key > importance. Since modules are part of bigger system with DAC synth, > specific value, like 7.150 was not im
Re: [volt-nuts] LTZ1000 project build
Hi, The product I designed to use the 2ppm/degC resistors is sealed in an enclosure where the LM399AH etc live, so I don't expect much problem with humidity. I use 2K2, 3.3K, 10K & 22K. It's the RG2012L series from Susumu:- http://datasheet.octopart.com/RG2012L-103-L-T05-Susumu-datasheet-13139546.pdf http://www.susumu.co.jp/common/pdf/n_catalog_partition01_en.pdf The 2nd PDF above is mainly for the 5ppm versions but does contains various graphs including humidity. Ian. On 03/06/2016 17:01, Lars Walenius wrote: Hi, Do you have any idea of the aging and humidity sensitivity of these 2ppm/C resistors? I have tested many resistors during the years and my opinion is that tempco is not the biggest problem for normal nuts use but of course may be if you sell a product. For a nut it is quite easy to check the tempco but not aging and humidity sensitivity. Of the resistors I have tested it is only hermetical sealed resistors that I don´t see a humidity sensitivity on. Even if they are bulk metal foil, wire wounds or metal films with low tempco´s they seem to have very varying humidity sensitivity if not sealed. Lower values of resistance (100-1k) normally seems to be better than 10-100kohm values that in all families may have up to 1-2ppm/%RH. Last year I bought wire wounds, to be used in a LTZ1000 based design, that were supposed to be insensitive to humidity but they were not. They were even worse than other brands of WW and had several months of time constant. After a while the manufacturer admitted that it was a problem in the manufacturing and they were humidity sensitive. So far I haven´t received a replacement. Of course if you have a resistor with 2ppm/%RH and a seasonal variation of 50%RH it is only maximum 1ppm seasonal variation on the output on the LTZ. That is not easy to measure for most of us nuts. Lars Från: i...@ianjohnston.com<mailto:i...@ianjohnston.com> Skickat: den 26 maj 2016 13:35 Till: il...@xdevs.com<mailto:il...@xdevs.com>; volt-nuts@febo.com<mailto:volt-nuts@febo.com> Ämne: Re: [volt-nuts] LTZ1000 project build Hi all, TiN, out of interest I spent ages fishing for low ppm/degC resistors for my own reference..ended up with 2ppm units from DigiKey. Yes, they are expensive and send the BOM into the clouds!..so I know how you feel! Ian. - Original Message - From: Illya Tsemenko [mailto:il...@xdevs.com] To: volt-nuts@febo.com Sent: Thu, 26 May 2016 18:42:54 +0800 Subject: [volt-nuts] LTZ1000 project build I'm glad my little project got so much attention. Worth to mention few points regarding my (or any other LTZ1000) unit: A. One should consider cost of whole BOM, not only LTZ chip. In my case resistors from VPG exceed cost of LTZ1000 by far. B. PCB I designed was targeted best to my own application need, where compact size and direct output (whatever is coming out of LTZ) were key importance. Since modules are part of bigger system with DAC synth, specific value, like 7.150 was not important for me. C. If there is large enough group to have this project going, I can actually modify original design (e.g. for easier hand soldering, or added buffer or ratio circuits, onboard LDO, you name it) and get PCBs made for interested nuts, at small cost. You can also find lot of temco/stability data with temperature/environment logs with mentioned LTZ modules on my site here: https://xdevs.com/services/ in LTZ1000 voltage reference tests section. RAW DSV data are available. I still have 3 modules at hand for any testing/checks (one was sold while ago, one was installed in HP 3245A source to replace LM399). Illya "TiN" Tsemenko ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. -- Ian Johnston www.ianjohnston.com https://www.youtube.com/user/IanScottJohnston /"Because it feels good to make stuff!"/ ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
Hi, Do you have any idea of the aging and humidity sensitivity of these 2ppm/C resistors? I have tested many resistors during the years and my opinion is that tempco is not the biggest problem for normal nuts use but of course may be if you sell a product. For a nut it is quite easy to check the tempco but not aging and humidity sensitivity. Of the resistors I have tested it is only hermetical sealed resistors that I don´t see a humidity sensitivity on. Even if they are bulk metal foil, wire wounds or metal films with low tempco´s they seem to have very varying humidity sensitivity if not sealed. Lower values of resistance (100-1k) normally seems to be better than 10-100kohm values that in all families may have up to 1-2ppm/%RH. Last year I bought wire wounds, to be used in a LTZ1000 based design, that were supposed to be insensitive to humidity but they were not. They were even worse than other brands of WW and had several months of time constant. After a while the manufacturer admitted that it was a problem in the manufacturing and they were humidity sensitive. So far I haven´t received a replacement. Of course if you have a resistor with 2ppm/%RH and a seasonal variation of 50%RH it is only maximum 1ppm seasonal variation on the output on the LTZ. That is not easy to measure for most of us nuts. Lars Från: i...@ianjohnston.com<mailto:i...@ianjohnston.com> Skickat: den 26 maj 2016 13:35 Till: il...@xdevs.com<mailto:il...@xdevs.com>; volt-nuts@febo.com<mailto:volt-nuts@febo.com> Ämne: Re: [volt-nuts] LTZ1000 project build Hi all, TiN, out of interest I spent ages fishing for low ppm/degC resistors for my own reference..ended up with 2ppm units from DigiKey. Yes, they are expensive and send the BOM into the clouds!..so I know how you feel! Ian. - Original Message - From: Illya Tsemenko [mailto:il...@xdevs.com] To: volt-nuts@febo.com Sent: Thu, 26 May 2016 18:42:54 +0800 Subject: [volt-nuts] LTZ1000 project build I'm glad my little project got so much attention. Worth to mention few points regarding my (or any other LTZ1000) unit: A. One should consider cost of whole BOM, not only LTZ chip. In my case resistors from VPG exceed cost of LTZ1000 by far. B. PCB I designed was targeted best to my own application need, where compact size and direct output (whatever is coming out of LTZ) were key importance. Since modules are part of bigger system with DAC synth, specific value, like 7.150 was not important for me. C. If there is large enough group to have this project going, I can actually modify original design (e.g. for easier hand soldering, or added buffer or ratio circuits, onboard LDO, you name it) and get PCBs made for interested nuts, at small cost. You can also find lot of temco/stability data with temperature/environment logs with mentioned LTZ modules on my site here: https://xdevs.com/services/ in LTZ1000 voltage reference tests section. RAW DSV data are available. I still have 3 modules at hand for any testing/checks (one was sold while ago, one was installed in HP 3245A source to replace LM399). Illya "TiN" Tsemenko ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] LTZ1000 project build
Hello, I monitor two DIY LTZ1000 references, standard circuit from LT datasheet, since about 7 years. These are running on 45°C, and are using 5 precision wirewound resistors (PWW), T.C. ~0.2 .. 4ppm/K. BOM cost are currently 42.85$ for the LTZ1000 (non A!!) directly from LT (recommended, beware of chinese fakes), and about 7-10$ for each resistor. So each reference may be built for about 100$ in total. By avoiding some pitfalls, like excessive heating, causing hysteresis effects, these reference will have an annual stability of typ. -0.8ppm/year, without any special treatment, like "ageing", slots in the PCB, or usage of extremely sophisticated and expensive metal foil resistors. One only has to monitor the drift of each reference if it's having this typical drift rate, or not. Andreas Jahn currently makes a great job in investigating on the overall T.C. performance of his LTZ references, in terms of selecting these PWW resistors, trimming by 400k resistor, and LTZ lead length. I'm doing that drift monitoring by comparison of a total of 4 references, that's an additional LTZ1000A in a 3458A, pimped to 65°C, and not running continuously (practically no drift), and two stacked SZA263 in a Fluke 5442A, which have typ. +1ppm/yr, but this instrument also is not powered continuously. The differences within this group are still within +1/-2ppm after 7 years. The absolute uncertainty (w/o traceability) is estimated to 1ppm, as given by the initial calibration of the 5442A in 2009, one check vs. a freshly calibrated 8508A from another volt-nut in 2013 (+0.5ppm difference), and vs. two new 34465A last year (<0.5 ppm difference on 10V range). So it's really less complicated and less expensive than expected, to build and qualify LTZ references. "Dr. Frank" ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
Illya, Very interesting. The temco/stability data does not plot but the X-Y format seems to be OK. Is it supposed to plot or do I have to plot it myself with the csv data? Thanks, Randy Evans On Thu, May 26, 2016 at 3:42 AM, Illya Tsemenkowrote: > I'm glad my little project got so much attention. Worth to mention few > points regarding my (or any other LTZ1000) unit: > > A. One should consider cost of whole BOM, not only LTZ chip. In my case > resistors from VPG exceed cost of LTZ1000 by far. > B. PCB I designed was targeted best to my own application need, where > compact size and direct output (whatever is coming out of LTZ) were key > importance. Since modules are part of bigger system with DAC synth, > specific value, like 7.150 was not important for me. > C. If there is large enough group to have this project going, I can > actually modify original design (e.g. for easier hand soldering, or added > buffer or ratio circuits, onboard LDO, you name it) and get PCBs made for > interested nuts, at small cost. > > You can also find lot of temco/stability data with temperature/environment > logs with mentioned LTZ modules on my site here: > https://xdevs.com/services/ in LTZ1000 voltage reference tests section. > RAW DSV data are available. I still have 3 modules at hand for any > testing/checks (one was sold while ago, one was installed in HP 3245A > source to replace LM399). > > Illya "TiN" Tsemenko > > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
Hi all, TiN, out of interest I spent ages fishing for low ppm/degC resistors for my own reference..ended up with 2ppm units from DigiKey. Yes, they are expensive and send the BOM into the clouds!..so I know how you feel! Ian. - Original Message - From: Illya Tsemenko [mailto:il...@xdevs.com] To: volt-nuts@febo.com Sent: Thu, 26 May 2016 18:42:54 +0800 Subject: [volt-nuts] LTZ1000 project build I'm glad my little project got so much attention. Worth to mention few points regarding my (or any other LTZ1000) unit: A. One should consider cost of whole BOM, not only LTZ chip. In my case resistors from VPG exceed cost of LTZ1000 by far. B. PCB I designed was targeted best to my own application need, where compact size and direct output (whatever is coming out of LTZ) were key importance. Since modules are part of bigger system with DAC synth, specific value, like 7.150 was not important for me. C. If there is large enough group to have this project going, I can actually modify original design (e.g. for easier hand soldering, or added buffer or ratio circuits, onboard LDO, you name it) and get PCBs made for interested nuts, at small cost. You can also find lot of temco/stability data with temperature/environment logs with mentioned LTZ modules on my site here: https://xdevs.com/services/ in LTZ1000 voltage reference tests section. RAW DSV data are available. I still have 3 modules at hand for any testing/checks (one was sold while ago, one was installed in HP 3245A source to replace LM399). Illya "TiN" Tsemenko ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] LTZ1000 project build
I'm glad my little project got so much attention. Worth to mention few points regarding my (or any other LTZ1000) unit: A. One should consider cost of whole BOM, not only LTZ chip. In my case resistors from VPG exceed cost of LTZ1000 by far. B. PCB I designed was targeted best to my own application need, where compact size and direct output (whatever is coming out of LTZ) were key importance. Since modules are part of bigger system with DAC synth, specific value, like 7.150 was not important for me. C. If there is large enough group to have this project going, I can actually modify original design (e.g. for easier hand soldering, or added buffer or ratio circuits, onboard LDO, you name it) and get PCBs made for interested nuts, at small cost. You can also find lot of temco/stability data with temperature/environment logs with mentioned LTZ modules on my site here: https://xdevs.com/services/ in LTZ1000 voltage reference tests section. RAW DSV data are available. I still have 3 modules at hand for any testing/checks (one was sold while ago, one was installed in HP 3245A source to replace LM399). Illya "TiN" Tsemenko ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
What prices are you seeing for 100 LTZ1000s? The main thing is they need to be cooked in then sorted for stability. A good data acquisition system used to do stats on all the units would really help. On Wed, May 25, 2016 at 10:17 PM, Andreas Jahnwrote: > Hello, > > there are further projects (including measurement results of Illya = TiN) > > in this thread: > > http://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/ > > and others in the metrology/projects section of eevblog > > > http://www.eevblog.com/forum/projects/project-kx-diy-calibrator-reference-sourcemeter/ > > With best regards > > Andreas > > > Am 25.05.2016 um 17:27 schrieb Eric Garner: > >> I ran across this: >> https://xdevs.com/article/kx-ref/ >> >> this morning and thought I'd share. >> >> > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > -- *John Phillips* ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
Hello, there are further projects (including measurement results of Illya = TiN) in this thread: http://www.eevblog.com/forum/metrology/ultra-precision-reference-ltz1000/ and others in the metrology/projects section of eevblog http://www.eevblog.com/forum/projects/project-kx-diy-calibrator-reference-sourcemeter/ With best regards Andreas Am 25.05.2016 um 17:27 schrieb Eric Garner: I ran across this: https://xdevs.com/article/kx-ref/ this morning and thought I'd share. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
Heck, if you’re going to make one, you might as well make one GREAT. I would be interested in doing a 4910 version, so to speak. Four in one thermal box and the ability to use them independently or summed and averaged. Similar to the Fluke 730a, I think? Oh, and one of them must be removable for transportation. : ) Of course, if properly designed one could build any number of them into whatever box they can podge together and use the boards as they please. I’m in!! I would find it difficult to believe we could not find 100 individuals that would buy at least one. The challenge is buying the LTZ’s in enough QTY to grade them like Solartron did in the 7071 and 81’s. David > On May 25, 2016, at 9:42 PM, bowneswrote: > > > As Dr K said, traceable and usefully calibrated are not necessarily > connected. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
As Dr K said, traceable and usefully calibrated are not necessarily connected. I can calibrate to any arbitrary standard I like. That standard need not be traceable if all that is important to me is consistency across all the instruments in my lab. If I, on the other hand, want to be consistent with someone else's lab, then we need to be traceable to a common source. Thus NIST. I presume most countries have a NIST like organization. How often to they cross check each other? Accreditation, on the other hand can be, as the good Dr. points out, pretty useless unless the accreditation body is, itself, held to some (professional) standard. And I'd also love to build one of these if there is enough interest. While I'm sure we can't get enough orders to get the 100pcs discount on the LTZ1000, it would be a great group project and I'd be willing to participate in bringing it to fruition. I'm also sure I can find a calibrated, traceable, reliable 3458 in the area code. The irony is that while I'm less than 15Km from the New York State Bureau of Weights & Measures Metrology group, which has all the traceable standards for the state, they cannot do high accuracy for time or voltage. My personal house standards are better than theirs for those two. By a lot. Adding one of these would add a few more orders of magnitude...:) > On May 25, 2016, at 17:06, Dr. David Kirkby (Kirkby Microwave Ltd) >wrote: > >> On 25 May 2016 at 19:24, Russ Ramirez wrote: >> >> A lot of great information Eric, thanks for sharing the link. >> >> Due to my ignorance in general on the subject of Metrology, I have the >> following question for the list. >> >> If one built a project with the LTZ1000, like the one described on xDevs >> and could set it to a value of 7.1500v at the NIST lab, and observed >> stability to 7 1/2 digits, would using that device to calibrate your own 7 >> 1/2 digit DMM be considered NIST traceable? Let's say your device is well >> insulated and battery powered, and your calibration was done at the same >> altitude and room temperature as at NIST, plus anything I left out that >> would make the conditions ideal. >> >> The above was not meant to be a trick question, and I may have asked it >> incorrectly, but I view the answers as instructive - or I hope they are. >> >> Russ > > As far as I can determine, as long as you can work out the uncertainties, > no matter how large they might b, the measurement is traceable. If you use > a 3.5 digit multimeter that is NIST traceable to calibrate a 7.5 digit > multimeter, the calibration is still NIST traceable. The calibration will > be pretty useless, and you may not be accredited, but it is still NIST > traceable. > > Or if you want to be accredited, get your mate down the local pub (bar) to > accredit you! > > On a more serious note, if people felt that design was good, and wanted to > produce the PCBs. and/or make parts available, I'd be interested. I only > have a 6.5 digit meter, but feel sure I could find someone with a 3458A in > the UK who could measure the voltage for me. > > Dave > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
Re: [volt-nuts] LTZ1000 project build
A lot of great information Eric, thanks for sharing the link. Due to my ignorance in general on the subject of Metrology, I have the following question for the list. If one built a project with the LTZ1000, like the one described on xDevs and could set it to a value of 7.1500v at the NIST lab, and observed stability to 7 1/2 digits, would using that device to calibrate your own 7 1/2 digit DMM be considered NIST traceable? Let's say your device is well insulated and battery powered, and your calibration was done at the same altitude and room temperature as at NIST, plus anything I left out that would make the conditions ideal. The above was not meant to be a trick question, and I may have asked it incorrectly, but I view the answers as instructive - or I hope they are. Russ On Wed, May 25, 2016 at 10:27 AM, Eric Garnerwrote: > I ran across this: > https://xdevs.com/article/kx-ref/ > > this morning and thought I'd share. > > -- > --Eric > _ > Eric Garner > ___ > volt-nuts mailing list -- volt-nuts@febo.com > To unsubscribe, go to > https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts > and follow the instructions there. > ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.
[volt-nuts] LTZ1000 project build
I ran across this: https://xdevs.com/article/kx-ref/ this morning and thought I'd share. -- --Eric _ Eric Garner ___ volt-nuts mailing list -- volt-nuts@febo.com To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/volt-nuts and follow the instructions there.