Re: [Histonet] Tissue Processor Down Question

2024-03-14 Thread Diana Martinez-Longoria via Histonet
- From: Gudrun Lang [mailto:gu.l...@gmx.at] Sent: Thursday, March 14, 2024 7:28 AM To: Diana Martinez-Longoria Cc: histonet@lists.utsouthwestern.edu Subject: AW: [Histonet] Tissue Processor Down Question Warning! This email was sent from outside your organization. Do not click or open attachments

Re: [Histonet] Tissue Processor Down Question

2024-03-14 Thread Gudrun Lang via Histonet
-Ursprüngliche Nachricht- Von: Diana Martinez-Longoria via Histonet [mailto:histonet@lists.utsouthwestern.edu] Gesendet: Donnerstag, 14. März 2024 13:59 An: histonet@lists.utsouthwestern.edu Betreff: [Histonet] Tissue Processor Down Question Wichtigkeit: Hoch Good morning, Our VIP5 tissue

[Histonet] Tissue Processor Down Question

2024-03-14 Thread Diana Martinez-Longoria via Histonet
Good morning, Our VIP5 tissue processor went down and our tissue cassettes were in clearing solution. What is the best media to put them while we wait for the tissue processor to be repaired, please let me know? Thank you, Diana Martinez-Longoria El Centro Regional Medical Center Lead Histotechn

Re: [Histonet] Tissue Processor Schedule Validations

2023-03-23 Thread Cooper, Brian via Histonet
: Thursday, March 23, 2023 11:54 AM To: histonet@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Schedule Validations (EXTERNAL EMAIL) CAUTION: BE CAREFUL WITH THIS MESSAGE* This email came from outside CHLA. Do not open attachments, click on links, or respond unless you expected

Re: [Histonet] Tissue Processor Schedule Validations

2023-03-23 Thread Paula Sicurello via Histonet
The one thing that bothers me about the preloaded programs is the paraffin temperature is set at 65 degrees C.  That's close to 10 degrees higher than the melting point of our paraffin. Does anyone know why the temp. is so high?  To me, a temp that high will cause problems with IHC and small bio

Re: [Histonet] Tissue Processor Schedule Validations

2023-03-23 Thread Thomas Podawiltz via Histonet
Sent: Thursday, March 23, 2023 1:54 PM To: histonet@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Schedule Validations WARNING: This email appears to have originated outside of the UW Health email system. DO NOT CLICK on links or attachments unless you recognize the sender and

Re: [Histonet] Tissue Processor Schedule Validations

2023-03-23 Thread Normington, Lacy via Histonet
, March 23, 2023 1:54 PM To: histonet@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Schedule Validations WARNING: This email appears to have originated outside of the UW Health email system. DO NOT CLICK on links or attachments unless you recognize the sender and know the content is

Re: [Histonet] Tissue Processor Schedule Validations

2023-03-23 Thread Whitaker, Bonnie via Histonet
: Thursday, March 23, 2023 2:54 PM To: histonet@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Schedule Validations Good afternoon Histonet, We're going to be validating a new tissue processor (Peloris 3) in the coming months, and I'm curious how people have validated sm

[Histonet] Tissue Processor Schedule Validations

2023-03-23 Thread Cooper, Brian via Histonet
Good afternoon Histonet, We're going to be validating a new tissue processor (Peloris 3) in the coming months, and I'm curious how people have validated small tissue processing protocols (GI bx's, liver/renal needle cores). Larger tissues are much easier to do because we can readily gross dupl

Re: [Histonet] Tissue processor advice

2020-09-17 Thread Bacon, Charles via Histonet
- From: Colleen Forster Sent: Wednesday, September 16, 2020 1:17 PM To: histonet-request Subject: [Histonet] Tissue processor advice HEllo Histoneters, I am looking at replacing my VIP2000, porr girl finally quit on me. The two processors I am looking at: Leica ASP300S Sakura VIP5 Any of you out

[Histonet] Tissue processor Pros/Cons

2020-09-16 Thread Dessasau,, Evan D. via Histonet
edde...@emory.edu -Original Message- From: Colleen Forster via Histonet Sent: Wednesday, September 16, 2020 1:17 PM To: histonet-request Subject: [External] [Histonet] Tissue processor advice HEllo Histoneters, I am looking at replacing my VIP2000, porr girl finally quit on me. The two

[Histonet] Tissue processor advice

2020-09-16 Thread Colleen Forster via Histonet
HEllo Histoneters, I am looking at replacing my VIP2000, porr girl finally quit on me. The two processors I am looking at: Leica ASP300S Sakura VIP5 Any of you out there who have used either of these or both, can you give me pros , cons, yes, noI am just looking for experiences those who ha

Re: [Histonet] Tissue processor validation

2020-09-02 Thread Simmons, Christopher via Histonet
.ms/o0ukef> From: Martha Ward-Pathology via Histonet Sent: Wednesday, September 2, 2020 3:51:42 PM To: histonet@lists.utsouthwestern.edu Subject: [Histonet] Tissue processor validation I am posting this question for my colleague in the Histology lab. She is work

[Histonet] Tissue processor validation

2020-09-02 Thread Martha Ward-Pathology via Histonet
I am posting this question for my colleague in the Histology lab. She is working on a plan for a new tissue processor validation, including special stains, and would like to get input from anyone that has already gone through it.This is the first time anyone here has ever been involved in

Re: [Histonet] Tissue processor errors, failures and what to do

2020-07-04 Thread E. Wayne Johnson via Histonet
Automation is a wonderful thing but it is only a replacement for what people used to do by hand. We have incubators that can be set to 60C and we have a Rube Goldberg-ized microwave oven with a thermal controller and relays (and the not-to-be-forgotten flyback diode) and a K-type thermal probe c

Re: [Histonet] Tissue processor errors, failures and what to do

2020-07-04 Thread Patpxs via Histonet
Hi Garrey, The answer is “it depends”. What you do when a processor fails depends on the failure point. If the tissue is still in dehydrant it gets treated differently than if it fails in the intermediate solvent. Paula Sent from my iPhone > On Jul 4, 2020, at 10:08 AM, Garrey Faller via

[Histonet] Tissue processor errors, failures and what to do

2020-07-04 Thread Garrey Faller via Histonet
Happy 4th to all. Does anyone have a procedure on what to do when a tissue processor fails or alarms. I want to learn more about the science behind tissue processing so I know what to do when the machine fails. This happened to a friend recently and I want to prevent my tissues/biopsies from be

[Histonet] Tissue Processor Validation

2019-09-05 Thread Ashwin Ashwin via Histonet
Hello Everyone, If we change a processing protocol and are in need to recalibrate our tissue processors, is it required to perform concordance study between the old protocol and the new? Please let me know Thanks Sent from my iPhone ___ Histonet mailin

Re: [Histonet] Tissue Processor

2018-02-05 Thread Terri Braud via Histonet
1648 Huntingdon Pike Meadowbrook, PA 19046 ph: 215-938-3689 fax: 215-938-3874 Care, Comfort, and Heal Today's Topics: 1. tissue processor (Lisa Brenner) -- Message: 1 From: "Lisa Brenner" Subject: [Histonet] tissue pr

Re: [Histonet] tissue processor

2018-02-02 Thread Cooper, Brian via Histonet
Lisa Brenner via Histonet [mailto:histonet@lists.utsouthwestern.edu] Sent: Friday, February 02, 2018 10:36 AM To: histonet@lists.utsouthwestern.edu Subject: [Histonet] tissue processor (EXTERNAL EMAIL) Hello, We are in the market for a new vacuum infiltrating tissue processor. Our old VIP has

[Histonet] tissue processor

2018-02-02 Thread Lisa Brenner via Histonet
Hello, We are in the market for a new vacuum infiltrating tissue processor. Our old VIP has been a work horse but needs replacing. What is everyone using? What works well? What's new? Lisa Brenner Confidentiality Notice: The information contained in this e-mail message, including any atta

[Histonet] Tissue Processor question

2016-10-21 Thread Terri Braud via Histonet
Hi fellow Histonetters- I'm looking for any feedback from working labs using Thermo's Excelsior Processor. If you can help, can you please also include the age and usage on your instrument? Thanks, Terri Terri L. Braud, HT(ASCP) Anatomic Pathology Supervisor Laboratory Holy Redeemer Hospital 16

[Histonet] Tissue processor advice for a small start-up

2015-09-29 Thread Caroline Miller via Histonet
Hi All, Our company has reached the size where we can justify buying our own processor! Yippee!! However that also leads me to the issue of which one to buy :) We will have very small batches of blocks to process, approx 10 per process maybe twice per week, but we need the machine to be very res

[Histonet] Tissue Processor Reviews

2015-09-21 Thread ODea, Elise via Histonet
Thank you all for your opinions and experiences of Excelsior and Sakura! Elise ___ Histonet mailing list Histonet@lists.utsouthwestern.edu http://lists.utsouthwestern.edu/mailman/listinfo/histonet

RE: [Histonet] Tissue Processor Validation Procedure

2015-01-31 Thread Jamal
@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Validation Procedure Hello to all in Histoland. Does anyone have a tissue processor validation procedure that they would be willing to share? Any help in this would be greatly appreciated. Allison Scott HT(ASCP) Histology Supervisor LBJ Hospital

[Histonet] Tissue Processor Validation Procedure

2015-01-30 Thread Johnson, Carole
Hello Allison, When we purchased a new processor, I chose tissue specimens representative of what we normally see in our lab (surgical/biopsy and necropsy tissues) from cases which had been signed out within the previous 48-72 hours. I then ran parallel samples on both processors (same protocol

[Histonet] Tissue Processor Validation Procedure

2015-01-29 Thread Scott, Allison D
Hello to all in Histoland. Does anyone have a tissue processor validation procedure that they would be willing to share? Any help in this would be greatly appreciated. Allison Scott HT(ASCP) Histology Supervisor LBJ Hospital 713-566-5287(Lab) 713-566-2148(Office) CONFIDENTIALITY NOTICE: If yo

[Histonet] Tissue Processor

2014-12-02 Thread Patti McDavid
We have used VIP in the past and it was adequate. We use the Leica Peloris as our main processor and love it. Patti L. McDavid Laboratory Manager 4500 Thirteenth Street P.O. Box 1810 Gulfport, MS 39502-1810 Phone 228-575-2340 Fax 228-865-3325 pmcda...@mhg.com [http:/

RE: [Histonet] Tissue Processor Validation Procedure

2014-11-08 Thread Jamal
@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Validation Procedure Hi, I am wondering if you guys in histoland will like to share your Tissue Processor Validation Procedure with me. Thanks, Adesuyi == CONFIDENTIALITY NOTICE: This e-mail

[Histonet] Tissue Processor Validation Procedure

2014-11-06 Thread Adesupo, Adesuyi (Banjo)
Hi, I am wondering if you guys in histoland will like to share your Tissue Processor Validation Procedure with me. Thanks, Adesuyi == CONFIDENTIALITY NOTICE: This e-mail communication and any attachments may contain confidential and privileged information

[Histonet] Tissue Processor Location in the Lab

2014-10-30 Thread Wheelock, Timothy R.
Hi Everyone: Even though many tissue processors are closed systems with fume filters, do people try to put them next to fume hoods or under some ventilation duct to add an extra layer of protection from solvent fumes? Thanks, Tim Wheelock Harvard Brain Tissue Resource Center McLean Hospital B

Re: [Histonet] Tissue Processor Validation:

2014-09-06 Thread Mike Andrews
e: +966 12 670 0099 | Fax: +966 12 676 4984 | > www.alborglaboratories.com > > > -Original Message- > From: histonet-boun...@lists.utsouthwestern.edu > [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Jb > Sent: Friday, September 05, 2014 9:1

RE: [Histonet] Tissue Processor Validation:

2014-09-06 Thread Jamal
@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Validation: Does anyone have a tissue processor validation form that they are willing to share? Thank you. Sent from my iPhone ___ Histonet mailing list Histonet@lists.utsouthwestern.edu http

[Histonet] Tissue Processor Validation:

2014-09-05 Thread Jb
Does anyone have a tissue processor validation form that they are willing to share? Thank you. Sent from my iPhone ___ Histonet mailing list Histonet@lists.utsouthwestern.edu http://lists.utsouthwestern.edu/mailman/listinfo/histonet

[Histonet] tissue processor regulations

2014-05-01 Thread Ronda Mire
Hello Histonet, I recently started work in a research lab after working in a clinical setting for 20 years. The lab that I was at had all the tissue processors in a separate negative pressure room. The lab that I work at now has the processors out in the main work area. I noticed the fumes i

[Histonet] Tissue Processor

2013-09-04 Thread Vickroy, Jim
Wondered if anyone had any thoughts or information on the comparison of two tissue processors that we are considering. The two instruments are the Leica ASP6025 and the Thermofisher Excelsior ES. We originally also considered another VIP. The Leica ASP6025 and the Thermofisher unit have a l

RE: [Histonet] Tissue Processor Validation

2013-09-01 Thread tahseen
:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Beth Brinegar Sent: Wednesday, August 28, 2013 8:40 AM To: Rathborne, Toni Cc: histonet@lists.utsouthwestern.edu; dianar...@aol.com Subject: Re: [Histonet] Tissue Processor Validation Yes, we do a few validation runs. Nothing too extreme, however we gross some

[Histonet] Tissue Processor

2013-08-29 Thread Sue
We always validated our processor but not formally.  CAP wants a more formal approach with procedures. What we did in the past was run placenta and colon and reviewed stained slides. We may have to be a little more critical and run various types of slides and programs. Susan T. Patur

RE: [Histonet] Tissue Processor Validation

2013-08-28 Thread joelle weaver
have to do a complete revalidation. Joelle Weaver MAOM, HTL (ASCP) QIHC > From: hor...@archildrens.org > To: bbrinegar...@gmail.com; trathbo...@somerset-healthcare.com > Date: Wed, 28 Aug 2013 09:10:39 -0500 > Subject: RE: [Histonet] Tissue Processor Validation > CC: histonet@lists.u

RE: [Histonet] Tissue Processor Validation

2013-08-28 Thread Horn, Hazel V
- From: histonet-boun...@lists.utsouthwestern.edu [mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Beth Brinegar Sent: Wednesday, August 28, 2013 8:40 AM To: Rathborne, Toni Cc: histonet@lists.utsouthwestern.edu; dianar...@aol.com Subject: Re: [Histonet] Tissue Processor Validatio

Re: [Histonet] Tissue Processor Validation

2013-08-28 Thread Beth Brinegar
tsouthwestern.edu > Subject: RE: [Histonet] Tissue Processor Validation > > Yes definitely, 5 programs, 25-50 sister, serial sections on tissue that > represents patient samples. > > > > > Joelle Weaver MAOM, HTL (ASCP) QIHC > > > From: dianar...

RE: [Histonet] Tissue Processor Validation

2013-08-28 Thread Rathborne, Toni
onet@lists.utsouthwestern.edu Subject: RE: [Histonet] Tissue Processor Validation Yes definitely, 5 programs, 25-50 sister, serial sections on tissue that represents patient samples. Joelle Weaver MAOM, HTL (ASCP) QIHC > From: dianar...@aol.com > Date: Tue, 27 Aug 2013 22:07:19 -0400 >

RE: [Histonet] Tissue Processor Validation

2013-08-28 Thread joelle weaver
Yes definitely, 5 programs, 25-50 sister, serial sections on tissue that represents patient samples. Joelle Weaver MAOM, HTL (ASCP) QIHC > From: dianar...@aol.com > Date: Tue, 27 Aug 2013 22:07:19 -0400 > To: histonet@lists.utsouthwestern.edu > Subject: [Histonet] Tiss

[Histonet] Tissue Processor Validation

2013-08-27 Thread DianaRip1
I am interested in knowing how many of you out there Validate your Tissue Processors before you put them into use and how extreme you go. Diana Ripley ___ Histonet mailing list Histonet@lists.utsouthwestern.edu http://lists.utsouthwestern.edu/mailman/

RE: [Histonet] tissue processor remote alarm systems

2013-02-14 Thread Dessoye, Michael J
rom: Morken, Timothy [mailto:timothy.mor...@ucsfmedctr.org] Sent: Thursday, February 14, 2013 11:01 AM To: Histonet Subject: [Histonet] tissue processor remote alarm systems I'd like to hear from labs that use a remote messaging system for their VIP5 tissue processors that are left unattended ov

[Histonet] tissue processor remote alarm systems

2013-02-14 Thread Morken, Timothy
I'd like to hear from labs that use a remote messaging system for their VIP5 tissue processors that are left unattended overnight. One I saw on old histonet posts is Sensaphone. I am wondering what kind of information you can get from the system. Does it just say there is a problem but no detai

RE: [Histonet] TISSUE PROCESSOR FOLLOW-UP QUESTION

2012-12-07 Thread Cynthia Pyse
...@lists.utsouthwestern.edu] On Behalf Of Tim Wheelock Sent: Friday, December 07, 2012 12:44 PM To: Histonet@lists.utsouthwestern.edu Subject: [Histonet] TISSUE PROCESSOR FOLLOW-UP QUESTION Hi All: First, thank you for all your feedback on the processors. I site-visited a VIP6 at a local hospital From my

[Histonet] TISSUE PROCESSOR FOLLOW-UP QUESTION

2012-12-07 Thread Tim Wheelock
Hi All: First, thank you for all your feedback on the processors. I site-visited a VIP6 at a local hospital From my understanding the VIP6 can rotate absolute alcohol and xylene (using the bulk reservoirs), as well as the paraffin stations, but it cannot rotate other concentrations of alcohol

Re: [Histonet] Tissue Processor

2012-08-08 Thread Pam Marcum
.  If you have any questions let me know. Pam Marcum UAMS - Original Message - From: "Karen - SMMC-SF Heckford" To: histonet@lists.utsouthwestern.edu Sent: Wednesday, August 8, 2012 7:11:13 AM Subject: [Histonet] Tissue Processor I am going to need to purchase a

Re: [Histonet] Tissue Processor

2012-08-08 Thread Rene J Buesa
Sakura René J. From: "Heckford, Karen - SMMC-SF" To: "histonet@lists.utsouthwestern.edu" Sent: Wednesday, August 8, 2012 8:11 AM Subject: [Histonet] Tissue Processor I am going to need to purchase a new tissue processor mine keeps breaking

[Histonet] Tissue Processor

2012-08-08 Thread Heckford, Karen - SMMC-SF
I am going to need to purchase a new tissue processor mine keeps breaking down. What tissue processor would you buy and why? I would greatly appreciate the help. Cheers, Karen Heckford HT ASCP CE Lead Histology Technician St. Mary's Medical Center 450 Stanyan St. San Francisco, Ca. 94117 415-

[Histonet] Tissue Processor

2012-04-16 Thread Gauch, Vicki
Hi everyone, I was wondering if anyone had any input on the Logos tissue processor by Milestone. Is anyone currently using it and what are the pros and cons, etc. . Any information would be greatly appreciated. Thanks, Vicki Gauch AMCH Albany, NY

[Histonet] Tissue Processor

2011-12-12 Thread Patricia Bourassa
Hello! We are looking to purchase a used tissue processor for a small biotech. We will be working with calcified bone so we will only use the machine for dehydration and will infiltrate with methacrylate by hand. I'm interested in suggestions. It's been a while since I've done histology and I'd

RE: Histonet Digest, Vol 95, Issue 19 [Histonet] Tissue processor

2011-10-14 Thread Steve McClain
financial interst in that company, but I have had great experience and I am biased toward small companies which deliver great service. Message: 8 Date: Tue, 04 Oct 2011 11:56:03 -0400 From: "Mike Tighe" Subject: [Histonet] Tissue processor To: Message-ID: <4e8a

[Histonet] Tissue Processor

2011-10-14 Thread Lori White
Hi Mike, We have both the VIP6 and TEC5 and are extremely happy with them. What you spend up front, you will save in repair costs, downtime and frustration. Lori -- Message: 8 Date: Tue, 04 Oct 2011 11:56:03 -0400 From: "Mike Tighe" Subject: [Histon

RE: [Histonet] Tissue processor

2011-10-04 Thread Marcum, Pamela A
Haas Sent: Tuesday, October 04, 2011 12:08 PM To: Mike Tighe; histonet@lists.utsouthwestern.edu Subject: Re: [Histonet] Tissue processor We purchased a processor this past year and went with the VIP6. We demo'd several but went with the VIP because of its reliability and quality. We own two

Re: [Histonet] Tissue processor

2011-10-04 Thread Sheila Haas
le to convince administration the VIP was worth the difference.   Hope this helps. Sheila Haas Laboratory Supervisor MicroPath Laboratories, Inc.   From: Mike Tighe To: histonet@lists.utsouthwestern.edu Sent: Tuesday, October 4, 2011 11:56 AM Subject: [Histonet] Tissue processor I am looking to pur

RE: [Histonet] Tissue processor

2011-10-04 Thread Breeden, Sara
Although I appreciate individual opinions, do not overlook the Leica line. Us "old timers" may recognize the names "Sakura" or "Thermo" (in its many permutations) more readily, I'd put my money on Leica every time (and I have - with processor, embedding center, microtome and autostainer). They do

Re: [Histonet] Tissue processor

2011-10-04 Thread Rene J Buesa
transportation? It is the same thing with histology instruments! René J. --- On Tue, 10/4/11, Mike Tighe wrote: From: Mike Tighe Subject: [Histonet] Tissue processor To: histonet@lists.utsouthwestern.edu Date: Tuesday, October 4, 2011, 11:56 AM I am looking to purchase a Processor and embedding

[Histonet] Tissue processor

2011-10-04 Thread Mike Tighe
I am looking to purchase a Processor and embedding center. I have received a quote for VIP6 and a TEC5 but they seem to be a bit out of our price range. Does anyone have another Processor/embedding center that they found to be reliable? Thanks Mike

RE: [Histonet] Tissue Processor Reagents

2011-01-04 Thread sgoebel
...@lists.utsouthwestern.edu] On Behalf Of Katherine Leonard Sent: Tuesday, January 04, 2011 12:32 PM To: Histonet@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Reagents Hello histonetters ~ Is there any consensus regarding how often to change processor reagents? I am using the TBS

[Histonet] Tissue Processor Reagents

2011-01-04 Thread Katherine Leonard
Hello histonetters ~ Is there any consensus regarding how often to change processor reagents? I am using the TBS ATP1 tissue processor and the tissue size ranges from biopsy to 1.5cm^3. -- Katherine Leonard Research Associate ___ Histonet mailing list

RE: [Histonet] Tissue Processor Advice

2010-10-22 Thread WILLIAM DESALVO
> Date: Fri, 22 Oct 2010 16:16:50 -0400 > From: caymanfl...@gmail.com > To: histonet@lists.utsouthwestern.edu > Subject: [Histonet] Tissue Processor Advice > > We are in need of some advice regarding rapid tissue processors. Models we > are considering: > > Sakura Xpre

RE: [Histonet] Tissue Processor Advice

2010-10-22 Thread Jesus Ellin
[mailto:histonet-boun...@lists.utsouthwestern.edu] On Behalf Of Feher, Stephen Sent: Friday, October 22, 2010 2:12 PM To: caymanfl...@gmail.com; histonet@lists.utsouthwestern.edu Subject: RE: [Histonet] Tissue Processor Advice We are using the Peloris with a 2 hr, 4 hr and 8 hr protocol. We run 2 hour protocols

RE: [Histonet] Tissue Processor Advice

2010-10-22 Thread Feher, Stephen
22, 2010 4:17 PM To: histonet@lists.utsouthwestern.edu Subject: [Histonet] Tissue Processor Advice We are in need of some advice regarding rapid tissue processors. Models we are considering: Sakura Xpress Leica Peloris Thermo STP 420 It seems none of these models are perfect in every respect

[Histonet] Tissue Processor Advice

2010-10-22 Thread caymanfl...@gmail.com
We are in need of some advice regarding rapid tissue processors. Models we are considering: Sakura Xpress Leica Peloris Thermo STP 420 It seems none of these models are perfect in every respect. I'm interested in anyone's opinions of these processors and your experience with them. All input is

[Histonet] tissue processor times

2009-12-04 Thread Carol Bryant
Hello everyone! We have always had one program (catch-all) for all our specimens that we ran overnight on an Excelsior tissue processor. We just purchased a new VIP 6 and I want to have a program for our larger fatty specimens such as breast cases and run our small biopsies on another program.

Re: [Histonet] Tissue Processor Advice

2009-08-05 Thread Anne van Binsbergen
Sakura Tissue Tek VIP5!! I have 2 of these sturdy workhorses and am very happy Have been associated with VIP5's for as long as they have been around - wonderful machines! AnnieinAbuDhabi aka.Annie out of Africa 2009/8/5 Igor Deyneko > Dear Histonetters! > I need your advice in Tissue Processors.

Re: [Histonet] Tissue Processor Advice

2009-08-04 Thread thecitan
Sender: histonet-boun...@lists.utsouthwestern.edu To: Histonet Subject: [Histonet] Tissue Processor Advice Sent: Aug 4, 2009 1:43 PM Dear Histonetters! I need your advice in Tissue Processors. The one we currently use, Tissue Tek VIP 3000, is archaic and has finally died. So we are looking into

[Histonet] Tissue Processor Advice

2009-08-04 Thread Igor Deyneko
Dear Histonetters! I need your advice in Tissue Processors. The one we currently use, Tissue Tek VIP 3000, is archaic and has finally died. So we are looking into buying a new one. I know that Thermo and Leica both have processors, as well as new Tissue teks, but I wanted to get opinions if you hav

Re: [Histonet] tissue processor

2009-06-23 Thread mari . ann . mailhiot
cc lists.utsouthwest ern.edu Subject [Histonet] tissue

[Histonet] tissue processor

2009-06-23 Thread Carol Bryant
If you were going to purchase a new tissue processor, which would you recommend for a small lab that runs ≈ 200 blocks per day?  The majority of our workload is biopsies (skins & GI) with a few breast masses.  We also do a fair volume of IHC and are reluctant to go to microwave processing for th

[Histonet] Tissue Processor

2009-04-02 Thread Pathology
FREE TBS Tissue Processor (you pay shipping). Kind of a lemon. Liesl Frasier 719-530-2200 x2195 ___ Histonet mailing list Histonet@lists.utsouthwestern.edu http://lists.utsouthwestern.edu/mailman/listinfo/histonet

Re: [Histonet] Tissue processor issues

2009-02-04 Thread Mary . Cheles
ern.edu cc 02/04/2009 08:40 Subject AM [His

[Histonet] Tissue processor issues

2009-02-04 Thread Witt, Dan
Hello, this is my use of Histonet. I have a Leica ASP 300 tissue processor that is constantly malfunctioning. The last 5 weks it has had a pressurise issue between stations 2 & 3. The service guy has installed multiple new parts with little success. Has anyone else had this issue or any other pr

Re: [Histonet] Tissue Processor

2008-10-02 Thread Mark Tarango
; Tarango > Sent: Wednesday, October 01, 2008 5:15 PM > To: histonet > Subject: [Histonet] Tissue Processor > > Does anyone happen to have a good used tissue processor for sale? I'm > also > interested in an embedding center and microtome. > > Thanks > >

RE: [Histonet] Tissue Processor

2008-10-02 Thread Ford Royer
: [EMAIL PROTECTED] [mailto:[EMAIL PROTECTED] On Behalf Of Mark Tarango Sent: Wednesday, October 01, 2008 5:15 PM To: histonet Subject: [Histonet] Tissue Processor Does anyone happen to have a good used tissue processor for sale? I'm also interested in an embedding center and microtome. Thanks

[Histonet] Tissue Processor

2008-10-01 Thread Mark Tarango
Does anyone happen to have a good used tissue processor for sale? I'm also interested in an embedding center and microtome. Thanks Mark ___ Histonet mailing list Histonet@lists.utsouthwestern.edu http://lists.utsouthwestern.edu/mailman/listinfo/histone