The issue with talk-only is you don't have remote control of instrument
settings hence talk only
Content by Scott
Typos by Siri
> On Feb 24, 2017, at 11:47 AM, James Peroulas wrote:
>
> Thanks for the suggestions Scott(s)!
>
> It ended up being something simple. With no
Thanks for the suggestions Scott(s)!
It ended up being something simple. With no signal coming into the 5328A,
it did not produce any output over GPIB. I would have expected a continuous
readout of 0Hz. Once I applied a signal, a simple GPIB "read" request
returned the measurements I needed. No
If your counter is showing up in the NI device tree, you should be pretty
close to getting it working.
I haven't used a 5328A, but I suspect the command you have shown above is
sample code for a basic program, the actual HPIB command may just be a 'T'.
On Thu, Feb 23, 2017 at 1:37 AM, James
'Talk Only' mode responds to serial poll mode which is querying all instruments
on the bus. And it's primary use is in automated test systems
You want to disable talk-only mode for normal interactive control of an
instrument using GPIB
Content by Scott
Typos by Siri
> On Feb 23, 2017, at
Does anyone have experience using older HP equipment with the NI
GPIB-ENET/1000 GPIB to ethernet box? Is this even possible with such an old
piece of equipment? I'm very familiar with using SCPI commands over telnet
but not so familiar with HPIB.
So far I've installed the NI 488.2 software which