This whole discussion does not really apply to the LTZ1000 circuit, that is the bare 7.15..V output, w/o step up amplifier, as all drift parameters were attenuated by orders of magnitude, depending on the resistors, R1-R5.

In the eevblog thread, several volt-nuts measured the real dependency, in contrast to the datsheet, which simply claims 1:100 attenuation for each of these resistors:


R1      R2      R3      R4      R5      author
120     70000   70000   12-15k  1k      
-1/770  -1/250  -1/1400         +1/100  
        janaf
-1/600  -1/238  -1/1000         +1/87   
        lars
-1/656  
        
        
        
        TiN
1/714   1/250   1/3333  1/105   
        bbs38hot
-1/670  -1/238  -1/1184         +1/74   -1/75   Andreas


Therefore, even humidity effects at most affect R4/R5, with 1/75 impact.

I also think, that it's possible for volt-nuts to measure such residual effects, by assembling several references, keeping the parameter of question (e.g. humidity or temperature) constant for the master reference, varying this paramater for a DUT, and evaluating the drift by monitoring the difference between the references.

Same goes for longterm drift, by using a group of many references (A man with one, two, three clocks..)

In the end, maintaining ultra precise references at first requires to care for controlled/constant ambient parameters, than to put too much effort to design references, which are immune to all possible drift factors.

Frank

_______________________________________________