Ever tried turning the handle of the cryostat, when it's at room temperature? Cryostats are tooled manufactured to operate at a low temperature. Since metal contracts at the low temperature, you'll find that you can't operate the microtome at the higher temperature. The handle will barely move.

RE: [Histonet] Countertop fume hoods

2013-04-12 Thread Harrison, Sandra C.
Dear Merissa, I have used the Labconco countertop fume hood in a few different Histology settings, both MOHS and general, and found it to be very compact, easy to change the filter, and very effective at pulling the fumes away from the user. It is also a ductless system. See

[Histonet] Leica 4020 linear stainer

2013-01-07 Thread Harrison, Sandra C.
Could I get some feedback on linear stainers?Has anyone had any experience with the Leica ST4020? We are going to purchase a linear stainer for a new MOHS laboratory. Thanks, Sandy Harrison Histology Supervisor VA Health Care Systems, Minneapolis 612-467-2449

[Histonet] procedure for destaining a cytology slide followed by restaining with a special stain, such as PAS?

2012-11-16 Thread Harrison, Sandra C.
Does anyone have a procedure for restaining a cytology slide? We experienced a loss of cells on an FNA aspirate of a lymph node, which had been originally stained with Diff Qwik. The procedure we used was as follows: 1) Destain using 1% Acid Alcohol (1 ml. concentrated HCL added to 99

[Histonet] H.Pylori

2012-10-17 Thread Harrison, Sandra C.
We do an H.P. routinely on stomach, body, fundus, antrum, pyloric, prepyloric, and gastric (Not on cardia area of stomach or GE jcn.) We use a kit from Newcomer Supply of a Steiner-Chapman modified Silver Stain. The pathologist's order an IHC for H.P. only about 5% of the time, if for example,

[Histonet] Are those sites using the Thermo Fisher Slide Mate slide printer also using Thermo slides?

2012-10-10 Thread Harrison, Sandra C.
I just purchased 6 Slide Mates. We just began using them yesterday. Are any of you using slides that are NOT Fisher slides on your SlideMates? When you respond, could you answer the following questions? a) Is the print job consistently satisfactory? b) How long have had

[Histonet] BOND IHC instrument- anyone experiencing multiple heater failures??

2012-08-06 Thread Harrison, Sandra C.
We have 2 Bond Max instruments and have experienced multiple heaters getting contaminated and corroded underneath, with a particulate matter, causing the heaters to fail. Our 2nd BOND Max is 3 years old and we have had to replace a total of 12 heaters (out of 30.) The only thing we do that is

RE: [Histonet] automated H+E and cover slipper

2012-07-18 Thread Harrison, Sandra C.
Leica Autostainer XL with CV5030 coverslipper and transfer station. This has been a real timesaver for us. It automatically moves the slides from the stainer to the coverslipper. It has been relatively trouble free. -Original Message- From: histonet-boun...@lists.utsouthwestern.edu

RE: [Histonet] Mohs

2012-07-03 Thread Harrison, Sandra C.
Depends ;-) All kidding aside, there are factors that weigh in, such as geography, availability, training, experience, speed, etc. 10 years ago, when I worked as a Mohs' tech in the Denver, Colorado area, I received $25.00 per hour, plus benefits. I usually only worked 3 days per week, but on

RE: [Histonet] Humidity Check

2012-07-03 Thread Harrison, Sandra C.
Yes, we have a wireless temperature monitoring system, called Checkpoint, and each room has a sensor that monitors temp and humidity. -Original Message- From: histonet-boun...@lists.utsouthwestern.edu [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of

RE: [Histonet] Paperless Histology

2012-06-29 Thread Harrison, Sandra C.
And what about a paperless procedure manual? We have an electronic system, called Proquis, for our manuals. It sends an e-mail to all the appropriate people, when there is a new procedure to review or when old procedures are due for review. Pretty cool. The only drawback is, most of the

[Histonet] compatibility of Thermo slide printers with Leica cassette printer

2012-05-29 Thread Harrison, Sandra C.
1) Does anyone have a Leica Cassette Printer on which they are printing 2-Dimensional barcodes? 2) Is anyone using a Thermo slide printer, with bar code reader, to read cassettes printed off of a Leica Cassette Printer with 2-D bar codes? Thanks for your input. Sandy C.

[Histonet] Chicago Sky Blue stain vs. KOH - striking comparison

2011-11-10 Thread Harrison, Sandra C.
Does anyone have ordering information for a stain called Chicago Sky Blue(CSB)? I just saw an article in the NOV issue of ASCP LABMEDICINE (page 649) comparing KOH vs. CSB and, wow, is that a beautiful stain. The results were dramatic, with about a 25-50% increase in the Dermatophytes

[Histonet] saving slides between levels for IHC?

2011-10-25 Thread Harrison, Sandra C.
Does anyone know if it is standard to save sections from between the levels for IP? Is it standard at your institution? Sandy C. Harrison, HTL (ASCP) Histology Supervisor Minneapolis VA 612-467-2449 ___ Histonet mailing list

[Histonet] Leica cassette and slide printer ink

2011-09-21 Thread Harrison, Sandra C.
I have been told that I can no longer buy only the replacement ink cartridges for Leica cassette and slide printers. The ink cartridge was already an outrageous $411.00 each. Now you have to by a kit for $566.00 each. The kit includes a replacement plate along with the ink. Why in the

[Histonet] gout processing procedure

2011-08-30 Thread Harrison, Sandra C.
Could someone share their gout processing procedure? The specimen is fresh, not in formalin. Thanks, Sandy C. Harrison, HTL (ASCP) Histology Supervisor Minneapolis VA 612-467-2449 ___ Histonet mailing list

RE: [Histonet] bluing

2011-06-24 Thread Harrison, Sandra C.
Hi Dorothy, Try Richard Allan Bluing Reagent. Here's what they say about their product: It is a buffered product that ensures the proper alkalinity (pH=8.0). Unlike ammonia and lithium carbonate, RA's Bluing Reagent does not allow for the pH shift which can affect the crispness of nuclear

RE: [Histonet] question on H pylori

2011-06-10 Thread Harrison, Sandra C.
We do either IHC or a Modified Steiner, depending on the Pathologist's preference. We switched to Newcomer Supply's Steiner-Chapman Modified Silver Stain Kit about a year ago, because it eliminated the use of Uranyl Nitrate. We have been delighted with the consistency of the stain. Sandy

RE: [Histonet] Input needed

2011-04-22 Thread Harrison, Sandra C.
Out of 6 recent hires, I hired two 50+ employees. One had been employed in another department for over 28 years. One was a fairly recent grad. of a Histology program. They were both good employees, though neither had much routine Histology experience. That being said, neither one of them are

RE: [Histonet] Control Slides

2011-03-14 Thread Harrison, Sandra C.
We dry the control slide at the same time that we dry the tissue being stained, since controls are supposed to be handled in the same manner as the test tissue. That being said, I guess if we were truly following that rule, we would cut the control on the same day we cut the test sample. :-)

RE: [Histonet] No Mail??

2010-12-17 Thread Harrison, Sandra C.
Wow...didn't realize how addicted to Histonet I'd become till this week! Sandy Harrison Histology Supervisor VA Medical Center Minneapolis, MN 55417 612-467-2449 -Original Message- From: histonet-boun...@lists.utsouthwestern.edu [mailto:histonet-boun...@lists.utsouthwestern.edu]

RE: [Histonet] CHECKING ALCOHOL CONCENTRATIONS

2010-11-23 Thread Harrison, Sandra C.
CBG Biotech. Their contact info is 1-800-941-9484 or you can e mail them at i...@cbgbiotech.com. I purchased my recycler through them and have ordered extra hydrometers from them, too. They give excellent customer service. Sandy Harrison, HTL(ASCP) Histology Supervisor, VA Minneapolis

RE: [Histonet] ThermoFisher's PrintMate (cassette printer) and SlideMate (slide labeler with scanner)

2010-11-15 Thread Harrison, Sandra C.
I'm looking at the Slide Mate and Print Mate, too. Currently, I have the Leica cassette and slide printers. They have really been great. Reliable and relatively problem free. But they are so large. I'd like to find a slide printer that has a small footprint, like the Slide Mate, so that we can

RE: [Histonet] Block release

2010-10-22 Thread Harrison, Sandra C.
Here's what the recently revised CAP checklist says about releasing blocks; scroll down to Note 2: **REVISED** 06/17/2010 ANP.12500 Record Retention Phase II Surgical pathology records and materials are retained for an appropriate period. NOTE 1: Minimum requirements for surgical pathology,

RE: [Histonet] Cassette Labeler

2010-09-17 Thread Harrison, Sandra C.
We have the Leica Cassette Labeler. It works just fine. I have looked at the Thermo Fisher Cassette Printer, which had a much smaller footprint than the Leica. I would be tempted to go with the Thermo if I have to replace my current labeler. These are expensive instruments but worth having.

RE: [Histonet] accu-edge low profile microtome blades

2010-09-01 Thread Harrison, Sandra C.
About 1/2 of my tech.'s wipe the oil off; otherwise it causes the first 2-3 ribbons to curl off the knife, rather than forming a nice ribbon. Sandy Harrison Histology Supervisor Minneapolis VA -Original Message- From: histonet-boun...@lists.utsouthwestern.edu

RE: [Histonet] QC on stained slides

2010-08-16 Thread Harrison, Sandra C.
You have included all the criteria; fixation, processing, embedding, microtomy, staining, coverslipping and labeling. If you were able to submit slides to CAP under their HQIP program, you would get graded evaluations. Short of that, could you perhaps send out duplicate slides to another local

[Histonet] labeling of recycled chemicals

2010-07-29 Thread Harrison, Sandra C.
Dear Histonetters, If you use a recycler for xylene and/or alcohol, how are you labeling the recycled containers? Do you assign a lot # to each carboy? Do you keep a log or just label any container filled from a carboy with the lot # and the %, in the case of alcohols? Thanks, Sandy

RE: [Histonet] Chemical outdates

2010-01-25 Thread Harrison, Sandra C.
If there are no expiration dates on the chemical: For chemicals distributed by Fisher, look at the Lot #. The first 2 numbers are the year it was manufactured. In general, chemicals are considered expired 3-5 years after the year they were manufactured. You can also check the certificate that

[Histonet] Histotech opening at Minneapolis VA

2009-10-30 Thread Harrison, Sandra C.
DEPT. OF VETERAN AFFAIRS MEDICAL CENTER, MINNEPOLIS, MN. Full time Histotech. opportunity at Minneapolis VA. BS or BA in Biology. HT cert. required, HTL preferred. Prefer 5 yr. exp. IHC experience a plus. Effective interpersonal skills required. Holiday, evenings and weekends off. Excellent

[Histonet] job opening, Minneapolis VA

2009-08-19 Thread Harrison, Sandra C.
DEPT. OF VETERAN AFFAIRS MEDICAL CENTER, MINNEPOLIS, MN. Full time Histotech. opportunity at Minneapolis VA. BS or BA in Biology. HT cert. required, HTL preferred. Prefer 5 yr. exp. IHC experience a plus. Effective interpersonal skills required. Holiday, evenings and weekends off. Excellent

RE: [Histonet] nuclear bubbling

2009-07-17 Thread Harrison, Sandra C.
How about microwaving to dry slides? Can that cause nuclear bubbling? When you say completely drained off, does that mean your slides have to be completely dry prior to placing them in the oven? Thanks, Sandy -Original Message- From: histonet-boun...@lists.utsouthwestern.edu

RE: [Histonet] Eosin in Alcohol

2009-07-14 Thread Harrison, Sandra C.
polycyclic aromatic flourescent compounds that in high concentrations I wouldn't think the 3 mls of eosin dropped in the last 95% alcohol could be considered high concentration but that's what keeps Histonet entertaining; I learn something new every day. -Original Message- From: